06/22/1994

Recalls and Field Corrections:  Foods -- Class II -- 06/22/1994

June 22, 1994                                                   94-25

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Vallone's Assorted Butter Style Cookies, in 1 pound, 2
               pound, and 5 pound boxes.  Recall #F-421-4.
CODE           Products with expiration dates from 12/93 through 5/94.
MANUFACTURER   Vallone's, Inc., Carteret, New Jersey.
RECALLED BY    Manufacturer, by letters in mid-November 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   New Jersey, Florida.
QUANTITY       4,000 pounds.
REASON         Product contains undeclared FD&C Yellow No. 5.

_______________
PRODUCT        Wheaties Honey Gold Wholesome, Sweetened Whole Wheat & Corn
               Cereal With Honey, 21.25 ounces.  Recall #F-426-4.
CODE           Better If Used By date of 21FEB95QQ____.
MANUFACTURER   General Mills, Inc., Covington, Georgia.
RECALLED BY    General Mills, Inc., Minneapolis, Minnesota, by electronic
               mail May 23, 1994, by press release May 20, 1994, and by
               issuing a notice to the Food Allergy Network May 23, 1994. 
               Firm-initiated recall ongoing. 
DISTRIBUTION   Alaska, California, Idaho, Montana, Oregon, Washington
               state.
QUANTITY       1,128 cases (14 boxes per case).
REASON         Product contains undeclared almonds._______________
PRODUCT        Ice Cream in 3 gallon bulk containers:  (a) Rocky Road; (b)
               Heavenly Hash.  Recall #F-427/428-4.
CODE           None other than the flavor of the ice cream written or
               stamped on container.
MANUFACTURER   Taylor Milk Company, Ambridge, Pennsylvania.
RECALLED BY    Tri-Point Ice Cream Sales Company, Ambridge, Pennsylvania,
               by letter March 25, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania, West Virginia, Ohio.
QUANTITY       Undetermined.
REASON         Products contained undeclared almonds.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Ice Cream in 3 gallon bulk containers:  (a) Butter Pecan;
               (b) Praline Pecan.  Recall #F-424/425-4.
CODE           None other than the flavor of the ice cream written or
               stamped on container.
MANUFACTURER   Taylor Milk Company, Ambridge, Pennsylvania.
RECALLED BY    Tri-Point Ice Cream Sales Company, Ambridge, Pennsylvania,
               by letter dated March 25, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Pennsylvania, West Virginia, Ohio.
QUANTITY       Undetermined.
REASON         Products contain undeclared pecans.

_______________
PRODUCT        Pasta products:  (a) Luigi Vitelli #9 Thin Spaghetti;
               (b) Luigi Vitelli #17 Linguine, unit weight 16 ounces. 
               Recall #F-422/423-4.
CODE           Lot numbers:  (a) 466352; (b) 466395.
MANUFACTURER   Maktas Makarnacilik, Izmir, Turkey.
RECALLED BY    Vitelli-Elvea Company, Inc., Hawthorne, New Jersey, by
               letter issued mid-December 1993.  Firm-initiated recall
               complete.
DISTRIBUTION   New Jersey, New York.
QUANTITY       Firm estimates none remains on the market.
REASON         Products are substandard for declared iron.

_______________
PRODUCT        Toasted Almond Fudge Ice Cream in 3 gallon containers. 
               Recall #F-429-4.
CODE           None other than the flavor of the contents written or
               stamped on the container.
MANUFACTURER   Taylor Milk Company, Ambridge, Pennsylvania.
RECALLED BY    Tri-Point Ice Cream Sales Company, Ambridge, Pennsylvania,
               by letter dated March 24, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Pennsylvania, West Virginia, Ohio.
QUANTITY       Undetermined.
REASON         Product contained undeclared almonds.

                                    -2-_______________
PRODUCT        Pistachio Almond Ice Cream, in 3 gallon bulk containers.  
               Recall #F-430-4.
CODE           None other than the flavor of the ice cream written or
               stamped on container.
MANUFACTURER   Taylor Milk Company, Ambridge, Pennsylvania.
RECALLED BY    Tri-Point Ice Cream Sales Company, Ambridge, Pennsylvania,
               by letter dated March 25, 1994.  Firm-Initiated recall
               complete.
DISTRIBUTION   Pennsylvania, West Virginia, Ohio.
QUANTITY       Undetermined.
REASON         Product contained undeclared Pistachios.

_______________
PRODUCT        Kern's Real Fruit Sodas, packed in 12 fluid ounce cans:  (a)
               Apricot Soda; (b) Guava Soda; (c) Mango-Orange Soda; (d)
               Peach-Passion Fruit Soda; (e) Strawberry-Kiwi Soda; (f)
               Tropical Soda.  Recall #F-431/436-4.
CODE           Although the contaminated cans were only noted in the codes
               indicated by an *, all of the following codes were recalled:
               Apricot Soda (Item 28500): 4035 LA, 4082 NA, 4083 NA, 4088
               NA, 4102 NA*, and 4107 NA*;
               Mango-Orange Soda (Item 28530): 4035 LA, 4085 NA, 4104 NA*,
               and 4105 NA*;
               Guava Soda (Item 28520): 4035 LA, 4084 NA, 4085 NA, 4099
               NA*, 4100 NA*, and 4101 NA*; 
               Strawberry-Kiwi Soda (Item 28550): 4035 LA, 4084 NA, 4087
               NA, 4088 NA, and 4106 NA*; 
               Peach-Passion Fruit (Item 28510): 4035 LA, 4083 NA, 4103 NA,
               and 4107 NA;
               Tropical Soda (Item 28560): 4035 LA, 4086 NA, and 4105 NA. 
MANUFACTURER   Arrowhead Mountain Spring Water Company, Ontario,
               California.
RECALLED BY    Nestle Beverage Company, San Francisco, California, by
               letter April 21, 1994.  Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       Approximately 15,145 cans were distributed; firm estimates
               none remains on the market.
REASON         Product was contaminated with mold.

_______________
PRODUCT        Plochman's Mild Yellow Mustard in 10.5 ounce yellow plastic
               squeeze bottles.  Recall #F-437-4.
CODE           Lot numbers 216C3 and 242C3.
MANUFACTURER   Plochman, Inc., Chicago, Illinois.
RECALLED BY    Manufacturer, by telephone April 15, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Massachusetts, Georgia, Illinois, South Carolina, Missouri,
               Michigan, Wisconsin, New York.
QUANTITY       3,566 cases (24 bottles per case) were distributed; firm
               estimates none remains on the market.
REASON         Product contained mold.

                                    -3-RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Various drug products:  (a) Acetic Acid (2%) & Aluminum
               Acetate Otic Solution, 60 ml, under the following labels: 
               Bausch & Lomb, H.L. Moore, Mayor, Rugby; (b) Neomycin &
               Polymyxin B Sulfates and Hydrocortisone Otic Solution, 10
               ml, under the following labels:  Bausch & Lomb, Akorn, Best,
               Butler, Clinical, Equipharm, Goldline, Harber, J.J. Balan,
               Major, OcuSoft, Parmed, Pedinol, Qualitest, Rugby, Teral,
               URL; (c) Neomycin Polymyxin B Sulfates and Gramicidin
               Ophthalmic Solution, 2 ml, and 10 ml, under the following
               labels:  Bausch & Lomb, Akorn, Best, Butler, Equipharm,
               Genetic, Goldline, Harber, H.L. Moore, J.J. Balan, Major,
               Medical Ophthalmics, OcuSoft, Parmed, Phoenix, Qualitest,
               Rugby, Schein, Spectrum, URL, Vedco; (d) Erythromycin
               Ethylsuccinate Oral Suspension, 200 mg/5 ml, 100 ml, under
               the following labels:  Bausch & Lomb, Major, Qualitest,
               Rugby, Schein, URL.  Recall #D-300/303-4.
CODE           All lots with four digit lot numbers.
MANUFACTURER   Pharmafair, Inc., Hauppauge, New York.
RECALLED BY    Pharmafair, Inc., c/o BLP, (Bausch & Lomb Pharmaceuticals),
               Tampa, Florida, by letters of April 22, 1994 and May 25,
               1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       5,335,745 total units were distributed; firm estimated that
               50% of this number remained on market at time of recall.
REASON         Products were marketed with inadequate documentation to
               support Abbreviated New Drug Applications.

_______________
PRODUCT        Liqui-Prep Bowel Evacuant Kit with Magnesium Citrate Oral
               Solution, USP, in 10 fluid ounce containers,
               nonprescription, for use in preparation of the patient for
               surgery, or for preparation of the colon for x-ray and
               endoscopic examination.  Recall #D-304-4.
CODE           Catalog #3066, lot numbers:  08243066A, 08743066, 08743066A,
               08843066, 08943066, 08143066A.
MANUFACTURER   Cumberland-Swan, Inc., Smyrna, Tennessee (Magnesium Citrate
               Oral Solution); E-Z-EM, Inc., Westbury, New York (kit).
RECALLED BY    E-Z-EM, Inc., Westbury, New York (repacker), by letter April
               20, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       500 cases (24 bottles per case) were distributed.
REASON         Some bottles of the Magnesium Citrate Solution Kit component
               contained glass particles.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        Dog Fennel Allergenic Extract (Eupatorium capillifolium), in
               10 ml, 30 ml, and 50 ml vials.  Recall #B-312-4.
CODE           Lot #F13083193 EXP 8/30/99.

                                    -4-MANUFACTURER   Antigen Laboratories, Inc., Liberty, Missouri.
RECALLED BY    Manufacturer, by telephone January 7, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia, South Carolina, Florida, Washington state.
QUANTITY       2 10-ml vials, 1 30-ml vial, and 2 50-ml vials.
REASON         Allergenic extracts which were labeled with incorrect
               species name were distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-318-4.
CODE           Unit numbers 82297744 and 82298208.
MANUFACTURER   North American Biologicals, Inc., doing business as Grand
               River Donor Center, Detroit, Michigan.
RECALLED BY    North American Biologicals, Inc., Miami Florida, by letter
               September 28, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Puerto Rico.
QUANTITY       2 units.
REASON         Blood products, which tested negative for the antibodies to
               human, immunodeficiency virus types 1 and 2 (anti-HIV-1/2),
               but were collected from a donor who previously tested
               repeatedly reactive for anti-HIV-1/2, were distributed.

_______________
PRODUCT        (a) Red Blood Cells for Further Manufacturer; (b) Recovered
               Plasma.  Recall #B-319/320-4.
CODE           Unit #03R69097.
MANUFACTURER   The American National Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by letters of November 5 and 15, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois, Switzerland.
QUANTITY       1 unit of each component.
REASON         Blood products, that tested negative for hepatitis B surface
               antigen (HBsAg), but were collected from a donor who
               previously tested repeatedly reactive for HBsAg, were
               distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen Plasma;
               (d) Platelets for Further Manufacture.  Recall #B-321/324-4.
CODE           Unit numbers:  (a) 03C53984, 03G47125; (b) 03G47125; (c)
               03C53984; (d) 03C53984.
MANUFACTURER   The American National Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by letter December 22, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia, Florida.
QUANTITY       (a) 2 units; (b) 1 unit; (c) 1 unit; (d) 1 unit.
REASON         Blood products collected from a donor who reported a history
               of hepatitis were distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-325-4.
CODE           Unit 158427.

                                    -5-MANUFACTURER   Beth Israel Hospital, Boston, Massachusetts.
RECALLED BY    Manufacturer, by telephone, fax, and letter April 14, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Massachusetts.
QUANTITY       1 unit.
REASON         Blood product, which tested negative for the antibody to
               hepatitis C virus encoded antigen (anti-HCV), but was
               collected from a donor who previously tested repeatedly
               reactive for anti-HCV, was distributed.

_______________
PRODUCT        Fresh Frozen Plasma.  Recall #B-326-4.
CODE           Unit #22GG51332.
MANUFACTURER   The American National Red Cross, Philadelphia, Pennsylvania.
RECALLED BY    Manufacturer, by telephone January 14, 1993, followed by
               letter January 26, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit.
REASON         Blood product, containing anti-D (Rho) antibodies, but
               labeled as negative for unexpected antibodies, was
               distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen Plasma. 
               Recall #B-330/332-4.
CODE           Unit numbers:  (a&b) 03FR02035, 03G25343, 03LM18668,
               03R81193; (c) 03LM18668.
MANUFACTURER   The American National Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by letter August 6, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia.
QUANTITY       (a) 4 units; (b) 4 units; (c) 1 unit.
REASON         Blood products, that tested negative for the antibody to the
               human immunodeficiency virus type 1 (anti-HIV-1), but were
               collected from a donor who previously tested repeatedly
               reactive for anti-HIV-1, were distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-333-4.
CODE           Unit #11471935.
MANUFACTURER   North American Biologicals, Inc., Kansas City, Missouri.
RECALLED BY    Manufacturer, by telephone January 7, 1993, followed by
               letter January 12, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Source Plasma, that tested repeatedly reactive for the
               antibody to hepatitis C virus encoded antigen (anti-HCV),
               was distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-334-4.
CODE           Unit #11N75587.

                                    -6-MANUFACTURER   The American National Red Cross, St. Louis, Missouri.
RECALLED BY    Manufacturer, by telephone December 2, 1992, followed by
               letter December 3, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Missouri.
QUANTITY       1 unit.
REASON         Blood product, that tested negative for the antibody to
               hepatitis B core antigen (anti-HBc), but was collected from
               a donor who previously tested repeatedly reactive for anti-
               HBc on more than one occasion, was distributed.

_______________
PRODUCT        Source Plasma.  Recall #B-335-4.
CODE           C-10176-691     C-10232-691    C-10432-691     C-10561-691  
               C-10653-691     C-10777-691    C-10867-691     C-11007-691  
               C-11143-691     C-11333-691    C-11411-691     C-11600-691  
               C-11691-691     C-11808-691    C-11893-691     C-12064-691  
               C-12161-691     C-12392-691    C-12514-691     C-12762-691  
               C-12866-691     C-13093-691    C-13095-691     C-13189-691  
               C-13309-691     C-13341-691    C-13403-691     C-13447-691.
MANUFACTURER   Premier Bio Resources, Inc., Gulfport, Mississippi.
RECALLED BY    Premier BioResources, Inc., Fort Worth, Texas, by letters
               dated March 4 and 7, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY       28 units.
REASON         Source Plasma, that either tested negative for the
               antibodies to human immunodeficiency virus types 1 and 2
               (anti-HIV-1/2), but were collected from a donor who
               previously tested repeatedly reactive for anti-HIV-1/2; or
               tested negative for hepatitis B surface antigen (HBsAg) and
               the antibody to hepatitis C virus encoded antigen (anti-
               HCV), but were collected from a donor who previously
               reported a history of hepatitis, were distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Cryoprecipitated AHF (Human).  Recall #B-315-4.
CODE           Unit numbers:  22GL96498, 22GL96511, 22GT51314, 22GT51316,
               22GX11292, 22GY43319, 22LV14533, 22LV14540, 22LQ26362.
MANUFACTURER   American Red Cross, Philadelphia, Pennsylvania.
RECALLED BY    Manufacturer, by telephone May 13, 1993, followed by letter
               May 21, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania, New Jersey.
QUANTITY       9 units.
REASON         Units of Cryoprecipitated AHF (Human), exposed to
               unacceptable temperature and allowed to thaw, were
               distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-327/328-4.
CODE           Unit numbers:  (a) 03FF21084; (b) 03FF21084.
MANUFACTURER   The American National Red Cross, Atlanta, Georgia.

                                    -7-RECALLED BY    Manufacturer, by telephone July 2, 1993, followed by letter
               July 30, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Georgia.
QUANTITY       1 unit of each component.
REASON         Blood products, untested for hepatitis B surface antigen
               (HBsAg), were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-329-4.
CODE           Unit #03K11304.
MANUFACTURER   The American National Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by telephone December 4, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia.
QUANTITY       1 unit.
REASON         Blood product, labeled with the incorrect expiration date,
               was distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-336-4.
CODE           Unit #03K01387.
MANUFACTURER   American Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by letter January 14, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Recovered Plasma, corresponding to a unit of Red Blood Cells
               contaminated with Alpha Streptococcus, was distributed.

_______________
PRODUCT        Cryoprecipitated AHF.  Recall #B-337-4.
CODE           Unit numbers:  11K64979, 11K65346, 11F86180.
MANUFACTURER   The American National Red Cross, St. Louis, Missouri.
RECALLED BY    Manufacturer, telephone December 2, 1992, followed by letter
               January 4, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Missouri.
QUANTITY       3 units.
REASON         Blood product, which was exposed to unacceptable storage
               temperatures and allowed to thaw, was distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-340-4.
CODE           Unit #03FT11965.
MANUFACTURER   The American National Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by letter March 3, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product collected from a donor who reported a history
               of cancer was distributed.


                                    -8-
RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Glucose Test Strips for use with Glucometer 3, for in-vitro
               (external) diagnostic use, under the following labels: 
               Family Pharmacy, Health Mart, Full Value, Good Neighbor
               Pharmacy, Valu Rite, Relief Plus, Quick Check 3, Perry
               Healthcare, and Longs.  Recall #Z-930-4.
CODE           Lot Numbers:  30152, 30173, 30187, 30190B, 30193, 30216,
               30228, 30238, 30242, 30244, 30251A, 30251B, 30259, 30263,
               30264, 30266, 30270, 30272, 30276, 30278, 30280, 30284,
               30285, 30293, 30294, 30300, 30301A, 30301B, 30303, 30310,
               30313, 30314, 30319, 30320, 30324, 30333, 30336, and 30356; 
MANUFACTURER   Diagnostic Solutions, Inc., Irvine, California.
RECALLED BY    Manufacturer, by letter April 4, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       37,458 bottles were distributed from 7/7/93 to 1/14/94.
REASON         The product does not meet its labeling claims.  The device
               also gives results which are significantly higher than those
               obtained by the reference method.

_______________
PRODUCT        Radiofocus Introducer B Kit, intended to be inserted
               percutaneously into a vein or artery to facilitate the
               insertion of angiographic, electrode, balloon or similar
               catheters.  Recall #Z-962-4.
CODE           8893I20A, 8893I20, 8893I23, 8893I25, 8893I25A.
MANUFACTURER   Terumo Corporation, Toyko, Japan.
RECALLED BY    Terumo Medical Corporation, Elkton, Maryland, by letters
               sent March 11, 1994, and April 6, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Massachusetts, Canada.
QUANTITY       16,600 kits were distributed.
REASON         The hemostasis valve within the introducer device may leak,
               depending upon the orientation (positioning) of the valve.

_______________
PRODUCT        Model 3000LE Ophthalmic Laser System, for use in ophthalmic
               surgery - posterior capsulotomies.  Recall #Z-966-4.
CODE           Model 3000LE.
MANUFACTURER   Alcon Surgical Inc., Irvine, California.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action
               plan May 20, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       56 units.
REASON         Joystick aiming control was poorly seated causing internal
               electrical wires to become prematurely worn and broken.


                                    -9-RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Simmons Plating System Locking Nut, Catalog #SP-1003, non-
               sterile, intended for sterilization by hospital, used in
               lumbar spinal fusions.  Recall #Z-940-4.
CODE           All lots.
MANUFACTURER   Smith & Nephew Richards, Inc., Memphis, Tennessee.
RECALLED BY    Manufacturer, by voice mail message March 2, 1994, followed
               by letter March 8, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       20,534 locking nuts were distributed from October 1990 until
               January 1994.
REASON         The nuts were manufactured from a type of 300 stainless
               steel other than type 316LVM (ASTM F138).

_______________
PRODUCT        Automated Chemiluminiscene System (ACS):Cuvette Part
               #672002, used with the ACS:180.  A cuvette is a tall plastic
               container which carries a patient sample and to which test
               reagents are dispensed where a test reaction takes place. 
               Recall #Z-943-4.
CODE           Lot numbers D01600 - D02175.
MANUFACTURER   Carrera Corporation, Export, Pennsylvania.
RECALLED BY    Ciba Corning Diagnostics Corporation, Norwood,
               Massachusetts, by letter March 28, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Undetermined.
REASON         The ASC:180 system had a track jam caused by a deformity in
               the neck of the cuvette.

_______________
PRODUCT        Automated Chemiluminiscene System (ACS) Multicalibrator A
               Lot CA27 and CA31 the Low and High Calibrators contained in
               the ACS Calibrator A are for calibrating the ACS
               immunoassays:  (a) Part #672170000; 
               (b) Part #672180000.  Recall #Z-944/945-4.
CODE           Lot numbers CA27 and CA31, EXP 27 Jul 94 packaged in ACS
               Multicalibrator kit lot numbers:
               (a) 22911, 22951, 23027, 23591, 23611, 23779, 23824; 
               (b) 22913, 22952, 23592, 23610, 23655.
MANUFACTURER   Ciba Corning Diagnostics Corporation, East Walpole,
               Massachusetts.
RECALLED BY    Manufacturer, by telephone February 10, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 741 kits; (b) 249 kits were distributed.
REASON         The theoretical concentrations of master curve standards
               were incorrectly assigned by quality control.  The values
               were not obtained via referenced standard values used in
               testing.

                                   -10-_______________
PRODUCT        Automated Chemiluminiscene System (ACS):180 SPT (System
               Performance Tests) Reagent, Part #672196 SPT used to
               evaluate the performance of the ACS:180 subassemblies at
               installation, after repair procedures, and as a
               troubleshooting aid.  Recall #Z-946-4.
CODE           Reagent Lot 2773 packaged in kit lot numbers 1506, 4005,
               4007, 4308, 4309, 4310, 6006, 01113; Wash Solid Phase lot
               LOP 2993 in kit lot numbers 14007, 16003, 16809; Reagent
               Dispense Check  lot LOL 0604 in kit lot #182302.
MANUFACTURER   Ciba Corning Diagnostics Corporation, East Walpole,
               Massachusetts.
RECALLED BY    Manufacturer, by telephone April 14, 1993, and by recall
               notice July 21, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       480 units of lot 2773, 270 units of lot 2993, and 42 units
               of lot 0604.
REASON         The lot of reagent 2773 was contaminated with lactobacillus
               Coryniformis.  The (SPT) reagents have degraded in their
               performance specifications resulting in check problems in
               the ACS operations.

_______________
PRODUCT        Automated Chemiluminiscene System (ACS) T4 Master Curve Card
               used with ACS T4/50 test and 300 test kits, intended for the
               quantitative determination of Thyroxine (T4) in serum using
               the ACS:180 system:  (a) Catalog #672200; (b) Catalog
               #672201.  Recall #Z-947/948-4.
CODE           Lot numbers:  (a) 02003; (b) 02004.
MANUFACTURER   Ciba Corning Diagnostics Corporation, East Walpole,
               Massachusetts.
RECALLED BY    Manufacturer, by telephone February 5, 1993, and by fax to
               international account January 3, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   New York, Michigan, Pennsylvania, Massachusetts, Ohio,
               California, Rhode Island, Oklahoma, Japan.
QUANTITY       (a) 350 kits; (b) 39 kits.
REASON         The standard value printed on the master curve card was
               incorrect.

_______________
PRODUCT        Cell Culture Tube and Drams, Catalog #71-217 D, used for
               isolation of virus in clinical specimens.  Recall #Z-949-4.
CODE           Lot numbers:  1C2818 (tubes), 1C2804 (drams).
MANUFACTURER   Whittaker Bioproducts, Inc., Walkersville, Maryland.
RECALLED BY    Manufacturer, by telephone October 18, 1991.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       670 units of lot 1C2804 and 3,755 units of 1C2818 were
               distributed.
REASON         Device was contaminated with IBR bovine herpes virus.

                                   -11-_______________
PRODUCT        EMEM Cell Culture Medium, in 100 ml bottles, used for the
               growth and maintenance of cells in culture.  
               Recall #Z-952-4.
CODE           Catalog #12-125A, Lot 1M0865.
MANUFACTURER   Whittaker Bioproducts, Inc., Walkersville, Maryland.
RECALLED BY    Manufacturer, by telephone on or about October 9, 1992. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Germany, Spain, France, Japan.
QUANTITY       1,226 units were distributed.
REASON         The device had a pH of 7.6 which is higher that the firm's
               established specification of 6.7 to 7.3.

_______________
PRODUCT        Rubecap-M Test Kit, an enzyme-linked antibody capture assay
               for Rubella Virus IgM antibody in human serum.  
               Recall #Z-953-4.
CODE           Catalog #31-002U, lot 1E2241.
MANUFACTURER   Mercia Diagnostics supplies the kit format and Whittaker
               Bioproducts labels and packages the kits.
RECALLED BY    Whittaker Bioproducts, Inc., Walkersville, Maryland, by
               telephone October 31, 1991, followed by letter November 30,
               1991.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Austria, Germany, France, England, Italy.
QUANTITY       70 kits were distributed.
REASON         Device had a pH of 7.6 which is higher than the firm's
               established specification of 6.7 to 7.3.  The control ranges
               for positive and negative controls included in the kit are
               incorrectly stated on the component label.

_______________
PRODUCT        QuinEvac IIa Smoke Evacuator, Model #SE-11111-E, used in
               hospital operating rooms where there is a large volume of
               smoke plume generated during laser surgery or electrosurgery
               products.  Recall #Z-967-4.
CODE           Serial numbers:  155300, 155302, 155309-18.
MANUFACTURER   Linemaster Switch Corporation, Woodstock, Connecticut.
RECALLED BY    Recto Molded Products, Inc., (doing business as Quinn
               Healthcare Products), Cincinnati, Ohio, by telephone and fax
               March 29, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       20 units.
REASON         The footswitch may stick in the open or closed position. 

_______________
PRODUCT        Triage brand Drugs of Abuse in-Vitro Diagnostic Test Kits, a
               urine test for screening seven classes of drugs: (a) Catalog
               #95000 (25 determinations); (b) 95004 (10 determinations). 
               Recall #Z-970/971-4.
CODE           Lot numbers:  (a) W4095 through W4377 EXP 11/29/94 through
               2/5/95; (b) W4467 through W4598 EXP 2/15/95 through 3/28/95.
MANUFACTURER   Biosite Diagnostics, Inc., San Diego, California.

                                   -12-RECALLED BY    Manufacturer, by letter April 22, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Canada, Japan, Singapore, China, Malaysia.
QUANTITY       (a) 11,462 kits; (b) 387 kits were distributed between
               December 1993 and April 1994.
REASON         The test results may indicate false positive for the THC
               (marijuana) portion of the panel.  The performance of the
               THC assay in the Triage Panel degrades at a faster rate than
               indicated by the expiration date on the product labeling.


MEDICAL DEVICE SAFETY ALERTS: 
============================================
_______________
PRODUCT        Kangaroo Enteral Feeding Pumps Model 224, Reorder #8884-
               322408, and Model 324, Reorder #8884-322407.
               Safety Alert #N-050/051-4.
CODE           All serial numbers.
MANUFACTURER   Sherwood Medical Watertown, New York.
ALERTED BY     Sherwood Medical Company, St. Louis, Missouri, by letter
               February 10, 1994.
DISTRIBUTION   Nationwide and international.
QUANTITY       35,048 Model 224 pumps and 71,588 Model 324 pumps since
               1988.
REASON         Arcing may occur between electrical components leading to
               electrical fires.

                                   -13-

END OF ENFORCEMENT REPORT FOR JUNE 22, 1994.  BLANK PAGES MAY FOLLOW.
                                   ####