 ENFORCEMENT REPORT FOR 04/03/96 


April 3, 1996                                      96-14

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT        Betty Crocker brand (a) Cheddar Cheese Melt
               Hamburger Helper, in 8 ounce boxes; (b)
               Homestyle Salisbury Hamburger Helper, in 7.25
               ounce boxes.  Recall #F-370/371-6.
CODE           Lot numbers:  (a) G529KA14, G529KB14,
               H602KA14, H602KB14, and H603KA14; (b) G529KA3,
               H602KA3, and H602KB3.
MANUFACTURER   Crest Food Company, Inc., Ashton, Illinois
               (hamburger helper); Borden, Inc., St. Louis,
               Missouri (pasta component).
RECALLED BY    General Mills, Inc., Minneapolis, Minnesota,
               by telephone on January 12, 1996, followed by
               letter January on January 13, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       5,000 cases (12 boxes per case) were
               distributed.
REASON         Product contains undeclared eggs.

_______________
PRODUCT        Knirps brand Limburger Cheese, in 6.5 ounce
               tins, soft ripened cheese from Germany. 
               Recall #F-372-6.
CODE           Item #922.  37-3 on shipping carton. 
               Individual foil-wrapped package has no code
               markings.
MANUFACTURER   J. Bauer, KG, Wasserburg, Germany.
RECALLED BY    Swissrose International, Inc., Moonachie, New
               Jersey, by letter on February 29, 1996.  FDA-
               initiated recall ongoing.
DISTRIBUTION   Florida, Indiana, Maine, Massachusetts, New
               Jersey, New York, Ohio, Utah, Virginia.
QUANTITY       200 cases (10 tins per case) were distributed.
REASON         Product is contaminated with Listeria
               monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
_______________
PRODUCT        Regular Lay's Potato Chips, in 6 ounce bags.
               Recall #F-362-6.
CODE           MAR12 4351C27 $149.
MANUFACTURER   Frito-Lay, Inc., Orlando, Florida.
RECALLED BY    Frito-Lay, Inc., Plano, Texas, by computer
               notification on February 13, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION   Florida.
QUANTITY       420 cases (12 bags per case) were distributed;
               firm estimates none remains on the market.
REASON         The product contains dairy ingredients but the
               label fails to disclose this fact.  The
               product bears an inappropriate identity
               statement.

_______________
PRODUCT        Gummi Rabbits, Easter candy in 8 ounce clear
               plastic bags.  Recall #F-365-6.
CODE           None.
MANUFACTURER   Trolli, Inc., Plantation, Florida.
RECALLED BY    Maggie Lyons, Inc., Norcross, Georgia, by
               telephone.  Firm-initiated recall ongoing.
DISTRIBUTION   Northeast and Southeast United States.
QUANTITY       480 pounds were distributed; firm estimates
               none remains on the market.
REASON         Product contained undeclared FD&C Yellow 
               No. 5.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
_______________
PRODUCT        Zinc Lozenges, dietary supplement.  
               Recall #F-361-6.
CODE           7698.
MANUFACTURER   Magno-Humphries, Inc., Tigard, Oregon.

                             -2-RECALLED BY    Manufacturer, by telephone and letter on
               January 10, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Washington state.
QUANTITY       298 bottles were distributed.
REASON         The "plain" zinc lozenges are packaged in
               containers labeled "Cherry flavored).

_______________
PRODUCT        On the Border brand Salsa in 16 and 32 ounce
               jars.  Recall #F-373-6.
CODE           Lot numbers:  031097A, 031197B, 032197B,
               032697A, 032797A/5358, 090197B, 5258/020897A,
               5336 and 5354. 
MANUFACTURER   National Food and Beverage, Inc. (NFBI),    
               Row, Dallas, Texas.
RECALLED BY    Manufacturer, by telephone and letter on
               October 18, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Texas, Arkansas.
QUANTITY       26,496 jars were distributed; firm estimates
               none remains on the market.
REASON         Product is contaminated with yeast.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Spacol Tablets, Rx anti-cholinergic agent in
               bottles of 100, packaged under the Dayton
               Laboratories label.  Recall #D-098-6.
CODE           Lot #70A1282.
MANUFACTURER   Wildflower Pharmacal Corporation, Mineola, New
               York.
RECALLED BY    Propharma, Inc., Miami, Florida, by telephone
               on March 12, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Florida, Puerto Rico.
QUANTITY       Firm estimates none remains on market. 
REASON         Product does not meet content uniformity
               specifications.

_______________
PRODUCT        TMP/SMZ Oral Suspension, Rx synthetic
               antibacterial combination product repackaged
               in the following bottle sizes:  100 ml, 120
               ml, 150 ml, 180 ml, and 200 ml.
               Recall #D-101-6.
CODE           Lot numbers:  RP5779-A, RA4032, RS3848 .
MANUFACTURER   CompuMed, Inc., Harahan, Louisiana (repacker).
RECALLED BY    Repacker, by letter on February 19, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Louisiana.

                             -3-QUANTITY       364 bottles were distributed; firm estimates
               none remains on the market.
REASON         Subpotency of trimethoprim ingredient.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Oxygen, USP, refrigerated liquid in a
               cryogenic cylinder.  Recall #D-102-6.
CODE           Lot #0616596A EXP 3/2001.
MANUFACTURER   National Welders Supply Company, Inc.,
               Knoxville, Tennessee.
RECALLED BY    Manufacturer, by visit on March 23, 1996. 
               FDA-initiated recall complete.
DISTRIBUTION   Tennessee.
QUANTITY       1 cylinder was distributed and retrieved.
REASON         Unit not tested for purity and identity.

_______________
PRODUCT        (a) Urodol Caplets, containing Phenazopyridine
               HCl, USP 200 mg per caplet, used to treat
               urinary tract infections, in bottles of 100
               and 30 caplets, Rx under the Dayton
               Laboratories label; (b) Serenitas Tablets, in
               bottles of 60, gives indications for aiding in
               sleep, OTC under the Propharma label.  
               Recall #D-099/100-6.
CODE           Lot numbers:  (a) 70E1305; (b) 70C1299.
MANUFACTURER   Wildflower Pharmacal Corporation, Mineola, New
               York.
RECALLED BY    Propharma, Inc., Miami, Florida, by telephone
               on March 12, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Florida and Puerto Rico.
QUANTITY       Firm estimates none remains on the market.
REASON         Urodol product do not meet assay and content
               uniformity specifications.  Serenitas product
               do not meet content uniformity specifications.

_______________
PRODUCT        Docusate Sodium 100 mg Soft Gelatin Capsules,
               non-prescription single ingredient stool
               softener under the Schein label.  
               Recall #D-103-6.
CODE           Lot #s: 67126 and 67127 labeled for Schein
               (only 67127 was distributed) EXP 3/98.
MANUFACTURER   Banner Pharmacaps, Inc., Elizabeth, New
               Jersey.
RECALLED BY    Manufacturer, by telephone on November 12,
               1995, followed by letters on December 1, 1995,
               and January 16, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   New York, Arizona.
                             -4-QUANTITY       12,492 bottles were distributed.
REASON         Discrepancy in label and carton text in
               directions section.

_______________
PRODUCT        Bactocill (oxacillin sodium for injection) 1
               gram in 20ml vial, Rx antibiotic. 
               Recall #D-104-6. 
CODE           Lot #94507DA EXP 11/1/96.
MANUFACTURER   SmithKline Beecham Pharmaceuticals,     
               Piscataway, New Jersey. 
RECALLED BY    Manufacturer, by letter sent December 18,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       15,380 vials were distributed.
REASON         Marginally subpotent.

_______________
PRODUCT        North O/H Pak Alcohol Wipes, wound cleanser,
               antiseptic containing 70% Isopropyl Alcohol,
               packaged for North Health Care, 100 foil
               packed wipes per carton.  Recall #D-105-6.
CODE           Lot #5F3880, item #03-25-25.
MANUFACTURER   Professional Disposables, Inc., Orangeburg,
               New York.
RECALLED BY    North Health Care, Rockford, Illinois
               (repacker), by telephone on August 24, 1995. 
               Firm-initiated field correction complete.
DISTRIBUTION   Hawaii, Michigan, Pennsylvania, Texas,
               Illinois, Louisiana.
QUANTITY       57 cartons were distributed.
REASON         Some intermediate packages were mislabeled as
               North O/H Cold-A-Rest Nasal Decongestant, Pain
               Reliever Tablets.

_______________
PRODUCT        Amphojel Suspension, aluminum hydroxide gel,
               without flavor, OTC antacid, in 12 fluid ounce
               bottles.  Recall #D-106-6.
CODE           Lot #3941126 EXP 12/96.
MANUFACTURER   Wyeth-Ayerst Laboratories, Rouses Point, New
               York.
RECALLED BY    Whitehall Robbins, Richmond, Virginia, by
               letter dated March 22, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       31,668 units were distributed.
REASON         Product does not meet potency specifications.

                             -5-_______________
UPDATE         Cyproheptadine HCl 2mg/5ml Syrup, Recall #D-
               096-6, manufactured by Halsey Drug Company,
               Brooklyn, New York,  which appeared in the
               March 27, 1996 Enforcement Report has been
               extended to include the following lot numbers: 
               4K10 (9/96), 4K19C (10/96), 4K19D (10/96), 
               4K28F (11/96), 4K28G (12/96), 5A12C (12/96),
               5B16B (3/97), 5B16C (5/97), 5F08B (6/97),
               5H08C (8/97), 5J07C (9/97), 5J07D (9/97) and
               5L06 (10/97).


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
_______________
PRODUCT        (a) Platelets. Recall #B-327-6.
CODE           Unit #39KX10113.
MANUFACTURER   American Red Cross Blood Services, Huntington,
               West Virginia.
RECALLED BY    Manufacturer, by letter dated September 6,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   West Virginia.
QUANTITY       1 unit.
REASON         Blood products, corresponding to a unit of Red
               Blood Cells from which micrococcus sp. was
               cultured, were distributed.

_______________
PRODUCT        Platelets. Recall #B-329-6.
CODE           Unit #39KX10791.
MANUFACTURER   American Red Cross Blood Services, Huntington,
               West Virginia.
RECALLED BY    Manufacturer, by letter dated October 14,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   West Virginia.
QUANTITY       1 unit.
REASON         Blood products, corresponding to a unit of Red
               Blood cells contaminated with Staphylococcus
               aureus, were distributed.

_______________
PRODUCT        (a) Whole Blood; (b) Red Blood Cells; 
               (c) Platelets; (d) Platelets, Pheresis; (e)
               Fresh Frozen Plasma.  Recall #B-336/340-6.
CODE           Contact FDA, Center for Biologics Evaluation
               and Research, Office of Compliance (301) 594-
               1070 for individual unit numbers recalled.
MANUFACTURER   American Red Cross Blood Services, Lansing,
               Michigan.
RECALLED BY    Manufacturer, by letter on May 10, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Michigan, California, Puerto Rico.

                             -6-QUANTITY       (a) 7 units; (b) 613 units; (c) 406 units;
               (d) 14 units; (e) 16 units.
REASON         Blood products, which were incorrectly tested
               for the antibody to the hepatitis C virus
               encoded antigen (anti-HCV), were distributed.

_______________
PRODUCT        (a) Red Blood Cells: (b) Platelets.
               Recall #B-344/345-6.
CODE           Unit #12415-0565.
MANUFACTURER   Blood Systems, Inc., doing business as United
               Blood Services, Albuquerque, New Mexico.
RECALLED BY    Manufacturer, by telephone on January 23,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   New Mexico.
QUANTITY       1 unit of each component.
REASON         Blood products, collected from a donor who
               received a tattoo within twelve months of
               donation.

_______________
PRODUCT        Autologous Red Blood Cells. Recall #B-346-6.
CODE           Unit #N29989.
MANUFACTURER   Mississippi Valley Regional Blood Center,
               Davenport, Iowa.
RECALLED BY    Manufacturer, by telephone on July 13, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Iowa.
QUANTITY       1 unit. 
REASON         Autologous blood product, not correctly
               labeled "For Autologous Use Only," was
               distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-351-6.
CODE           Unit numbers:  36242-8173-1, 36242-8173-2,
               36243-4376, 36244-2882.
MANUFACTURER   Blood Systems, Inc., doing business as United
               Blood Services, Ventura, California.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               letter dated January 23, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   California, Illinois, Missouri. 
QUANTITY       4 units.
REASON         Blood products, collected from a donor whose
               spouse tested positive for the hepatitis B
               surface antigen (HBsAg), were distributed.

_______________
PRODUCT        Platelets.  Recall #B-357-6.
CODE           Unit #11313 1029.
MANUFACTURER   Blood Systems, Inc., doing business as United
               Blood Services, El Paso, Texas.
                             -7-RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               letter dated February 21, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood product, which was collected from a
               donor who exceeded the donation collection
               time, was distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT        Recovered Plasma.  Recall #B-328-6.
CODE           Unit #39KX10113.
MANUFACTURER   American Red Cross Blood Services, Huntington,
               West Virginia.
RECALLED BY    Manufacturer, by letter on August 31, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood products, corresponding to a unit of Red
               Blood Cells from which micrococcus sp. was
               cultured, were distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-330-6.
CODE           Unit #39KX10791.
MANUFACTURER   American Red Cross Blood Services, Huntington,
               West Virginia.
RECALLED BY    Manufacturer, by letter on October 14, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood products, corresponding to a unit of Red
               Blood Cells contaminated with Staphylococcus
               aureus, were distributed.

_______________
PRODUCT        (a) Red Blood Cells, for further manufacture;
               (b) Platelets, for further manufacture; (c)
               Recovered Plasma.  Recall #B-341/343-6.
CODE           Contact FDA, Center for Biologics Evaluation
               and Research, Office of Compliance (301) 594-
               1070 for individual unit numbers recalled.
MANUFACTURER   American Red Cross Blood Services, Lansing,
               Michigan.
RECALLED BY    Manufacturer, by letter on May 10, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Michigan, California, Puerto Rico.
QUANTITY       (a) 1 unit; (b) 78 units; (c) 631 units.

                             -8-REASON         Blood products, which were incorrectly tested
               for the antibody to the hepatitis C virus
               encoded antigen (anti-HCV), were distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-347-6.
CODE           Unit #N29989.
MANUFACTURER   Mississippi Valley Regional Blood Center,
               Davenport, Iowa.
RECALLED BY    Manufacturer, by telephone on July 14, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit. 
REASON         Autologous blood product, not correctly
               labeled "For Autologous Use Only," was
               distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c)
               Recovered Plasma.  Recall #B-348/350-6.
CODE           Unit numbers:  (a) 10213-5598           
               10201-6825       10193-6705      10231-4988
               01198-6632       10206-5924      10210-6553
               10210-1919       10241-4203      10229-0618
               10152-6999       10493-5336
               (b) 10206-5924   10213-5598      10231-4988
               (c) 10198-6632   10210-1919      10201-6825
               10152-6999       10493-5336      10241-4203
               10229-0618       10220-6553      10213-5598
               10206-5924       10193-6705      10231-4988.
MANUFACTURER   Blood Systems, Inc., doing business as United
               Blood Services, Scottsdale, Arizona.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               letter dated December 28, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Florida, Massachusetts, Virginia, Texas, North
               Carolina, Arizona, Switzerland.
QUANTITY       (a) 12 units; (b) 3 units; (c) 12 units.
REASON         Blood products, which tested negative for the
               hepatitis B surface antigen (HBsAg), and the
               antibody to the hepatitis B core antigen
               (anti-HBc), but were collected from a donor
               who previously tested positive for HBsAg, were
               distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        Latex Free Outlet Hoses for use with Ohmeda
               Anesthesia Systems:
               (a) Latex Free Outlet Hoses for Part No.
               0236-0042-800;

                             -9-               (b) Latex Free Outlet Hoses for Part No.
               0236-0042-801;
               (c) Latex Free Outlet Hoses for Part No.
               1010-8057-000;
               (d) Latex Free Outlet Hoses for Part No.
               1010-7064-000.  Recall Z-538/541-6.
CODE           Hoses manufactured since September 22, 1995.
MANUFACTURER   Ohmeda Medical Systems Division, Madison,
               Wisconsin.
RECALLED BY    Manufacturer, by letter dated March 1, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,661 hoses were distributed.
REASON         The inner wall of this hose may be cut during
               insertion of the hose connectors during
               manufacturing which results in restricted gas
               flow during the oxygen flush.

_______________
PRODUCT        Biad 89 and Triad 88 Whole body Spect Nuclear
               Imaging Systems, used by technologists for
               diagnostic and research purposes.  
               Recall #Z-546/547-6.
CODE           Biad 89 with serial numbers  102 - 190;
               Triad 88 with serial numbers 101 - 197.
MANUFACTURER   Trionix Research Laboratory, Inc., Twinsburg,
               Ohio.
RECALLED BY    Manufacturer, by letter May 6, 1995.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       133 units were distributed.
REASON         Due to a gantry code software error in
               calculating the collimator thickness, the
               device's detector head may move closer to the
               patient than necessary and may even touch the
               patient, which could result in potential
               injury to patients.

_______________
PRODUCT        One-Step Armboard Surgical Table Accessory,
               Part #426637-542, Model numbers F-ABSA-A1, F-
               ABSA-A2, F-ABSA-A3, designed for use on the
               AMSCO 1080, AMSCO 2080 and AMSCO QUANTUM 3080
               Surgical Tables.  Recall #Z-551-6.
CODE           Serial numbers 29724 through 40852.
MANUFACTURER   AMATECH Corporation, Acton, Massachusetts
               (armboard); AMSCO, American Sterilizer
               Company, Montgomery, Alabama (tables).
RECALLED BY    AMSCO, American Sterilizer Company, Erie,
               Pennsylvania, by memorandum issued on December
               18, 1995, January 20, 1996, and February 21,
               1996.  Firm-initiated recall ongoing.

                            -10-DISTRIBUTION   Nationwide and international.
QUANTITY       4,498 units were distributed.
REASON         The armboard may detach from the surgical
               table, and possibly injure the patient.

_______________
PRODUCT        Midmark Model 7100 General Surgical Tables. 
               Recall #Z-552-6.
CODE           Serial Numbers:  TDW-1081 through TDW-1092,
               TDW-1094 through TDW-1115.
MANUFACTURER   M. Schaerer AG, Moosseedorf, Switzerland.
RECALLED BY    Midmark Corporation, Versailles, Ohio, by
               telephone on September 27, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Virginia.
QUANTITY       2 units were distributed.
REASON         There is a potential for the surgical table
               become unstable due to faulty weld of the
               table column.

_______________
PRODUCT        SYNERGYTM Anterior Spinal System Implants:
               (a) Part No. 4105 Implant Sterilization Case
               and SYNERGYTM Anterior Spinal System Implants
               including hex nuts, integral screws, closed
               screws, rods, anterior washers, and crosslink
               assemblies
               (b) Part No. 4106 Instrument Sterilization
               Case # 1 
               (c) Part No. 4107 Instrument Sterilization
               Case # 2.  Recall #Z-553/555-6.
CODE           All code.
MANUFACTURER   Cross Medical Products, Inc., Columbus, Ohio.
RECALLED BY    Manufacturer, by letter dated March 10, 1995. 
               Firm-initiated field correction complete.
DISTRIBUTION   Kentucky, Utah, Alabama, Georgia, Indiana,
               Minnesota, North Carolina, Florida, Italy,
               Japan, Turkey, Korea, Singapore, Belgium.
QUANTITY       5,726 implantable devices were distributed.
REASON         The package insert, Part No. EXP003-1,
               specifies an inadequate sterilization time
               that does not ensure that the required
               sterility assurance level is achieved.

_______________
PRODUCT        First-Medic 710 Semi-Automatic Defibrillator. 
               Recall #Z-556-6.
CODE           Serial Numbers  560-00032 to 560-000857 (non-
               continuous).
MANUFACTURER   Physio Control Corporation, Redmond,
               Washington.

                            -11-RECALLED BY    Manufacturer, by visit beginning on February
               2, 1996 and hand delivered letter.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       109 units were distributed.
REASON         The units may inadvertently switch from the
               semi-automatic mode to the manual mode during
               operation of the device, due to defective
               software.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        IL Testtm PT-Fibrinogen for the ACL System,
               used for the simultaneous determination of
               Prothrombin Time and Fibrinogenon the ACl (tm)
               System.  Recall #Z-532-6.
CODE           Catalog #97567-10, lot numbers NO342966 and
               N0443403.
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York.
RECALLED BY    Instrumentation Laboratory Company (ILC),
               Lexington, Massachusetts, by letter dated
               September 15, 1994 and by telephone December
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Michigan, Missouri, Florida, Arkansas,
               Louisiana, Vermont, Arizona, Iowa, Texas,
               Virginia, New York, Oregon, California.
QUANTITY       4200 kits were distributed.
REASON         The recovery values were not in the expected
               range.

_______________
PRODUCT        ReferrILtm 3/C Calibrator for the IL Genesis
               Analyzer, and the IL Monarch and Multistat
               Systems, for in-vitro diagnostic use.
               Recall #Z-533-6.
CODE           Catalog #35222-12, lot #N1132194.
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York.
RECALLED BY    Instrumentation Laboratory Company (ILC),
               Lexington, Massachusetts, by letter dated May
               15, 1995, followed by telephone during mid-
               December 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,745 packages were distributed.
REASON         The T. Bilirubin recoveries were high.

                            -12-_______________
PRODUCT        IL Testtm ISE Reference and Buffer Solution
               for the Monarch and Phoenix Chemistry Systems,
               for in-vitro diagnostic use: (a) ISE Reference
               Catalog numbers: 3337204; (b) ISE Buffer
               Catalog #3337404.  Recall #Z-534/535-6.
CODE           Lot Numbers: (a) N0642924, N0732923, N0532065,
               N0432064; (b) N1142995, N0142072, N1232071,
               N1133554, N0732070, N0632089, N0432068,
               N1121692.
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York.
RECALLED BY    Instrumentation Laboratory Company (ILC),
               Lexington, Massachusetts, by letters dated
               April 11, 1995 and September 29, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide, Puerto Rico, Antigua, British West
               Indies.
QUANTITY       Lot  No.        Quantity Released
               (a) N0842824      3000
               N0732923          2541
               N0532085          2549
               N0432084          2547
               (b) N0142072      5182
               N1232071          5088
               N1133554          5162
               N0732070          3078
               N0632089          3839
               N0432068          2048
               N1121692          811.
REASON         The bacteriostat/fungistat potency may
               decrease after one year which is short of the
               expiration date.
_______________
PRODUCT        (a) IL Testtm APTT (Activated Partial
               Thromboplastin Time), for in-vitro diagnostic
               use; (b) IL Testtm Pt-Fibrinogen reagents for
               the ACO System, for in-vitro diagnostic use. 
               Recall #Z-536/537-6.
CODE           (a) Catalog No. 200060, 
               Kit Lot Nos.            APPT Reagent Lot Nos.
               N0242975 (5/95)         N0242974 (5/95)
               N0543374 (11/96)        N0543369 (11/96)
               N0742946 (12/95)        N0543340 (12/95)
               N0742947 (1/96)         N0743371 (1/96)
               N0742948 (1/96)         N0743244 (1/96)
               N1043312 (3/96)         N1043245 (3/96
               (b) Catalog No. 97567-10, Lot Nos. N0543405
               (5/97), N01143411 (11/97), N01043409 (10/97),
               N0743406 (7/97).
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York.
                            -13-RECALLED BY    Instrumentation Laboratory Company (ILC),
               Lexington, Massachusetts, by letter dated
               January 25, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       (a) Kit Lot #N0242975 = 817 kits
               Kit Lot #N0543374 = 4643 
               Kit Lot #N0742946 = 5045  
               Kit Lot #N0742947 = 5374
               Kit Lot #N0742948 = 7851
               Kit Lot #N1043312 = 8712
               (b) PT-FIB: Lot #N0543405 = 3864
               Lot #N1143411 = 3831
               Lot #N1043409 = 3733
               Lot #N0743406 = 3791.
REASON         The normal range values may have an upward
               shift of 0.5 to 1.0 seconds for the IL TEST
               PT-FIB and 2 to 4 seconds for the APTT. The
               APTT may also have an increased sensitivity to
               lupus-like inhibitors.

_______________
PRODUCT        Victory series Orthodontic 2nd Bicuspid
               Brackets:
               (a) Miniature Metal Wing, Part# 017-433;
               (b) Mini Mesh Alexander Rx, Part# 2017-433;
               (c) APC Miniature Metal Single, Part# 017-460.
               Recall #Z-557/559-6.
CODE           None.  Recall limited to specific shipments of
               product to specific customers who were sent
               the recall notification letter.
MANUFACTURER   3M Unitek, 3M Dental Products Division,
               Monrovia, California.
RECALLED BY    Manufacturer, by letter dated November 17,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Colorado, Indiana, Missouri, Ohio,
               Oklahoma, South Dakota, Tennessee, Texas,
               Utah, and international.
QUANTITY       3,162 brackets subject to recall.
REASON         The packaging labels were labeled with a -17°
               bracket torque when the packages actually
               contained products with a -11° bracket torque.

_______________
PRODUCT        Bacto Streptolysin O Reagent, desiccated
               streptolysin O, in vials labeled 10 ml, an in-
               vitro diagnostic used for determining the
               amount of antistreptolysin O titers in sera. 
               Recall #Z-563-6.
CODE           Lot #69762JA.
MANUFACTURER   Difco Laboratories, Livonia, Michigan.
RECALLED BY    Manufacturer, by letter sent on November 20,
               1995.  Firm-initiated recall complete.
                            -14-DISTRIBUTION   California, Delaware, Illinois, Kansas,
               Kentucky, Mississippi, Missouri, Ohio, Puerto
               Rico, Texas, Canada, England, Finland
               Honduras, Turkey.
QUANTITY       135 packages were distributed.
REASON         The 25 ml vials containing reagent were
               incorrectly labeled to rehydrate with 10 ml of
               distilled water.

_______________
PRODUCT        Dade Actin Activated Cephaloplastin Reagent,
               an in-vitro diagnostic, under the following
               private labels:  Baxter Healthcare, Baxter
               Diagnostics, Baxter S.A., Baxter S.p.A.,
               Baxter NV/S, Baxter S.A. BP 56, Baxter
               Deutschland GmbH, Baxter Healthcare Ltd,
               Baxter Medical AB, Baxter Dade A.G.
               Recall #Z-566-6.
CODE           Lot numbers:  APAC-611 A, B, C, D, E and F;
               APAC-614 A, B, C, D and E; APAC-618 A, B, C,
               and D.
MANUFACTURER   Dade International of Puerto Rico, Aguada,
               Puerto Rico.
RECALLED BY    Manufacturer, by letter dated December 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       52,217 packages were distributed.
REASON         The Activated Cephaloplastin Reagent gave a
               prolonged activated partial thromboplastin
               times with normal and abnormal controls and
               normal patient samples. 

INFORMATION: ================================================

AGAINST        Roger Ferris and Anthony Ferris, Utica, New
               York.
CHARGED        August 17, 1995 - One-Count Criminal
               Misdemeanor Information; Cr. #95-CR-369(DNH),
               U.S. District Court for the Northern District
               of New York.  On or about August 16, 1995, in
               the Northern District of New York, the
               defendants did forcibly assault, resist,
               oppose, impede, intimidate, and interfere with
               a government investigator while engaged in or
               on account of the performance of official
               duties, namely as an Investigator of the U.S.
               Food and Drug Administration conducting a
               routine sanitary inspection of R. Ferris &
               Sons, Inc., a wholesale produce distributor,
               in Utica, New York.  October 17, 1995 - Plea
               Agreement Under Rule 11 of the  Federal Rules
               of Criminal Procedure; Cr. #95-CR-369(DNH); 

                            -15-               U.S. District Court for the Northern District
               of New York.  The defendants each plead guilty
               to a One-Count Misdemeanor Information, as
               charged.
SENTENCE       The defendants were fined $1,500 each and
               sentenced to serve 100 hours of community
               service, and placed on probation for one year.

                            -16-

              END OF ENFORCEMENT REPORT FOR APRIL 3, 1996.  