The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
05/29/1996

ENFORCEMENT REPORT FOR 05/29/1996 
^

May 29, 1996                                   96-22

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT        Mountain Man Nut & Fruit Co. Sesame Nut Mix,
               packaged in 1 pound clear plastic bags. 
               Recall #F-566-6.
CODE           None.  All product lacking cashews and
               soybeans in the ingredient statement.
MANUFACTURER   Mountain Man Nut & Fruit Company, Parker,
               Colorado.
RECALLED BY    Manufacturer, by telephone beginning April 10,
               1996, followed by letter, and by press release
               on April 18, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       2,915 pounds were distributed.
REASON         Product contains undeclared cashews and
               soybeans.


RECALLS AND FIELD CORRECTIONS:  COSMETICS -- CLASS II =======
_______________
PRODUCT        Various children's cologne and cologne gift
               sets in 0.5 ounce bottles:
                    Product Code          Product
               (a) 20002.000.000   Barbie Comb/Mirror/Cologne
               (b) 20003.000.000   Barbie Carded Cologne
               (c) 20004.000.000   Barbie Carded Wing Rack 
               (d) 20005.000.000   Barbie Floorstand (Inc.
                                   Fragrance Fun Set)
               (e) 20006.000.000   Barbie Assorted Cosmetic
                                   Rack 
               (f) 20010.000.000   Barbie Fragrance Fun Set
               (g) 20010.000.101   Barbie Fragrance Fun Set
               (h) 20041.000.000   Barbie Assortment
               (i) 20079.000.000   Barbie Cologne Keepsake
               (j) 21021.000.000   The Little Mermaid
                                   Comb/Mirror/Cologne
               (k) 21035.000.000   The Little Mermaid Carded
                                   Cosmetic Rack
               (l) 21045.000.000   The Little Mermaid
                                   Fragrance Set
               (m) 21045.000.101   The Little Mermaid
                                   Fragrance Set
               (n) 21073.000.000   The Little Mermaid
                                   Assorted Cosmetic Sets
               (o) 21073.100.000   The Little Mermaid Shelf
                                   Tray Assortment
               (p) 21080.000.000   The Little Mermaid Luau
                                   Dress Up Set
               (q) 21085.000.000   The Little Mermaid Prepack
               (r) 23067.000.000   Fragrance Set Assortment
               (s) 23068.000.000   Assorted Cosmetic Wing
                                   Rack
               (t) 23231.000.000   Girls Toiletry Floorstand
               Recall #F-546/565-6.
CODE           Those containing 0.5 ounce bottles of lot #
               L138 cologne under "Barbie" or "The Little
               Mermaid" label.
MANUFACTURER   Chemaid Laboratories, Inc., Saddle Brook, New
               Jersey (cologne).
RECALLED BY    Tsumura International, Secaucus, New Jersey,
               by letter on March 14, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       289,500 bottles were distributed; firm
               estimated 20,000 bottles remained on market at
               time of recall initiation.
REASON         The cologne is contaminated with Pseudomonas
               aeruginosa.
 

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Dilantin (Extended Phenytoin Sodium), USP, 30
               mg, in bottles of 100, anticonvulsant.  
               Recall #D-142-6.
CODE           Lot #27835L EXP 3/97.
MANUFACTURER   Warner Lambert Company, Lititz, Pennsylvania.
RECALLED BY    Manufacturer, by letter on May 6, 1996.  Firm-
               initiated recall ongoing.
                             -2-DISTRIBUTION   Nationwide.
QUANTITY       28,397 bottles were distributed.
REASON         Product does not meet dissolution
               specifications.

_______________
PRODUCT        Puritan Bennett (a) Carbon Dioxide, USP, 99%
               in the following cylinder sizes: B, D, E, P,
               F, M, G, T, and XL-45 refrigerated liquid
               container;
               (b) Carbon Dioxide 5%, Oxygen 95% Mixture in
               the following cylinder sizes: E, G, and H.
               Recall #D-143/144-6.
CODE           (a) Carbon Dioxide, U.S.P., manufactured at
               Linthicum Heights, MD (referred to as the
               Baltimore Branch)
               BATCH 54  BATCH 55  BATCH 56  BATCH 57
               AA29D54   AB06D55   AB15D56   AB21D57
               AA30D54   AB07D55   AB16D56   AB22D57
               AA31D54   AB08D55   AB02D56   AB23D57
               AB01D54   AB12D55             AB26D57
               AB02D54   AB14D55

               BATCH 58  BATCH 59  BATCH 60  BATCH 61
               AB27D58   AC05D59   AC14D60   AC26D61
               AB28D58   AC06D59   AC15D60   AC27D61
               AB29D58   AC07D59   AC18D60   AC28D61
               AC01D58   AC08D59   AC20D60   AC29D61
                         AC11D59   AC21D60   AD01D61
                         AC12D59   AC22D60   AD02D61
                         AC13D59   AC25D60

               BATCH 62  BATCH 63  BATCH 64
               AD03D62   AD11D63   AD17D64
               AD04D62   AD12D63   AD18D64
               AD05D62   AD15D63
               AD08D62
               AD09D62
               AD10D62
               Carbon Dioxide, U.S.P., manufactured in
               Bellmawr, NJ (referred to as the Philadelphia
               Branch)
               BATCH 79
               XD11D79
               (b) Carbon Dioxide 5%, Oxygen 95% mixture,
               manufactured at Linthicum Heights, MD.  Carbon
               dioxide used in mixtures is supplied to the
               manufacturing system in cylinders and not
               direct from the bulk supply.  Therefore, the
               following lot numbers of CO2/O2 mixture were
               manufactured using cylinders from the listed
               batches and not produced direct from the bulk
               supply:
                             -3-               BATCH 55  BATCH 59  BATCH 60  BATCH 62
               AB09D62   AC19D64   AD04D66   AD08D67
               AB21D62   AC19D65             AD17D68
               AB23D63.
MANUFACTURERS  Puritan-Bennett Corporation, a Subsidiary of
               Nellcor Puritan Bennett Inc., Linthicum
               Heights, Maryland; 
               Puritan-Bennett Corporation, a Subsidiary of
               Nellcor Puritan Bennett Inc., Bellmawr, New
               Jersey. 
RECALLED BY    Puritan-Bennett Corporation, a Subsidiary of
               Nellcor Puritan Bennett Inc., Gas Products
               Division, Overland Park, Kansas, by telephone
               and visit beginning April 19, 1996, followed
               by letter May 8-14, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Pennsylvania, Maryland, New Jersey,
               Washington, D.C.
QUANTITY       The Baltimore Branch distributed 1,253
               cylinders or XL-45's CO2 between Philadelphia
               1/29/96-4/19/96; Branch distributed 5
               cylinders CO2 between 4/11-15/96.
               The Baltimore Branch distributed 106 cylinders
               of CO2/O2 mixture between 2/9/96-4/18/96.
REASON         Presence of trace quantities of HCN (Hydrogen
               Cyanide). 


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Pharmics brand Polytinic, Rx diet supplement,
               in bottles of 100.  Recall #D-141-6.
CODE           Lot #3053.
MANUFACTURER   Gull Laboratories, Inc., doing business as
               USANA, Inc., Murray, Utah.
RECALLED BY    Pharmics, Inc., Salt Lake City, Utah, by
               letter faxed beginning March 25, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Missouri, Utah, Idaho, Nevada, Colorado, Iowa,
               Ohio, Florida, Wyoming.
QUANTITY       646 bottles were distributed; firm estimated
               that less than 100 bottles remained on market
               at time of recall initiation.
REASON         Subpotency of the folic acid ingredient.

_______________
PRODUCT        (a) Caffeine Tablets, manufactured for
               Specialty Products Co., Inc., in white plastic
               bottles; (b) Cough Suppressant liquid in 8
               ounce clear plastic bottles, under Ion Labs
               label.  Recall #D-145/146-6.
CODE           Lot numbers:  (a) 400577 10-96; 
               (b) 500458-6-97.
                             -4-MANUFACTURER   Ion Labs, Inc., Pinellas Park, Florida.
RECALLED BY    Manufacturer, by telephone on March 18, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   New Jersey.
QUANTITY       (a) 2,300 bottles; (b) 468 bottles were
               distributed.
REASON         Caffeine product does not meet content
               uniformity specifications.  The Sorbutuss
               product is subpotent.

_______________
PRODUCT        Loxitane (Loxapine Succinate) Capsules, 25 mg,
               in bottles of 100 and 1000, Rx antipsychotic
               drug.  Recall #D-147-6.
CODE           Control numbers 342-351 (exp. date 09/97) and
               374-435 (exp. date 01/99) were packaged in
               bottles of 1000 capsules; 
               Control numbers 350-338 (exp. date 10/97),
               360-359 (exp. date 11/97), 376-413 (exp. date
               01/99) and 388-455 (exp. date 09/99) were
               packaged in bottles of 100 capsules. 
MANUFACTURER   Wyeth-Ayerst/Lederle Laboratories, Inc., Pearl
               River, New York.
RECALLED BY    Whitehall Robins, Inc., Division of Wyeth-
               Ayerst/Lederle Laboratories, Inc., Richmond,
               Virginia, by letter on May 8, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and Australia.
QUANTITY       29,093 bottles were distributed.
REASON         Potency not assured through expiration date.

_______________
PRODUCT        Tricolan Antimicrobial Soap, in 32 ounce
               bottles.  Recall #D-148-6.
CODE           Lot #019-96.
MANUFACTURER   Perfecto Products Manufacturing, Inc.,
               Atlanta, Georgia.
RECALLED BY    Manufacturer, by telephone on April 8, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,365 bottles were distributed; firm estimated
               that 30-40% of the lot remained on market at
               time of recall initiation.
REASON         Subpotency.


HUMAN TISSUES MANDATORY RECALLS =============================
_______________
PRODUCT        Human Tissues for Transplant:  Human Eye
               Tissues.  Recall #B-410-6.
CODE           None.
MANUFACTURER   Mid-South Eye Bank, Memphis, Tennessee.

                             -5-RECALLED BY    Manufacturer, by letter dated January 23,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   Tennessee, Arkansas.
QUANTITY       112 shipments.
REASON         Human tissues, not tested for HIV-2 were
               distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
_______________
PRODUCT        Platelets, Pheresis.  Recall #B-402-6.
CODE           Unit #FF11472.
MANUFACTURER   Irwin Memorial Blood Center, San Francisco,
               California.
RECALLED BY    Manufacturer, by letter dated January 11,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product was collected from a donor who
               previously tested repeat reactive for HBsAg
               and was not properly reentered.

_______________
PRODUCT        (a) Red Blood Cells; (b) Fresh Frozen Plasma. 
               Recall #B-411/412-6.
CODE           Unit numbers:  (a) 40FG26902, 40FG27376,
               40FG27688, 40FG27992; (b) 40FG27376.
MANUFACTURER   American Red Cross Blood Services, Rock Falls,
               Illinois.
RECALLED BY    American Red Cross Blood Services, Peoria,
               Illinois, by telephone on October 17, 1995,
               followed by letter on October 24, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Illinois, Missouri, California, Tennessee.
QUANTITY       (a) 4 units; (b) 1 unit.
REASON         Blood products, collected from a donor who
               travelled to an area designated as endemic to
               malaria, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-414-6.
CODE           Unit #13FM00749.
MANUFACTURER   American Red Cross, Detroit, Michigan.
RECALLED BY    Manufacturer, by letter dated October 24,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY       1 unit.
REASON         Blood product, which was collected from a
               donor with a medical history of hepatitis, was
               distributed.

                             -6-_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh
               Frozen Plasma.  Recall #B-415/417-6.
CODE           Unit numbers:  (a) 38N03281, 38M10808; (b)
               38NO3281; (c) 38M10808.
MANUFACTURER   American Red Cross, Fort Wayne, Indiana.
RECALLED BY    Manufacturer, by letters dated November 21,
               1995, February 2, 1996, and March 11, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Indiana.
QUANTITY       (a) 2 units; (b) 1 unit; (c) 1 unit.
REASON         Blood products, which tested negative for all
               infectious diseases, but were collected from a
               donor who previously tested repeat reactive
               for anti-HIV-1, and was never properly
               reentered, were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh
               Frozen Plasma.  Recall #B-418/420-6.
CODE           Unit #13FL40372.
MANUFACTURER   American Red Cross, Detroit, Michigan.
RECALLED BY    Manufacturer, by letter dated October 26,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY       1 unit of each component.
REASON         Blood products, collected from a donor taking
               the drug, Proscar were distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-421-6.
CODE           Unit numbers:  XM70573, XM75407, XM76739,
               XM77060, XM77677, XM78015, XM81068, XM81661,
               XM82218, XM83940, XM84260.
MANUFACTURER   Community Bio-Resources, Inc., Grand Rapids,
               Michigan.
RECALLED BY    Manufacturer, by letter dated October 16,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Austria.
QUANTITY       11 units.
REASON         Blood products, which were collected from a
               donor who reported his lips pierced, were
               distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c)
               Recovered Plasma.  Recall #B-422/424-6.
CODE           Unit numbers:  (a) C29773, C33026; (b) C29773,
               C29913; (c) C29773, C29913, C33026.
MANUFACTURER   Carlisle Hospital Blood Bank, Carlisle,
               Pennsylvania.

                             -7-RECALLED BY    Manufacturer, by letters dated August 21,
               1995, September 11, 1995, and October 19,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       (a) 2 units; (b) 2 units; (c) 3 units.
REASON         Blood products, collected from donors taking
               the drug, Proscar were distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT        AS-1 Red Blood Cells.  Recall #B-403-6.
CODE           Unit #J86475.
MANUFACTURER   Irwin Memorial Blood Center, San Francisco,
               California.
RECALLED BY    Manufacturer, by letter dated January 11,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product was shipped with an extended
               expiration.

_______________
PRODUCT        (a) AS-1 Red Blood Cells; (b) Platelets. 
               Recall #B-404/405-6.
CODE           Unit #1371694.
MANUFACTURER   LifeSource, Glenview, Illinois.
RECALLED BY    Manufacturer, by telephone on September 29,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit of each component.
REASON         Blood products, collected from a donor who had
               received a MMR vaccination, twenty-five days
               prior to the donation, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-426-6.
CODE           Unit #17178-6918.
MANUFACTURER   United Blood Service, McAllen, Texas.
RECALLED BY    Blood Services, Inc., Scottsdale, Arizona, by
               telephone on February 20, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Red Blood Cells, remained at room temperature
               for an unknown amount of time, were
               distributed.


                             -8-RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS I ===========
_______________
PRODUCT        Synthetic Serum Substitute, Catalog #99193, in
               12 ml vials, designed for use as a tissue
               culture protein supplement.  Recall #Z-709-6.
CODE           Lot #9919350921 EXP 9/96.
MANUFACTURER   Irvine Scientific Sales Company, Inc., Santa
               Ana, California.
RECALLED BY    Manufacturer, by letter dated December 14,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Alabama, California, Connecticut, Florida,
               Hawaii, Illinois, Massachusetts, Missouri, New
               Jersey, Nevada, New York, Ohio, Pennsylvania,
               Tennessee, Texas, Washington state, Japan.
QUANTITY       1,669 vials were distributed.
REASON         The device is contaminated with Penicillium
               sp. mold.  The device labeling does not meet
               the labeling requirements for in vitro
               diagnostic devices in 21 CFR 809.10.  


RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II ==========
_______________
PRODUCT        Harvey Vapo-Steril Solution, a sterilizing
               compound for use with Harvey Chemiclave
               sterilizers, for sterilizing dental and
               surgical instruments.  Recall #Z-725-6.
CODE           Lot numbers:  2794, 2395, 2595.
MANUFACTURER   MDT Biologics Company, Rancho Dominguez,
               California.
RECALLED BY    Manufacturer, by letter on February 1, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, New Zealand, Taiwan.
QUANTITY       82,606 bottles were distributed.
REASON         The product may not adequately sterilize since
               an active ingredient, formaldehyde, was found
               to be less than declared in some lots, and
               some bottles in one lot contained no
               formaldehyde at all.

_______________
PRODUCT        Transducer Protector in Custom Sterile Trays,
               a component of a sterile custom tray
               identified with the user's name and a unique
               work order number:
               (a) Dualex Transducer Protector;
               (b) Dualex Plus Transducer Protector;
               (c) Dualex Ultra Transducer Protector.
               Recall #Z-761/763-6.
CODE           The above devices are contained in the
               following NMC Item Numbers (Catalog Nos.):
               NMC-97-1030-00,      NMC-97-4030-00,        

                             -9-               NMC-97-0650-00,      NMC-97-7910-00,         
               NMC-97-9370-00,      NMC-97-7870,
               NMC-97-0010-00,      NMC-97-0240-00,
               NMC-97-8030-00,      NMC-97-9806-00,
               NMC-97-0480-00,      NMC-97-0570-00,
               NMC-97-9041-00,      NMC-97-4360-00,
               NMC-97-7380-00,      NMC-97-9820-00,
               NMC-97-0650-01,      NMC-97-0710-00,
               NMC-97-0371-00,      NMC-97-4422-00,
               NMC-97-8235-00,      NMC-0250-00,
               NMC-97-0380-00,      NMC-97-0400-00,
               NMC-97-0520-00,      NMC-97-0580-00,
               NMC-97-4841-00,      NMC-97-4361-00,
               NMC-97-4952-00,      NMC-97-0500-00,
               NMC-97-8200-00,      NMC-97-0720-00,
               NMC-97-0410-00,      NMC-97-7761-00,
               NMC-97-8350-00,      NMC-97-0430-00,
               NMC-97-0390-00,      NMC-97-0460-00,
               NMC-97-0530-00,      NMC-97-0601-00,
               NMC-97-4850-00,      NMC-97-4453-00,
               NMC-97-7831-00,      NMC-97-0500-01,
               NMC-97-9942-00,      NMC-97-0730-00,
               NMC-97-0420-00,      NMC-97-8163-00,
               NMC-97-0270-00,      NMC-97-0420-00,
               NMC-97-8163-00,      NMC-97-0270-00, 
               NMC-97-0440-00,      NMC-97-0260-00,
               NMC-97-0470-00,      NMC-97-0560-00, 
               NMC-97-0360-00,      NMC-97-5000-00.
MANUFACTURER   Millipore Corporation, Bedford, Massachusetts
               (transducers).
RECALLED BY    Sterile Design, Division of Maxxim Medical,
               Clearwater, Florida, by letter dated February
               21, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   New Jersey.
QUANTITY       490,578 custom kits were distributed.
REASON         The transducer protectors may exhibit air
               leakage at the female luer lock connection to
               the male luer of the hemodialysis machine
               under normal operating conditions. 

_______________
PRODUCT        BactiCard Neisseria, a test kit for use in the
               rapid, presumptive identification of
               pathogenic Neisseria species isolated from
               primary culture of clinical specimens on
               selective media.  Recall #Z-779-6.
CODE           Catalog #21-110, Lot #5502, EXP 2/28/96.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone on July 28 & 31,
               1995.  Firm-initiated recall complete.

                            -10-DISTRIBUTION   California, Oklahoma, Michigan, Virginia,
               Georgia, Wisconsin, Vermont, Tennessee,
               Minnesota, South Carolina, Ohio, Missouri,
               Maryland, Florida, New Mexico, Louisiana, New
               York, Illinois, Rhode Island.
QUANTITY       84 units were distributed.
REASON         The tests resulted in a positive GLUT reaction
               with Neisseria gonorrhoeae.  The expected
               reaction should be negative.  As a result
               Neisseria gonorrhoeae isolate would be
               misidentified as a false negative.

_______________
PRODUCT        Orthopedic Bone Screws with Cancellous or
               Cortical Thread Designs: 
               (a) Cortical Locking Screw, All Lengths, Part
               No. 1020-XX;
               (b) Cancellous Locking Screw, All Lengths,
               Part No. 1030-XX;
               (c) 5.5 mm Cortical Screw, All Lengths, Part
               No. 1515-XX;
               (d) 4.8 mm Cannulated Lag Screw, All Lengths,
               Part No. 1717-XX;
               (e) 4.5 mm Cortical Bone Screw, All Lengths,
               Part No. 14022-XX;
               (f) Cannulated Hip Screw, All Lengths, Part
               No. 14088-XX;
               (g) S.C.F.E. Screw, All Lengths, Part No.
               14188-XX;
               (h) 8.0 mm Cannulated Cancellous Screw, 24 mm
               Thread, All Lengths, Part No. 14192-XX;
               (i) 6.5 mm Cannulated Cancellous Lag Screw, 22
               mm Thread, All Lengths, Part No. 14196-XX;
               (j) 6.5 mm Cannulated Cancellous Lag Screw, 40
               mm Thread, All Lengths, Part No. 14197-XX;
               (k) Cannulated Cortical Bone Screw, All
               Lengths, Part No. 14200-XX;
               (l) 5.0 mm Cannulated Cancellous Lag Screw,
               All Lengths, Part No. 14225-XX;
               (m) Cannulated Cortical Lag Screw, All
               Lengths, Part No. 14250-XX;
               (n) 4.0 mm Solid Cancellous Screw, Full
               Thread, All Lengths, Part No. 14369-XX;
               (o) 4.0 mm Cannulated Cancellous Bone Screw,
               16 mm Thread, All Lengths, Part No. 14370-XX;
               (p) 4.0 mm Cannulated Cancellous Bone Screw,
               32 mm Thread, All Lengths, Part No. 14372-XX;
               (q) 3.5 mm Cannulated Cortical Bone Screw, All
               Lengths, Part No. 14375-XX;
               (r) 3.5 mm Cortical Bone Screw, All Lengths,
               Part No. 14377-XX.  Recall #Z-791/808-6.

                            -11-CODE           All lot numbers of the devices listed above
               distributed from September 1, 1994 to April 6,
               1996. 
MANUFACTURER   Ace Medical Company, El Segundo, California.
RECALLED BY    Manufacturer, by notice titled "urgent
               notification" on June 20, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 105,195 bone screws were
               distributed; firm estimates none of the
               misbranded product remains on the market
 REASON        The device labels did not contain the
               following required Warning Statement: 
               "WARNING:  This device  is not approved for
               screw attachment or fixation to the posterior
               elements (pedicles) of the cervical, thoracic,
               or lumbar spine."

_______________
PRODUCT        Stryker Rugged Ambulance Cot Models 6060 and
               6070, for ambulance loading/unloading.
               Recall #Z-809/810-6.
CODE           Serial numbers:  960339013, 960339016,
               960339017, 960339020 through 96339031,
               960339042 through 96339099, 960339300 through
               96339325, and 960339336.
MANUFACTURER   Stryker Medical, Division of Stryker
               Corporation, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by telephone between April 2 and
               15, 1996.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       99 cots.
REASON         The siderail mounting holes on the outer
               support rail were drilled incorrectly.  This
               defect could result in a decrease in the
               fatigue strength of the cot and collapse and
               cause more injury to patient.

_______________
PRODUCT        SeraChem Plus Clinical Chemistry Control Serum
               (Bovine), Unassayed Level II, intended for
               monitoring the precision of assays for
               digoxin, theophylline, phenobarbital,
               phenytoin, gentamicin, and tobramycin.  
               Recall #Z-815-6.
CODE           Catalog #3161-84, Lot #N0543560 EXP 6/96.
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York.
RECALLED BY    Instrumentation Laboratory Company (ILC),
               Lexington, Massachusetts, by telephone on
               September 25-26, 1996.  Firm-initiated recall
               complete.

                            -12-DISTRIBUTION   Pennsylvania, Kentucky, Tennessee, Florida,
               Washington state, Oregon, Montana, Texas,
               California, Idaho, Michigan, Missouri,
               Connecticut, Wisconsin.
QUANTITY       Approximately 1,051 units were distributed.
REASON         The reconstituted control serum had turbidity
               and a drop in calcium, creatine and other
               analytes.

_______________
PRODUCT        Hip Prosthesis Stem:
               (a) Apollo Hip Non Porous Stem, Size 2,
               Catalog Number 7370-00-02, 
               (b) Natural Hip Non Porous CoCr Stem, Size 1,
               Catalog Number 7352-35-001, 
               (c) Natural Hip Porous Collarless Stem, Size 2
               Left, Catalog Number 7354-01-102. 
               Recall #Z-816/818-6. 
CODE           Lot numbers:  (a) 1236826; (b) 1235461 and
               1235464; (c) 1234318.
MANUFACTURER   Intermedics Orthopedics, Inc. (IOI), Austin,
               Texas.
RECALLED BY    Manufacturer, by telephone on April 17, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Missouri, Ohio, Pennsylvania, South Dakota,
               Tennessee.
QUANTITY       19 units were distributed; firm estimates none
               remains on the market.
REASON         The devices were mislabeled.

_______________
PRODUCT        Disposable Steri-Shield Togas, used during
               surgical procedures in order to help protect
               the wearer against contamination and/or
               exposure to body fluids and microorganisms:
               (a) Part No. 400-410 Toga, Pullover, Regular;
               (b) Part No. 400-420 Toga, Pullover, Large;
               (c) Part No. 400-430 Toga, Pullover, Extra
               Large;
               (d) Part No. 400-412 Toga, Pullover, Regular;
               (e) Part No. 400-422 Toga, Pullover, Large;
               (f) Part No. 400-432 Toga, Pullover, Extra
               Large;
               (g) Part No. 400-510 Toga, Zipper, Regular;
               (h) Part No. 400-520 Toga, Zipper, Large;
               (i) Part No. 400-530 Toga, Zipper, Extra
               Large;
               (j) Part No. 400-512 Toga, Zipper, Regular;
               (k) Part No. 400-522 Toga, Zipper, Large;
               (l) Part No. 400-532 Toga, Zipper, Extra
               Large.  Recall #Z-819/830-6.
CODE           All lots bearing a lot number ending in 
               "-1-S".
                            -13-MANUFACTURER   BioMedical Devices, Irvine, California.
RECALLED BY    Stryker Instruments, Division of Stryker
               Corporation, Kalamazoo, Michigan, by letter
               issued on or about April 16, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       79,530 togas were distributed.
REASON         The sterility of the devices have been
               compromised due to loss of package integrity.

_______________
PRODUCT        ECG Pre Amp PC Board Assembly for the ECG
               Option of the Siemens Gamma Camera System. 
               Recall #Z-831-6.
CODE           Assembly #820824270.  All boards shipped on or
               after November 22, 1995.
MANUFACTURER   Solectron Technology, Inc., Charlotte, North
               Carolina (board).
RECALLED BY    Siemens Medical Systems, Inc., Nuclear
               Medicine Group, Hoffman Estates, Illinois, by
               letter dated April 9, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       67 units were distributed.
REASON         The circuit board exceeds the maximum leakage
               current safety requirement for patient
               connected devices.

_______________
PRODUCT        7200 Series Microprocessor Ventilators,
               indicated for delivery of gas to patients
               dependent on artificial respiration:
               (a) Model No. 7200ae shipped domestically;
               (b) Model No. 7200e shipped internationally;
               (c) Model No. 7200sp shipped internationally.
               Recall #Z-832/834-6.
CODE           8,402 ventilators each with its own unique
               nonsequential serial number.
MANUFACTURER   Nellcor Puritan Bennett (NPB), Carlsbad,
California.
RECALLED BY    Manufacturer, by letter October 2, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       8,402 ventilators were distributed.
REASON         Ventilators can enter the SVO State (Stopped
               Ventilating) when they are first turned on due
               to defective software.

_______________
PRODUCT        Power Wheelchairs marketed under the following
               brand names:
               (a) Ranger X Power Wheelchair;

                            -14-               (b) Ranger II Power Wheelchair;
               (c) Arrow Power Wheelchair;
               (d) Arrow XT Power Wheelchair;
               (e) Jaguar Power Wheelchair;
               (f) Power 9000 Action Cat.  
               Recall #Z-835/840-6.
CODE           All codes beginning with 93K through 94C and
               followed by any five digit combination.
MANUFACTURER   Invacare Corporation, Elyria, Ohio
               (wheelchairs); Gould Electronics, Mexico
               (component).
RECALLED BY    Invacare Corporation, Elyria, Ohio, by letter
               dated November 22, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       4,761 chairs were distributed.
REASON         A wire harness connecting the battery to the
               controller contains a fuse holder that may
               cause the wheelchair to lose power due to a
               faulty soldering weld.

_______________
PRODUCT        Laser Systems, used as optical oscillators for
               highly specialized laboratory applications:
               (a) Model KerBBQS-EAFB Laser System;
               (b) Model CYLF-KTP1 Optical Parametric
               Oscillator (OPO).  Recall #Z-850/851-6.
CODE           None.
MANUFACTURER   Schwartz Electro-Optics, Inc., Orlando,
               Florida.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan on May 3, 1996.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Southeast and Southwest region of the United
               States.
QUANTITY       1 unit each.
REASON         The laser systems were found to be in
               noncompliance with 21 CFR 1010.2 and 1010.3,
               1040.10(f),(g), and (h)(1) in that they failed
               to have identification, certification, warning
               logotype, aperture labels, and labels for
               defeatably interlocked protective housings. 
               The products also failed to incorporate beam
               attenuators, emission indicators, and
               defeatable safety interlocks.  There was no
               user information provided with the products.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        Urea Broth for AFB, 1.5 ml tubed media, for
               in-vitro diagnostic use.  Recall #Z-769-6.
CODE           Catalog #06-5222, Lot #5504 EXP 8/31/96.
                            -15-MANUFACTURER   Remel Limited Partnership, Lexena, Kansas.
RECALLED BY    Manufacturer, by telephone on November 7-8,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   California, Virginia, Connecticut, Wisconsin,
               Michigan, Pennsylvania, Arkansas, Kansas, New
               York, Florida, Ohio, Indiana, Puerto Rico.
QUANTITY       540 units were distributed; firm estimates
               none remains on the market.
REASON         The broth was pink in color while being stored
               in refrigeration at 2-8 centigrade compared to
               another lot that was yellow.

_______________
PRODUCT        Phenol Red Broth with 1% Sucrose, tubed media,
               for in-vitro diagnostic use.  Recall #Z-770-6.
CODE           Catalog #06-2432, Lot #5503 EXP 10/31/96.
MANUFACTURER   Remel Limited Partnership, Lexena, Kansas.
RECALLED BY    Manufacturer, by telephone on February 7,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   Ohio, California, Nebraska, Illinois, New
               York, Missouri.
QUANTITY       260 units were distributed; firm estimates
               none remains on the market.
REASON         The product appeared as a semi-solid that was
               light pink in color.  The expected appearance
               of the product is as a broth that is
               orange-red in color.

_______________
PRODUCT        Mueller Hinton Agar with 4% NaCL with 6 mcg/ml
               Oxacillin, plated media, for in-vitro
               diagnostic use.  Recall #Z-773-6.
CODE           Catalog #01-626, Lot #3542 EXP 10/27/94.
MANUFACTURER   Remel Limited Partnership, Lexena, Kansas.
RECALLED BY    Manufacturer, by telephone on October 4-6,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Texas, Ohio, Georgia, Florida, Colorado,
               California, Virginia, Connecticut, Wisconsin,
               Kansas.
QUANTITY       Approximately 2,030 units were distributed;
               firm estimates none remains on the market.
REASON         The device does not support growth, including
               growth of the positive control organism S.
               Aureu.

_______________
PRODUCT        Cystine Heart Agar without Blood, 20 ml tubed
               media, for in-vitro diagnostic use.  
               Recall #Z-774-6.
CODE           Catalog #09-308, Lot #5502 EXP 2/28/96.
MANUFACTURER   Remel Limited Partnership, Lexena, Kansas.
RECALLED BY    Manufacturer, by telephone on November 7,
               1995.  Firm-initiated recall complete.
                            -16-DISTRIBUTION   Kansas, Oklahoma.
QUANTITY       40 units were distributed.
REASON         The device was made with Heart Infusion broth
               instead of Heart Infusion Agar causing it to
               be liquid instead of a solid.

_______________
PRODUCT        Glycerol Buffered Saline, for in-vitro
               diagnostic use in qualitative procedures for
               the preservation of enteric pathogens and the
               transportation of fecal specimens to the
               laboratory:
               (a) Glycerol Buffered Saline, Fecal Transport
               System, 15ml vials, packaged 12 vials per box
               Product/Catalog No. 21-650; 
               (b) Cary Blair w/ Indicator, Glycerol Buffered
               Saline, Fecal Transport System, 15ml vials,
               packaged 6 vials per box.  Product/Catalog No.
               21-611. Recall #Z-775/776-6.
CODE           Lot 4542. 
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone on January 23,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   Nebraska, Indiana, Oklahoma.
QUANTITY       (a) 39 boxes; (b) 2 boxes were distributed.
REASON         The devices contained black particles within
               the saline solution.  The black particles were
               actually fungi contamination.

_______________
PRODUCT        Buffered CYE with CCVC (Charcoal Yeast Extract
               Agar with Cycloheximide, Cephalothin,
               Vancomycin, and Colistin), for in-vitro
               diagnostic use.  Recall #Z-777-6.
CODE           Catalog No. 01-344, Lot #5510 EXP 11/29/95.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone on October 9, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Illinois, New York.
QUANTITY       70 units were distributed.
REASON         The BCYE with CCVC was inadvertently released
               into distribution with poor growth of
               Legionella pheumophila.  

_______________
PRODUCT        SureCaltm 1 Calibrator, for use in the
               calibration of the IL Phoenixtm
               Chemistry/Eletrolyte Analyzer, for in-vitro
               diagnostic use.  Recall #Z-780-6.
CODE           Catalog #33365, Lot #N0553652 EXP 4/96.
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York.
                            -17-RECALLED BY    Instrumentation Laboratory Company (ILC),
               Lexington, Massachusetts, by letter dated
               October 23, 1995, followed by telephone on
               October 24 & 25, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   New York, New Jersey, North Carolina, South
               Carolina, Mississippi, Florida, Minnesota,
               Missouri.
QUANTITY       Approximately 103 cartons (12 5-ml bottles)
               were distributed.
REASON         Chloride channel for the IL Phoenix Analyzer
               would not calibrate, or the Chloride
               recoveries were high for patient values.

_______________
PRODUCT        Cutter Perfourm Light C.D. Vinyl Polysiloxane
               Impression Material Type 1, Low Viscosity Kit,
               intended as a dental impression material for
               professional use by dentists.  
               Recall #Z-787-6.
CODE           Catalog #35044, Kit Lot #041614, Catalyst Lot
               #041563.
MANUFACTURER   Heraeus Kulzer GmbH & Company, Kg, Dormagen,
               Germany.
RECALLED BY    Heraeus Kulzer, Inc., South Bend, Indiana, by
               letter dated April 3, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       1,327 kits were distributed.
REASON         The device fails to set within the specified
               time frame.

_______________
PRODUCT        ReferrIL B Calibrator and IL Test TCO2
               Reagent: (a) ReferrIL B Calibrators, Catalog
               No. 35262,(b) IL Test TCO2 Reagent, Catalog
               No. 182548.  Products used in combination are
               intended for the in-vitro diagnostic
               determination of TC02 Reagent in human serum. 
               Recall #Z-811/812-6.
CODE           Lot numbers:  (a) N1041015, N0153605,
               N1054504; (b) N0554298, N0554299.
MANUFACTURER   (a) Instrumentation Laboratory Company,
               Orangeburg, New York; (b) Medical Analysis
               Systems, Inc., Camarillo, California.
RECALLED BY    Instrumentation Laboratory Co., Lexington,
               Massachusetts, by letter on March 11, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 2,046 units of ReferrIL B and
               502 units of TC02 reagent (Monarch) were
               distributed.

                            -18-REASON         The TCO2 recoveries were lower than labeled on
               the Monarch system using a new liquid form of
               reagents.

_______________
PRODUCT        SeraChem Assayed Control Sera for Monitoring
               the ILab 500, ILab 900, ILab 1800, Monarch,
               and Phoenix Chemistry Systems              
               and the IL 943 Flame Photometer: 
               (a) Level 1, Catalog No. 181624-12,
               (b) Level 2, Catalog No. 181625-12.  
               Recall #Z-813/814-6.
CODE           Lot numbers:  (a) I1230549; (b) I1230550.
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York.
RECALLED BY    Instrumentation Laboratory Company, Lexington,
               Massachusetts, by letter dated November 22,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   New York, New Jersey, Massachusetts, Florida,
               Wisconsin, California, Maryland, Virginia,
               Arkansas, Oklahoma.
QUANTITY       Approximately 21 (12 5-ml vials) packages were
               distributed.
REASON         The label inserts had the incorrect amylase
               and triglyceride values.

_______________
PRODUCT        Dade Protein C Clotting Assay Kit, in-vitro
               diagnostic product used for the quantitative
               determination of the anticoagulant activity of
               protein C in plasma.  Recall #Z-846-6.
CODE           Catalog #B4233-70, Lot #KPC-17M.
MANUFACTURER   Dade International, Inc., Miami, Florida.
RECALLED BY    Manufacturer, by letter dated February 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Texas, Maryland, Wisconsin, Illinois,
               Kentucky, Michigan, Ohio, California,
               Australia.
QUANTITY       82 kits were distributed.
REASON         The kit was packaged using Actin Activated
               Cephaloplastin Reagent, which has been found
               to perform outside its specifications.  This
               APTT reagent generates long coagulation times
               when used with control material and patient
               specimen.

_______________
PRODUCT        Isolab's HemoCard S and A&S Kits, in-vitro
               diagnostic test kits for the detection of
               hemoglobin variants (Hb A & S) or (Hb S).
               (a) Code No. HC-1000; (b) Code No. HC-2030;
               (c) Code No. HC-2000. Recall #Z-847/849-6.

                            -19-CODE           Lot numbers:  (a) 510080, 512110;
               (b) 510084; (c) 510084.
MANUFACTURER   Isolab, Inc., Akron, Ohio.
RECALLED BY    Manufacturer, by letter on April 15, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Puerto Rico, Belgium, Israel,
               Portugal, Canada, England.
QUANTITY       186 kits were distributed.
REASON         The signal for Hb A and Hb S has weakened and
               may produce a negative signal on a positive
               control or patient sample prior to the kits'
               expiration date.  The absence of a strong
               positive signal when running the recommended
               controls could result in a misdiagnosis of any
               test performed with these kits. 

_______________
UPDATE         Recall #Z-728-6, Stratus brand CK-MB
               Fluorometric Enzyme Immunoassay, which
               appeared in the May 22, 1996 Enforcement
               Report should read:
               REASON:  The lot numbers for the calibrator
               components "E" and "F" are incorrect on the
               box label.


MEDICAL DEVICE SAFETY ALERTS: ===============================
_______________
PRODUCT        Fresenius Hemodialysis Machines, 2008 Series:
               (a) A2008 Dialysis System (models 2008C and
               2008D); (b) A1008D Dialysis System (not sold
               in U.S.); (c) A2008 Expansion Model (not sold
               in U.S.); (d) A2008E Dialysis System (model
               2008E); (e) 3008/FDS08 Dialysis System (model
               2008H).  Safety Alert #N-015/019-6.
CODE           All serial numbers.
MANUFACTURER   Fresenius USA, Inc., Walnut Creek, California.
ALERTED BY     Manufacturer, by letter dated December 26,
               1995.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 25,000 units were distributed.
REASON         The heaters have been found to have pit
               corrosion problems, resulting in cracks
               sufficient to cause the ground fault
               interrupter (GFI) to trip. This may be due to
               bleach used to clean the machine when it is
               not always sufficiently rinsed out after use.

                            -20-


END OF ENFORCEMENT REPORT FOR MAY 29, 1996.  BLANK PAGES MAY
FOLLOW.
 
####