The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
06/15/1994

Recalls and Field Corrections:  Foods -- Class II -- 06/15/1994

June 15, 1994                                                    94-24
  
RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Strawberry Toppings under the following labels and package
               sizes:  (a) Hardee's Strawberry Topping, in 26.5 fluid ounce
               pouches; (b) Strawberry Sundae Topping, under the Quaker
               Oats Company label for exclusive use by McDonald's
               Restaurants, in 32 fluid ounce pouches; (c) Sysco Classic
               Strawberry Topping, in 46 fluid ounce cans.  These are food
               service items, not for retail sale.  Recall #F-415-4.
CODE           (a) Use thru Dec 31 94; (b) Use thru Dec 31 94, lot #9412;
               and Use thru Jan 08 95, lot #9431; (c) 4Q15D.
MANUFACTURER   The Quaker Oats Company, doing business as Richardson Foods
               Corporation, Macedon, New York.
RECALLED BY    Manufacturer, by telephone May 3, 1994, followed by letter
               May 11, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 724 cases (6 pouches per case); (b) 2,147 cases (12
               pouches per case); (c) 628 cases (6 cans per case).
REASON         Product contains pieces of plastic.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Gatorade, in 32 fluid ounce plastic bottles.  Flavors are:
               (a) Orange; (b) Iced Tea Cooler (lemon and tea);
               (c) Grape; (d) Lemon-Lime; (e) Fruit Punch.
               Recall #F-416/420-4.CODE           Flavor         Case Code      Bottle Code          
               Orange         03-24-94       024C4* (see note)    
               Lemon-Lime     03-24-94       024C4*              
               Grape          03-25-94       025C4*              
               Fruit Punch    03-25-94       025C4*              
               Iced Tea       03-26-94       026C4*              
                 (Note:  Bottle code could end in A,B,C,D,E, or H) 
               Warehouse Club Shipments:
               Flavor         Pallet Code    
               Lemon-Lime     04-06-94      
               Orange         040894        
MANUFACTURER   Quaker Oats Company, Oakland, California.
RECALLED BY    Manufacturer, by letter April 21, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Arizona, California, Hawaii, Idaho, Montana, Nevada, Oregon,
               Utah, Washington state.
QUANTITY       33,217 cases (24 bottles per case) were distributed.
REASON         Product is contaminated with mold.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Amoxicillin Oral Suspension, in 100 ml bottles, Rx oral
               antibiotic: (a) 125 mg/5 ml; (b) 250 mg/5 ml.  
               Recall #D-295/296-4.
CODE           Lot numbers:  (a) 3270075, 3279117; (b) 3134029, 3298063,
               3099022.
MANUFACTURER   Biocraft Laboratories, Elmwood Park, New Jersey (responsible
               firm).
RECALLED BY    Allscrips Pharmaceuticals, Inc., Vernon Hills, Illinois
               (relabeler), by letter May 18, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 723 bottles; (b) 2,020 bottles were distributed.
REASON         Product does not meet stability specifications.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Xenon Xe 133 Gas - radioactive diagnostic for inhalation
               administration:  (a) 10 millicuries (mCi), Catalog #N097; 
               (b) 20 millicuries (mCi), Catalog #N098.  
               Recall #D-297/298-4.
CODE           Lot numbers:  (a) 0974016; (b) 0984016.
MANUFACTURER   Mallinckrodt Medical, Inc., Maryland Heights, Missouri.
RECALLED BY    Manufacturer, by telephone May 5, 1994, followed by letter
               May 9, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Puerto Rico, Buenos Aires, Canada, Mexico.
QUANTITY       Approximately (a) 1,270 vials; (b) 1,195 vials were
               distributed.
REASON         Low assay of radioactive component.

                                    -2-_______________
PRODUCT        Calplus brand of Calcium Chloride Dihyrate, USP, in 110
               pound drums, a bulk tested additive chemical for
               manufacturing, processing, or repacking food or drug
               products. Recall #D-299-4.
CODE           Lot #SLM018.
MANUFACTURER   Nichea, Tokyo, Japan.
REPACKER       Clinton Universal Warehouse, Newark, New Jersey.
RECALLED BY    Mallinckrodt Chemical, Inc., St. Louis, Missouri, by
               telephone April 29, 1994, followed by letter May 9, 1994. 
               Firm-initiated recall ongoing.
DISTRIBUTION   New Jersey, California.
QUANTITY       97 drums were distributed.
REASON         Current good manufacturing practice deficiency by recalling
               firm.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        Fresh Frozen Plasma.  Recall #B-317-4.
CODE           Unit #03T97357.
MANUFACTURER   The American National Red Cross, Atlanta, Georgia.
RECALLED BY    Manufacturer, by telephone August 17, 1993, followed by
               letter August 23, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Georgia.
QUANTITY       1 unit.
REASON         Blood product collected from a donor with a history of
               cancer was distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
UPDATE         Recall #B-275/281-4, manufactured by American Red Cross
               Blood Services, Dedham, Massachusetts, which appeared in the
               June 8, 1994 should read:
               PRODUCT: (a) Red Blood Cells; (b) Platelets; (c) Platelets,
               Expired; (d) Red Blood Cells, For Further Manufacture; (e)
               Cryoprecipitated AHF; (f) Fresh Frozen Plasma; (g) Recovered
               Plasma.  Recall #B-275/281-4.
               CODE Unit numbers:  (a) 04C17643, 04F06443, 04FC51677,
               04GF56801, 04GF58444, 04GF60232, 04GF64890, 04KK87617,
               04KK90263; (b) 04F06443, 04GF58444, 04GF60232, 04KK90263;
               (c) 04C17643, 04GF63060; (d) 04GF3060; (e) 04KK87617,
               04KK90263; (f) 04F06443, 04C17643; (g) 04FC51677, 04GF56801,
               04GF58444, 04GF60232, 04GF63060, 04KK87617, 04KK90263.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        FT-Ramean Accessory, used to perform laser-excited Ramean
               spectroscopy on chemical samples.  Recall #Z-593-4.
CODE           None.

                                    -3-MANUFACTURER   Digilab Division, Bio-Rad, Cambridge, Massachusetts.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action
               plan March 25, 1994.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       10 units.
REASON         The laser product failed to provide either multiple
               interlocks or a fail-safe design for the required safety
               interlocks.

_______________
PRODUCT        Hycor Coated Tube RIA, HS-TSH Immunoradiometric Assay,
               Thyroid Stimulating Hormone (TSH), intended for use in
               immunoradiometric assay (IRMA) for the quantitative
               determination of TSH.  Recall #Z-719-4.
CODE           Lot #24857 EXP 9/14/94.
MANUFACTURER   Hycor Biomedical, Inc., Garden Grove, California.
RECALLED BY    Manufacturer, March 8, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   United Kingdom.
QUANTITY       90,575 tubes were distributed; FDA estimates that little if
               any product remains on the market.
REASON         Not all tubes contain the antibody coating that is indicated
               on the label which may result in neonates with
               hypothyroidism (cretinism) being misdiagnosed as normal.

_______________
PRODUCT        Ferno Washington Ambulance Transportation Cots: 
               (a) Model 35-A; (b) Model 35-NM; (c) 35-IT.
               Recall #Z-742/744-4.
CODE           All lots.
MANUFACTURER   Ferno Washington, Inc., Wilmington, Ohio.
RECALLED BY    Manufacturer, by letter January 10, 1994.  Firm-initiated
               field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 13,000 units were distributed.
REASON         Products have an unsafe sliding area in which an attendant
               (bystander), while assisting in loading the units can
               position the hands near the X-frame slides in such a way as
               to cause serious finger injuries.

_______________
PRODUCT        Conseal brand Colostomy Continence Systems:
               (a) Part No. 1235; (b) Part No. 1245; (c) Part No. 1285; 
               (d) Part No. 1295; (e) Part No. 1435; (f) Part No. 1445; 
               (g) Part No. 1485; (h) Part No. 1495. Recall #Z-833/840-4.
CODE           All codes starting with 37788 through 84255.
MANUFACTURER   Coloplast A/S, Espergaerde, Denmark.
RECALLED BY    Coloplast, Inc., Tampa, Florida, by letter December 3, 1993.
               Firm-initiated recall ongoing. 
DISTRIBUTION   Nationwide.
QUANTITY       500 to 600 boxes of 10 units each.

                                    -4-REASON         The foam liner may separate from the plug and remain behind
               in the stoma when the plug is removed.

_______________
PRODUCT        Intraocular Lens, Model 815A, 11.50 and 21.50 diopters, used
               to replace the natural lens in the visual correction of
               aphakia in patients 60 years of age or older.  
               Recall #Z-854-4.
CODE           Lenses sterility lot #031293A.
MANUFACTURER   Pharmacia Ophthalmics, Inc., Monrovia, California.
RECALLED BY    Manufacturer, by letters of December 10, 1993, and February
               16, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Arizona, California, Colorado, Florida, Iowa, Illinois,
               Louisiana, Maryland, Minnesota, New York, Ohio,
               Pennsylvania, South Dakota, Texas, Utah, Vermont, Australia.
QUANTITY       Firm estimates no unimplanted lenses remain on the market.
REASON         Devices may be labeled with the incorrect dioptric power,
               which would result in a refractive error if implanted.

_______________
PRODUCT        Disposable Laryngoscope Blades, used with the Laryngoscope
               for the ocular examination of the larynx of adults and
               pediatrics: (a) Macintosh size 2 blades, Product Nos.
               9942-00 & 9982-00; 
               (b) Macintosh size 3 blades, Product Nos. 9943-00 & 9983-00;
               (c) Macintosh size 4 blades, Product Nos. 9944-00 & 9984-00;
               (d) Miller size 0 blades, Product Nos. 9960-00 & 9967-00;
               (e) Miller size 1 blades, Product Nos. 9961-00 & 9968-00;
               (f) Miller size 2 blades, Product Nos. 9962-00 & 9969-00;
               (g) Miller size 3 blades, Product Nos. 9963-00 & 9971-00;
               (h) Cabinet, with Macintosh blades, Product No. 9945-00;
               (i) Cabinet, with Miller blades, Product #9946-00;
               (j) Laryngoscope set with plastic handle, Product #9939-00;
               (k) Laryngoscope set with metal handle, Product #9941-00.
               Recall #Z-861/871-4.
CODE           Lot Nos. 120293 through 030494.
MANUFACTURER   Anesthesia Medical Specialties, Santa Fe Springs,
               California.
RECALLED BY    Manufacturer, by letter March 11, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Canada, Germany.
QUANTITY       57,526 blades were distributed.
REASON         The blades in some cases have demonstrated a propensity for
               breakage during clinical usage.

_______________
PRODUCT        Disposable Laryngoscope Blades, used with the Laryngoscope
               for the visual examination of the laryngeal orifice of
               adults and pediatrics:  (a) Macintosh size 3; (b) Miller
               Size 1; (c) Miller Size 3; (d) Variety Pack; (e)
               Laryngoscope Set.  Recall #Z-872/876-4.

                                    -5-CODE           Catalog No.        Batch No.
               (a) 17643          036 through 045
               (b) 17661          027
               (c) 17663          026
               (d) 176-VAR        002, 003, 004
               (e) 17686          018 through 021.
MANUFACTURER   Anesthesia Medical Specialties, Santa Fe Springs,
               California.
RECALLED BY    Respiratory Support Products, Irvine, California, by letter
               March 14, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, England, France, Netherlands, Taiwan.
QUANTITY       16.785 blades and 335 sets were distributed.
REASON         Blades in some cases have demonstrated a propensity for
               breakage during clinical usage.

_______________
PRODUCT        Grice Abdominal Lift Reddick-Saye Screw Modification,
               designed for  lifting the abdomen for laparoscopic surgery
               under the navel.  Recall #Z-897-4.
CODE           All lots distributed between 10/13/92 and 6/21/93.
MANUFACTURER   Ideas for Medicine, Inc., Clearwater, Florida.
RECALLED BY    Manufacturer, by verbal notification June 22, 1993.   Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,077 devices were distributed.
REASON         The device fails to function as intended due to lack of
               adequate design and quality controls.

_______________
PRODUCT        Twenty-one In Vitro Diagnostic products as listed below; 
               (a) Lyme Disease, Catalog Nos. 283-390, 783-390; 
               (b) Toxo. IgG, Catalog Nos. 283-300, 783-300; 
               (c) Rubella IgG, Catalog Nos. 283-300, 783-300; 
               (d) HSV 1, Catalog Nos. 283-340, 783-340; 
               (e) VZV, Catalog Nos. 283-380, 783-380; 
               (f) CMV IgG, Catalog Nos. 283-320, 783-320; 
               (g) HSV 2, Catalog Nos. 283-350, 783-350; 
               (h) Rubella IgM, Catalog Nos. 283-370, 783-370; 
               (i) CMV IgM, Catalog Nos. 283-330, 783-330; 
               (j) Scl-70, Catalog No. 283-280; 
               (k) RNP, Catalog Nos. 283-270, 783-270; 
               (l) SSA, Catalog Nos. 283-250, 783-250; 
               (m) Sm, Catalog Nos. 283-240, 783-240; 
               (n) Toxo. IgM, Catalog Nos. 283-310, 783-310; 
               (o) Measles, Catalog No. 283-520; 
               (p) DNP, Catalog Nos. 283-230, 783-230; 
               (q) IC, Catalog No. 783-210; 
               (r) SSB, Catalog No. 283-260; 
               (s) DNA, Catalog Nos. 283-200, 783-200; 
               (t) RF, Catalog Nos. 283-220, 783-220; 
               (u) Amebiasis, Catalog No. 783-120. Recall #Z-902/922-4.

                                    -6-CODE           (a) 40903F EXP 4/94, 40903FA EXP 4/94, 60603F EXP 6/96,
               60603FA EXP 6/94, 40903 EXP 4/94, 42703 EXP 4/94, 50703 EXP
               5/94, 60603 EXP 6/94, 70303 7/94, 70303F 7/94, 42703F 4/94,
               40903A 4/94; (b) 41703F EXP 4/94, 43803 EXP 4/94, 60703F EXP
               6/94, 43803 EXP 4/94, 41703 EXP 4/94, 60703 EXP 6/94; (c)
               40603F EXP 4/94, 43003F EXP 4/94, 41603 EXP 4/94, 41203 EXP
               4/94, 43003 EXP 4/94, 52203 EXP 5/94, 52203F EXP 5/94; (d)
               41103F EXP 4/94, 43503F EXP 4/94, 62703F EXP 6/94, 41103 EXP
               4/94, 43503 EXP 4/94, 62703 EXP 6/94; (e) 40803F EXP 4/94,
               43603F EXP 4/94, 61403F EXP 6/94, 40803 EXP 4/94, 43603 EXP
               4/94, 61403 EXP 6/94; (f) 41403F EXP 4/94, 43703F EXP 4/94,
               61003F EXP 6/94, 70703F EXP 7/94, 41403 EXP 4/94, 43703 EXP
               4/94, 61003 EXP 6/94; (g) 41903F EXP 4/94, 51403F EXP 5/94,
               41903 EXP 4/94, 51403 EXP 5/94, 42803 EXP 4/94; (h) 40403F
               EXP 4/94, 40303 EXP 4/94, 60903 EXP 6/94; (i) 41803F EXP
               4/94, 61903F EXP 6/94, 41803 EXP 4/94, 43303 EXP 4/94, 61903
               EXP 6/94; (j) 22603F EXP 2/94; (k) 21203F EXP 2/94, 42203F
               EXP 4/94, 62303 EXP 6/94, 21203 EXP 2/94, 62303 EXP 6/94;
               (l) 31503F EXP 3/94, 20703 EXP 2/94, 21603F EXP 2/94, 51303F
               EXP 5/94, 51303 EXP 5/94, 31403 EXP 3/94; (m) 21503F EXP
               2/94, 42103F EXP 4/94, 62903 EXP 6/94; (n) 43203F EXP 4/94,
               62503F EXP 6/94, 40703F EXP 4/94, 41303 EXP 4/94, 43203 EXP
               4/94, 60503 EXP 6/94, 70403 EXP 6/94; (o) 31103F EXP 3/94;
               (p) 30503F EXP 3/94, 30503 EXP 3/94; (q) 31003 EXP 3/94; (r)
               51203F EXP 5/94; (s) 20503 EXP 2/94, 20503F EXP 2/94, 42003
               EXP 4/94; (t) 70503 EXP 7/94, 20803F EXP 2/94; (u) 21703 EXP
               2/94.
MANUFACTURER   Diamedix Corporation, Miami, Florida.
RECALLED BY    Manufacturer, by letter February 2, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       8,833 kits were distributed.
REASON         The Cal-Verifier that is shipped with the device kits lacks
               a 510(k) approval. 
 
_______________
PRODUCT        Bard 9.5 Fr 40cc Double Lumen Intra-Aortic Balloon Catheter,
               single use device intended for temporary cardiac assistance
               employing the principle of counterpulsation.
               Recall #Z-923-4.
CODE           Catalog #H-8240, lot numbers: 28LDQ033, 28LDQ035, 28AEQ019,
               28AEQ030, 28BEQ019, 28BEQ020, 28IDQ501, 28IDQ502, 28HDQ501.
MANUFACTURER   Bard Cardiopulomary Division, C.R. Bard, Inc., Tweksbury,
               Massachusetts. 
RECALLED BY    Bard Cardiopulmonary, Haverhill, Massachusetts, by letter
               March 23, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       488 units.
REASON         An unknown number of brittle (defective) "Y" bifurcates  
               could leak, resulting in a loss of helium and the inability
               to achieve vacuum configuration.

                                    -7-_______________
PRODUCT        Oral Endotracheal Tube Holders, used to anchor and stabilize
               the tube, inhibiting extubation and eliminating taping
               procedures:
               (a) Secure Easy Endotracheal Tube Holder (padded faceplate
               with bite block, headgear and neckstrap), Catalog numbers
               242003, 242006;
               (b) Quickstrap Endotracheal Tube Holder (padded faceplate
               with bite block and non-elastic neckstrap), Catalog number
               242007.  Recall #Z-925/926-4.
CODE           Lot numbers:  (a) C093, C095; (b) C075, C088, C095.
MANUFACTURER   IPI Medical Products, Chicago, Illinois.
RECALLED BY    Manufacturer, by letter April 23, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide, France.
QUANTITY       5,430 units; firm estimates none remains on the market.
REASON         The tube retainer strap may crack when excessive force is
               applied.

_______________
PRODUCT        Night Vision Image Intensifiers, Models T3C-2 and T3C-3,
               used in viewing objects and persons in subdued lighting.  
               Recall #Z-927/928-4.
CODE           None.
MANUFACTURER   Novosibirsk Instrument Making Plant Novosibirsk, Russian
               Federation.
RECALLED BY    Intertech (Advanced International Technologies), North
               Kingstown, Rhode Island (importer).  FDA approved the firm's
               corrective action plan May 17, 1994.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       79 units.
REASON         Products emitted x-radiation that is unnecessary to the
               accomplishment of the purpose of the product, and creates a
               risk of injury to users. 

_______________
PRODUCT        Polycitra K Crystals, (potassium citrate and citric acid for
               oral solution), with Maintenance propHilter Urinary pH
               Testing Kit (envelope) consisting of a one month diary, and
               urinary pH indicator test strips packaged in a folding
               carton containing 100 packets of the Polycitra Crystals, and
               Polycitra K Crystals (potassium citrate and citric acid for
               oral solution) propHiler Urinary pH Testing Kit (envelope
               consisting of four starter samples of Polycitra K Crystals,
               two week Progress Diary, information brochure titled
               "Preventing Kidney Stone Recurrence with Polycitra K, and
               urinary pH indicator test strips, an alkalinizing agent for
               control of uric acid and cystine calculi of the urinary
               tract.  
               Recall #Z-929-4.
CODE           All maintenance propHiler Kit lots shipped from 4/5/93 to
               2/17/94.
                                    -8-MANUFACTURER   Baker Norton Pharmaceuticals, Inc., Miami, Florida, by
               letter March 4, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 17,138 (10/100 packet) cases, and a small
               quantity of physician starter samples.
REASON         The urinary test kit component is considered to be
               misbranded as the pH indicator test strips included do not
               have 510(k) approval.

_______________
PRODUCT        Sorin BCD ADV C Blood Cardioplegia Delivery Sets, designed
               to mix arterial blood from an oxygenator with asanquineous
               cardioplegic solution, and deliver it to the patient.  
               Recall #Z-941-4.
CODE           Lot numbers for BCD ADV C sets sold individually, packed in
               cartons:  
               9300120300     9300168800     9300312800     9400132200 
               9300144300     9300255400     9400102300     9400140900 
               9300152200     9300288400     9400111100     9400122300 
               Lot numbers for BCD ADV C sets within SCPS kits: 
               SCPS Lot Numbers:              BCD ADV C Lot Numbers: 
               OMS4982-4983   9400139700      9300288400     9400140900 
               9300328600     9400168700      9300312800     9400102300 
               9400114900     9400170700 
               9400128900     9400170600 
               9400139600 
               Lot numbers for BCD ADV C sets within PTS kits within SCPS
               kits: 
               SCPS Lot Numbers:  PTS Lot Numbers:   BCD ADV C Lot Numbers:
               9400177600          9400128800            9400122300 
               9400108700          9300308600            9300255400 
               OMS-5166.
MANUFACTURER   Sorin Biomedical, Inc., Irvine, California.
RECALLED BY    Manufacturer, by letter April 22, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Alabama, California, Florida, Illinois, Indiana, Michigan,
               Mississippi, North Carolina, New Mexico, Ohio, Oklahoma,
               Pennsylvania, Tennessee, Texas, Utah, Washington state,
               Wisconsin, West Virginia, Canada, Korea.
QUANTITY       Approximately 5,116 units were distributed; firm estimated
               500-700 units remained on market at time of recall.
REASON         The blood lines and cardioplegia solution lines for an
               unknown number of devices across several device lots were
               reversed, resulting in an incorrect blood to cardioplegia
               solution configuration.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        L-Cath Procedural Tray for L-Cath Peel Away Placement Sets
               with Catheter Extension Set:
               (a) L-Cath Procedural Tray Prod. No. PE16PIC20K;   

                                    -9-               (b) L-Cath Procedural Tray Prod. No. PE18PIC20K; 
               (c) L-Cath Procedural Tray Prod. No. PE20PIC20K; 
               (d) L-Cath Procedural Tray Prod. No. PE16PIC56K; 
               (e) L-Cath Procedural Tray Prod. No. PE18PIC56K; 
               (f) L-Cath Procedural Tray Prod. No. PE20PIC56K; 
               (g) L-Cath Procedural Tray Prod. No. PIC-03; 
               (h) L-Cath Procedural Tray Prod. No. PIC-03A.
               Recall #Z-877/884-4.
CODE           All lot numbers of the PIC-03 or PIC-03A procedural trays
               either sold separately or in kit form.  All product numbers
               involved.
MANUFACTURER   Tri-State Hospital Supply Corporation, Howell, Michigan.
RECALLED BY    Luther Medical Products, Inc, Tustin, California (own label
               distributor/responsible firm), by letters of December 2 and
               15, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Minnesota.
QUANTITY       4,479 trays were distributed; firm estimates none remains on
               the market.
REASON         The male luer of the extension set leaks.

_______________
PRODUCT        Stratus Free Thyroxine FT4 Fluorometric Enzyme Immunoassay
               Kit IVD, intended for the determination of free thyroxine
               concentrations in serum or plasma.  Recall #Z-892-4.
CODE           Kit catalog #B5700-48, lot #KXT4-869M.
MANUFACTURER   Baxter Diagnostics, Inc., Miami, Florida.  With the
               exception of the substrate (manufactured at the Aguada,
               Puerto Rico Facility) all components were made at the Miami
               plant.
RECALLED BY    Baxter Diagnostics, Inc., Miami, Florida, by telephone
               November 1-3, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Argentina.
QUANTITY       107 kits were distributed.
REASON         The antibody tabs did not have spotting solution and were
               not imprinted with a lot number.

_______________
PRODUCT        Stratus Ultra-Sensitive hTSH Enzyme Immunoassay Kits,
               intended for the determination of human thyroid stimulating
               hormone (hTSH) concentrations in serum or plasma.  
               Recall #Z-893-4.
CODE           Catalog #B5700, Kit lots KXS-24M through KXTS-322M, except
               KXTS-229M and KXTS-252M.
MANUFACTURER   Baxter Diagnostics, Inc., Aguada, Puerto Rico.
RECALLED BY    Baxter Diagnostics, Inc., Miami, Florida, by letter dated
               October 1993, mailed October 20, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 31,000 kits.

                                   -10-REASON         One of the kit components, the Ultra-Sensitive hTSH
               Conjugate, may have contained aggregates that could cause
               calibrator, control, or patient sample values to be higher
               than they should have been.

_______________
PRODUCT        Deknatel Custom Procedure Pak, V-Pak suture sleeve which
               includes 11 different products packaged in sealed Tyvek
               pouch.  Recall #Z-894-4.
CODE           Catalog #V-2307, lot #02591.
MANUFACTURER   Deknatel, Inc., Fall River, Massachusetts.
RECALLED BY    Manufacturer, by letter March 18, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Arkansas.
QUANTITY       60 pouches were distributed.
REASON         Some of the individual pouches did not contain the needles
               labeled.

_______________
PRODUCT        Cottony II Dacron Orthopaedic Non-Absorbable Surgical Suture
               Green.  Recall #Z-895-4.
CODE           Catalog #89-5039M5, lot #02447.
MANUFACTURER   Deknatel, Inc., Fall River, Massachusetts.
RECALLED BY    Manufacturer, by telephone, followed by letter March 18,
               1994.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Utah, Georgia.
QUANTITY       132 pouches were distributed.
REASON         The primary suture package (carrier) may be incorrectly
               labeled as Product Code 89-5032M5.

_______________
PRODUCT        Custom Procedure Pak Non-Absorbable Surgical Suture, 5
               green, 5 white braided surgical sutures packaged in a V-Pack
               suture sleeve.  Recall #Z-896-4.
CODE           Catalog #VF69-7101 GW, lot #02291.
MANUFACTURER   Deknatel, Inc., Fall River, Massachusetts.
RECALLED BY    Manufacturer, by telephone, followed by letter March 18,
               1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Wisconsin, Florida, Utah, Canada.
QUANTITY       175 pouches were distributed.
REASON         The shelf box and pouch labels identify the pledgets that
               are stitched onto the suture as "soft", when they are
               actually "firm".

_______________
PRODUCT        Arthrex Tunnel Notcher, used in arthroscopic surgery,
               Catalog #AR-1844.  Recall #Z-924-4.
CODE           All units manufactured with a design that has a visible
               outline of the notch within the shaft where the notch is
               brazed onto the shaft.
MANUFACTURER   Ultrexx, Inc., Ultra Precision Machining, Avilla, Indiana.

                                   -11-RECALLED BY    Arthrex, Inc., Naples, Florida, by letter sent during the
               first week of November 1993.  Firm-initiated recall
               complete.
DISTRIBUTION   California, Texas, Tennessee, Maryland, New York, Virginia,
               Michigan, Arizona, Florida, Colorado.
QUANTITY       103 units were distributed.
REASON         A piece of the tunnel notcher, the part that cuts the notch,
               can break off from the rest of the device.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS III
==========
_______________
PRODUCT        Spot Balm, in 2 fluid ounce plastic bottles, liquid topical
               OTC animal drugs containing sulfanilamide.  Recall #V-052-4.
CODE           Lot numbers:  120192, 30193, 40193, 60193, 80193, 90193,
               100193.
MANUFACTURER   Happy Jack, Inc., Snow Hill, North Carolina.
RECALLED BY    Manufacturer, by letter dated February 9, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       615 bottles of lot 120192, 622 bottles of lot 30193, 624
               bottles of lot 40193, 626 bottles of lot 60193, 626 bottles
               of lot 80193, 625 bottles of lot 90193, 449 bottles of lot
               100193.
REASON         No assurance that the product met labeled specifications for
               potency of the active ingredient (sulfanilamide).


SEIZURES:
=================================================================
_______________
PRODUCT        Canned Chunk Light Tuna in Spring Water (93-651-183).
CHARGE         Misbranded - The article purports to be and is represented 
               as canned tuna, a food for which a standard fill of
               container has been prescribed, and the article falls below
               such standard. 
LOCATION       Starkist Foods, Inc., doing business as Heinz Pet Products,
               Pascagoula, Mississippi. 
FILED          March 29, 1994; U.S. District Court for the Southern
               District of Mississippi, Southern Division;
               Civil #1:94cv164RR; FDC #66927.
SEIZED         April 4, 1994 - goods valued at approximately
               $152,000.

_______________
PRODUCT        Canned Tuna in Water (94-656-773/4).
CHARGE         Adulterated - The articles consist in part of decomposed
               seafood.
LOCATION       The Belt's Corporation, Baltimore, Maryland.
FILED          May 6, 1994; U.S. District Court for the District of
               Maryland; Civil #HAR-94-1199; FDC #66956.
SEIZED         May 20, 1994 - goods valued at approximately $73,000.

                                   -12-_______________
PRODUCT        Oxygen, USP (93-607-680).
CHARGE         Adulterated - The methods used in, and the facilities and 
               controls used for, the article's manufacture, processing,
               packing, and holding do not conform to and are not operated
               and administered in conformity with current good
               manufacturing practice requirements for drugs.
LOCATION       Butler Gas Products Co., New Brighton, Pennsylvania.
FILED          May 5, 1994; U.S. District Court for the Western
               District of Pennsylvania; Civil #94-0743; FDC #66958.
SEIZED         May 17, 1994 - goods valued at approximately $10,000.

_______________
PRODUCT        Oxygen, USP (93-713-727).
CHARGE         Adulterated - The methods used in, and the facilities and 
               controls used for, the article's manufacture, processing,
               packing, and holding do not conform to and are not operated
               and administered in conformity with current good
               manufacturing practice requirements.
LOCATION       Metro-Med, Inc., Burbank, California.
FILED          May 5, 1994; U.S. District Court for the Central District 
               of California; Civil #94-2939RSWL(GHKx); FDC #66946.
SEIZED         May 12, 1994 - goods valued at approximately $50,000.

_______________
PRODUCT        Oxygen, USP (93-607-839).
CHARGE         Adulterated - The methods used in, and the facilities and 
               controls used for, the article's manufacture, processing,
               packing, and holding do not conform to and are not operated
               and administered in conformity with current good
               manufacturing practice requirements.
LOCATION       Medi-Home-Care, Butler, Pennsylvania.
FILED          April 15, 1994; U.S. District Court for the Western
               District of Pennsylvania; Civil #94-0620; FDC #66904.
SEIZED         May 17, 1994 - goods valued at approximately $3,500.

                                   -13-
END OF ENFORCEMENT REPORT FOR JUNE 15, 1994.  BLANK PAGES MAY FOLLOW.
                                   ####