01/13/1993

Recalls and Field Corrections:  Drugs -- Class II -- 01/13/1993


January 13, 1993                                               93-2

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
              
PRODUCT        Lortab 7.5/500 Tablets CIII (7.5 mg hydrocodone bitartrate and   
               500 mg acetaminophen).  RecallD-093-3.
CO           920409D EXP 4/94.
MANUFACTURER   Mikart, Inc., Atlanta, Georgia.
RECALLED BY    Whitby Pharmaceuticals, Inc., Richmond, Virginia (responsible    
               firm), by letter December 4, 1992.  Firm-initiated recall        
               ongoing 
DISTRIBUTION   Nationwide.
QUANTITY       4,696 100-tablet unit dose packages (each containing 4           
               25-tablet packages) were distributed.
REASON         A carton labeled as 5 mg Lortab contained correctly labeled      
               labeled unit dose 7.5 mg.

              
PRODUCT        Medical Oxygen Cylinders, USP, in D, E and F (F same as C)       
               sized metal cylinders.  RecallD-099-3.
CODE           LotA02 C344, EXP 12/97.
MANUFACTURER   Mountain Airgas Medical Division, Salt Lake City, Utah.
RECALLED BY    Manufacturer, by telephone December 23, 1992, followed by        
               letter.  Firm-initiated recall ongoing.
DISTRIBUTION   Utah.
                                       
QUANTITY       149 cylinders were distributed; firm estimates that 40% of the   
               lot may still remain on market.
REASON         One cylinder contained trace quantities of chlorine.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
              
PRODUCT        Lasix (furosemide) Oral Solution, 60 ml, in brown glass bottle   
               10 mg/ml, a Rx diuretic used in treatment of edema.
               RecallD-094-3.
CODE           Lot0630011 EXP 4/94.
MANUFACTURER   Hoechst-Roussel Pharmaceuticals, Inc., Somerville, New Jersey.
RECALLED BY    Manufacturer, by letter dated April 27, 1992.  Firm-initiated    
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       28,225 bottles were distributed.  Firm estimates none remains    
               on the market.
REASON         Presence of cardboard particles.

              
PRODUCT        Ventolin (Albuterol, USP), Inhalation Aerosol, 6.8 g, 80         
               metered inhalations, a professional sample for use with the      
               blue Ventolin Inhalation Aerosol Actuator.  RecallD-095-3.
CODE           Lot numbers:  Z31572KS EXP 11/94, Z31582KS EXP 12/94.
MANUFACTURER   Glaxo, Inc., Zebulon, North Carolina.
RECALLED BY    Manufacturer, by letter November 24, 1992.  Firm-initiated       
               field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       87,503 cans of lot Z31572KS and 87,452 cans of lot Z31582KS      
               were distributed.
REASON         Some units were packaged with a Beclovent actuator.

              
PRODUCT        Cortisone Acetate, 25 mg, packaged under the following labels:   
               Richlyn, in bottles of 100; Rugby, in bottles of 100 and 500;    
               and URL, in bottles of 100.  RecallD-096-3.
CODE           Lot 44783 packaged under Richlyn, Rugby and URL labels and lot   
               44784 packaged only under the Rugby label.
MANUFACTURER   Richlyn Laboratories, Inc., Philadelphia, Pennsylvania.
RECALLED BY    Manufacturer, by letter September 28, 1992.  Firm-initiated      
               recall ongoing.
DISTRIBUTION   New York, Florida, New Jersey, Connecticut, Illinois,            
               California, Ohio, Pennsylvania, Georgia.
QUANTITY       448,300 tablets of lot 44783 and 468,500 tablets of lot 44784    
               were distributed; firm estimates that very little of this        
               product remains on market.
REASON         Some tablets are debossed with an incorrect NDC number.

              
PRODUCT        Ibuprofen 200 mg, in 1000 tablet containers, an analgesic        
               drug.  RecallD-097-3.
CODE           LotJ03292 EXP 8/95.

                                      -2-
MANUFACTURER   Interpharm, Inc., Plainview, New York.
RECALLED BY    Manufacturer, by telephone and letter November 17, 1992.         
               Firm-initiated recall ongoing.
DISTRIBUTION   Texas, Ohio.
QUANTITY       200 bottles were distributed.
REASON         Label declares an incorrect lot number.

              
PRODUCT        Ibuprofen, 600 mg, in 500 tablet bottles, an analgesic drug.     
               RecallD-098-3.
CODE           LotA00291 EXP 12/93.
MANUFACTURER   Interpharm, Inc., Plainview, New York.
RECALLED BY    Manufacturer, by telephone November 17, 1992.  Firm-initiated    
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       989 bottles were distributed.
REASON         Product exceeds ANDA specifications for degradation product.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II

               
PRODUCT        (a) Red Blood Cells; (b) Plasma.  RecallB-050/051-3.
CODE           Unit numbers 20GG42606, 20GG42610.
MANUFACTURER   American Red Cross, Boise, Idaho.
RECALLED BY    Manufacturer, by telephone September 14, 1992 and by letter      
               dated September 16, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Idaho, California.
QUANTITY       2 units of each component.
REASON         Blood products which tested initially reactive for the           
               hepatitis B surface antigen (HBsAg) were distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
               
PRODUCT        (a) Red Blood Cells; (b) Plasma.  RecallB-052/053-3.
CODE           Unit numbers:  (a) 20GG42605, 20GG42607, 20GG42608, 20GG42609,   
               20GG42611, 20GG42613, 20GG42614, 20GG42619;
               (b) 20GG42605, 20GG42607, 20GG42608, 20GG42609, 20GG42613,       
               20GG42614, 20GG42619;
MANUFACTURER   American Red Cross, Boise, Idaho.
RECALLED BY    Manufacturer, by telephone September 14, 1992 and by letter      
               dated September 16, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Idaho, California.
QUANTITY       (a) 8 units; (b) 7 units.
REASON         Blood products which tested initially reactive for the           
               hepatitis B surface antigen (HBsAg) were distributed.

              
PRODUCT        (a) Whole Blood; (b) Recovered Plasma.  RecallB-102/103-3.
CODE           Unit numbers:  (a) 1150827; (b) 1148318.
MANUFACTURER   Houchin Community Blood Bank, Bakersfield, California.

                                      -3-
RECALLED BY    Manufacturer, (a) by telephone July 21, 1992; (b) by letter      
               dated September 17, 1992.  Firm-initiated recall complete.
DISTRIBUTION   (a) California; (b) Delaware.
QUANTITY       1 unit of each component.
REASON         Blood products labeled with the incorrect expiration date, or    
               collected from an augologous donor who did not meet the          
               criteria for homologous donation, were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
              
PRODUCT        Dermabrader Centrifugal Chuck, Part1360-4, used to "sand or    
               plane" the elevated surface of the skin in order to minimize     
               the effects of a scar or blemish.  RecallZ-178-3.
CODE           All lots produced in April 1992.
MANUFACTURER   Stryker Corporation, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter September 23, 1992.  Firm-initiated      
               recall ongoing.
DISTRIBUTION   Nationwide, Ecuador, The Netherlands, United Kingdom.
QUANTITY       10 units were distributed.
REASON         The device may not be packaged with the correct warning insert.

              
PRODUCT        Lorad M-III Breast Imaging System -- Mammography system device,  
               a radiological device intended for diagnostic breast imaging.    
               RecallZ-179-3.
CODE           All M-III units manufactured before 9/23/91.
MANUFACTURER   Lorad Medical Systems, Danbury, Connecticut.
RECALLED BY    Manufacturer, by sending field modification instruction (FMI)    
               during the period of mid September 1991, to mid October 1991,    
               to all dealer home offices which purchased and M-III unit which  
               was manufactured before 9/23/91.  Firm-initiated field           
               correction ongoing.
DISTRIBUTION   Nationwide, Canada, Australia, Sweden.
QUANTITY       137 units were distributed.
REASON         The breast compression device continues to drive downward after  
               failure of a device fuse.

              
PRODUCT        Atrium - Water Seal, Chest Drainage Unit, (multiple purpose      
               system), a sterile, Rx medical device intended to help drain     
               fluids and air from the pleural cavity or mediastinum during or  
               after thoracic surgery.  RecallZ-185-3.
CODE           Model2060-000, lot260283-000.
MANUFACTURER   Atrium Medical Corporation, Hollis, New York.
RECALLED BY    Manufacturer, by letter October 28, 1992.  Firm-initiated        
               recall complete.
DISTRIBUTION   South Carolina, Kentucky, Minnesota.
QUANTITY       132 units were distributed.
REASON         Some units were assembled with a body with two openings for      
               patient drain tubes instead of one.


                                      -4-
              
PRODUCT        Transcutaneous Electrical Nerve Stimulators (TENS devices),      
               promoted for auriculotherapy:
               (a) Stim Flex Model 400; (b) Stem Flex Model 200;
               (c) Stem Flex Model 80.  RecallZ-187/189-3. 
CODE           All serial numbers.
MANUFACTURER   Electro Medical, Inc., Tulsa, Oklahoma.
RECALLED BY    Manufacturer, by letter October 28, 1992.  Firm-initiated        
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 950 units were distributed.
REASON         Products are prescription devices which were distributed         
               without prescription legends.  Models 200 and 400 are            
               preamendment devices.  Model 80 is being distributed without     
               premarket approval authority.

              
PRODUCT        CM5000 Airway Pressure Monitor, also referred to as CM5000 APM,  
               intended to monitor pressurized breathing circuits as in lung    
               ventilator systems.  RecallZ-190-3.
CODE           LotL959, serial numbers:  5060, 5195, 5426, 6280, 7268, 7354,  
               7273-7282, 7284-7292.
MANUFACTURER   Canadian Monaghan, Ltd., London, Ontario Canada.
RECALLED BY    Monaghan Medical Corporation, Plattsburgh, New York, by          
               telephone June 3, 1992, followed by letter.  Firm-initiated      
               recall complete.
DISTRIBUTION   Pennsylvania, Ohio, New York, Massachusetts, North Carolina.
QUANTITY       20 units were distributed; firm estimates none remains on        
               market. 
REASON         Incorrect wiring may cause the remote alarm not to function.

              
PRODUCT        Ritter Examination Tables:  (a) Model 98; (b) Model 99.
               RecallZ-196/197-3.
CODE           Serial numbers:  CN1000 through CN1399.
MANUFACTURER   Midmark Corporation, Versailles, Ohio.
RECALLED BY    Manufacturer, by letter August 31, 1992.  Firm-initiated recall  
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       400 tables were distributed; 374 of the tables are still in use.
REASON         Table may not be able to safely support a heavy patient.

              
UPDATE         McGaw Hyperformer System, RecallZ-144-3, which appeared in     
               the December 30, 1992, Enforcement Report should read:
               CODE:  Units within the serial number range G1000 through        
               G1403.  Only units with software versions E and EC1 are          
               involved in the recall.

                                      -5-
MEDICAL DEVICE SAFETY ALERTS:
              
PRODUCT        ACS Spectrum Coronary Dilatation catheters:
               (a)  Catalog No. 10070-107, 1.5 mm diameter;
               (b)  Catalog No. 10071-107, 2.0 mm diameter;