FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
11/02/1994

Recalls and Field Corrections:  Foods -- Class I -- 11/02/1994

November 2, 1994                                                94-44

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        Peony brand Dried Potatoes, in 8 ounce packages.
               Recall #F-027-5.
CODE           None.
MANUFACTURER   Product of the People's Republic of China.
RECALLED BY    Tai Wing Hong Importer, Inc., Brooklyn, New York (importer),
               by telephone followed by letter September 9, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   New York, Illinois, California, Massachusetts, Virginia.
QUANTITY       500 cartons (100 packages per carton) were distributed.
REASON         The product contains undeclared sulfites.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Dudek Foods, Inc. Potato/Cheddar Pierogi's, in 16 ounce
               plastic bags and various flavored Nalesniki Crepes, net
               weight 16 ounces:  (a) Potato/Cheddar Pierogi's; (b)
               Nalesniki Cheese Crepes; (c) Nalesniki Strawberry Crepes;
               (d) Nalesniki Apple Crepes; (e) Nalesniki Blueberry Crepes. 
               Recall #F-017/021-5.
CODE           All lots.
MANUFACTURER   Dudek Foods, Inc., Hamtramck, Michigan.
RECALLED BY    Manufacturer, by placing stickers on products beginning
               September 29, 1994.  Firm-initiated field correction
               (relabeling) complete.
DISTRIBUTION   Michigan, Ohio, Indiana.
QUANTITY       Approximately 600 pounds (100 cases) of Pierogi's and 180
               pounds (30 cases) of crepes were distributed.
REASON         Product contains undeclared FD&C Yellow no. 5.

_______________
PRODUCT        Spice Supreme Candy Sprinkles, in 6 ounce opaque bottles. 
               Recall #F-025-5.
CODE           Lot #IDE20.
MANUFACTURER   Gel Spice Company, Inc., Bayonne, New Jersey
               (repacker/responsible firm).
RECALLED BY    Repacker, by letter September 1, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, British West Indies.
QUANTITY       Undetermined.
REASON         Product contains undeclared FD&C Yellow No. 5.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Minute Maid Orange Juice, in 16 fluid ounce glass bottles. 
               Recall #F-022-5.
CODE           04047Y over CT124 and 04110Y over CT124.
MANUFACTURER   Knouse Foods, Inc., Gardners, Pennsylvania.
RECALLED BY    Coca Cola Foods, Inc., Houston, Texas, by telephone May 20,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       29,067 cases (24 bottles per case) were distributed.
REASON         Product is contaminated with Lactobacillus fermentum.

_______________
PRODUCT        ECEE Plus Tablets, a dietary supplement for oral
               administration, in 100 tablet bottles.  Recall #F-023-5.
CODE           Lot ALFM EXP 6/95.
MANUFACTURER   Manufacturing Chemists, Inc., Indianapolis, Indiana.
RECALLED BY    Manufacturer, by letter September 22, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Mississippi.
QUANTITY       1,122 bottles were distributed.
REASON         Product has an off odor and its color is fading.

_______________
PRODUCT        Koala Springs Sparkling Fruit Juice Beverages, all flavors,
               in 10 ounce and 25.4 ounce bottles.  Recall #F-024-5.
CODE           All codes beginning with WN followed by Julian date.
MANUFACTURER   Hiram Walker British Columbia, Canada (bottling firm).
RECALLED BY    Koala Springs International, Inc., Fremont, California, by
               press release October 7, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Alaska, Oregon, Washington State.
QUANTITY       20,000 to 25,000 cases were distributed.
REASON         Product is contaminated with yeast and Lactobacillus
               bacteria.

                                    -2-RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        SoloPak Heparin Lock Flush Solution, USP, Preservative,
               Free, 10 USP units/ml packaged in 1, 3 and 5 ml pre-filled
               syringes; for maintenance of patency of intravenous
               injection devices only, not to be used for anticoagulant
               therapy; (a) Catalog #06005: 5 ml Hy-Pod Syringe, contained
               in the Lok-Pak Heparin Lock Flush Procedure Pack, needle not
               included, 200 per case;
               (b) Catalog #10671: 1 ml Hy-Pod Syringe and 25G x 5/8"
               needle, individually wrapped, 120 per carton;
               (c) Catalog #10673: 3 ml Hy-Pod Syringe and 25G x 5/8"
               needle, individually wrapped, 120 per carton;
               (d) Catalog #10675: 5 ml Hy-Pod Syringe and 25G x 5/8"
               needle, individually wrapped, 120 per carton;
               (e) Catalog #10681: 1 ml Hy-Pod Syringe (needles not
               included), individually wrapped, 120 per carton,
               (f) Catalog #10683: 3 ml Hy-Pod Syringe (needles not
               included), individually wrapped, 120 per carton; 
               (g) Catalog #10685: 5 ml Hy-Pod Syringe (needles not
               included), individually wrapped, 120 per carton; 
               (h) Catalog #11771: 1 ml Hy-Pod Syringe, contained in the
               Lok-Pak-N Heparin Lock Flush Procedure Pack, 200 per case;
               (i) Catalog #11773: 3 ml Hy-Pod Syringe, contained in the
               Lok-Pak-N Heparin Lock Flush Procedure Pack, 200 per case;
               (j) catalog #11775: 5 ml Hy-Pod Syringe, contained in the
               Lok-Pak-N Heparin Lock Flush Procedure Pack, 200 per case.
               Recall #D-018-5.
CODE           (a) Catalog #06005: lot #94G008C 
               (b) Catalog #10671: lot #94E001C 
               (c) Catalog #10673: lot #94D004C, 94E005C, 94F004C 
               (d) Catalog #10675: lot #94D007C, 94F008B 
               (e) Catalog #10681: lot #94G001B 
               (f) Catalog #10683: lot #94E006B, 94G004B 
               (g) Catalog #10685: lot #94F008C, 94G008D 
               (h) Catalog #11771: lot #94D001B, 94E001B 
               (i) Catalog #11773: lot #94D004B, 94E005B, 94F004B 
               (j) Catalog #11775: lot #94D007B, 94G008B.
MANUFACTURER   SoloPak Medical Products, Inc., Franklin Park, Illinois.
RECALLED BY    SoloPak Medical Products, Inc., Elk Grove Village, Illinois,
               by letter October 11, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       589,200 were distributed; firm estimated that less than 25
               percent of the product remained on market at time of recall.
REASON         Potency cannot be assured through expiration date.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Dossitol Stool Softener, 100 mg, in bottles of 1000
               softgels.  Recall #D-019-5.
CODE           Lot numbers 255239 and 297339 EXP 9/96.

                                    -3-MANUFACTURER   Chase Laboratories, Inc., Newark, New Jersey.
RECALLED BY    Manufacturer, by letter April 20, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Undetermined.
QUANTITY       242 bottles were distributed.
REASON         Front panel label incorrectly states Docusate Calcium.  Left
               panel label correctly states Docusate Sodium.

_______________
PRODUCT        Levothyroxine Sodium for Injection, (a) 200 microgram (mcg)
               and (b) 500 mcg, a thyroid hormone, under the VHA and UDL
               labels.  Recall #D-020/021-5.
CODE           Ben Venue Lot No.  Customer Lot No.   EXP Date
               (a) 328-27-0008     FA559              1/95
                   328-27-0009     FC575              3/95
                   328-27-0010     FH597              8/95
                   328-27-0022     GD686              4/96
                                   GD686A             4/96
               (b) 328-27-0010     FL646             12/95
                                   FL646B            12/95.
MANUFACTURER   Ben Venue Laboratories, Bedford, Ohio.
RECALLED BY    Manufacturer, by letter October 5, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 42,503 vials; (b) 8,420 vials were distributed.
REASON         Product does not meet pH specifications on reconstitution.

_______________
PRODUCT        Lindane Lotion, USP 1%, an Rx topical, external use only,
               ectoparasiticide and ovicide, used for the treatment of
               patients infested with scabies under the following labels: 
               Pennex, in 2 fluid ounce bottles; PBI, in 2 fluid ounce
               bottles; Goldline, in 2 fluid ounce and in 1 pint bottles; 
               GG, distributed by Geneva Pharmaceuticals, Inc., in 2 fluid
               ounce bottles, and in 1 pint bottles;
               Major, in 1 pint bottles; Qualitest, in 2 fluid ounce
               bottles; Rugby, in 2 fluid ounce bottles; Schein, in 2 fluid
               ounce bottles, and in 1 pint bottles.  Recall #D-022-5.
CODE           Lot numbers 19890, 19901, 19944.
MANUFACTURER   Pennex Pharmaceuticals, Inc., formerly known as
               Pharmaceutical Basics, Inc., Morton Grove, Illinois.
RECALLED BY    Pennex Pharmaceuticals, Inc., now known as Morton Grove
               Pharmaceuticals, Inc., Morton Grove, Illinois, by letter
               October 13, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       50,340 2-fluid ounce and 14,919 pint bottles were
               distributed; firm estimated that less than 10 percent of the
               product remained on the market at time of recall.
REASON         Product does not meet compendial release criteria.


                                    -4-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        Red Blood Cells.  Recall #B-019-5.
CODE           Unit #15KG14596.
MANUFACTURER   American National Red Cross, Great Falls, Montana.
RECALLED BY    Manufacturer, by telephone March 3, 1993, followed by
               letter.  Firm-initiated recall complete.
DISTRIBUTION   Montana.
QUANTITY       1 unit.
REASON         A unit of Red Blood Cells, which tested repeatedly reactive
               for antibody to the hepatitis C virus encoded antigen(anti-
               HCV), was distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-021-5.
CODE           Unit #MU36132.
MANUFACTURER   East Texas Blood Center, Nacogdoches, Texas.
RECALLED BY    Manufacturer, by telephone January 25, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Florida.
QUANTITY       1 unit.
REASON         Blood product, that tested repeatedly reactive for the
               antibody to hepatitis C virus encoded antigen (anti-HCV),
               was distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Recovered Plasma.  Recall #B-015-5.
CODE           Unit #587097.
MANUFACTURER   Central Kentucky Blood Center, Lexington, Kentucky.
RECALLED BY    Manufacturer, by letter April 8, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit.
REASON         Blood product, which tested reactive for syphilis, was
               distributed.

_______________
PRODUCT        Whole Blood.  Recall #B-020-5.
CODE           Unit #18Q31209.
MANUFACTURER   The American National Red Cross, Lansing, Michigan.
RECALLED BY    Manufacturer, by telephone May 30, 1993, followed by letter
               June 4, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY       1 unit.
REASON         Blood product, labeled with the incorrect expiration date,
               was distributed.


                                    -5-RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Apnea Monitors, multiple models.  Recall #Z-891-4.
CODE           All models with unprotected (exposed pins style) lead wires
               and unprotected patient cables.
MANUFACTURER   Arvee Medical, Austin, Texas.
RECALLED BY    Manufacturer, by letter May 5, 1994.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Device labeling fails to provide adequate directions for use
               and the devices pose an unwarranted risk of injury to
               patients, because the product is not labeled to warn against
               the hazards of unprotected electrode lead wires being
               connected to electrical power sources.

_______________
PRODUCT        Electrode lead wires and patient cables, multiple models,
               used with apnea monitors.  Recall #Z-1084/1085-4.
CODE           All models of unprotected (exposed pin style), lead wires
               and unprotected patient cables.
MANUFACTURER   Del Mar Avionics, Irvine, California.
RECALLED BY    Manufacturer, by letter June 9, 1994.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Undetermined.
REASON         Undetermined.
REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the products are not labeled to contraindicate use
               with apnea monitors.

_______________
PRODUCT        Marquette male-connector lead wires and cables.  The
               respiration monitors are in the following monitoring systems
               and modules:  Marquette 7000 Series Monitoring System,
               Marquette ECG/Respiration Modules, Tram AR Modules, Tram III
               Modules, Tram A Modules, Tram SL Modules, Eagle (w/Resp)
               Monitors, and IVY Monitors.  Recall #Z-1106-4.
CODE           All serial numbers of the Marquette Modules and monitoring
               systems, capable of monitoring respiration, and all lots of
               Marquette male-connector leadwires and cables.
MANUFACTURER   Marquette Electronics, Inc., Milwaukee, Wisconsin.
RECALLED BY    Manufacturer, by letters of September 14, 1994, and October
               3, 1994.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 13,000 monitors.
REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the product is not labeled to warn against the
               hazards of unprotected electrode lead wires being connected
               to electrical power sources.

                                    -6-_______________
PRODUCT        Electrode Lead Wires and Patient Cables, multiple models,
               used with apnea monitors.  Recall #Z-1318/1319-4.
CODE           All models of unprotected (exposed pin style) lead wires and
               unprotected patient cables.
MANUFACTURER   Pace Tech Medical Monitors, Inc., Clearwater, Florida.
RECALLED BY    Manufacturer, by letter March 15, 1994.  Firm-initiated
               field correction ongoing.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the products are not labeled to contraindicate use
               with apnea monitors.

______________
PRODUCT        Electrode Lead Wires, used with Holter monitor recorders. 
               Recall #Z-1320-4.
CODE           All marketed in the past three years.
MANUFACTURER   Scole Engineering Company, Inc., Culver City, California.
RECALLED BY    Manufacturer by letter February 28, 1994.  Firm-initiated
               field correction ongoing.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the products are not labeled to contraindicate their
               use with apnea monitors.

_______________
PRODUCT        Electrode Lead Wires, used with muscle stimulators and TENS
               devices.  Recall #Z-1321-4.
CODE           All marketed in the past 3 years.
MANUFACTURER   Jace Systems, Inc., Moorestown, New Jersey.
RECALLED BY    Manufacturer, by letter April 7, 1994.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the products are not labeled to contraindicate use
               with apnea monitors.

_______________
PRODUCT        Apnea Monitors, multiple models.  Recall #Z-1383-4.
CODE           All models with unprotected (exposed pin style) lead wires
               and unprotected patient cables.
MANUFACTURER   Siemens Medical Systems, Inc., Danvers, Massachusetts.
RECALLED BY    Manufacturer, by letters of May 25 & 27, 1994.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.

                                    -7-REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the product is not labeled to warn against the
               hazards of unprotected electrode lead wires being connected
               to electrical power sources.

_______________
PRODUCT        Electrode Lead Wires, used with apnea monitors.  
               Recall Z-1384-4.
CODE           All units marketed in the past 3 years.
MANUFACTURER   Contour Medical Technology, Inc., LaVergne, Tennessee.
RECALLED BY    Manufacturer, by letter July 19, 1994.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the products are not labeled to contraindicate use
               with apnea monitors.

_______________
PRODUCT        Electrode Lead Wires, multiple models, used with
               electrocardiograph monitors, and other similar devices. 
               Recall #Z-1411-4.
CODE           All models with unprotected exposed metal pins.
MANUFACTURER   Sentry Medical Products, Irvine, California.
RECALLED BY    Manufacturer, by letter September 24, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Labeling fails to provide adequate directions for use and
               the devices pose an unwarranted risk of injury to patients,
               because the products are not labeled to contraindicate use
               with apnea monitors.

_______________
PRODUCT        IVAC Model 2863A Infusion Pump.  Recall #Z-008-5.
CODE           510 individual serial numbers.
MANUFACTURER   IVAC Corporation, San Diego, California.
RECALLED BY    Manufacturer, by letter July 19, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       433 units were distributed.
REASON         Unintended changes in the device's settings may occur due to
               a capacitor rubbing on the keyboard, which may result in
               erroneous infusion rates.

_______________
PRODUCT        Angiographic contrast media injectors, used to inject
               contrast media directly into the circulatory system:
               (a) Medrad Mark IV; (b) Mark IV CT 202 Injectors.  
               Recall #Z-032/033-5.

                                    -8-CODE           All serial numbers.
MANUFACTURER   Medrad, Inc., Indianola, Pennsylvania.
RECALLED BY    Manufacturer, by instructions sent on October 24, 1991. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Product is no longer being manufactured, but firm estimated
               that approximately 4,000 units might still possibly be in
               commerce.
REASON         Grease in the push button switches congealed causing the
               devices to deliver contrast media at higher rates or volumes
               than desired.

_______________
PRODUCT        Ophthalmic surgical packs:
               (a) Deluxe Posterior Vitrectomy Pack with Reflux, Catalog
                   No. DP4800;
               (b) Deluxe Phaco Pack without Reflux, Catalog No. DP4310; 
               (c) Deluxe Phaco Pack, Catalog No. DP4330.  
               Recall #Z-046/048-5.
CODES          (a) 1742A     1742B     1882B     1912B     2102B     2522A
                   2552A     2612A     2792B     2822B     2882B     3222B
                   3452A     3492A     3512A     0113A     0183A     0433B
                   0503B     0533A     0633A     0743A     1063B     1173B
                   1413A     1463B     1663AF    1723AF    1753AF    2093AF
                   2093BF    2103AF    2103BF    2143AF    2143BF    2443AF
                   2573BF    2653AF    2773BF    2993BF    3063AF    3093AF
                   3143AF    3363AF    1113B     2752B     3222A.
               (b) 0893B     1672A     1672B     1692B     1702B     1972B 
                   2032A     2202A     2232A     2252A     2252B     2262A 
                   2262B     2302B     2312A     2312B     2322A     2322B 
                   2332A     2332B     2342A     2352A     2392A     2392B 
                   2402A     2402B     2452B     2472B     2482B     2542B 
                   2592B     2612B     2652B     2662B     2672A     2682A 
                   2692B     2732A     2752A     2792A     2812A     2932A 
                   2932B     2952A     2952B     2962B     3002B     3012B 
                   3062A     3062B     3102A     3102B     3152A     3292B 
                   3422B     3452B     3462B     3502B     3522B     3562B 
                   3632B     3642B     3662B                  
                   0073A     0073B     0113B     0143B     0153B     0193B 
                   2022B     0213B     0223B     0273B     0293B     0323A 
                   0343B     0363B     0393A     0403B     0413A     0423A 
                   0483A     0493A     0633B     0673B     0713B     0783B 
                   0823B     0843B     0923A     0953B     1023B     1043A 
                   1043B     1093A     1123A     1163A     1313A     1313B 
                   1333A     1333B     1373A     1383B     1403B     1543A 
                   1543B     1583A     1613A     1663B     1693B     1743B 
                   1793B     1833A     1833B     1893B     1903A     1943B 
                   1953A     1963B     1973A     2233B     2253A     2283B 
                   2303B     2313A     2353A     2353B     2373A     2503A
                   2523B     2533A     2603B     3013B     3063B     3083B
                   3123B     3153B     3203B     3223B     3373A     3373B 
                   3423A     3443A     3483A     0114A 
               (c) 2943B     2953B     3273B     3343B     3623A.
                                    -9-MANUFACTURER   Storz Instrument Company, St. Louis, Missouri. 
RECALLED BY    Manufacturer, by letter May 12, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 60,528 packs; (b) 202,005 packs; (c) 6,210 packs were
               distributed.
REASON         The sterility of the devices may be compromised due to
               packaging defects.


MEDICAL DEVICE SAFETY ALERTS: 
============================================
_______________
PRODUCT        Electrode lead wires, used with Dynography Physiological
               Recorder.  Safety Alert #N-055-4.
CODE           All marketed since 1983 with unprotected (exposed pin style)
               lead wires.
MANUFACTURER   Sensormedics Corporation, Yorba Linda, California.
ALERTED BY     Manufacturer, by letter November 22, 1994.
DISTRIBUTION   Undetermined.
QUANTITY       Approximately 1,500 units.
REASON         Electrocution hazard associated with unprotected (exposed
               pin style) lead wires.

                                   -10-

END OF ENFORCEMENT REPORT FOR NOVEMBER 2, 1994.  BLANK PAGES MAY
FOLLOW.
                                   ####