The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
01/24/1996

 ENFORCEMENT REPORT FOR 01/24/95 


January 24, 1996                                       96-04

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT        Ground Round Restaurant Apple Crisp, packaged
               in 2 pound aluminum tray with cardboard cover. 
               Recall #F-146-6.
CODE           5421 and 5383 ink jet sprayed on master
               cartons only.
MANUFACTURER   Basic American Frozen Foods, Brockton,
               Massachusetts.
RECALLED BY    Manufacturer, by electronic mail on October
               25, 1995.  Firm-initiated field correction
               (relabeling) complete.
DISTRIBUTION   Northeast states to Midwestern states.
QUANTITY       1,084 cases were distributed.
REASON         Product contains undeclared walnuts.

_______________
PRODUCT        Eden's Whole Pitted Sun Dried Malatya
               Apricots, packaged in 16 ounce clear poly
               bags.  Recall #F-147-6.
CODE           None.
MANUFACTURER   Elmas a.s., Istanbul, Turkey.
RECALLED BY    Eden International, Inc., Blue Island,
               Illinois, by letter dated November 30, 1995. 
               Firm-initiated field correction (relabeling)
               ongoing.
DISTRIBUTION   Ohio, Illinois, Florida, Michigan, Colombia.
QUANTITY       Approximately 56-84 one-pound bags were
               distributed.
REASON         Product contains undeclared sulfites.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
_______________
PRODUCT        Holiday Assorted Butter Cookies, in 3 pound 8
               ounce metal tins with a picture of "Santa's
               Workshop" on the side and a red lid.  
               Recall #F-145-6.
CODE           UPC #70453-11123.  All tins with red lid and
               "Santa's Workshop" depicted on the side of the
               tin.
MANUFACTURER   Maurice Lennell Cooky Company, Chicago,
               Illinois.
RECALLED BY    Manufacturer, by telephone on December 14,
               1995, followed by visit.  Firm-initiated field
               correction (relabeling) ongoing.
DISTRIBUTION   Northeastern United States.
QUANTITY       600 cases (10 tins per case) were distributed;
               firm estimated that 150 tins remained on
               market at time of recall initiation.
REASON         Product contains undeclared food colors (FD&C
               Yellow #5, #6, and #40, and Blue #1).


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
_______________
PRODUCT        Waldbaum's Peppermint Starlights Candy, in 10
               ounce plastic bags.  Recall #F-144-6.
CODE           None.
MANUFACTURER   Cendee, Plainview, New York.
RECALLED BY    Waldbaum's, Inc., Central Islip, New York, by
               memorandum December 18, 1995.  FDA-initiated
               recall ongoing.
DISTRIBUTION   New York.
QUANTITY       Undetermined.
REASON         Product contains undeclared FD&C Red #40 and
               falsely represents the product as containing
               FD&C Yellow #5 and #6 when in fact it does
               not.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Myochrysine (Gold Sodium Thiomalate) Injection
               50 mg/ml, in 10 ml multi-dose vials, used for
               rheumatoid arthritis, under the Merck and King
               Pharmaceuticals label.  Recall #D-077-6.

                             -2-CODE           Lot numbers:  0481T EXP 1/96, 1049T EXP 4/96,
               0250V EXP 11/96, 1373V EXP 2/97, 0775W EXP
               10/97, 0367A EXP 6/98, 0697B EXP 8/99, 1013A
               EXP 2/99 (King label), 0482T EXP 1/96, 1085T
               EXP 5/96, 0249V EXP 11/96, 0113W EXP 8/97,
               1665W EXP 9/98, 0912A EXP 5/99, 0820B EXP
               9/99, 1544A EXP 6/99 (King label), 0738T EXP
               2/96, 1780T EXP 3/96, 0778V EXP 11/96, 0594W
               EXP 8/97, 0133A EXP 11/98, 0017B EXP 2/99,
               0249B EXP 8/99.
MANUFACTURER   Merck and Company, Inc., West Point,
               Pennsylvania.
RECALLED BY    Manufacturer, by letter dated December 18,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Spain, Peru.
QUANTITY       Approximately 90,000 unit packages remained on
               market at time of recall initiation.
REASON         Product does not meet antimicrobial
               effectiveness test.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Sodium Iodide I-131 Solution for Oral
               Therapeutic Use, in 100 mCi and 50 mCi vials,
               used for the treatment of hyperthyroidism and
               selected cases of carcinoma of the thyroid. 
               Recall #D-078-6.
CODE           Lot #K049 EXP 1/17/96.
MANUFACTURER   CIS Bio International, Subsidiary of Compagnie
               ORIS Industrie SA, Cedex, France.
RECALLED BY    CIS-US, Inc., subsidiary of CIS Bio
               International, Bedford, Massachusetts, by
               letter dated December 15, 1995.  Firm-
               initiated field correction complete.
DISTRIBUTION   Nationwide.
QUANTITY       71 100-mCi vials and 55 50-mCi vials were
               distributed.
REASON         Some packages contained a package insert for
               Sodium Iodide I-131 Capsules.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        I.V. Extension Sets with "T" Connectors:
               (a) Abbott Extension Set with "T" Connector,
               Syringe Administration Set, List 4612;
               (b) Abbott LifeShield Extension Set with "T"
               Connector and Pre-pierced Reseal Injection
               Site, List 11572;

                             -3-               (c) Abbott LifeShield (LAV) Microbore
               Extension Set with Luer Activated Valve and
               "T" Connector, List 11749, 
               (d) Abbott Extension Set with "T" Connector,
               List Rx 17084;
               (e) Abbott LifeShield Extension Set 6 Inch
               with Pre-pierced Reseal "T" Connector and Luer
               Activated Valve, List Rx 11869;
               Recall #Z-321/325-6.
CODE           Lot numbers: (a)  02-169-HG, 03-213-HG,
               03-214-HG, 03-215-HG, 03-216-HG, 03-128-HG,
               03-193-HG, 04-007-HG, 04-056-HG, 04-164-HG,
               04-165-HG, 04-166-HG, 04-167-HG, 05-188-HG,
               05-189-HG, 05-190-HG, 05-202-HG, 05-300-HG,
               05-301-HG, 06-140-HG, 06-189-HG;
               (b) 01-175-HG, 02-058-HG, 02-126-HG,
               03-006-HG, 03-122-HG, 04-026-HG, 05-007-HG,
               05-227-HG, 05-191-HG, 05-293-HG, 05-294-HG,
               06-086-HG, 06-134-HG, 07-013-HG;
               (c) 04-195-HG; 
               (d) 04-193-HG, 80-235-H1, 91-078-H1;
               (e) 03-194-HG, 04-188-HG, 05-249-HG,
               06-084-HG, 08-027-HG, 10-233-HG, 93-279-HG.
MANUFACTURER   Abbott Laboratories, Laurinburg, North
               Carolina (HG lot suffix); Abbott Health
               Products, Barceloneta, Puerto Rico (H1 lot
               suffix).
RECALLED BY    Abbott Laboratories, Abbott Park, Illinois, by
               letters dated December 8, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       2,120,180 units were distributed; firm
               estimated that 135,000 - 150,000 units
               remained on market at time of recall
               initiation.
REASON         Some of the extension sets leak and disconnect
               between the male fitment of the device and the
               luer fitment of the I.V. access device.

_______________
PRODUCT        System 97, a helium charged, mobile Intra-
               Aortic Balloon Pump (IABP), Catalog #0998-00-
               0104-X, for use as a patient aid during pre-,
               intra-, or post- operative open heart surgery;
               use within patients demonstrating unstable
               angina; use within patients with left main
               artery occlusion or poor left ventricle
               function:  (a) Power Supply; (b) Front End
               Board; (c) Solenoid Driver Board.  
               Recall #Z-332/334-6.

                             -4-CODE           Serial numbers:  (a) 1020, 1045, 1104, 1123,
               1125, 1126, 1128, 1129, 1134-1143, 1145-1147,
               1149, 1152, 1154, 1155, 1157, 1158, 1160-1173,
               1175, 1176, 1178, 1181-1190, 1193-1203, 1205,
               1206, 1228, 1230-1254, 1256-1292, 1294-1314,
               1316-1371, 1373, 1374, 1380-1384, 1386, 1392,
               1411, 1442, 1444, 1454, 1456, 1461, 1463,
               1509, 1550, 1571, 2008, 2047, 2048, 1389,
               1392, 1400, 1441, 1447, P0011, P0013, P0014,
               P0016-P0018, P0022, P0023;
               (b) P0001-P0024 and 1001-1608 with the
               following exceptions:  P0003-P0006, P0012,
               1005, 1007, 1080, 1100, 1129, 1159, 1293,
               1297, 1347, 1443, 1453, 1459, 1474, 1476,
               1479, 1496, 1499-1505, 1507-1534, 1527,
               1530-1539, 1545-1555;
               (c) 1551, 1553-1556, 1559-1586, 1588-1623,
               2000-2067, 2069, 2070, 2072-2077, 2080, 2084,
               2088-2092, 2126, 2139.
MANUFACTURER   Datascope Corporation, Paramus, New Jersey.
RECALLED BY    Manufacturer, by letter beginning August 31,
               1995, followed by visit.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 259 units; (b) 554 units; (c) 153 units
               were distributed;
REASON         The system may fail because of the following
               component failures:
               Power Supply:  Failure of a zener diode,
               designated at VR9 in the system power supply,
               which can result in a "Power On" condition
               which cannot be switched off by normal means.
               Front End Board:  Intermittent contact in the
               solder joints of certain surface mount trim
               resistors, may lead to such problems as loss
               of blood pressure wave form, inability to zero
               the blood pressure channel, loss or inaccuracy
               of the shuttle transducer gain, no ECG wave
               form, no pacer detection, inaccuracy of the
               external pressure output or an analog to
               digital (A/D) converter failure.
               Solenoid Driver Board:  A solder flux left
               beneath the integrated circuit, created the
               possibility of a conductive path which can
               result in a short circuit of the board.
 
_______________
PRODUCT        3M Sarns brand Perfusion System 9000 Gas Flow
               System, used to provide and monitor CO2 flow
               to the patient during cardiopulmonary bypass.
               Recall #Z-335-6.

                             -5-CODE           All serial numbers.
MANUFACTURER   Sarns 3M Health Care, Ann Arbor, Michigan.
RECALLED BY    Manufacturer, by letter sent on or about July
               24, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 500 units.
REASON         The relay controlling the CO2 flow can
               experience random failure and latch in the
               "on" position, causing CO2 gas to continue to
               flow after completion of the pre-bypass CO2
               flush of the perfusion circuit.

_______________
PRODUCT        Duopulse and Unipulse Dental Laser System,
               used in dental surgery.  Recall #Z-336/337-6.
CODE           None.
MANUFACTURER   Excel/Quantronix Corporation, Hauppauge, New
               York.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan December 28, 1995. 
               Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide, Germany, Korea.
QUANTITY       213 units were distributed.
REASON         Noncompliance with performance standards for
               laser products in that the operator's manuals
               lacked adequate calibration procedures and the
               devices had several labeling noncompliances.

_______________
PRODUCT        Tempo/Tingle Timers Installed with Diagnostic
               X-Ray Systems, used to regulate the duration,
               and to some extent, the amount of exposure to
               x-radiation.  Recall #Z-346-6.
CODE           None.
MANUFACTURER   Tingle X-Ray Products, Inc., Vance, Alabama.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan January 3, 1996.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       17 timers.
REASON         Devices were not tested for accuracy in
               accordance with the specifications, and did
               not bear proper certification and
               identification labels, as required by 21 CFR
               1020.30(d).

_______________
PRODUCT        ImmunoCard Helicobacter pylori Test Kit,
               Catalog #710030, a rapid enzyme immunoassay
               for the detection of IgG antibodies to
               Helicobacter pylori in human serum and plasma.
               Recall #Z-347-6.

                             -6-CODE           Lot numbers and EXP dates:  Lot Nos.
               710030.007 9/26/96 and 710030.008 11/29/96 (on
               boxes);
               7851.007 10/10/96, 7581.008 12/13/96
               (conjugate enzymes).
MANUFACTURER   Meridian Diagnostics, Inc., Cincinnati, Ohio.
RECALLED BY    Manufacturer, by telephone on November 14,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Italy, Russia, Puerto Rico.
QUANTITY       175 test kits of lot #710030.007 and 57 test
               kits of lot #710030.008 were distributed.
REASON         The enzyme conjugate in the kit was prepared
               at the wrong dilution, resulting in an
               improper sensitivity of the tests.

_______________
PRODUCT        Captia Rubella-M EIA Test Kit, 
               Product #801-165.  Recall #Z-348-6.
CODE           Lot #34K0104.
MANUFACTURER   Centocor, UK Ltd., United Kingdom.
RECALLED BY    Centocor, Inc., Malvern, Pennsylvania, by
               letter June 23, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       200 kits were distributed.
REASON         There is a higher incidence of
               equivocal/positive results than anticipated.

_______________
PRODUCT        Critikon Brand Protectiv and Protectiv-Plus
               I.V. Catheter Safety System, various gauge and
               lengths of IV needle sets, packed in Tyvek
               blister packs which act as sterility barriers:
               (a) Product No. 3042; (b) Product No. 3050;
               (c) Product No. 3055; (d) Product No. 3056;
               (e) Product No. 3057; (f) Product No. 3060;
               (g) Product No. 3065; (h) Product No. 3066;
               (i) Product No. 3067.  Recall #Z-349/357-6.
CODE           This recall affects ONLY those products where
               each individual package's seven character
               code, located on the peel tab, has a code
               which falls into the following ranges:
               "3124Y9X" through "3654Y9X" and "0015Y9X"
               through "0705Y9X" where the first four digits
               are the Julian date, "Y" can be any letter,
               and "X" can be any number. 
MANUFACTURER   Johnson and Johnson Company, formerly known as
               Critikon, Inc., Southington, Connecticut.
RECALLED BY    Johnson and Johnson, formerly Critikon, Inc.,
               Tampa, Florida, by letter mailed on April 10,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.

                             -7-QUANTITY       778,191 catheters were distributed.
REASON         An opening may be present in the clear plastic
               portion of the blister package, therefore
               compromising the sterile barrier of the
               package.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III =======
_______________
PRODUCT        Cellfree Interleukin-2 Receptor (IL-2R) Bead
               Assay Kit, Catalog #AK3120, for the
               quantitative measurement of Interleukin-2
               Receptor Level in Human Serum.  
               Recall #Z-358-6.
CODE           Lot #6029 EXP 3/13/94.
MANUFACTURER   T Cell Diagnostics, Inc., Cambridge,
               Massachusetts.
RECALLED BY    Manufacturer, by letter dated January 28,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Massachusetts, Texas, Connecticut, Louisiana,
               Greece.
QUANTITY       20 kits were distributed.
REASON         Controls provided with the product were not
               meeting the assigned ranges and fell below
               specified limits.

_______________
PRODUCT        Glucometer Encore Test Strips, used to measure
               glucose in whole blood.  Recall #Z-360-6.
CODE           Product Code 2250, Lot numbers:  A90C5, B95C5,
               C95C5.
MANUFACTURER   Bayer Corporation, Elkhart, Indiana.
RECALLED BY    Manufacturer, by letter July 7, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       45,451 cartons.
REASON         Subject lots may fail to give blood glucose
               reading, and the Glucometer Encore meter will
               display an E-3 error code.


RECALLS AND FIELD CORRECTIONS: VETERINARY PRODUCTS -- CLASS I
_______________
PRODUCT        Warbex Famphur Pour-On for Cattle, for control
               of cattle grubs and reduction of lice
               infestations.  Recall #V-004-6.
CODE           Lot #950658.
MANUFACTURER   PM Resources, Bridgeton, Missouri.
RECALLED BY    Mallinckrodt Veterinary, Inc., Mundelein,
               Illinois, by telephone on November 20, 1995,
               and by press release.  Firm-initiated recall
               ongoing.

                             -8-DISTRIBUTION   Nebraska, Iowa, North Dakota, South Dakota,
               Colorado, Kansas, Wyoming, Texas.
QUANTITY       1,996 (499 cases) were distributed.
REASON         Product is contaminated with isazophos, an
               organophosphate turf insecticide.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS III
========
_______________
UPDATE         The correct recall number for Baciferm 50
               Medicated Bacitracin Zinc Premix Type A
               Medicated Article, in 50 pound bags which
               appeared in the January 10, 1996 Enforcement
               Report is V-005-6.

                             -9-


END OF ENFORCEMENT REPORT FOR JANUARY 24, 1996.  BLANK PAGES MAY
FOLLOW.


                                   ####