The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
11/16/1994

Recalls and Field Corrections:  Foods -- Class II -- 11/16/1994

November 16, 1994                                             94-46

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Dried Preserved Prunes, in 2.5 ounce jars.  Recall #F-091-5.
CODE           None.
MANUFACTURER   Mei Yuan Preserved Foods Company, Ltd., China.
RECALLED BY    Great Rich Trading Company, Inc., Brooklyn, New York, by
               telephone on or about March 7, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   New York.
QUANTITY       Firm estimates none remains on the market.
REASON         Product contains undeclared FD&C Yellow No. 5, FD&C Yellow
               No. 6, and Ponceau 4R.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Hawthorne Berry, cut, dried for use as a tea, in 8 ounce
               plastic bags.  Recall #F-092-5.
CODE           Lot #08469 packaged 8/11/94.
MANUFACTURER   Indiana Botanic Gardens, Hobart, Indiana.
RECALLED BY    Manufacturer, by letter October 5, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       200 units.
REASON         Product contains insect contamination.

_______________
PRODUCT        Suncorn Enriched Corn Meal packaged in 2 pound bags.  
               Recall #F-093-5.
CODE           None.
MANUFACTURER   Intermares Foods, Inc., Catano, Puerto Rico.
RECALLED BY    Manufacturer, by visit beginning on or about July 28, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Puerto Rico.
QUANTITY       252 bags were distributed; firm estimates none remains on
               the market.
REASON         Product is mislabeled.  The ingredients listed on the label
               were not added to the product.

_______________
PRODUCT        Tofutti brand Vanilla Almond Bark Frozen Desserts, in 3
               gallon containers.  Recall #F-094-5.
CODE           All codes manufactured in the past 6 months.
MANUFACTURER   Kemp's Foods, Inc., Lancaster, Pennsylvania.
RECALLED BY    Manufacturer, by letters of September 2 and 9, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Pennsylvania, New Jersey, Delaware, New York, Maryland,
               Virginia, Connecticut.
QUANTITY       Firm estimates very little if any product remains on the
               market.
REASON         Product contains undeclared almonds.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Slo-Bid Gyrocaps (Theophylline) 50 mg, extended release
               capsules used in treatment of asthma.  Recall #D-024-5.
CODE           Lot #CN93251 EXP 7/95.
MANUFACTURER   Rhone Poulenc Rorer, Fort Washington, Pennsylvania.
RECALLED BY    Rhone Poulenc Rorer, Collegeville, Pennsylvania, by letter
               dated October 20, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Firm estimates less than 145 units remain on the market.
REASON         Product does not meet dissolution specifications.

_______________
UPDATE         Dexamethasone Sodium, in 5 ml bottles, Recall #D-009-5 which
               appeared in the October 19, 1994, Enforcement Report is
               being amended to correct the list of private labels. 
               Product was distributed under the following labels:   Akorn,
               Bausch & Lomb, Butler, Equipharm, Goldline, Infinity, Major,
               Ocusoft, Phoenix, Pro-Vet, Rugby, Qualitest, Vedco, Webster,
               Richmond and Pharmascience.


                                    -2-RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Lozol Tablets, (Indapamide), 1.25 mg, packaged in bottles of
               100, an oral anti-hypertensive.  Recall #D-023-5.
CODE           Lot #MN0885 EXP 4/96.
MANUFACTURER   Rhone Poulenc Rorer, Manati, Puerto Rico.
RECALLED BY    Rhone Poulenc Rorer, Collegeville, Pennsylvania, by letter
               June 30, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       4,563 bottles were distributed.
REASON         Product is slightly below specification for dissolution.

_______________
PRODUCT        Slo-Bid Gyrocaps (Theophylline) 50 mg, extended release
               capsules used in treatment of asthma, in unit dose packs of
               10 capsules per card and 10 cards per carton.  
               Recall #D-025-5.
CODE           Lot #CN93755 EXP 5/95.
MANUFACTURER   Rhone Poulenc Rorer, Fort Washington, Pennsylvania.
RECALLED BY    Rhone Poulenc Rorer, Collegeville, Pennsylvania, by letter
               September 22, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Firm estimates 210 cartons of 10 x 10 blister packs remain
               on the market.
REASON         A 75 mg Slo-Bid Gyrocap was found in a unit dose blister
               cavity.

_______________
PRODUCT        Dupont Merck brand Naltrexone HCl Tablets, 50 mg, Rx, in
               amber glass bottles of 14 and 50, Rx, indicated to provide
               blockage of pharmacologic effect of exogenously administered
               opioids.  Recall #D-026-5.
CODE           Lot numbers:  EE156A EXP 7/97, EE156B EXP 7/97, EF172A EXP
               7/97.
MANUFACTURER   DuPont Merck Pharmaceutical Company, Garden City, New York.
RECALLED BY    Manufacturer, by telephone October 7, 1994, followed by
               letter October 12, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Germany, France, Spain, Switzerland, Portugal, Austria.
QUANTITY       28,00 tablets of lot EE156A, 17,600 tablets of lot EE156B,
               and 52,318 tablets of lot EF172A were distributed.
REASON         Product does not meet total degradation specifications.

_______________
PRODUCT        Sinclair Pharmacal brand Boroleum Ointment, OTC, used to
               relieve nasal discomfort due to head colds, hayfever, sinus
               and minor nasal and skin irritations, packaged in 0.6 ounce
               tubes.  Recall #D-027-5.
CODE           Lot numbers:  S284 and 0054. 
MANUFACTURER   Sinclair Pharmacal Company, Inc., Fishers Island, New York.
RECALLED BY    Manufacturer, by telephone October 7, 1994, followed by fax. 
               Firm-initiated recall ongoing.
DISTRIBUTION   New York, New Jersey, Florida, Virginia, Illinois.

                                    -3-QUANTITY       4,824 tubes were distributed.
REASON         Presence of small amounts of water in the tubes.

_______________
PRODUCT        Naproxyn (naproxen) Suspension, 16 ounce bottles, a
               nonsteroidal anti-inflammatory.  Recall #D-028-5.
CODE           Lot numbers:  40217 EXP 4/97, 40220 EXP 4/97, 40221 EXP
               4/97.
MANUFACTURER   Syntex, Inc., Ontario, Canada.
RECALLED BY    Syntex Puerto Rico, Inc., Humacao, Puerto Rico, by letter
               October 17, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       9,858 bottles were distributed.
REASON         Some units were superpotent.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS I
======================
_______________
PRODUCT        Red Blood Cells.  Recall #B-024-5.
CODE           Unit 18Z65722.
MANUFACTURER   The American National Red Cross, Lansing, Michigan.
RECALLED BY    Manufacturer, by telephone August 5, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Michigan.
QUANTITY       1 unit.
REASON         Blood product contaminated with coagulase negative
               Streptococcus agalactiae Group B was distributed for
               transfusion.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        (a) Summit Sample Handling System; (b) Elisa Computer
               Software, Version 5.2; (c) Midas Computer Software, Version
               1.2; (d) Adams Computer Software, Version 1.5, 1.6, and 1.8. 
               Recall #B-008/011-5.
CODE           (a) All serial numbers; (b) Software version 5.2; (c)
               Software version 1.2; (d) Software version 1.5, 1.6, 1.8.
MANUFACTURER   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
RECALLED BY    Manufacturer, (a) by letter in August 1992; (b) by letters
               of June 1, 1992 and July 2, 1992; (c) telephone February
               1993; (d) by letter November 8, 1993.  Firm-initiated field
               correction complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 200 units; (b) 137 packages; (c) 161 packages;
               (d) 355 packages.
REASON         Computer software with design defects that could cause the
               inadvertent distribution of untested blood and blood
               products was distributed.

                                    -4-_______________
PRODUCT        Red Blood Cells.  Recall #B-012-5.
CODE           Unit numbers:  15G70212, 15KG11359, 15KG06468, 15KG16129,
               15KG20309, 15KG19594.
MANUFACTURER   American National Red Cross (The), Boise, Idaho.
RECALLED BY    Manufacturer, by letter October 29, 1993, November 10, 1993,
               and December 2, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Michigan, Montana, Arkansas, Florida.
QUANTITY       6 units.
REASON         Blood products, collected from donors who had visited areas
               designated as endemic for malaria, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall B-025-5.
CODE           Unit #18Z65719.
MANUFACTURER   The American National Red Cross, Lansing, Michigan.
RECALLED BY    Manufacturer, by telephone August 5, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Michigan.
QUANTITY       1 unit.
REASON         Blood product, corresponding to Platelets, Pooled
               contaminated with coagulase negative Streptococcus
               agalactiae Group B, was distributed for transfusion.

_______________
PRODUCT        Red Blood Cells.  Recall #B-035-5.
CODE           Unit #9401569.
MANUFACTURER   Lorain County Blood Bank, Inc., Elyria, Ohio.
RECALLED BY    Manufacturer, by telephone on March 2, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Ohio.
QUANTITY       1 unit. 
REASON         Blood product contaminated with coagulase negative
               Staphylococcus epidermidis was distributed. 

_______________
PRODUCT        Blood grouping reagents:
               (a) Anti-A, Anti-B, Anti-D (Anti-RH1), (Monoclonal),
               Control, Reverse Diluent Ortho BioVue System, Product Code
               707100; (b) Anti-A, Anti-B, Anti-AB, Anti-D, Anti-CDE,
               (Monoclonal), Control, Ortho BioVue System, Product Code
               707190; (c) Anti-C, Anti-D, Anti-E, Anti-e, (Monoclonal),
               Control, Ortho BioVue System, Product Code 707200; (d) Anti-
               C, Anti-E, Anti-c, Anti-e, Anti-K, (Monoclonal), Control,
               Ortho BioVue System, Product Code 707280.  
               Recall #B-037/040-5.
CODE           Lot numbers:  ABE601A, ABE601B, ABE601C, ABE601D, ABE602A, 
               ABE603A, ABE603B, ABE603C1, ABE603D1, ABE603E1, 
               ABE603F2, ABE603G2, ABE603H2, ABE603I2, ABE604A, 
               ABE604B, ABE604C, ABE604D, ABE604E, ABE604F, 
               ABE605A, ABE605B, ABE605C, ABE606A, ABR301A, 

                                    -5-               ABR302A, DEV700A, DEV700B, DEV700C, RHP701A,
               RHP701B, RHP701C, RHP702A, RHP702B, RHP702C, 
               RHP702D, RHP702E, RHP703A, RHP704A, RHP704B, 
               RHP705A, RHP705B, RHP705C, RHP705D.
MANUFACTURER   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
RECALLED BY    Manufacturer, by letter May 4, 1994.  Firm-initiated recall
               complete. 
DISTRIBUTION   International.
QUANTITY       Firm estimates that all of the product has been destroyed or
               replaced.
REASON         Blood grouping reagents with mislabeled cassettes were
               distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Cryoprecipitated AHF; (c) Recovered
               Plasma.  Recall #B-016/018-5.
CODE           Unit numbers:  (a) 16J71558, 16J85871, 16J88934, 16F36127,
               16J24182; (b) 16J71558, 16J85871, 16J88934, 16F36127; (c)
               16J71558 16J85871, 16J88934, 16F36127, 16J24182.
MANUFACTURER   American National Red Cross (The), Columbus, Ohio.
RECALLED BY    Manufacturer, by letter July 26, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Ohio, California, Switzerland.
QUANTITY       (a) 5 units; (b) 4 units; (c) 5 units.
REASON         Blood products, which were collected from a donor who had
               previously tested repeatedly reactive for the antibody to
               the human immunodeficiency virus type 1 (anti-HIV-1), were
               distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-026-5.
CODE           Unit numbers:  18Z65719, 18Z65722, 18Z65723, 18Z65727,
               18Z65736, 18Z65746.
MANUFACTURER   The American National Red Cross, Lansing, Michigan.
RECALLED BY    Manufacturer, by telephone August 5, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Michigan.
QUANTITY       6 units.
REASON         Blood products, corresponding to Platelets, Pooled
               contaminated with coagulase negative Streptococcus
               agalactiae Group B, were distributed for further
               manufacture.

_______________
PRODUCT        Recovered Plasma.  Recall #B-036-5.
CODE           Unit #9401569.
MANUFACTURER   Lorain County Blood Bank, Inc., Elyria, Ohio.
RECALLED BY    Manufacturer, by letter June 24, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Florida.

                                    -6-QUANTITY       1 unit.
REASON         Blood product contaminated with coagulase negative
               Staphylococcus epidermidis was distributed for further
               manufacture.

_______________
PRODUCT        Platelets.  Recall #B-045-5.
CODE           Unit numbers:  Unit numbers:  16FN08909, 16FN08911,
               16GL07852, 16GL07857, 16GL07858, 16GL07863, 16GL07865,
               16GL07866, 16GL07867, 16GL07868, 16GL07869, 16GL07871,
               16GL07878, 16GL07884, 16GL07885, 16GL07890, 16GL07891,
               16GL07894, 16GL07895, 16GL07896, 16GL07898, 16GL07899,
               16GL07901, 16GL07902, 16GL07904, 16GL07907, 16GL07908,
               16GL07911, 16GL07912, 16GL07913, 16GL07914, 16GL07915,
               16GL07916, 16GL07917, 16GL01919, 16GL07921, 16M74952,
               16M74953, 16M74954.
MANUFACTURER   American Red Cross Blood Services, Columbus, Ohio.
RECALLED BY    Manufacturer, by letter March 16 and 17, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Ohio.
QUANTITY       39 units.
REASON         Blood products, incorrectly tested for the antibodies to
               human immunodeficiency virus, types 1 and 2 (anti-HIV-1/2),
               were distributed.

RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Adult Manual Resuscitators with Variable Volume Reservoir:
               (a) Catalog #008006 
               (b) Catalog #008006S 
               (c) Catalog #008006SP  
               (d) Catalog #008009 
               (e) Catalog #008106 
               (f) Catalog #44008-006
               (g) Catalog #RESC-1007
               (h) Catalog #RESC-1016
               (i) Catalog #RESC-1030.  Recall #Z-079/087-5.
  CODE         Lot Numbers: (a) 4A488, 4C694, 4D680 
               (b) 4A561, 4C345, 4D679
               (c) 4A628 
               (d) 4A302, 4A489, 4B461 
               (e) 4A562
               (f) 4A626, 4E610 
               (g) 4A485, 4B380, 4B523, 4C516, 4D291, 4D455, 4D566, 4E247 
               (h) 4B192, 4D350, 4E254, 4E564 
               (i) 4B381, 4D346. 
MANUFACTURER   Intertech Resources, Inc., Fort Myers, Florida.
RECALLED BY    Manufacturer, by letter August 15, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, South Africa, Canada, AbuDhabi U.A.
QUANTITY       Undetermined.

                                    -7-REASON         Associated oxygen tubing may be occluded by excessive
               bonding material, thereby preventing or limiting the flow of
               oxygen.

SEIZURES:
=================================================================
_______________
PRODUCT        Frozen Shrimp (94-557-490 et al).
CHARGE         Adulterated - The article consists in part of a decomposed
               substance by reason of the presence therein of decomposed
               shrimp.
LOCATIONS      Americold Cold Storage, Tampa, Florida.
FILED          July 19, 1994, U.S. District Court for the Middle District
               of Florida, Tampa Division, Civil #94-1131-CIV-T-17C;
               FDC #66994.
SEIZED         August 1, 1994 - goods valued at approximately $65,541.

_______________
PRODUCT        Ultra Slim Diet Nutrition Bars (94-711-111).
CHARGES        Adulterated - Valuable constituents, namely Vitamin C and
               Riboflavin, have been partly omitted from the article.
               Misbranded - The article's labeling is false and misleading
               because it represents and suggests that the article provides
               30 percent of the U.S. Recommended Daily Allowance (RDA) of
               Vitamin C and 33 percent of the U.S. RDA of Riboflavin per
               serving; whereas, the article provides from 4.07 to 8.2
               percent of the U.S. RDA of Vitamin C and from 16 to 16.6
               percent of the U.S. RDA of Riboflavin per serving; and the
               nutrition information, ingredients statement, and name and
               place of business of the distributor are not prominently
               placed on the label with such conspicuousness as to render
               them likely to be read by the ordinary individual under
               customary conditions of purchase and use because they do not
               appear together on the principal display panel or together
               without intervening material on the information panel, and
               they appear in a type size of less than 1/16-inch in height.
LOCATION       Dietary Enterprises, Ltd., doing business as The Diet Store,
               Ltd., Niles, Illinois.
FILED          November 3, 1994; U.S. District Court for the Northern
               District of Illinois; Civil #94C 6593; FDC #67005.
SEIZED         November 3, 1994 - goods valued at approximately $9,000.

_______________
PRODUCT        Sphygmomanometers and components (94-614-086/093).
CHARGES        Adulterated - The strength, purity, or quality of the
               sphygmomanometer (Model SB-A900, LOT 001) falls below that
               which it purports or is represented to possess, in that the
               manometer was inaccurate, exceeding the limitation set
               forth; and the methods used in, and the facilities and
               controls used for, the articles' (all lots) manufacture,
               packing, and storage are not in conformance with the current
               good manufacturing practice requirements for devices.

                                    -8-LOCATION       Surgical Instrument Company of America, Ridgefield, 
               New Jersey.
FILED          November 2, 1994; U.S. District Court for the District of
               New Jersey; Civil #94-5355; FDC #66950.
SEIZED         November 3, 1994 - goods valued at approximately $150,000.

_______________
PRODUCT        Nifedipine Capsules, USP, 10 and 20 mg (93-675-794, et al).
CHARGES        New drug - No approval of a supplement to an application is
               in effect for the drug.  Adulterated - The methods used in,
               and the facilities and controls used for, the article's
               manufacture, processing, packing, and holding do not conform
               to and are not operated and administered in conformity with
               current good manufacturing practice requirements.
LOCATION       Chase Laboratories, Newark, New Jersey.
FILED          October 14, 1994; U.S. District Court for the District of
               New Jersey; Civil #94-4880(NHP); FDC #66914.
SEIZED         October 24, 1994 - goods valued at approximately $900,000.

                                    -9-

END OF ENFORCEMENT REPORT FOR NOVEMBER 16, 1994.  BLANK PAGES MAY