Recalls and Field Corrections:  Foods -- Class II -- 08/31/1994

August 31, 1994                                              94-35

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=======================
_______________
PRODUCT        Oxygen USP in size D & E Cylinders.  Recall #D-352-4.
CODE           Lot numbers:  032894-25, 032994-29, 032994-30, 032994-30,
               032994-31, 032994-32, 032994-33, 053194-2.
MANUFACTURER   Medical Park Pharmacy, Morehead, North Carolina.
RECALLED BY    Manufacturer, by telephone June 17, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Current good manufacturing practice deficiencies.

_______________
PRODUCT        Risperdal (Risperidone) Tablets, labeled as containing 2 mg
               tablets, an oral anti-psychotic drug, in bottles of 60. 
               Recall #D-373-4.
CODE           Lot #93L425E EXP 6/95.
MANUFACTURER   Johnson and Johnson Pharmaceutical Partners, Gurabo, Puerto
               Rico.
RECALLED BY    Janssen Pharmaceutica, Titusville, New Jersey, by letter
               July 15, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       16,536 bottles were distributed; firm estimated that 1000
               bottles remained on market at time of recall.
REASON         A bottle labeled as 2 mg tablets contained 1 mg tablets.
_______________
PRODUCT        Prednisone Tablets, USP, (a) 5 mg; (b) 10 mg, in unit dose
               blister packs in cartons of 100 tablets, Rx oral
               glucocorticoid anti-inflammatory.  Recall #D-374/375-4.
CODE           Lot numbers:  (a) 2P425, 3A822, 3F173, 3K572, 4B244, 4I681,;
               (b) 2P394, 2V694, 3C002, 3J413, 3P738, 4B207.
MANUFACTURER   Purepac Pharmaceutical Company, Division of Kalipharma,
               Inc., Elizabeth, New Jersey.
RECALLED BY    UDL Laboratories, Inc., Rockford, Illinois (repacker), by
               letter August 12, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 20,215 cartons of 100 unit dose; (b) 13,199 cartons of
               100 unit dose tablets were distributed. Firm estimates 35%
               of product remains on market.
REASON         Product may not meet dissolution specifications through
               expiration dates.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Bulk pharmaceuticals:
               1.   Activated Charcoal Powder, USP, (125 grams & 500 grams)
               2.   Ammonium Carbonate, Chips, NF, (125 grams, 500 grams &
                    x 500 grams); Ammonium Carbonate, Powder, NF (500
                    grams, 12 x 500 grams & 300 pounds) 
               3.   Benzyl Alcohol, NF, (475 pounds) 
               4.   Butylparaben, NF (2.5 kilograms) 
               5.   Calcium Chloride, Dihydrate, Granular, USP, (pyrogen
                    tested - 12 kilograms, 70 pounds & 100 pounds) (not
                    pyrogen tested - 500 grams, 100 pounds & 300 pounds) 
               6.   Edetate Disodium, USP, (500 grams & 12 kilograms) 
               7.   Hydrogen Peroxide, Topical Solution, 2.5% - 3.5%, USP, 
                    (500 milliliters & 4 liters) 
               8.   Iodine Tincture, 2%, USP, (500 milliliters) 
               9.   Lanolin, Hydrous (lanolin, modified), USP, (454 grams) 
               10.  Magnesium Chloride, Hexahydrate, USP, (12 kilograms &
                    90 pounds) 
               11.  Potassium Iodide, Granular, USP, (125 grams, 500 grams,
                    2.5 kilograms, 12 kilograms & 100 pounds) 
               12.  Potassium Phosphate, monobasic, crystal, pyrogen
                    tested, USP (12 kilograms & 100 pounds) 
               13.  Salicylic Acid, fine crystal, USP (125 grams, 500 grams
                    & 4 x 500 grams); and, salicylic acid, powder, USP, 
                    (125 grams, 12 x 125 grams, 500 grams & 4 x 500 grams) 
               14.  Silver Nitrate, crystal, USP, (30 grams, 125 grams &
                    500 grams) 
               15.  Sodium Bicarbonate, powder, pyrogen tested, USP, (12
                    kilograms & 100 pounds); Sodium Bicarbonate, powder,
                    [not pyrogen tested], USP, (500 grams, 12 x 500 grams,
                    2.5  kilograms, 4 x 2.5 kilograms, 12 kilograms & 100
                    pounds) 
               16.  Sodium Carbonate, anhydrous, granular, NF, (500 grams,
                    100 pounds & 200 pounds)
                                    -2-                         
               17.  Sodium Iodide, Crystal, USP, (125 grams & 500 grams) 
               18.  Sodium Phosphate, dibasic, pyrogen tested, USP (12
                    kilograms & 200 pounds), sodium phosphate, dibasic,
                    [not pyrogen tested], USP, (500 grams, 2.5 kilograms,
                    25 pounds, 100 pounds & 300 pounds) 
               19.  Sodium Thiosulphate, pentahydrate, USP, (500 grams, 
                    12 x 500 grams, 2.5 kilograms, 4 x 2.5 kilograms & 110
                    pounds) 
               20.  Sucrose, Crystal, NF, (12 kilograms).  
               Recall #D-353/372-4. 
CODE           All lots bearing lot numbers C26001 (July 1989) through
               H26001 (July 1994).
MANUFACTURER   J.T. Baker, Inc., Jackson, Tennessee.
RECALLED BY    J.T. Baker, Inc., Phillipsburg, New Jersey, by letter July
               13, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and Puerto Rico.
QUANTITY       Approximately 1.2 million pounds of the powder/crystal
               products and approximately 1,500 liters of the liquid
               products were distributed.
REASON         Insufficient stability data to support expiration dates.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
====================
______________
PRODUCT        LifeTec Community Blood Center Systems Software Revision
               1.60.  Recall #B-423-4.
CODE           Version 1.60 only.
MANUFACTURER   SysTec Computer Associates, Inc., Smithtown, New York.
RECALLED BY    Manufacturer, by letter February 24, 1993, followed by the
               program revisions on March 1, 1993.  Firm-initiated field
               correction complete.
DISTRIBUTION   South Carolina, Tennessee, Louisiana, Georgia, Mississippi.
QUANTITY       7 pieces.
REASON         Computer software, containing design defects which resulted
               in the release of unsuitable blood products, was distributed
               for use in blood banks.
______________
PRODUCT        Source Plasma.  Recall #B-428-4.
CODE           Unit numbers:  XE02549, XE02592, XE02710, XE03245, XE03327,
               XE03363, XE03809, XE03877, XE03948, XE04015, XE04349,
               XE04374, XE04570, XE04761, XE05103, XE05133, XE05211,
               XE05266, XE05325, XE05404, XE05675, XE05715, XE05788,
               XE05853, XE05916, XE06077, XE06120, XE06178, XE06221,
               XE09195, XE09264, XE09308, XE09394, XE09896, XE09959,
               XE10180, XE10243, XE10314, XE10665, XE10742, XE10838,
               XE10879, XE12010, XE12114, XE12239, XE12324, XE12460,
               XE12545, XE12741, XE12928, XE12999, XE13110, XE13181,
               XE13301, XE13371, XE13669, XE13743, XE02569, XE02794,
               XE05556, XE06025, XE06422, XE06869, XE07256, XE15802,
               XE16171, XE16259, XE16425, XE16647, XE16715, XE16918,
               XE17338, XE18641, XE20609, XE20933, XE20952, XE21054,
               XE21118, XE21140, XE21569, XE21650, XE21777, XE21846, 
               XE23240, XE23240, XE23522, XE23522, XE24333, XE24333, 
                                    -3-
               XE22183, XE22292, XE22393, XE22431, XE22673, XE23168,   
               XE24415, XE24415, XE24789, XE24961, XE25363, XE25790,
               XE26404, XE26607, XE26718, XE27407, XE27537, XE28433,
               XE29031, XE29074, XE29182, XE29226, XE29540, XE29769,
               XE29853, XE29952, XE30159, XE30513, XE30672, XE30827,
               XE31663, XE31746, XE32311, XE32428, XE32713, XE32824,
               XE32952, XE33406, XE33471, XE33703, XE34045, XE34149,
               XE34339, XE35034, XE38347, XE39043, XE39945, XE40240,
               XE40924, XE41655, XE41968, XE42054, XE42342, XE42595,
               XE42903, XE43058, XE43282, XE43292, XE43654, XE44164,
               XE44400, XE45115, XE45372, XE45700, XE46012, XE48467,
               XE70185, XE86637.
MANUFACTURER   Plasma Products of Virginia, Inc., Richmond, Virginia;
               Community Bio Resources, Inc., Richmond, Virginia.
RECALLED BY    Community Bio Resources, Inc., Birmingham, Alabama, by fax
               April 14, 1994, and by letter April 29, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Michigan, Austria.
QUANTITY       157 units.
REASON         Blood products which were non-reactive for hepatitis B
               surface antigen (HBsAg), but collected from donors who
               previously tested repeatedly reactive for HBsAg were
               distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-429-4.
CODE           Unit numbers:  PE48018, PE48121, PE48430, PE49639, PE48927,
               PE49121, PE49407, PE49568, PE49867, PE50032, PE50297,
               PE50444, PE50732, PE50930, PE51220, PE51395, PE51707,
               PE51979, PE52164, PE52381, PE53840, PE52935, PE53272,
               PE53612, PE54029, PE54467, PE55826.
MANUFACTURER   Community Bio Resources, Inc., Richmond, Virginia.
RECALLED BY    Community Bio Resources, Inc., Birmingham, Alabama, by
               letter April 26, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Austria.
QUANTITY       27 units.
REASON         Blood products collected from a donor who reported a history
               of behavior known to increase the risk for transmission of
               human immunodeficiency virus (HIV) were distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-430-4.
CODE           Unit numbers:  XC92459, XC92922, XC93387, XC93965, XC94382.
MANUFACTURER   Community Bio Resources, Inc., Richmond, Virginia.
RECALLED BY    Community Bio Resources, Inc., Birmingham, Alabama, by
               letter April 26, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Austria.
QUANTITY       5 units.
REASON         Blood products collected from a donor who had previously
               been deferred for suspected intravenous drug use were
               distributed.


                                    -4-
_______________
PRODUCT        Red Blood Cells.  Recall #B-435-4.
CODE           Unit #6730953.
MANUFACTURER   Blood Center of Southeast Texas, Beaumont, Texas.
RECALLED BY    Manufacturer, by telephone January 20, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood product collected from an ineligible donor due to ear
               piercing by a non-sterile method less than twelve months
               prior to donation was distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-434-4.
CODE           Unit #90-48037.
MANUFACTURER   Medic, Inc., Knoxville, Tennessee.
RECALLED BY    Manufacturer, by telephone October 21, 1993 with follow-up
               letter November 12, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Tennessee.
QUANTITY       1 unit.
REASON         Blood product that tested negative for the antibody to
               hepatitis B core antigen (anti-HBc) was collected from a
               donor who previously tested reactive for anti-HBc on two
               separate occasions was distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-436-4.
CODE           Unit numbers:  PE2466, PE2631, PE2759, PE2935, PE3078,
               PE3237, PE3401, PE3544, PE3658, PE3809, PE3929, PE4088,
               PE4209, PE4350, PE4445, PE4612, PE4704, PE4819, PE5204,
               PE5297, PE5352, PE5521, PE5619, PE5771, PE5913, PE6042,
               PE6192, PE6299, PE6458, PE6602, PE6715, PE7101, PE7245,
               PE7364, PE7647, PE7987, PE8133, PE8458, PE8607, PE9066,
               PE9301, PE11073, PE11157, PE11263, 0300145, 0300369,
               E4036443, E4036772, E4037104, E4039829, E4039992, E4040210,
               E4040401, E4040596, E4040751, E4040997, E4041174, E4041427,
               E4041610, E4041843, E4042007, E4042242, E4042390, E4042640,
               E4042815, E4043093, E4043265, E4043517, E4043668, E4044015,
               E4044259, E4044693, E4044913, E4045489, E4045619, E4045880,
               E4046047, E4046158, E4046407, E4046541, E4046779, E4046968,
               E4047201, E4047456, E4047666, E4047849, E4048099, E4048333,
               E4048623, E4048851, E4049419, E4049616, E4049805, E4050047,
               E4050267, E4050520, E4050735, E4050945.
MANUFACTURER   Plasma Products of Virginia, Inc., Richmond, Virginia.
RECALLED BY    Community Bio Resources, Inc., Birmingham, Alabama by letter
               April 26, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Austria.
QUANTITY       98 units.
REASON         Blood products collected from a donor who had previously
               been deferred for intravenous drug use were distributed.
                                    -5-_______________
PRODUCT        Red Blood Cells.  Recall #B-437-4.
CODE           Unit #128654.
MANUFACTURER   Blood Bank of Alaska, Inc., Anchorage, Alaska.
RECALLED BY    Manufacturer, by letter April 6, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Alaska.
QUANTITY       1 unit.
REASON         Blood product which tested repeatedly reactive for hepatitis
               B surface antigen (HBsAg) was distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-438-4.
CODE           Unit numbers:  0045247, 0045460, 0045725, 0045819, 0047192,
               0047474, 0052807, 0053105, 0064462, 0064917, 0064945,
               0065281, 0065282, 0065306, 0065622, 0066338, 0066339.
MANUFACTURER   Stillwater Plasma Center, Inc., Stillwater, Oklahoma.
RECALLED BY    Manufacturer, by letter December 7, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   New York and international.
QUANTITY       17 units.
REASON         Blood products, which either tested negative for the
               antibody to human immunodeficiency virus type 1 (anti-HIV-
               1), but were collected from a donor who previously tested
               repeatedly reactive for anti-HIV-1; or were collected from a
               donor who tested negative for anti-HIV-1, but who reported
               behavior that increased the risk of HIV infection, were
               distributed.

_______________
PRODUCT        Source Plasma. Recall #B-448-4.
CODE           Unit numbers:  XF79127, XF79391.
MANUFACTURER   Community Bio-Resources, Inc., New Orleans.
RECALLED BY    Manufacturer, by letter January 17, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Austria.
QUANTITY       2 units.
REASON         Blood product which was collected from a donor who had a
               history of intravenous drug (IV) use was distributed.

_______________
PRODUCT        Anti-Fyb Blood Grouping Reagent.  Recall #B-451-4.
CODE           Lot numbers FYB25A3 and FYB25B EXP 10/8/94.
MANUFACTURER   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
RECALLED BY    Manufacturer, by letter April 14, 1994, and by fax April 15,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       517 packages of lot FYB25A3 and 183 packages of lot FYB25B.
REASON         Anti-Fyb blood grouping reagent, exhibiting decreased
               reactivity with Fyb positive red blood cells was
               distributed.


                                    -6-
_______________
PRODUCT        Red Blood Cells.  Recall #B-452-4.
CODE           Unit #FR00467.
MANUFACTURER   Central Pennsylvania Blood Bank, Ephrata, Pennsylvania.
RECALLED BY    Central Pennsylvania Blood Bank, Hummelstown, Pennsylvania,
               by telephone January 25, 1994, followed by letter January
               26, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit.
REASON         Blood product, that tested negative for the antibodies to
               human immunodeficiency virus types 1 and 2 (anti-HIV-1/2),
               but was collected from a donor who previously tested
               repeatedly reactive for anti-HIV-1/2, was distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        LifeTec Community Blood Center Systems Software Version 2.0. 
               Recall #B-424-4.
CODE           Version 2.0.
MANUFACTURER   SysTec Computer Associates, Inc., Smithtown, New York.
RECALLED BY    Manufacturer, by letter February 24, 1993, followed by the
               program revisions on March 1, 1993.  Firm-initiated field
               correction complete.
DISTRIBUTION   Iowa.
QUANTITY       1 piece.
REASON         Computer software requiring subsequent modifications, was
               distributed for use in blood banks.

_______________
PRODUCT        Platelets.  Recall #B-439-4.
CODE           Unit #6180941.
MANUFACTURER   Carter Blood Center, Fort Worth, Texas.
RECALLED BY    Manufacturer, by telephone April 27, 1994, followed by
               letter.  Firm-initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood product corresponding to Red Blood Cells possibly
               contaminated with Propionibacterium acnes, was distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-442-4.
CODE           Unit numbers:  42FR25628, 42FR25636, 42FR25644.
MANUFACTURER   American Red Cross, Cleveland, Ohio.
RECALLED BY    American Red Cross, Charlotte, North Carolina, by telephone
               February 21, 1994, and letters dated February 22 and 24,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   North Carolina.
QUANTITY       3 units.
REASON         Blood products shipped under unacceptable temperature were
               distributed.


                                    -7-
RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Vas-Cath Soft-Cell Catheters, used for hemodialysis,
               apheresis, hemofiltratio and hemoperfusion access via the
               subclavian, jugular or femoral vein:  (a) Soft-Cell Lumen
               Catheters (straight): (b) Soft-Cell Dual Lumen Catheters
               (Curved); (c) Peritoneal Dialysis Catheters (Curved); (d)
               Peritoneal Dialysis Catheters (Straight).  
               Recall #Z-1243/1246-4.
CODE           (a) Catalog Numbers:  (a) PDLC-5512, PDLC-5519, PDLC-5523,
               lot numbers:  All lots below lot 152-### or with a sterility
               date earlier than 9/13/91.
               (b) Catalog numbers:  PDLC-5519PC, PDLC-5523PC; Lot numbers: 
               All lots below lot 246-### or with a sterility date earlier
               than 5/12/92.
               (c) Catalog numbers:  CCPD-11060, CCPD-11056, CCPD-12056,
               CCPD-12062; Lot numbers:  All lots below 210-## or with a
               sterility date earlier than 1/27/92;
               (d) Catalog numbers:  CPD-11031, CPD-11037, CPD-11042, CPD-
               11046, CPD-11541, CPD-11546, CPD-12031, CPD-12032, CPD-
               12037, CPD-12042, CPD-12047; Lot numbers:  All lots below
               lot 179-### or with a sterility date earlier than 11/19/91.
MANUFACTURER   Vas-Cath, Inc., Mississauga, Ontario, Canada.
RECALLED BY    Manufacturer, by telephone December 7, 1993, and by fax
               December 9, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Indiana, Texas, New Jersey, Kansas, Alabama, Illinois,
               Michigan, Nebraska, Utah, Washington state, California,
               Minnesota, Arizona.
QUANTITY       Soft-Cell Dual Lumen Catheters:
               PDLC-5512 -   400 units         PDLC-5519 -   4,290 units
               PDLC-5523 -   3,270 units       PDLC-5519PC - 2,740 units
               PDLC-5523PC - 1,745 units
               Peritoneal Dialysis Catheters:
               CCPD-11060 - 30 units           CCPD-11056 - 30 units
               CCPD-12056 - 95 units           CCPD-12062 - 30 units
               CPD-11031 -   5 units           CPD-11037 -  20 units
               CPD-11042 -   5 units           CPD-11046 - 5 units
               CPD-11541 - 30 units            CPD-11546 - 20 units
               CPD-12032 - 10 units            CPD-12037 - 15 units
               CPD-12042 - 50 units            CPD-12047 - 70 units.
REASON         The cuff may detach due to the breakdown of adhesive bonding
               after long term implantation.

_______________
PRODUCT        Multiple Arterial and Venous Bloodlines, Rx medical device
               products used in hemodialysis treatment:
               (a) Catalog #03-9252-2 - Arterial Bloodline;
               (b) Catalog #03-9253-0 - Arterial Bloodline;
               (c) Catalog #03-9254-8 - Arterial Bloodline;
               (d) Catalog #03-9265-4 - Arterial Bloodline;
               (e) Catalog #03-9452-8 - Venous Bloodline;
               (f) Catalog #03-9460-1 - Venous Bloodline;
               (g) Catalog #03-9463-5 - Venous Bloodline;
                                    -8-
               (h) Catalog #03-9464-3 - Venous Bloodline;
               (i) Catalog #03-9483-3 - Venous Bloodline.  
               Recall #Z-1249/1257-4.
CODE           Catalog Numbers                  Lot Numbers 
               03-9252-2                   M2C002, M2D001, M2D002 
               03-9252-2                   M2D003, M2L003 
               03-9253-0                   M2N002, M2N003 
               03-9254-8                   M2L004, M2L005 
               03-9265-4                   M2N005 
               03-9452-8                   M2K001, M2K002, M2K003 
               03-9452-8                   M2K004, M2K005, M2K006 
               03-9452-8                   M2K007, M2K008, M2K009 
               03-9452-8                   M2K010 
               03-9460-1                   M2S004 
               03-9463-5                   M3A004 
               03-9464-3                   M3A005, M3D003 
               03-9483-3                   M2L001, M2L002.
MANUFACTURER   National Medical Care, Medical Products Division, McAllen,
               Texas, and satellite plant at Reynosa, Mexico.
RECALLED BY    National Medical Care, Medical Products Division, Rockleigh,
               New Jersey (responsible firm), by letter December 8, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide and Ireland.
QUANTITY       285,564 devices were distributed; firm estimates that little
               product remains on the market.
REASON         Certain bloodlines were manufactured without sufficient
               documentation to assure compliance with device
               specifications and also exhibit a tendency to cause
               microbubbles.

_______________
PRODUCT        Ohmeda Vaporizers:  (a) Ohmeda Tec 6 (desflurane) Vaporizers
               for use with Ohmeda Anesthesia Systems; (b) Ohmeda Tec 6
               (desflurane) Vaporizers for use with North American Drager
               Anesthesia Systems (also known as the Tec 6 NAD Variant. 
               Recall #Z-1261/1262-4.
CODE           Serial numbers:  ACTV32001 to ACTV 51502, ACTW01001 to
               ACTW30093; (b) ACWW14001 to ACWW30288, ACWW31001 to
               ACWW51200, ACWX01001 to ACW19010.
MANUFACTURER   Ohmeda Medical Systems Division, Steeton, West Yorkshire,
               England.
RECALLED BY    Ohmeda Medical Systems Division, Madison, Wisconsin, by
               letters dated July 8, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       12,700 units were distributed.
REASON         Devices delivered higher concentrations of the anesthetic
               agent than indicated by the vaporizer setting.  In addition,
               the Tec 6 NAD Variant Vaporizer has developed a significant
               fresh gas leak when the vaporizer was turned off after use.

_______________
PRODUCT        Neonatal Monitoring System and Central Display Unit.
               Recall #Z-1273-4.
                                    -9-
CODE           All serial numbers.
MANUFACTURER   S&W Medico Teknik, Abyro, Denmark (responsible firm).
RECALLED BY    Air Shields, Inc., Hatboro, Pennsylvania, by letter June 2,
               1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Puerto Rico
QUANTITY       Approximately 40 units.
REASON         The 105 degree centigrade capacitor, when exposed to
               temperatures higher than it is rated, will dry out and lose
               its electrical capacity prematurely.  The capacitor failure
               will cause an interruption of the monitor network
               communication and loss of display functions.

_______________
PRODUCT        Hardware Trigger Board, Catalog #0005-1030, a component of
               the Abiomed BVS 5000 Bi-Ventricular Support System, which is
               an automated bi-ventricular support device intended to
               provide complete short term support of the left and/or right
               sides of the heart.  Recall #Z-1274-4.
CODE           1149, 1160, 1162, 1163, 1164, 1165, 1166, 1167, 1168, 1169,
               1170, 1171, 1172, 1173, 1174, 1175, 1176, 1177, 1178, 1179,
               1180, 1182, 1183, 1184, 1185, 1186, 1187, 1188, 1189, 1190,
               1191, 1192, 1193, 1194, 1195, 1196, 1197, 1198 (consoles
               located at the manufacturer at time of recall);
               1121, 1129, 1146, 1147, 1148, 1150, 1151, 1152, 1153, 1154,
               1155, 1156, 1157, 1158, 1161, 1181, 1159 (consoles in
               distribution).
MANUFACTURER   Electronic Associates, Inc., (EAI), West Long Branch, New
               Jersey.
RECALLED BY    Abiomed, Inc., Danvers, Massachusetts, by letter January 28,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   California, Wisconsin, Arizona, New York, Texas, Minnesota,
               Virginia, Louisiana., Kuwait.
QUANTITY       65 units.
REASON         When power to the unit is interrupted, the emergency system
               (battery backup) may not engage in the specified time.  The
               likely result of such an event would be temporary impairment
               of circulation.

_______________
PRODUCT        Digital I/O Board, part #0005-1050, a component of the
               Abiomed BVS 5000 Bi-Ventricular Support System Consoles. 
               The BVS 5000 Bi-Ventricular Support System is an automated
               bi-ventricular support device intended to provide complete
               short term support of the left and/or right sides of the
               heart.  Recall #Z-1275-4.
CODE           Serial numbers:  1144, 1149, 1162, 1163, 1169, 1171, 1172,
               1173, 1175, 1176, 1177, 1178, 1179, 1180, 1182, 1183, 1184,
               1185, 1186, 1187, 1188, 1189, 1190, 1191, 1192, 1193, 1195,
               1196 (units at manufacturer at time of recall);
               1139, 1140, 1141, 1142, 1143, 1145, 1146, 1147, 1148, 1150.
               1151, 1152, 1153, 1154, 1155, 1157, 1158, 1160, 1161, 1164,
               1165, 1166, 1167, 1168, 1170, 1181, 1159, 1174 (units
               distributed).
                                   -10-
MANUFACTURER   Electronic Associates Inc., (EAI), West Long Branch, New
               Jersey.
RECALLED BY    Abiomed, Inc., Danvers, Massachusetts, by letter February
               18, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Illinois, Georgia, Louisiana, Massachusetts, Minnesota, New
               Jersey, Ohio, Pennsylvania, Arkansas, California, New York,
               Texas, Virginia, Wisconsin, Liechtenstein, The Netherlands.
QUANTITY       57 units.
REASON         The potential exists for a capacitor at C3 on the Digital
               I/O board to be damaged by coming in contact with the
               circuit board.  If this occurred, there would be a shorting
               of the capacitor, an insulation split, and an eventual
               shutdown of the unit.

_______________
PRODUCT        Howmedica Luhr Fixation Systems Micro/Pan Screwdriver, used
               to attach plates and screws to the crainofacial area during
               surgery.  Recall #Z-1276-4.
CODE           Catalog #6518-2-410.
MANUFACTURER   Howmedica GMBH, Kiel, Germany.
RECALLED BY    Howmedica, Inc., Rutherford, New Jersey (responsible firm),
               by letter June 15, 1994, followed by telephone on or about
               July 15, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide
QUANTITY       286 units were distributed; firm estimates 276 units remain
               on the market.
REASON         The device contains gold electroplating which has been found
               to peel and flake under normal conditions of use.

_______________
PRODUCT        Custom Cataract Trays containing Laparotomy Sponges used to
               absorb blood and other body fluids during cataract eye
               surgery:  (a) 12" x 12" Lap Sponges, Catalog numbers
               CCESC01LAC, TSS01SL; (b) 18" x 18" Lap Sponges, Catalog
               numbers AVMC06LC, CCESC01LAC, GNWMH01LH, TRH01TT, TEC01TO.
               Recall #Z-1280/1281-4.
CODE           All lots.
MANUFACTURER   Sterile Design, Inc., Tampa, Florida.
RECALLED BY    Iolab, Inc., Claremont, California, by telephone November
               12, 1993.  Firm-initiated recall complete.
DISTRIBUTION   California, Hawaii, Texas, Louisiana, South Carolina,
               Oklahoma.
QUANTITY       8,698 trays were distributed; firm estimates none remains on
               the market.
REASON         The lap sponges were found to be non-sterile.

_______________
PRODUCT        Various single use custom sterile procedure kits and trays
               packaged and sterilized by Sterling Medical Products:
               1.   Laceration Trays - Catalog Nos. 661, STD-670, 675, 679,
                    786, 823-FH, 828, 923, 949, 956,  966; 
               2.   Suture Removal Trays - Reorder Nos. 101L, 116, 117;
                    Catalog Nos. STD-702, STD-706L, STD-713,  STD-718,
                                   -11-
                    STD-736, STD-741; 
               3.   Dressing Trays - Catalog Nos. 538, 755, 758, 779, 889A,
                    5091; 
               4.   Suture Trays - Catalog Nos. STD-621, STD-690, 985; 
               5.   Circumcision Trays - Catalog Nos. 901, 904; 
               6.   Towel Clamps, 2-1/2" - Catalog No. 526;   
               7.   Raytec Sponge Kit - Catalog No. 762; 
               8.   CVL Tubing Change Tray - Catalog No. 890; 
               9.   Renal Acute Care Cath Tray - Catalog No. 1573; 
               10.  Scissors, Sharp/Blunt - Catalog No. SS30; 
               11.  Hemostat, Kelly Curved - Catalog No. SS41; 
               12.  Baxter Sharp/Blunt Scissors, 5-1/2" - Catalog No.
                    35048-150 (distributed by Baxter Healthcare Corp.,
                    Hospital Supply Div.); 
               13.  Baxter Iris Scissors, Straight - Catalog No.  35048-045
                    (distributed by Baxter Healthcare Corp., Hospital
                    Supply Div.).   Recall #Z-1301/1313-4.
CODE           Sterilization/production lot 478.
MANUFACTURER   Sterling Medical Products, Prophetstown, Illinois.
RECALLED BY    Manufacturer, by telephone August 9, 1994, followed by
               letter.  Firm-initiated recall ongoing.
DISTRIBUTION   Illinois, Pennsylvania, New York, Massachusetts, Alabama,
               Virginia, Wisconsin, Texas, Arkansas.
QUANTITY       430 cases were distributed; firm estimates 50 percent
               remains on the market.
REASON         Sterility testing indicated the presence of mold in some of
               the devices.

_______________
PRODUCT        Portable Medical Oxygen Cylinders with Thermo Oxygen Valves. 
               Recall #Z-1315-4.
CODE           Lot numbers 5304 1/92 or 5304 4/92 embossed on the side of
               the valve.
MANUFACTURER   Thermo Valves, Inc., Santa Rosa, California (responsible
               firm).
RECALLED BY    Contemporary Products, Inc., Irvine, California, by letter
               May 11, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Illinois, Michigan, Wyoming.
QUANTITY       1,147 valves mated to cylinders and 35 valves were
               distributed.
REASON         The post valves have defective seating, resulting in leakage
               of the oxygen.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Triglycerides Reagent, used along with patient samples and
               controls to determine the triglyceride level in the blood
               using a spectrophotometer.  Recall #Z-1192-4.
CODE           Lot numbers:  J14NK (kit), GIN (14 ml fill bottle).
MANUFACTURER   Ciba Corning Diagnostics Corporation, Oberlin, Ohio.
RECALLED BY    Manufacturer, by letter January 11, 1993.  Firm-initiated
               recall ongoing.
                                   -12-
DISTRIBUTION   Nationwide and international.
QUANTITY       410 kits (20 14-ml bottles per kit) were distributed.
REASON         The wrong stopper was used with the bottle of reagent
               causing the reagent to become unstable.

_______________
PRODUCT        Probes used with Diagnostic Ultrasound BladderScan BVI 2500,
               used to scan a patient's bladder using ultrasound and to
               calculate the amount of urine in the bladder.  
               Recall #Z-1242-4.
CODE           Probe serial numbers: 04154, 10471, 10513, 10564.
MANUFACTURER   Diagnostic Ultrasound Corporation, Redmond, Washington.
RECALLED BY    Manufacturer, by telephone June 21, 1994, followed by letter
               June 23, 1994.  Firm-initiated recall complete.
DISTRIBUTION   New Jersey, California.
QUANTITY       4 probes were distributed.
REASON         Probes do not meet in-process test limits.

_______________
PRODUCT        Steinman Pin, Plain, Single Diamond Point, Plain End - 3/16"
               x 9", Catalog #5009115S, an orthopedic implant.
               Recall #Z-1314-4.
CODE           Lot #024307.
MANUFACTURER   Kirschner Medical Corporation, Fair Lawn, New Jersey.
RECALLED BY    Manufacturer, by telephone between April 11, 1994 and June
               22, 1994, and by letters April 18, 1994, May 2, 1994, and 
               May 17, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Virginia, North Carolina.
QUANTITY       80 pins were distributed; firm estimates none remains on
               market.
REASON         The product is incorrectly labeled as plain when, in fact,
               the Steinman Pin is threaded.

                                   ####


END OF ENFORCEMENT REPORT FOR AUGUST 31, 1994.  BLANK PAGES MAY
FOLLOW.