FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
06/29/1994

Recalls and Field Corrections:  Foods -- Class I -- 06/29/1994

June 29, 1994                                                 94-26
 
RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        Frookie All Natural Chocolate Chip Cookies, 6.5 ounces. 
               Recall #F-704-4.
CODE           Lot numbers:  31184P, 21194P, 31194P, 21254P, 31254P,
               11264P, 21314P, 31314P, 21374P, 11374P, 11444P, 11194.
MANUFACTURER   Imperial Baking, St. Louis, Missouri.
RECALLED BY    R.W. Frookies, Inc., Sag Harbor, New York, by letters of
               April 5, 1994, and May 23, 1994, and by press release April
               7, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Firm estimates that fewer than 3,500 cases (12 boxes per
               case) remain on market.
REASON         Product contains pecans which are not declared on the label.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Tops brand Non-Dairy Creamer, frozen, 32 ounces.  
               Recall #F-706-4.
CODE           4055 (C, D, G, or H).
MANUFACTURER   Rich Products Corporation, Buffalo, New York.
RECALLED BY    Manufacturer, by telephone May 19, 1994, followed by letter
               May 19, 1994.  Firm-initiated recall complete.
DISTRIBUTION   New York, Pennsylvania.
QUANTITY       671 cases (12 units per case) were distributed between
               February and May 1994.
REASON         Product contains undeclared sodium caseinate.

_______________
PRODUCT        Mint Chip Ice Cream, pure chocolate chips in mint flavored
               ice cream packaged in half gallon round paperboard tubs. 
               Recall #F-705-4.
CODE           Lot #07394.
MANUFACTURER   Gillette Dairy, Norfolk, Nebraska.
RECALLED BY    Manufacturer, by telephone May 11, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nebraska, Iowa, Missouri.
QUANTITY       Approximately 1,518 units were distributed.
REASON         Product contains undeclared FD&C Yellow #5.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Pitressin (Vasopressin Injection, USP) Steri-Vial 10 Pressor
               Units, in 0.5 ml, Synthetic 25/0.5 ml glass vials.  
               Recall #D-305-4.
CODE           Lot numbers:  022N2PN EXP 3/94, 03792PH EXP 1/94.
MANUFACTURER   Warner-Lambert Company, Parke-Davis Division, Rochester,
               Michigan.
RECALLED BY    Parke-Davis Division of Warner-Lambert Company, Morris
               Plains, New Jersey, by letter January 20, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       791 cartons of 25 vials of lot 022N2PN, 1,024 cartons of 25
               vials of lot 03792PH were distributed.
REASON         Portion of carton label bears incorrect route of
               administration.

_______________
PRODUCT        Procan SR Tablets (Procaine Hydrochloride Extended Release
               Tablets), indicated for the treatment of documented
               ventricular arrhythmias:  (a) 250 mg; (b) 500 mg; (c) 750
               mg.  Recall #D-312/314-4.
CODE           Lot numbers:  (a) 01872VA EXP 6/94, 01872VB EXP 6/94; (b)
               08442VB EXP 4/94, 04052VB EXP 4/94, 04142VB EXP 4/94,
               01972VA EXP 6/94, 10382VA EXP 7/94, 08092VA EXP 8/94,
               08192VA EXP 8/94, (c) 05852VA EXP 5/94, 08872VA EXP 7/94,
               07392VA EXP 8/94, 03523VA EXP 1/95, 03623VA EXP 1/95,
               03723VA EXP 1/95, 03823VA 1/95, 04353VA EXP 4/95, 04453VA
               EXP 4/95, 03873VA EXP 6/94, 04052V EXP 4/94, 05852V EXP
               5/94, 07392V EXP 8/94, 04453VA 4/95.  The following lots
               were shipped to Parke-Davis in Brockville, Canada for
               Canadian distribution:  08442V, 04052V, 10382V (500 mg);
               05852V, 03523V, 03623V (750 mg).

                                    -2-MANUFACTURER   Warner Lambert, Vega Baja, Puerto Rico.
RECALLED BY    Parke-Davis, Division of Warner Lambert, Morris Plain, New
               Jersey, by letter April 29, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       Firm estimates little product remains on the market.
REASON         Abbreviated New Drug Application deficiencies.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Parke-Davis Consumer Health Products Group brand Benadryl
               Tablets 25 mg, OTC antihistamine, in bottles of 100.  
               Recall #D-306-4.
CODE           Lot #09033V EXP 8/94.
MANUFACTURER   Warner-Lambert Company, Vega Baja, Puerto Rico.
RECALLED BY    Warner-Wellcome, Consumer Health Product, Morris Plains, New
               Jersey, by letter mailed week of April 25, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       23,808 bottles were distributed.
REASON         Subpotency.

_______________
PRODUCT        Methenamine Mandelate 1.0 Tablets, oral solid dosage form
               packaged in 100 and 1,000 tablet bottles, Rx antiseptic for
               urinary tract infections.  Recall #D-307-4.
CODE           3053A EXP 6/95, 3167A EXP 12/95.
MANUFACTURER   Amide Pharmaceutical, Inc., Little Falls, New Jersey.
RECALLED BY    Manufacturer, by letter April 7, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   New York, Florida, California, Indiana.
QUANTITY       48 bottles of 1000 and 288 bottles of 100 tablets were
               distributed.
REASON         Non-validated manufacturing process.

_______________
PRODUCT        Duragesic Fentanyl Transdermal System controlled substance
               Rx drugs: (a) 25 mcg/hr; (b) 50mcg/hr. Recall #D-308/309-4.
CODE           Lot numbers:  (a) 131305, 131306; (b) 132306, 132307.
MANUFACTURER   Alza Corporation, Vacaville, California. 
RECALLED BY    Janssen Pharmaceutica, Titusville, New Jersey, by letter on
               or about February 25, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       66,653 cartons of lot 131305, 19,061 cartons of lot 131306,
               35,230 cartons of lot 132306, and 15,159 cartons of lot
               132307 were distributed.
REASON         Release rate does not meet product specifications.

                                    -3-_______________
PRODUCT        Flurazepam Hydrochloride Capsules, USP, 30 mg, Rx schedule
               IV narcotic oral hypnotic for the treatment of insomnia.
               Recall #D-310-4.
CODE           Lot 4A154 EXP 7/95.
MANUFACTURER   Mylan Pharmaceuticals, Morgantown, West Virginia.
RECALLED BY    UDL Laboratories, Rockford, Illinois (repacker/responsible
               firm), by letter June 15, 1994.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       2,190 unit cartons were distributed.
REASON         Some unit cartons may be mislabeled as Temazepam Capsules 15
               mg.  The unit dose packs are correct.

_______________
PRODUCT        Tagamet (brand of Cimetidine) 800 mg, in bottles of 30,
               indicated for short-term treatment of active duodenal ulcer. 
               Recall #D-311-4.
CODE           Lot #8034T27 EXP 1/3/196.
MANUFACTURER   Smith Kline Beecham Pharmaceuticals, Cidra, Puerto Rico.
RECALLED BY    Manufacturer, by letter May 23, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       61,992 units were distributed.
REASON         A tablet of Relafen 750 mg was found in a bottle of Tagamet 
               800 mg.

_______________
PRODUCT        Miochol (Acetylcholine Chloride) 1:100 Intraocular, 10
               mg/2ml, in 2 ml vials, used to obtain complete miosis of the
               iris in seconds after delivery of the lens in cataract
               surgery.  Recall #D-315-4.
CODE           Lot numbers:  R4807D EXP 11/96 (Iocare Steri-Tags), R4807E
               EXP 11/96, R4808E EXP 12/96 (System Paks).
MANUFACTURER   Iolab, Inc., San German, Puerto Rico.
RECALLED BY    Iolab, Claremont, California, by letter May 20, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       12,225 units were distributed; firm estimates none remains
               on market.
REASON         Product does not meet pH specifications.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
               Recall #B-308/310-4.
CODE           Unit #22LF63110.
MANUFACTURER   The American National Red Cross, Philadelphia, Pennsylvania.
RECALLED BY    Manufacturer, by letter October 19, 1993.  Firm-initiated
               recall ongoing.

                                    -4-DISTRIBUTION   Pennsylvania, California.
QUANTITY       1 unit of each component.
REASON         Blood products collected from a donor taking Proscar were
               distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-339-4.
CODE           Unit #03LT10561.
MANUFACTURER   American Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by letter August 8, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia.
QUANTITY       1 unit.
REASON         Red Blood Cells, collected from a donor who traveled to an
               endemic area for malaria and had taken prophylactic anti-
               malarial medication prior to donation, was distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Cryoprecipitated
               AHF; (d) Fresh Frozen Plasma; (e) Platelet, Expired; (f)
               Recovered Plasma; (g) Recovered Human Serum (Pool).  
               Recall #B-341/347-4.
CODE           Unit numbers:  (a) 31G44005, 31GE01404, 31GE02485,
               31GE04447, 31T12697, 31T16663, 31T18895, 31T20598, 31T22927,
               31V32870, 31W73945; (b) 31G44005, 31GE04447, 31T12697,
               31T16663, 31T18895, 31T22927, 31V32870, 31W73945; (c)
               31G44005, 31T22927, 31GE01404; (d) 31T12697, 31T16663,
               31T18895, 31T20598, 31W73945; (e) 31T20598; (f) 31GE01404,
               31GE02485, 31GE04447, 31T22927, 31V35314; (g) 31GE01404
               (9RB1512-1-S), 31GE04447 (90RB03017), 31T16663 (92RB3602S),
               31T18895 (93RB3060S), 31T20598 (93RB3115S), 31T22927
               (93RB3182S), 31W73945 (92RB3523S).
MANUFACTURER   American Red Cross Blood Services, Buffalo, New York.
RECALLED BY    Manufacturer, by telephone from September 15, 1993 through
               December 7, 1993, and by letters dated from June 1, 1993,
               through December 12, 1993. Firm-initiated recall ongoing.
DISTRIBUTION   New York, Pennsylvania, California, South Carolina,
               Oklahoma, Alabama, Ohio, Georgia, Massachusetts,
               Switzerland.
QUANTITY       (a) 11 units; (b) 8 units; (c) 3 units; (d) 5 units; (e) 1
               unit; (f) 5 units; (g) 7 units.
REASON         Blood products which were collected from a donor with a
               history of cancer; or tested negative but were collected
               from donors who previously tested repeatedly reactive for
               the antibody to the human immunodeficiency virus type 1
               (anti-HIV-1), were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets, Expired; (c) Recovered
               Plasma.  Recall #B-350/352-4.
CODE           Unit numbers 1098037, 5070552.
MANUFACTURER   Northwest Florida Blood Center, Inc., Pensacola, Florida.

                                    -5-RECALLED BY    Manufacturer, (a) by letter May 25, 1993 and by telephone
               January 4, 1993; (b&c) telephone December 30, 1992, and by
               letter May 25, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Florida.
QUANTITY       2 units of each component.
REASON         Blood products, which tested negative for all viral markers,
               but were collected from donors who participated in high risk
               behavior or experienced an occupational needle stick
               exposure less than 12 months prior to donation, were
               distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Influenza Virus Vaccine Trivalent types A and B, Subvirion
               Formula 1993-1994, Connaught Flu Vaccine, in prefilled 0.05
               ml unit dose syringe.  Recall #B-338-4.
CODE           Lot #3M41109.
MANUFACTURER   Connaught Laboratories, Inc., Swiftwater, Pennsylvania.
RECALLED BY    Manufacturer, by letter March 29, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Argentina.
QUANTITY       30,000 units were distributed.
REASON         In translating labeling from English to Spanish, a portion
               of the indication section was omitted.

_______________
PRODUCT        (a) Granulocytes Apheresis; (b) Recovered Plasma.  
               Recall #B-348/349-4.
CODE           Unit numbers: (a) 31GJ04300; (b) 31G44005, 31V32870.
MANUFACTURER   American Red Cross Blood Services, Buffalo, New York.
RECALLED BY    Manufacturer, by telephone September 15, 1993, November 12,
               1993, and December 7, 1993, and by letters dated June 1,
               1993, September 17 and 24, 1993, November 22, 1993, and
               December 10, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   New York, California, Massachusetts, Switzerland.
QUANTITY       (a) 1 unit; (b) 2 units.
REASON         Blood products which were collected from a donor with a
               history of cancer or stored at an unacceptable temperature
               were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Prowess 2000 Radiation Treatment Planning Software Package,
               Version 2.08.  Recall #Z-740-4.
CODE           None.
MANUFACTURER   Small Systems Group, Inc., Chico, California.
RECALLED BY    Manufacturer, by letter January 13, 1993.  Firm-initiated
               field correction complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       96 units.
                                    -6-REASON         Calculations in the brachytherapy module produced dose
               errors of 48 percent.

_______________
PRODUCT        AC Power Adaptor or Transformer labeled "Component Power
               Supply Class 2, Part #394-4010-001, intended for use with
               medical and dental equipment.  Recall #Z-793-4.
CODE           All lots except revision E.
MANUFACTURER   Ault, Inc., Minneapolis, Minnesota.
RECALLED BY    CNS, Inc., Chanhassen Minnesota, by letter November 8, 1993. 
               Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide and canada.
QUANTITY       333 units were distributed.
REASON         Some adaptors exceeded leakage current specifications, due
               to loose fuses or fuses in close proximity to the
               transformer lamination stack.

_______________
PRODUCT        Sterile ophthalmic solutions for use with contact lenses:
               (a) Sherman Pharmaceutical Stay-Wet 3 preserved wetting and
               "in eye" lubricating solution and rewetting drop for use
               with fluoro/silicone crylate and silicone acrylate rigid gas
               permeable (RPG) contact lenses, .33 ounces (10 ml) sample
               size; 
               (b) Sherman Pharmaceutical Stay-Wet 3 preserved wetting and
               "in eye" lubricating solution and rewetting drop for use
               with fluoro/silicone acrylate and silicone acrylate RPG
               contact lenses, 1 fluid ounce (30 ml) size; 
               (c) Wukvision Stay-Wet 3 preserved wetting and "in eye"
               lubricating solution and rewetting drop for use with
               fluoro/silicone acrylate and silicone acrylate RPG contact
               lenses, 10 ml size; 
               (d) Sherman Pharmaceutical de Stat 3 cleaning, disinfecting
               and storage solution for use with fluoro/silicone acrylate
               and silicone acrylate RPG contact lenses, 1 fluid ounce (30
               ml) sample size; 
               (e) Sherman Pharmaceutical de Stat 3 cleaning, disinfecting
               and storage solution for use with fluoro/silicone acrylate
               and silicone acrylate RPG contact lenses, 4 fluid ounces
               (118 ml) size; 
               (f) Wukvision de Stat 3 cleaning, disinfecting and storage
               solution for use with fluoro/silicone acrylate and silicone
               acrylate RPG contact lenses, 118 ml size;
               (g) Wukvision de Stat 3 cleaning, disinfecting and storage
               solution for use with fluoro/silicone acrylate and silicone
               acrylate RPG contact lenses, 30 ml size; 
               (h) Sherman Pharmaceutical Sof/Pro-Clean cleaning solution
               for use with all soft (hydrophilic) contact lenses in heat
               (thermal) and chemical (not heat) lens care systems, 1 fluid
               ounce (30 ml) size; 

                                    -7-               (i) Choongwae Pharma Corporation Sof/Prp-Clean cleaning
               solution for use with all soft (hydrophilic) contact lenses
               in heat (thermal) and chemical (not heat) lens care systems,
               30 ml size.   Recall #Z-931/939-4.
CODE           Lot numbers:  (a) 1793; (b) 363; (c) 1793; (d) 3452, 3642;
               (e) 2332, 1603; (f) 1603; (g) 3642; (h) 1392, 1842, 1852;
               (i) 1842, 1852.
MANUFACTURER   Sherman Pharmaceuticals, Inc., Mandeville, Louisiana.
RECALLED BY    Manufacturer, by letter May 9, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 26,371 units of Stay-Wet 3; approximately
               45,906 units of de-Stat; and approximately 63,786 units
               Sof/Pro-Clean from all lot numbers involved were
               distributed.
REASON         Product sterility may have been compromised due to potential
               inadequate sterilization.

_______________
PRODUCT        IVAC Core-Check Tympanic Thermometer System, Model 2090A, a
               hand-held thermometer which, when positioned in the external
               auditory canal, measures the infrared energy emitted from
               the tympanic membrane.  Recall #Z-963-4.
CODE           All lot numbers.
MANUFACTURER   IVAC Corporation, San Diego, California.
RECALLED BY    Manufacturer, by issuing a service bulletin in February
               1992, and by letter February 14, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       11,518 units were distributed from August 1991 to October
               1992; firm estimated 10,255 units remained on market at time
               of recall.
REASON         Inadequate shielding allowing high levels of electromagnetic
               or radio interference to possibly cause device malfunction.

_______________
PRODUCT        IPI Respiratory Therapy Fluid Transfer Sets with Luer Tip
               and Spring Loaded Clamp, Catalog #1122, a sterile
               prescription device used to transfer respiratory therapy
               solutions to the desired respiratory device.  
               Recall #Z-965-4.
CODE           Lot #D040.
MANUFACTURER   IPI Medical Products, Chicago, Illinois.
RECALLED BY    Manufacturer, by telephone March 10, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Illinois, Ohio, North Carolina, Massachusetts, Arizona,
               Georgia.
QUANTITY       37 cases were distributed; firm estimated that 7 cases
               remained on market at time of recall.
REASON         The clamp on the sets were glued in the open position.

                                    -8-_______________
PRODUCT        MDE Defibrillator Model 19E and MDE Defibrillator with Pacer
               Model 20E, options used with the Escort 300 Series full
               parameter transport monitor.  Both options are low energy DC
               defibrillators which are able to deliver an electrical shock
               with a range of 2 to 360 joules of energy.  
               Recall #Z-968/969-4.
CODE           Serial numbers 1198 through 1307.
MANUFACTURER   Medical Data Electronics, Inc., Arleta, California.
RECALLED BY    Manufacturer, by letter December 2, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Canada, Hong Kong, Korea, the Philippines.
QUANTITY       110 units were distributed.
REASON         The energy selector switch can be set between the detented
               increments of the switch causing higher than selected energy
               to be delivered.

_______________
PRODUCT        Hot Choice Microwave Oven, Model 122000, used in heating
               frozen prepared pizza or chicken using convection/microwave
               energy.  Recall #Z-978-4.
CODE           Serial numbers:  Undetermined.
MANUFACTURER   Kaiser Electroprecision, Irvine, California.
RECALLED BY    Manufacturer.  FDA Approved the firm's corrective action
               plan June 16, 1994.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   California.
QUANTITY       30 units.
REASON         Product failed to meet the certification labeling
               requirements.

_______________
PRODUCT        First Medic Defibrillators (a) Model 90510 (FM 510); (b)
               Model 90526 (FM 610).  Recall #Z-979/980-4.
CODE           Serial numbers:  (a) 510B00312 - 510B00458, 510C00314 -
               510C01074; (b) 526B01865 - 526B01915, 526C01681 - 526C02426.
MANUFACTURER   SpaceLabs Medical, Inc., Redmond, Washington.
RECALLED BY    Manufacturer, by letter April 26, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 610 units; (b) 483 units were distributed.
REASON         The material (tin) used for the plating of the isolation
               module shield is susceptible to metal fiber growth that, if
               oriented in a certain fashion, can make contact with the
               connector pins and result in a small electrical short
               circuit condition for the device.

_______________
PRODUCT        Mevatron Linear Accelerator and Mevasim Therapy Simulator -
               Overhead Suspension Hand Control.  Recall #Z-981-4.
CODE           Overhead Suspension Assembly-A12, Part #94 10 309.  Used in
               conjunction with all conventional and digital Mevatron
               Linear Accelerators and Mevasim Therapy Simulators.
                                    -9-MANUFACTURER   Siemens Medical Systems, Inc., Concord, California.
RECALLED BY    Manufacturer, by letter October 29, 1994.  Firm-initiated
               field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,141 units.
REASON         The devices were distributed with a defective weld that can
               fail and cause the overhead support plate (which weighs 45
               pounds) to fall.

_______________
PRODUCT        Mevatron Linear Accelerator Wedge Filter Assembly, used with
               the Mevatron Therapeutic X-ray Machine.  Recall #Z-982-4.
CODE           All codes.
MANUFACTURER   Siemens Medical Systems, Inc., Concord, California.
RECALLED BY    Manufacturer, by letter October 29, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 4,000 wedges were distributed 1976-1993.
REASON         The devices were distributed with defective connection
               screws between the wedge filters and the aluminum trays
               which could allow the wedge to fall from the mounting tray.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Calibration Gas Mixture, in UD size cylinders, used as a
               calibration or check standard for anesthesia monitoring. 
               Recall #Z-942-4.
CODE           Lot #131403 EXP 5/95.
MANUFACTURER   Scott Medical Products, Plumsteadville, Pennsylvania.
RECALLED BY    Manufacturer, by telephone April 22 and 27, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Texas, Indiana, Iowa, California.
QUANTITY       5 cylinders were distributed.
REASON         The product which is 30% oxygen is incorrectly labeled as
               containing 21% oxygen. 

_______________
PRODUCT        Cynomologus Monkey Kidney tubes, Catalog #70-177-D, used for
               the isolation of virus in clinical specimens.  
               Recall #Z-950-4.
CODE           Lot #2C0087.
MANUFACTURER   Whittaker Bioproducts, Inc., Walkersville, Maryland.
RECALLED BY    Manufacturer, by telephone January 20, 1992.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       125 units were distributed.
REASON         Product was contaminated with gram positive Bacillus
               species.

                                   -10-_______________
PRODUCT        Rhesus Monkey Kidney Tubes with SV5 and SV40 Antiserum cell
               cultures which is used for isolation of virus in clinical
               specimens.  Recall #Z-951-4.
CODE           Catalog #70-103D, Lot #2C0449.
MANUFACTURER   Whittaker Bioproducts, Inc., Walkersville, Maryland.
RECALLED BY    Manufacturer, by telephone March 4, 1992.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,474 units were distributed; firm estimates none remains on
               the market.
REASON         Product was contaminated with Micrococcus, Bacillus, and
               Coryneform bacteria.

_______________
PRODUCT        Skull Mount Kit A-1330, consisting of sterile gloves, PVP
               solutions, small prep tray, razor, sponge gauze, towels, CSR
               wraps, needles, zylocaine, syringe, scalpel, drape, tray,
               skull ring, bone screws, screw driver, drill bit, and hand
               drill.  Recall #Z-964-4.
CODE           Sterile lot #22063A, Kit lot #38027893.
MANUFACTURER   Ohio Medical Instruments, Cincinnati, Ohio.
RECALLED BY    Manufacturer, by telephone December 8, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Idaho.
QUANTITY       1 kit was distributed.
REASON         The product had no sterilization validation.

_______________
PRODUCT        Human Neonatal Kidney (HNK) Cell Cultures in tubes, 50 per
               rack, living cell cultures for viral isolation procedures in
               clinical virology labs.  Recall #Z-973-4.
CODE           Catalog #70-151-D, Lot #1C3503, 1C3493.
MANUFACTURER   Whittaker Bioproducts, Inc., Walkersville, Maryland.
RECALLED BY    Manufacturer, by telephone December 18, 1991.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       3,380 units were distributed; firm estimates none remains on
               the market.
REASON         The device was contaminated with yeast.

_______________
PRODUCT        Human Neonatal Kidney (HNK) Cell Cultures in tubes, 50 per
               rack, living cell cultures for viral isolation procedures in
               clinical virology labs.  Recall #Z-974-4.
CODE           Catalog #70-151-D, Lot #1C3568.
MANUFACTURER   Whittaker Bioproducts, Inc., Walkersville, Maryland.
RECALLED BY    Manufacturer, by telephone December 27, 1991.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.

                                   -11-QUANTITY       1,995 units were distributed; firm estimates none remains on
               the market.
REASON         Device was contaminated with bacteria.

_______________
PRODUCT        SeeQuence 2/Medalist Trial Contact Lenses, Not For Sale,
               intended for use as trial lenses only, and remain on the
               premises of the eye care practitioner.  Recall #Z-977-4.
CODE           Lot #I092336.
MANUFACTURER   Bausch & Lomb, Inc., Contact Lens Division, Rochester, 
               New York.
RECALLED BY    Manufacturer, by letter April 14, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       195 cartons were distributed between 12/20/93 and 3/25/94.
REASON         While the blister packs are correctly labeled as -4.50
               diopter, the cartons are mislabeled as -4.75 diopter.

_______________
UPDATE         The catalog number and lot number for Baxter's Stratus
               Ultra-Sensitive hTsH Enzyme Immunoassay Kits, 
               Recall #Z-893-4, should be Catalog No. B5700-44, Kit lots
               Nos. KXTS-224M through KXTS-252M except for Lots KXTS-229M
               and KXTS-252M and not catalog B5700, Kit lots KXS-24M
               through KXTS-322M as erroneously stated in the June 15, 1994
               Enforcement Report. 

                                   -12-


END OF ENFORCEMENT REPORT FOR JUNE 29, 1994.  BLANK PAGES MAY FOLLOW.
                                   ####