FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
11/24/1993

Recalls and Field Corrections:  Foods -- Class II -- 11/24/1993

November 24, 1993                                                93-47

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        All types and sizes of bags of corn meal:
               (a) Sholar's River Bank Buttermilk Hush Puppy Mix, 16
               ounces;
               (b) Sholar's River Bank Hushpuppy Mix, 5 pounds;
               (c) Sholar's Shur-Rise Stone Ground Air Sifted White Self
               Rising Corn Meal Mix, 24-ounce and 5-pound size; 
               (d) Stone Ground Air Sifted White Sholar's Medium ground
               Corn Meal Mix, in 2-pound and 5-pound bags;
               (e) Stone Ground Air Sifted White Sholar's Fine Ground Corn
               Meal Mix, in 2-pound and 5-pound bags;
               (f) Sikes' Corn Meal, in 5-pound bags.  Recall #F-096/101-4.
CODE           All lots manufactured through JUL 13, 1993.
MANUFACTURER   C & C Corn Products, Inc., Pelham, Georgia.
RECALLED BY    Manufacturer, by visit beginning July 19, 1993.
               Firm-initiated recall ongoing.
DISTRIBUTION   Georgia.
QUANTITY       Firm estimates 5,125 pounds of all products remain on the
               market.
REASON         The products are contaminated with aflatoxin.

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Ultravate (halobetasol proprionate) Ointment, 0.05%, 50 g
               tubes, Rx topical corticosteroid.  Recall #D-026-4.
CODE           Lot #145F016 EXP 5/96.
MANUFACTURER   Westwood Squibb, Buffalo, New York.
RECALLED BY    Manufacturer, by telephone followed by letters dated October
               27, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       394 cases (24 units per case) were distributed.
REASON         A portion of the lot was adulterated with an equipment
               cleaning solution.

_______________
PRODUCT        Transfilled Compressed Medical Oxygen in C, D, E, M, and S
               size cylinders.  Recall #D-027-4.
CODE           All fill dates.
MANUFACTURER   Robinson's Industrial Gas and Equipment Corporation,
               Amityville, New York.
RECALLED BY    Manufacturer, by visit November 8, 1993, to be followed by
               letters.  Firm-initiated recall ongoing.
DISTRIBUTION   New York.
QUANTITY       Approximately 86 cylinders were distributed.
REASON         Current good manufacturing practice deficiencies.

_______________
PRODUCT        Transfilled Compressed Medical Oxygen in D, C, E, M, and 244
               cubic feet cylinders.  Recall #D-028-4.
CODE           All fill dates.
MANUFACTURER   Anderson Fire Equipment Company, Inc., Batshore, New York.
RECALLED BY    Manufacturer, by letter delivered beginning the week of
               November 8, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   New York.
QUANTITY       304 cylinders were distributed.
REASON         Current good manufacturing practice deficiencies.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Percodan Tablets, in blister packs of 25, used for the
               relief of moderate to moderately severe pain.  
               Recall #D-029-4.
CODE           Lot EEA-00A1 EXP 12/94.
MANUFACTURER   Dupont Merck Pharm, Manati, Puerto Rico.
RECALLED BY    Manufacturer, by telephone September 24, 1993, followed by
               letter.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and Puerto Rico.
QUANTITY       5,691 boxes containing 10 blister packs were distributed;
               firm estimates 75 boxes remain on the market.
REASON         Aspirin component does not meet stability specifications for
               dissolution.

                                    -2-RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS I
========================
_______________
PRODUCT        Pacemaker/Defibrillators:  (a) Zoll PD 1200
               Pacemaker/Defibrillator; (b) Zoll D 900 Defibrillator, for
               the purposes of converting ventricular fibrillation (VF), to
               sinus rhythm or other cardiac rhythms capable of producing
               hemodynamically significant heart beats.  
               Recall #Z-037/038-4.
CODE           Serial numbers:  (a) 4663 to 6191 inclusive; (b) 2009 to
               2637 inclusive.
MANUFACTURER   Omron, Japan (manufacturers the K1-G6B component relay).
RECALLED BY    Zoll Medical Corporation, Burlington, Massachusetts, by
               letter June 15, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately (a) 1,467 units; (b) 622 units.
REASON         A defective high voltage board component and a
               malfunctioning relay switch makes the defibrillation
               function inoperative.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
======================
_______________
PRODUCT        System 1000 Thyroid Uptake System using Software Version
               2.30, used to determine the activity of the thyroid gland by
               measuring the amount of radionucleide taken up by the gland. 
               Recall #Z-043-4.
CODE           All units.
MANUFACTURER   Canberra Industries, Meriden, Connecticut (software
               developer).
RECALLED BY    Capintec, Inc., Pittsburgh, Pennsylvania, by letter June 14,
               1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, Saudi Arabia, France.
QUANTITY       165 units.
REASON         Software Version 2.30 contains a defect whereby incorrect
               capsule counts occurred during a Thyroid Uptake Test when
               the initial count was taken between 7:00 pm and 12:00
               midnight.

_______________
PRODUCT        Software upgrade revision 0.1 used in Ultrasound Imaging
               Systems:  (a) Sonos 500 Ultrasound Imaging System Model
               77025A; (b) Sonos 1000 Ultrasound Imaging System Model
               77030A; (c) Sonos 1500 Ultrasound Imaging System Model
               77035A; (d) Sonos or Ultrasound Imaging System Model 77040A. 
               Recall #Z-079/082-4.
CODE           HP Sonos Software Revision 0.1.
MANUFACTURER   Hewlett-Packard Company, Imaging Systems, Andover,
               Massachusetts.
RECALLED BY    Manufacturer, by letters of March 11 and 19, 1993.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.

                                    -3-QUANTITY       665 systems or upgrade kits with Revision 0.1.
REASON         The 0.1 software revision used in these devices was released
               without being validated to the original design
               specifications resulting in an error in the mean pressure
               gradient.

_______________
PRODUCT        Acryl-X System Ultrasound Bone Cement Removal System. 
               Recall #Z-083-4.
CODE           A varied range of combination serial numbers.
MANUFACTURER   Lydo Precision Product, Mountain Lakes, New York.
RECALLED BY    Sonokinetics Group, Hoboken, New Jersey, by letter January
               21, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       23 units; firm estimates none remains on the market.
REASON         If the Acryl-X handpiece is improperly autoclaved or cleaned
               with Cidex, the patient or operating room personnel could be
               exposed to above normal levels of electrical current during
               surgery.

_______________
PRODUCT        Advertisements for "Dry Docs Disposable Protective Shoe
               Covers" used to eliminate the risk of contamination and
               provide security against contagious disease and "Economy dry
               Docs", used to eliminate the risk of contamination.  Recall
               #Z-088-4.
CODE           None.
MANUFACTURER   Hydro-Med Products, Inc., subsidiary of The Spectrum
               Companies, Dallas, Texas.
RECALLED BY    Hydro-Med Products, Inc., subsidiary of The Spectrum
               Companies, Houston, Texas, by letters dated September 30,
               1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Alabama, California, Florida, Georgia, Illinois, Kansas,
               Michigan, Missouri, New Jersey, New York, Oregon, Texas,
               Wisconsin, Canada.
QUANTITY       Undetermined.
REASON         The advertised devices were labeled for medical use without
               an approved 510(k).

RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Hearing Aids:  (a) Starkey CE 7 and CE 8 in-the-ear Hearing
               Aids; (b) SA 57 and SA S8 Custom Canal - Starkey SecretEar
               Mini in-the-ear Hearing Aids; (c) Trilogy - Programmable
               Hearing Aids.  Recall #Z-114/116-4.
CODE           All lots.
MANUFACTURER   Starkey Labs, Inc., Eden Prairie, Minnesota.
RECALLED BY    Manufacturer, by memorandum May 7, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       20,000 to 30,000 units per year are distributed.

                                    -4-REASON         Unwarranted claims.  The use of the devices does not cause
               injury in itself; however, lack of appropriate medical
               evaluation may lead to potential delay or error in
               treatment.

_______________
PRODUCT        Miracle Ear Hearing Aids.  Recall #Z-117-4.
CODE           All lots.
MANUFACTURER   Dahlberg, Inc., Golden Valley, Minnesota.
RECALLED BY    Manufacturer, by letters of April 22 and 29, 1993.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       20,000 to 30,000 units per year are distributed.
REASON         Unwarranted claims.  The use of the device does not cause
               injury in itself; however, lack of appropriate medical
               evaluation may lead to potential delay or error in
               treatment.

                                    -5-

END OF ENFORCEMENT REPORT FOR NOVEMBER 24, 1993.  BLANK PAGES MAY
FOLLOW.
                                   ####