04/05/1995

Recalls and Field Corrections:  Foods -- Class I -- 04/05/1995

April 5, 1995                                                95-14

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        Smoked Salmon, vacuum packed in various weights.  
               Recall #F-568-5.
CODE           12 22 94 on package and 269 on case carton.
MANUFACTURER   Tony's Smoke House and Cannery, Oregon City, Oregon.
RECALLED BY    Manufacturer, by telephone February 11-12, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Oregon, Washington state.
QUANTITY       857 pounds were distributed.
REASON         Product is contaminated with Listeria monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Alba Dairy Shake Mix, Artificial Vanilla Flavor, Alba 77,
               available in retail cartons containing 8 x 0.75 ounce
               pouches of product.  Recall #F-530-5.
CODE           Lot numbers:  H2374, H238, H239, H2404, H2414.
MANUFACTURER   Near East Food Products, Leominster, Massachusetts.
RECALLED BY    Heinz U.S.A., Pittsburgh, Pennsylvania, by fax September 27,
               1994, and October 4, 1994 and by letter October 17, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Undetermined
REASON         Product contains undeclared FD&C Yellow No. 5. 

_______________
PRODUCT        Power Ade Beverage, in 64 ounce plastic bottles, bottled for
               the Coca-Cola Company, Atlanta, Georgia, by Hi Country,
               Selah, Washington.  Recall #F-577-5.
CODE           3257****2  (**** = Time) is stamped on some of the bottles. 
               Majority of the bottles were not coded.
MANUFACTURER   Johnson Controls Inc., Tacoma, Washington.
RECALLED BY    Manufacturer, by meeting held March 3, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Washington state.
QUANTITY       14,000 bottles were distributed.
REASON         Product contains undeclared FD&C Yellow No. 5.

_______________
PRODUCT        Frozen, preformed cookie dough:
               (a) Mrs. Rich's Chocolate Chip Cookie Dough, 22 pound 8
               ounce case, approximately 240 cookies per case; (b) Mrs.
               Rich's Oatmeal Raisin Cookie Dough, 22 pound 8 ounce case,
               Approximately 240 cookies per case; (c) Marquee Macadamia
               Nut with White Chunks, Ready to Bake Gourmet Cookie Dough,
               11 pound 4 ounce case, (12 15-ounce boxes per case); (d)
               Marquee Honey & Nut Oatmeal Raisin, Ready to Bake Gourmet
               Cookie Dough, 11 pound 4 ounce case (12 15-ounce boxes per
               case); (e) Marquee Chocolate Chunk, Ready to Bake Gourmet
               Cookie Dough, 11 pound 4 ounce case (12 15-ounce per case). 
               Recall #F-596/600-5.
CODE           Product Code              Lot Numbers
               (a) 39642                 M5051, M5052, M5056, M5058
               (b) 39677                 M5045, M5055
               (c) 11205-45577-39407     M5047, M5054
               (d) 11205-45575-39415     M5048, M5056
               (e) 11205-45576-39423     M5040, M5055.
MANUFACTURER   Rich Products Corporation, Murfreesboro, Tennessee.
RECALLED BY    Rich Products Corporation, Buffalo, New York, by letter
               March 2, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       1,779 cases were distributed.
REASON         The products may be contaminated with aluminum wire.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Alba Dairy Shake Mix, Artificial Strawberry Flavor, in
               individual serving envelopes of 0.75 ounces each.  
               Recall #F-529/530-5.
CODE           Lot numbers:  _H1974 J2, _H2314 J2, _H2324 J2, _H2574 J2,
               _H2594 J2 (first character of code is an Alpha character
               designating hour of day).
MANUFACTURER   Near East Food Products, Leominster, Massachusetts.

                                    -2-RECALLED BY    Heinz U.S.A., Pittsburgh, Pennsylvania, by fax September 27,
               1994, and October 4, 1994 and by letter October 17, 1994. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Undetermined
REASON         Product is misbranded -- incorrect ingredient statement.

_______________
PRODUCT        Various tomatoes packed in olive oil:
               (a) Bella Sun Luci Sun Dried Tomatoes in Olive Oil and
               Herbs, 35 ounces; (b) Trader Joe's Sun Dried Tomato Pesto in
               Olive Oil, 8.5 ounces;
               (c) Bella Sun Luci Sun Dried Tomato Pesto in Pure Olive Oil,
               6.5 ounces.  Recall #F-593/595-5.
CODE           None.
MANUFACTURER   Mooney Farms, Chico, California.
RECALLED BY    Manufacturer, by telephone April 4, 1994; (b & c) by
               telephone February 22, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Alaska, Arizona, California, Colorado, Connecticut, Florida,
               Hawaii, Maryland, Massachusetts, Montana, Nevada, New
               Hampshire, New Jersey, New Mexico, New York, Oregon,
               Vermont, Washington state, Canada.
QUANTITY       Undetermined.
REASON         The products are misbranded in that the label states the
               products are packed in olive oil instead of packed in a
               blend of soybean oil and olive oil.


RECALLS AND FIELD CORRECTIONS:  COSMETICS -- CLASS II
=====================
_______________
PRODUCT        Vinegar and Water Extra Mild Douche, in 6 fluid ounce
               plastic bottles, under the following brand names:  Fresh
               Gards, Freddy's Rx, and Kroger Disposable Douche.  
               Recall #F-578-5.
CODE           Lot numbers:  4245C, 4246C, 4249C, 4250C.
MANUFACTURER   Hospital Specialty Company, Cleveland, Ohio.
RECALLED BY    Manufacturer, by letter December 2, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and The Philippines.
QUANTITY       Approximately 2,300 cases (12 retail boxes per case) were
               distributed.
REASON         Product is contaminated with the Aeurobasadium pullulan (an
               airborne mold).


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Preparation H Hemorrhoidal Ointment, in 1 ounce tubes.
               Recall #D-094-5.
CODE           Lot #93-34 EXP 9/95.
MANUFACTURER   Whitehall Laboratories, Madison, New Jersey.

                                    -3-RECALLED BY    Manufacturer, by letters of May 16, 1994 and July 15, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       3,596 dozen 1-ounce tubes were distributed; firm estimates
               none remains on the market.
REASON         Presence of mold.

_______________
PRODUCT        Searle Calan Tablets, verapamil hydrochloride 40 mg, in 100
               tablet bottles, an immediate action Rx oral calcium slow-
               channel blocker for the treatment of angina, arrhythmias and
               essential hypertension.  Recall #D-105-5.
CODE           Lot numbers:  3J043, 3L045, 3N046, 3P047, 4A048, 4A049.
MANUFACTURER   G.D. Searle & Company, Caguas, Puerto Rico.
RECALLED BY    G.D. Searle & Company, Skokie, Illinois, by letters dated
               March 22, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       147,698 bottles were distributed; firm estimated that 20% of
               the product remained on market at time of recall initiation.
REASON         Product does not meet dissolution specifications.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Claforan (Cefotaxime Sodium) Sterile Powder for Injection
               (IM/IV), Rx broad spectrum cephalosporin antibiotic:  
               (a) 500 mg vials; (b) 1.0 gram vials; (c) 2.0 gram vials. 
               Recall #D-095/097-5.
CODE           All lots within expiration date.
MANUFACTURER   Hoechst Aktiengesellschaft, Federal Republic of Germany or
               Roussel Uclaf Plant 4, Romainville, France.
RECALLED BY    Hoechst Roussel Pharmaceuticals, Inc., Somerville, New
               Jersey by letter February 20, 1995.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Firm estimated that 740,000 vials remained on the market at
               time of recall initiation.
REASON         Subpotency/stability problems.  Field correction labeling
               change to assure withdrawable assay specifications after
               reconstitution.

_______________
PRODUCT        Hydrocortisone Spray, 0.5% in 1.5 fluid ounce (45ml) plastic
               bottles, OTC, used for the temporary relief of itching,
               under the following labels:  Superx Drugs, Goldline, Rugby.
               Recall #D-098-5.
CODE           Lot #U393 EXP 5/95.
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY    Manufacturer, by letter mailed March 16, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Florida.

                                    -4-QUANTITY       Approximately 2,841 45ml-units were distributed.
REASON         Product does not meet pH specifications.

_______________
PRODUCT        Facial Soap (liquid scrub), in 8 ounce plastic bottles, OTC,
               used for the control of acne.  Recall #D-099-5.
CODE           Lot numbers:  T148 EXP 7/95, V775 EXP 6/96.
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY    Manufacturer, by letter mailed March 10, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   California and Tennessee.
QUANTITY       Approximately 2,199 bottles were distributed.
REASON         Subpotency.

_______________
PRODUCT        Benzoyl Peroxide 5% Gel, 45 grams (1.5 ounce) plastic tubes,
               OTC, used for the treatment of acne bacteria on skin. 
               Recall #D-100-5.
CODE           Lot numbers:  S285 EXP 3/95, T960 EXP 10/95, V238 EXP 3/96,
               V770 EXP 6/96, W180 EXP 8/96, W832 EXP 11/96, W868 EXP
               10/96, X043 EXP 11/96, X052 EXP 11/96.
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY    Manufacturer, by letter mailed March 14, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   California, Ohio, Georgia, New York, Kentucky, Florida,
               Michigan, Maine, Maryland, Alabama, Pennsylvania, West
               Indies.
QUANTITY       Approximately 37,998 units were distributed.
REASON         Product separation.

_______________
PRODUCT        Medadyne Liquid, in 0.5 ounce and 1 ounce glass bottles, OTC
               used for the treatment of canker sores, fever blisters, and
               cold sores, under the Vetco and Dal-Med labels.
               Recall #D-101-5.
CODE           Lot numbers:  T247 EXP 7/95, U202 EXP 11/95.
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY    Manufacturer, by letter mailed March 16, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   New York.
QUANTITY       29,456 bottles were distributed.
REASON         Superpotency.

_______________
PRODUCT        (a) Qualitest brand Promethazine with Codeine Cough Syrup,
               in 4 fluid ounce  bottles, indicated for the temporary
               relief of coughs and upper respiratory symptoms associated
               with allergy or the common cold;
               (b) Qualitest brand Promethazine VC with Codeine Cough
               Syrup, in 4 fluid ounce bottles, indicated for the temporary
               relief of coughs and upper respiratory symptoms.
               Recall #D-102/103-5.

                                    -5-CODE           Lot numbers:  RA5051, RA5048; 
               Catalog numbers:  L-0603-1578-54, L-0603-1581-54.
MANUFACTURER   Barre National, Inc., Baltimore, Maryland.
RECALLED BY    Manufacturer, by letter March 1, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Alabama.
QUANTITY       12,024 bottles of lot RA5051, and 2,340 bottles of lot
               RA5048.
REASON         Label does not bear name and address of manufacturer or
               distributor.

_______________
PRODUCT        Selenium Sulfide Lotion/Shampoo, USP 2.5%, in 4 ounce
               plastic bottles, Rx, used for the treatment of dandruff and
               seborrheic dermatitis of the scalp.  Recall #D-104-5.
CODE           Lot #P819 EXP 8/95.
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY    Manufacturer, by letter to be mailed on April 17, 1995.
               Firm-initiated recall ongoing.
DISTRIBUTION   New York, Texas, Florida, Iowa, Ohio, Mississippi, Nebraska,
               New Jersey, Tennessee, California.
QUANTITY       Approximately 1,151 bottles were distributed.
REASON         Use of an unapproved raw material source.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Lifestyles Assorted Color Condoms with Spermicidal
               Lubricant.  Recall #Z-519-5.
CODE           Lot numbers:  410046500, 410049100, 410049200. 
MANUFACTURER   Ansell, Inc., Dothan, Alabama.
RECALLED BY    Manufacturer, by telephone December 2, 1994, followed by fax
               December 3, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Florida.
QUANTITY       189,000 condoms were distributed.
REASON         The subject lots had been recorded as passing the
               accelerated aging tests for tensile strength, when, in fact
               they had failed this testing.  The failing test results may
               indicate that the product does not have the shelf life which
               is claimed on its labeling.

_______________
PRODUCT        Drake Willock System 1000 Single Patient Delivery System for
               single patient hemodialysis.  Recall #Z-553-5.
CODE           Units within serial number block from 1001 to 3873. 
               However, each unit within this block does not require
               replacement of all of the suspect parts.  
MANUFACTURER   Althin Medical, Inc., Portland, Oregon.
RECALLED BY    Manufacturer, by telephone starting in June 1994.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.

                                    -6-QUANTITY       2,865 units are subject to field correction.
REASON         The CPVC may swell and impede the flow of liquid through the
               system.  The defect may affect the ultrafiltration flow
               meter, the flow equalizer, supply regulator, rinse block
               fitting, concentrate connectors, ultrafiltration removal
               regulator and dialyzer fittings.

_______________
PRODUCT        Drake Willock Systems 1000 Single Patient Delivery System
               fir single patient hemodialysis.  Recall #Z-556-5.
CODE           Units within serial number block 1001 to 3439.
MANUFACTURER   Althin Medical, Inc., Portland, Oregon.
RECALLED BY    Manufacturer, by telephone beginning June 1994.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       2,428 units are subject to field correction.
REASON         The pump head is cutting or abrading the blood tubing
               causing blood leaks.

_______________
PRODUCT        Laser Sensing Devices:  (a) Autosense Vehicle Sensors;
               (b) Miles ATS-101 Aerial Target System; (c) Ocean Wave Laser
               System (OWLS).  Recall #Z-566/568-5.
CODE           None.
MANUFACTURER   Schwartz Electro-Optics, Inc., Orlando, Florida.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action
               plan February 24, 1995.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 9 units; (b) 2 units; (c) 25 units.
REASON         Units exceeded the Class I accessible emission limits.

_______________
PRODUCT        Storz +4.0 Posterior Chamber D Intraocular Lenses:
               (a) Model 121UV; (b) Model P431UV.  Recall #Z-570/571-5.
CODE           Serial numbers:  (a) 19UP01 to 19UP32. 19UP34 to 19UP45,
               19UP82 to 19UP85; (b) 19LV15, 19LV16 and 19LV18.
MANUFACTURER   Storz Ophthalmics, Inc., Clearwater, Florida.
RECALLED BY    Manufacturer, by letter January 3, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Arizona, Florida, New York, Colorado, Wisconsin, Kansas,
               Maryland, South Dakota, California, Texas, Ohio,
               Pennsylvania, Missouri, Arizona, Massachusetts, Tennessee,
               Iowa, Utah, Puerto Rico, Canada, Taiwan.
QUANTITY       47 lenses were distributed.
REASON         Devices may be labeled with the incorrect diopter.

_______________
PRODUCT        Oxford Medilog Sleep and Respiration Interface (SRI) which
               monitors respiration.  Recall #Z-572-5.
CODE           All serial numbers.
MANUFACTURER   Oxford Medilog Systems, Abingdon, England.

                                    -7-RECALLED BY    Oxford Instruments Medical Systems, Division, Clearwater,
               Florida, by telephone October 7, 1993, followed by letter
               October 19, 1993.  Firm-initiated field correction complete.
DISTRIBUTION   Florida, Alabama, Georgia, Utah, Missouri, Massachusetts,
               Pennsylvania, Washington State, Louisiana, Virginia, Puerto
               Rico, Canada, Mexico, Brazil.
QUANTITY       41 units were distributed.
REASON         If the electrodes are plugged into the SRI incorrectly,
               excessive current may be transmitted to the patient causing
               a tingling sensation.

________________
PRODUCT        1.  Intermate SV 50 ml/HR, product code 044050; 
               2.  Intermate SV 100 ml/HR, product code 044120; 
               3.  Intermate SV 200 ml/HR, product code 044220; 
               4.  Intermate LV 50 ml/HR, product code 049050; 
               5.  Intermate LV 100 ml/HR, product code 049100; 
               6.  Intermate LV 250 ml/HR, product code 049250; 
               7.  Italian Intermate SV 200 ml/HR, product code 42200; 
               8.  French Intermate SV 50 ml/HR, product code 43050; 
               9.  French Intermate SV 100 ml/HR, product code 43100; 
               10. French Intermate SV 200 ml/HR, product code 43200; 
               11. French Intermate LV 50 ml/HR, product code 439050; 
               12. French Intermate LV 100 ml/HR, product code 439100; 
               13. French Intermate LV 250 ml/HR, product code 439250; 
               14. Australian Intermate LV 250 ml/HR, product code 2C1006; 
               15. Australian Intermate SV 200 ml/HR, product code 2C1017; 
               16. Australian Intermate LV 50 ml/HR, product code 2C1018; 
               17. Australian Intermate LV 100 ml/HR, product code 2C1019; 
               18. Australian Intermate SV 50 ml/HR, product code 2C1028; 
               19. Australian Intermate SV 100 ml/HR, product code 2C1029.
               Recall #Z-573/591-5.
CODE           Product Code:       Lot Numbers: 
               044050              94G010, 94H015, 94J002, 94J014, 94J030, 
                                   94M008, AND 94N003 
               044120              94G002, 94G003, 94G004, 94H002, 94H003, 
                                   94H004, 94J003, 94J004, 94J005, 94J011, 
                                   94K001, 94K002, 94K003, 94M001, 94M002, 
                                   94M003, 94M004, AND 94N001 
               044220              94H013, 94H014, 94H016, 94H017, 94J007, 
                                   94J015, 94J028, 94K004, 94W001, 94W002,
                                   94W003, 94W004
               049050              94G015, 94J013 AND 94N002 
               049100              94H012, 94H019, 94J018, AND 94M009 
               049250              94G014, 94H007, 94H018, 94J010, 94J017, 
                                   94K005, AND 94M005 
               42200               94D006 AND 94E015 
               43050               94C009 AND 94D011 
               43100               94C005, 94D014, 94J001, 94J009, 94K006, 
                                   94K007, 94K008, 94M011 
               43200               94C007, 94D003 AND 94K009 
               439050              94E011 

                                    -8-               439100              94E006, 94J019, AND 94M011 
               439250              94E004 AND 94J020 
               2C1006              94F014 
               2C1017              94F019 
               2C1018              94G016 
               2C1019              94F017 
               2C1028              94G011 
               2C1029              94F018. 
MANUFACTURER   I.V. Systems Division, Baxter Healthcare Corporation,
               Irvine, California.
RECALLED BY    Manufacturer, by letter dated December 6, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       621,072 units were distributed; FDA estimated that 150,000
               units remained on market at time of recall initiation.
REASON         The adhesive which bonds the flow restricting stainless
               steel cannula to the acrylic stress member fails and allows
               fluid to bypass the flow restricting cannula and can result
               in solution infusion rates higher than intended.

_______________
PRODUCT        Bear-1000 Ventilator, Catalog #50-0850-00, indicated for the
               delivery of medical gases to pediatric and adult patients. 
               Recall #Z-593-5.
CODE           136 units containing EPROM release #7 bearing a unique
               serial number were manufactured.
MANUFACTURER   Bear Medical Systems, Inc., Riverside, California.
RECALLED BY    Manufacturer, by memorandum November 4, 1994, and by letter
               on November 8, 1994. Firm-initiated field correction
               complete.
DISTRIBUTION   California, Delaware, Florida, Kansas, Louisiana, Maine,
               Michigan, North Carolina, New Jersey, New York,
               Pennsylvania, South Dakota, Tennessee, Utah, Virginia, and
               internationally.
QUANTITY       115 units were distributed.
REASON         The ventilator does not deliver the lower tidal volumes in
               increments of 1 ml as described in the labeling due to
               defective software.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Drake Willock System 1000 Single Patient Delivery System, a
               single-patient delivery hemodialysis delivery system.  
               Recall #Z-554-5.
CODE           The following serial numbers are for domestic machines that 
               had pumps replaced in the field: 3583-3584, 3587-3590, 
               4084-4088, 4096-4101, 4104, 4107, 4182, 4194, 4203.  The 
               following serial numbers are for overseas machines that had 
               pumps replaced in the field:  4089, 4090, 4102, 4111-4124,
               4132-4137, 4144-4147, 4166-4170, 4175-4181, 4234, 4241,
               4242.  

                                    -9-               The following serial numbers are for machines that were     
               returned to the manufacturing facility for pump replacement: 
               4091-4095, 4103, 4105, 4106, 4148-4165, 4171-4174,
               4183-4193, 4195-4202, 4204-4207, 4226, 4227, 4229-4233,
               4238-4240.   
MANUFACTURER   Micropump, Inc., Vancouver, Washington (component).
RECALLED BY    Althin Medical, Inc., Portland, Oregon.  On August 29, 1994,
               it was decided that no more System 1000 machines would be
               installed, and that all shipped machines would be tracked. 
               On this day, shipping hold #94-002 was instituted.  On
               August 30, 1994, all undelivered domestic machines were
               located and stopped.  Firm-initiated field correction
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       127 machines were subject to field correction.
REASON         Mercury was found in the pump head.

_______________
PRODUCT        Drake Willock System 1000 Single Patient Delivery for single
               patient hemodialysis.  Recall #Z-555-5.
CODE           1001 to 2870 (non-inclusive).
MANUFACTURER   Althin Medical, Inc., Portland, Oregon.
RECALLED BY    Manufacturer, by telephone starting in June 1994 to arrange
               for retrofitting.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide and international. 
QUANTITY       1,857 units are subject to field correction.
REASON         The air trap float which controls the incoming water supply
               absorbs water and becomes deformed.

_______________
PRODUCT        Clinical Chemistry Analyzer, Summit Series 7000.
               Recall #Z-569-5.
CODE           Serial numbers 5306, 8319, 5315.
MANUFACTURER   Integrated Clinical Systems, Inc., Indianapolis, Indiana.
RECALLED BY    Manufacturer, by sending an update bulletin and field repair
               kits on January 14, 1994, and by fax January 23, 1995. 
               Firm-initiated field correction complete.
DISTRIBUTION   Tennessee, France, England.
QUANTITY       3 units were distributed.
REASON         If a serum tube is not fully seated in the primary tube
               sampler (an optional unit designed to alert the instrument
               operator when a tube does not contain enough blood serum to
               run all of the chemistry test specified), it can interfere
               with the drive bar stops causing the glass tube to break,
               which may cause injury.

_______________
PRODUCT        OPUS Myoglobin Test Modules, an in-vitro diagnostic device
               used with the OPUS and OPUS Plus Immunoassay System in the
               quantitative measurement of myoglobin in serum as an aid in
               the diagnosis of acute myocardial infarction.  
               Recall #Z-592-5.

                                   -10-CODE           Catalog #700-50, Lot #MYB8 EXP 11/24/94.
MANUFACTURER   Behring Diagnostics, Inc., Westwood, Massachusetts.
RECALLED BY    Manufacturer, by telephone between April 28, 1994 and May 3,
               1994, and by letter May 4, 1994.  Firm-initiated field
               correction complete.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       209 boxes (containing 50 modules each) were distributed;
               firm estimates none remains on the market.
REASON         Defective OPUS Myoglobin Test Modules may cause an inability
               by the user to perform or correctly read test results.

_______________
UPDATE         Beckman Synchron CX7 Clinical Analyzer, computer enhanced,
               CX4CE/CX5CE, Recall #Z-520-5 which appeared in the March 22,
               1995 Enforcement Report should read:
               RECALLED BY:  Beckman Instruments, Inc., Brea, California,
               by letter dated July 14, 1994.  Firm-initiated field
               correction ongoing.
               REASON:  Defective software may cause patient ammonia test
               results to be inaccurate.

                                   -11-


END OF ENFORCEMENT REPORT FOR APRIL 5, 1995.  BLANK PAGES MAY FOLLOW.



                                   ####