The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
03/23/1994

Recalls and Field Corrections:  Cosmetics -- Class II -- 03/23/1994

March 23, 1994                                                    93-12

RECALLS AND FIELD CORRECTIONS:  COSMETICS -- CLASS II
=====================
_______________
PRODUCT        Formula 405 Soapless Cleansing Lotion in 16 fluid ounce
               bottles.  Recall #F-318-4.
CODE           All lots except lot 3510A and 31109, or any other lot
               starting with 4.
MANUFACTURER   Doak Pharmacal Company, Inc., Westbury, New York.
RECALLED BY    Manufacturer, by letter January 13, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 450 units were distributed.
REASON         Product contained a color additive not authorized for use in
               cosmetics (D&C Red #19).


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        SoloPak Heparin Lock Flush Solution, USP, Preservative    
               Free, 10 USP units/ml packaged in 1, 3 and 5 ml syringes;
               for maintenance of patency of intravenous injection devices
               only, not to be used for anticoagulant therapy:
               (a) Catalog #10671: 1 ml fill in 1.5 ml Hy-Pod Syringe and
               25G x 5/8" needle, individually wrapped, 120 per carton;
               (b) Catalog #10673: 3 ml Hy-Pod Syringe and 25G x 5/8"
               needle, individually wrapped, 120 per carton;
               (c) Catalog #10675: 5 ml Hy-Pod Syringe and 25G x 5/8"
               needle, individually wrapped, 120 per carton;
               (d) Catalog #10681: 1 ml Hy-Pod Syringe (needles not
               included), individually wrapped, 120 per carton;
               (e) Catalog #10683: 3 ml Hy-Pod Syringe (needles not
               included), individually wrapped, 120 per carton;
               (f) Catalog #10685: 5 ml Hy-Pod Syringe (needles not
               included), individually wrapped, 120 per carton;
               (g) Catalog #11771: 1 ml fill in 1.5 ml Hy-Pod Syringe,
               contained in the Lok-Pak-N Heparin Flush Kit, 200 per case;
               (h) Catalog #11773: 3 ml Hy-Pod Syringe, contained in the
               Lok-Pak-N Heparin Flush Kit, 200 per case;
               (i) Catalog #11775: 5 ml Hy-Pod Syringe, contained in the 
               Lok-Pak-N Heparin Flush Kit, 200 per case.  Recall #D-196-4.
CODE           (a) Catalog #10671: lot #93E059C, 93E060C, 93G001C, 93G002C;
               (b) Catalog #10673: lot #93E002C, 93F005C, 93F005D, 93F006C, 
               93H006C, 93L003C;
               (c) Catalog #10675: lot #93F011C, 93H012B;
               (d) Catalog #10681: lot #93J073B; 
               (e) Catalog #10683: lot #93H071B, 93K071B;
               (f) Catalog #10685: lot #93H074B, 93L007C;
               (g) Catalog #11771: lot #93E059B, 93E060B, 93F001B, 93G001B,
               93G002B; 
               (h) Catalog #11773: lot #93E002B, 93F005B, 93F006B, 93H006B,
               93L003B; 
               (i) Catalog #11775: lot #93E003B, 93F011B, 93L007B.
MANUFACTURER   SoloPak Medical Products Inc., formerly known as Smith &
               Nephew SoloPak, Franklin Park, Illinois.
RECALLED BY    SoloPak Medical Products Inc., Elk Grove Village, Illinois,
               by letter February 22, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       997,800 syringes were distributed; firm estimates 48,890
               syringes remain on the market.
REASON         Potency not assured through expiration date.

_______________
PRODUCT        S2 Inhalant, Racepinephrine HCl USP 2.25% Inhalation
               Solution, 15 ml glass bottles.  Recall #D-200-4.
CODE           Lot #R11911 EXP 11/94.
MANUFACTURER   Nephron Pharmaceuticals Corporation, Orlando, Florida.
RECALLED BY    Manufacturer, by letter January 28, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       402 bottles were distributed; firm estimates none remains on
               the market.
REASON         Subpotency.

_______________
PRODUCT        Pancrelipase Capsules, in 100 and 250 count bottles,
               indicated for patients with exocrine pancreatic enzyme
               deficiency, under the Trinity and Qualitest labels.  
               Recall #D-203/204-4.
CODE           Lot #K108A EXP 10/93.
MANUFACTURER   Pharmaceutical Delivery Systems, Inc., Creve Coeur,
               Missouri.
                                    -2-RECALLED BY    Trinity Technologies Corporation, Roseville, Michigan, by
               telephone August 4, 1993, followed by letter September 13,
               1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,470 100-tablet bottles and 1,198 250-tablet bottles were
               distributed.
REASON         Subpotency of amylase ingredient.

_______________
PRODUCT        Oxygen, USP in size D & E compressed gas aluminum cylinders.
               Recall #D-205-4.
CODE           All lots.  Lot numbers all begin with #14 followed by the
               date in which an H size cylinder was added to the tank bank.
MANUFACTURER   Medical Services of America, Inc., doing business as Medi-
               Home Care, Roanoke, Virginia.
RECALLED BY    Manufacturer, by telephone March 7, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Virginia.
QUANTITY       9 size E cylinders and 3 size D cylinders were distributed.
REASON         Current good manufacturing practice deficiencies.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Motrin IB Caplets (Ibuprofen) in foil packets of two, in
               dispenser of 100 packets and in bulk containers of 2,500
               packets.  Recall  #D-197-4.
CODE           Lot #115HT EXP 4/30/97 for both containers.
MANUFACTURER   The Upjohn Company, Kalamozoo, Michigan.
RECALLED BY    Manufacturer, by telephone and memorandum April 14, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       9,968 units were distributed.
REASON         Trace quantities of the lacquer used on exterior of foil
               package transferred to interior surface.

_______________
PRODUCT        Patient brochure for Depo-Provera Contraceptive Injection. 
               Recall #D-198-4.
CODE           Brochures with print date of November 1992 or February 1993
               distributed since 12/15/92.
MANUFACTURER   The Upjohn Company, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter.  Distribution of the incorrect
               patient brochure was discontinued 3/31/93 and new brochures
               began to be shipped 4/14/93.  Firm-initiated field
               correction complete.
DISTRIBUTION   Nationwide.
QUANTITY       238,600 brochures were distributed.
REASON         Incomplete printing of the brochure.

                                    -3-_______________
PRODUCT        Hydrocodone Bitartrate/Acetaminophen (HCB/APAP), 5/500 mg
               tablets, under the following labels:  H.N Nortcon
               Co./Luchem, 100's and 500's; Goldline 100's and 500's; Major
               500's.  Recall #D-199-4.
CODE           All lots.
MANUFACTURER   H.N. Norton Company, Shreveport, Louisiana.
RECALLED BY    Manufacturer, by letter February 2, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Undetermined.
REASON         Product does not meet dissolution specifications through
               expiration date.

_______________
PRODUCT        S2 Inhalant, Racepinephrine HCl USP 2.25% Inhalation
               Solution, 15 ml glass bottles, .  Recall #D-201-4.
CODE           Lot numbers:  99219, R109308 EXP 8/95, 10/95.
MANUFACTURER   Nephron Pharmaceuticals Corporation, Orlando, Florida.
RECALLED BY    Manufacturer, by letter February 4, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       6,741 bottles of lot 99219 and 8,604 bottles of lot R109308
               were distributed.
REASON         Product does not meet USP specifications for color.

_______________
PRODUCT        Medrol brand of methylprednisolone, USP, 4 mg tablets, in
               blister packs of 100, a corticosteroid.  Recall #D-202-4.
CODE           Lot #995JP EXP 12/31/97.
MANUFACTURER   The Upjohn Company, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter June 16, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,582 tablets were distributed.
REASON         Carton incorrectly declares "Methylprednisolone" as
               "Methylprednisone".  Blisters and insert are correct.

_______________
PRODUCT        Urinary Antiseptic Tablets, Rx indicated for the relief of
               discomfort of the lower urinary tract, in 100 and 1,000
               count bottles, under Trinity, Qualitest, Rugby, Moore, and
               Goldline labels.  Recall #D-204-4.
CODE           Lot No.    EXP Date
               A108D      7/94 
               A119D      7/94 
               B115D      7/94 
               C114D      9/94 
               D109C      4/94 
               D112D      10/94 
               D113D      10/94 
               E110C      5/94 

                                    -4-               F109C      6/94 
               G115C      7/94 
               J101C      9/94 
               J118C      9/94 
               L123C      5/94 
               M122C      6/94 
               F109D      10/94 
               J102C      9/94 
               L122C      5/94. 
MANUFACTURER   Trinity Technologies Corporation, Roseville, Michigan.
RECALLED BY    Manufacturer, by letter July 21, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       106,534 units wre distributed.
REASON         Product does not meet potency specifications.


RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS II
======================
_______________
PRODUCT        Red Blood Cells.  Recall #B-097-4.
CODE           Unit numbers:  04K28065, 04K28066, 04K28097, 04K28099,
               04R78705, 04R78707, 04R78709, 04R78711, 04R78715, 04S43640,
               04FK13491.
MANUFACTURER   American Red Cross Blood Services, Dedham, Massachusetts.
RECALLED BY    Manufacturer, by telephone July 28, 1992, followed by letter
               August 8, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Massachusetts.
QUANTITY       11 units.
REASON         Blood products, held at uncontrolled storage conditions were
               distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-170-4.
CODE           Unit #4052631.
MANUFACTURER   Community Blood Center of Greater Kansas City, Kansas City,
               Missouri.
RECALLED BY    Manufacturer.  Firm-initiated recall complete.
DISTRIBUTION   Missouri.
QUANTITY       1 unit.
REASON         Blood product, that tested repeatedly reactive for the
               antibody to hepatitis B core antigen (anti-HBC), was
               distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-172-4.
CODE           Unit #6283895.
MANUFACTURER   BloodCare, Dallas, Texas.
RECALLED BY    Manufacturer, by telephone September 7, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.

                                    -5-REASON         Red Blood Cells, collected from a donor currently taking the
               drug Proscar, were distributed for transfusion.

_______________
PRODUCT        Red Blood Cells.  Recall #B-173-4.
CODE           Unit 647102.
MANUFACTURER   W.E. and Lela I. Stewart Regional Blood Center, Tyler,
               Texas.
RECALLED BY    Manufacturer, by telephone February 25, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood product, collected from a donor who reported taking
               the medication Tegison, was distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Fresh Frozen Plasma.  Recall #B-174-4.
CODE           Unit #J27557.
MANUFACTURER   Texoma Regional Blood Center, Sherman, Texas.
RECALLED BY    Manufacturer, by visit January 6, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood product labeled with an incorrect expiration date, was
               distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Naturalyte Acid Concentrate for Bicarbonate Dialysis, for
               use in dialysis equipment during dialysis treatment:
               (a) Catalog No. 13-4004-1, Series 4000
               (b) Catalog No. 13-4010-8, Series 4000
               (c) Catalog No. 13-4002-5, Series 4000
               (d) Catalog No. 13-4018-1, Series 4000
               (e) Catalog No. 13-4022-3, Series 4000
               (f) Catalog No. 13-6608-7, Series 6000
               (g) Catalog No. 13-9027-7, Series 9000
               (h) Catalog No. 13-9029-3, Series 9000.
               Recall #Z-161/168-4.
CODE           Lot numbers:  (a) E2S027; (b) E2S028, (c) E2S029; 
               (d) E2S030; (e) E2S031; (f) E2S032; (g) E2S033;
               (h) E2S034; (i) E2S026.
MANUFACTURER   National Medical Care, Medical Product Division, Livingston,
               California.
RECALLED BY    National Medical Care (NMC), Medical Product Division,
               Rockleigh, New Jersey, by letter dated June 17, 1993, and by
               telephone beginning on or about May 28, 1993 or June 1,
               1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.

                                    -6-QUANTITY       434 55-gallon drums were distributed;
REASON         The products do not meet the NMC pyrogen specification for
               hemodialysate of 5.0 eu/ml.

_______________
PRODUCT        O.R. Saracap Respiratory Monitor and Capnograph, Model
               200009, for use during surgery and general anesthesia to
               monitor airway levels of carbon dioxide, oxygen, and nitrous
               oxide.  Recall #Z-345-4.
CODE           Software version 2.04.
MANUFACTURER   PPG Industries, Inc., Lenexa, Kansas.
RECALLED BY    Manufacturer, by visit beginning January 10, 1994.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       2,433 units remain in distribution.
REASON         The device contains invalidated software.

_______________
PRODUCT        Promatek ElectroStim Ultra 3F Therapy Ultrasound Systems,
               ultrasonic generators and applicators.  Recall #Z-402-4.
CODE           All serial numbers, 5108 through 5794.
MANUFACTURER   Diter-Elektroniikka Oy, Turku, Finland.
RECALLED BY    Promatek Medical Systems, Inc., Joliet, Illinois.  FDA
               approved the firm's corrective action plan May 10, 1993. 
               Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       106 units were distributed; firm estimates 68 units remain
               on the market.
REASON         The units were found in non-compliance with the performance
               standard for ultrasonic therapy products, in that:
               The error in the indication of power output exceeds      
               the +/- 20 percent error allowed by the performance          
               standard for the following tests: 
               - continuous wave power measurements 
               - pulsed power measurements 
               - +/- 10 percent line voltage power measurements 
               - short term life test power measurements; 
               Discrepancies are found in the operation control labeling; 
               Incorrect values are listed for Rtpa in the user
               information; 
               The measured effective radiating area (ERA) exceeds the
               listed ERA for the 2 MHz large applicator in the user
               information; 
               The measured beam non-uniformity ratio (BNR) exceeds the
               labeled maximum BNR for the 0.86 and 3.0 MHz applicators;
               and 
               The unit indicates output when none exists, and vice versa
               under certain conditions. 

_______________
PRODUCT        Eight French Single Lumen Groshong High Performance
               Catheters:

                                    -7-               (a) Product Code 7710922; (b) Product Code 7710924;
               (c) Product Code 7711922; (d) Product Code 7711924;
               (e) Product Code 7711932; (f) Product Code 7711934.
               Recall #Z-404/409-4.
CODE           All lots.
MANUFACTURER   Bard Access Systems, Inc., Salt Lake City, Utah.
RECALLED BY    Manufacturer, by telephone beginning January 13, 1994,
               followed by letters sent during the week of January 17-21,
               1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 3,550 units.
REASON         The thrombosed polyurethane liner of the distal silicone
               catheter segment was extruding through the Groshong valve
               opening out of the lumen of the silicone catheter into the
               lumen of the patient's central venous system.

_______________
PRODUCT        DeRoyal Surgical Laparotomy Sponges, pre-washed x-ray
               detectable: (a) 18" x 18", Product #9-1818;
               (b) 12" x 12" Lap Sponges, Product #1-1212.
               Recall #Z-439/440-4.
CODE           Lot numbers:  (a) 16325; (b) 16580.
MANUFACTURER   Qualtex, Inc., Division of DeRoyal Industries, Rose Hill,
               Virginia (packager/sterilizer).
RECALLED BY    DeRoyal Industries, Inc., Powell, Tennessee, by letters of
               February 1, 1994, and March 9, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   (a) Alabama, Maryland, Texas, North Carolina, Tennessee; (b) 
               California, Maryland, New Jersey, South Carolina, Texas,
               Pennsylvania, Tennessee.
QUANTITY       (a) 47 cases; (b) 43 cases.
REASON         The products were found to be non-sterile.

_______________
PRODUCT        Diagnostic Magnetic Resonance Imaging Equipment, intended
               for general diagnostic use to produce images used in
               diagnosing disease:  (a) Model SMT-50X; (b) Model SMT-100X.
               Recall #Z-441/442-4.
CODE           Serial numbers:  682001, 628002, 628004, 628021, 628591,
               628631, 628633; (b) 628841, 628844, 628846, 628885.
MANUFACTURER   Shimadzu Corporation, Koyoto, Japan.
RECALLED BY    Shimadzu Medical Systems, Torrance, California, by letter
               June 16, 1993.  Firm-initiated field correction complete.
DISTRIBUTION   Alaska, Arkansas, Florida, Louisiana, New Jersey, New York,
               Pennsylvania.
QUANTITY       (a) 8 systems; (b) 4 systems.
REASON         Software versions 2.0, 2.1, or 2.2 do not correctly store
               inverted image annotations on the disk.  The annotations
               will be reversed upon retrieval, for example, head to foot,
               posterior to anterior, left to right, which could result in
               misdiagnosis.

                                    -8-_______________
PRODUCT        Sunlamp Tanning Product, Sunquest 5000, Sunward I, and Tan
               AM 4000.  Recall #Z-460-4.
CODE           Model STI 1000.
MANUFACTURER   The Daavlin Distributing Company, Bryan, Ohio.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action
               plan February 18, 1994.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       544 units.
REASON         Digital timers could be reset to exceed maximum recommended
               time (MRET).


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Blood Chemistry Vacutainer Collection Tubes, Reorder #6440,
               an evacuated blood collection tube.  Recall #Z-205-4.
CODE           Lot #2C708 EXP 4/94.
MANUFACTURER   Becton Dickinson Vacutainer Systems, Sumpter, South
               Carolina.
RECALLED BY    Becton Dickinson and Company, Becton Dickinson Division,
               Franklin Lakes, New Jersey, by letter August 26, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       203,400 units were distributed; firm estimates less than 10%
               of the amount shipped remains on the market.
REASON         The Vacutainer Branch Tubes were defective causing red cells
               to adhere to the glass surface of the tubes.

_______________
PRODUCT        (a) Coulter Bilirubin, Total (B), Catalog #2907076; (b)
               Coulter Bilirubin, Direct (A+C) DACOS Reagent System,
               Catalog #2907075, intended for the in-vitro quantitative
               determination of Total and Direct Bilirubin with the DACOS
               DACOS XL chemistry analyzers. Recall #Z-378/379-4.
CODE           Lot numbers:  (a) 15672, 15703; (b) 15671, 15739, 15740.
MANUFACTURER   Trace America, Inc., Australia (reagents).
RECALLED BY    Coulter Corporation, Miami, Florida, by telephone followed
               by letter May 3, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Louisiana, Pennsylvania, Texas, California, Georgia,
               Colorado, Oklahoma, Oregon, Florida, Illinois, Minnesota,
               Washington state, North Carolina, Michigan, Ohio,
               Massachusetts, Kansas.
QUANTITY       Firm estimates none remains on the market.
REASON         The firm released a new bilirubin methodology with a new
               line of DACO chemistry reagents, that are not
               interchangeable.  The labeling failed to identify that they
               could not be interchanged.

                                    -9-_______________
PRODUCT        Coulter Optichem Chemistry Control, Level II, for in-vitro
               diagnostic use.  Recall #Z-380-4.
CODE           Vial lot #165001, kit lot numbers:  160006K3 and 160006K4
               EXP 10/14/94.
MANUFACTURER   Ciba Corning Diagnostics Corporation, Irvine, California.
RECALLED BY    Coulter Corporation, Miami, Florida, by letter October 7,
               1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, Turkey, United Kingdom, Germany, Puerto
               Rico, France.
QUANTITY       239 kits of lot 160006k3 and 10 kits of lot 160006K4 were
               distributed.
REASON         The tables of assigned values had typographical errors.

_______________
PRODUCT        Coulter Chemistry Control Level II, 10 x 15 ml, Part
               #7546995, used as a quality control material for the Coulter
               DACOS Analyzer, DACOS XL and Optichem Chemistry Systems. 
               Recall #Z-388-4.
CODE           Kit lot numbers:  160006K, 160006K1, 160006K2.
MANUFACTURER   Trace America, Inc., Australia.
RECALLED BY    Coulter Corporation, Miami, Florida, by letter June 25,
               1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Bermuda, Europe, Canada, United Kingdom, Spain,
               Germany.
QUANTITY       336 kits of lot 160006K, 165 kits of lot 160006K1, and 150
               kits of lot 160006K2 were distributed.
REASON         There was a change in the DACAL Calibrator assigned values.  
               This resulted in a shift in the mean and low recovery values
               when using new Coulter Dacos Uric Acid reagent.

_______________
PRODUCT        Coulter ISE Module, a component of Coulter Opti-Chem 120 and
               180 Systems, used to perform clinical chemistry analyses:
               (a) Coulter OptiChem 120 & 180 Systems Software Catalog No.
               6604263; 
               (b) Coulter OptiChem 120 & 180 Systems Software Catalog No.
               6604264; 
               (c) Coulter OptiChem 120 & 180 Systems Software Catalog No.
               6604271; 
               (d) Coulter OptiChem 120 & 180 Systems Software Catalog No.
               6604273.  Recall #Z-389/392-4.
CODE           All serial numbers.
MANUFACTURER   Kone Instruments, Helsinki, Finland.
RECALLED BY    Coulter Corporation, Miami, Florida, by "important product
               information" letter dated May 1, 1992.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 185 units ramain on the market.
REASON         There was an error in the ISE module software that allowed
               test results to be reported outside the validated range
               which may not be accurate.

                                   -10-_______________
PRODUCT        4C Plus Normal Cell Control, for in-vitro diagnostic use:
               (a) 4C Plus Normal Cell Control, Part No. 7546923; 
               (b) 4C Plus Triple Cell Control 9 x 3.3 mL, Part No.
               7546771;
               (c) 4C Plus Triple Cell Control 6 x 3,3 mL, Part No.
               7546968.  Recall #Z-393/395-4.
CODE           Lot numbers:  (a) 776F; (b) 103434K; (c) 108159K.
MANUFACTURER   Coulter Corporation, Diagnostics Group, Hialeah, Florida.
RECALLED BY    Coulter Corporation, Miami, Florida, by letter beginning
               November 8, 1991.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       Firm estimates none remains on the market.
REASON         Customers reported high Hgb recovery when using these lots
               of control, on Coulter T series, S880, and JT instruments.
               Assay values printed in the package insert were unable to be
               retrieved.

_______________
PRODUCT        Centermark Dual Lumen Peripherally Inserted Central Venous
               Catheter.  Recall #Z-403-4.
CODE           Catalog #C2-202-58G, Lot #308241.
MANUFACTURER   Menlo Care, Inc., Menlo Park, California.
RECALLED BY    Manufacturer, by telephone and by letter in October 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Firm estimates none remains on the market.
REASON         There was a potential for crossover leakage between the two
               lumens during blood withdrawal.

_______________
PRODUCT        Irrijet Irrigation System, Product #55-0504, used for
               external wound irrigation.  Recall #Z-410-4.
CODE           Lot numbers:  3514 and 3529.
MANUFACTURER   Ackrad Laboratories, Cranford, New Jersey.
RECALLED BY    Manufacturer, by telephone and by letter February 26, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       560 units were distributed.
REASON         The disposable tips may be plugged resulting in lack of
               fluid flow.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
===========
_______________
PRODUCT        (a) Horse and Pony Feed (740); (b) Phase Two (751) and Phase
               Three Horse Feed (752); (c) Horse and Pony Feed (747); 
               (d) Rabbit Pellets (711).  Recall #V-037/038-4.
CODE           Dates of productions: 11/9/93, 11/11/93, 11/23/93, 12/10/93,
               12/30/93, 1/14/93.
MANUFACTURER   Molinos de Puerto Rico, Inc., Guaynabo, Puerto Rico.

                                   -11-RECALLED BY    Manufacturer, by February 7, 1994 inspection.  Firm-
               initiated recall complete.
DISTRIBUTION   Puerto Rico.
QUANTITY       (a) 48,000 pounds; (b) 155,000 pounds; (c) 12,280 pounds;
               (d) 20,000 pounds.
REASON         Possible cross contamination of horse feed with lincomycin.  
               Possible cross contamination of rabbit pellets with tylosin
               and sulfamethazine. 

_______________
PRODUCT        Stepsaver WDC, medicated swine feed containing carbadox and
               pyrantel tartrate.  Recall #V-039-4.
CODE           LK8974.
MANUFACTURER   Moorman Manufacturing Company, LaMars, Iowa.
RECALLED BY    Moorman Manufacturing Company, Quincy, Illinois, by
               telephone December 31, 1993.  Firm-initiated recall
               complete.  See also FDA talk paper T94-10, February 2, 1994.
DISTRIBUTION   Wisconsin.
QUANTITY       2 tons were distributed; firm estimates none remains on the
               market.
REASON         Unlabeled bulk swine feed containing carbadox and pyrantel
               tartrate erroneously shipped to a dairy farmer.


MEDICAL DEVICE SAFETY ALERTS:
=============================================
_______________
PRODUCT        Bipolar Forceps: (a) 60-5120-001, Cushing Bayonet;
               (b) 60-5121-001, Scoville-Greenwood Bayonet;
               (c) 60-5122-001, Jewelers, Straight Tip;
               (d) 60-5123-001, Semkin, Micro-Tip.
               Safety Alert #N-040/043-4.
CODE           None.
MANUFACTURER   Aspen Laboratories, Inc., Englewood, Colorado.
ALERTED BY     Manufacturer, by letter August 25, 1993.
DISTRIBUTION   Nationwide and international.
QUANTITY       351 units.
REASON         Inadequate cleaning of the bipolar forceps may cause an
               unintended change in power setting and/or difficulty in
               activation of the Aspen Bistat electrosurgical unit.


SEIZURE:
==================================================================
_______________
PRODUCT        Frozen Yellow Fin Tuna Loins (94-630-112).
CHARGE         Adulterated - The article bears or contains a poisonous and
               deleterious substance, histamine, which ordinarily renders
               it injurious to health; and, it consists in part of a
               decomposed substance by reason of the presence of decomposed
               fish. 
LOCATIONS      Cityice Cold Storage Company, Seattle, Washington.


                                   -12-FILED          March 3, 1994; U.S. District Court for the Western 
               District of Washington; Civil #C 94-316R; FDC #66936.
SEIZED         March 9, 1994 - goods valued at approximately $33,150.

                                   -13-

END OF ENFORCEMENT REPORT FOR MARCH 23, 1994.  BLANK PAGES MAY
FOLLOW.
                                   ####