FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
12/22/1992

RECALLS AND FIELD CORRECTIONS: December 23, 1992

                                FOODS -- CLASS II
                                                                92-52
              
PRODUCT        Healthy Choice Nonfat Pasteurized Process Cheese Product, 2      
               pound loaves, all flavors, regular and Mexican.
               Recall #F-132-3.
CODE           All lots.
MANUFACTURER   Beatrice Cheese, Inc., Waukesha, Wisconsin.
RECALLED BY    Manufacturer, by telephone November 2, 1992.  Firm-initiated     
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 200,000 loaves were distributed.
REASON         The product has a remote potential for supporting the growth of  
               Clostridium Botulinum and toxin production.


              
PRODUCT        Shucks Buttered Popcorn, packed in 28 ounce clear cellophane     
               bags.  Recall #F-133-3.
CODE           None.
MANUFACTURER   Golden Harvest Popcorn Company, Overland Park, Kansas.
RECALLED BY    Manufacturer, by telephone November 3, 1992.  Firm-initiated     
               recall complete.
                                       
DISTRIBUTION   Local distribution.
QUANTITY       Approximately 288 bags were distributed.
REASON         Product contains undeclared FD&C Yellow #2.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
              
PRODUCT        Platelets, Pheresis.  Recall #B-092-3.
CODE           Units Q15485 and Q18718BB.
MANUFACTURER   Carolina-Georgia Blood Center, Greenville, South Carolina.
RECALLED BY    Consignees notified firm by telephone April 30, 1992, and July   
               20, 1992.  Firm-initiated recall complete.
DISTRIBUTION   South Carolina and Missouri.
QUANTITY       2 units.
REASON         Products labeled with incorrect expiration dates were            
               distributed.

                 
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  Recall #B-093/094-3.
CODE           Unit #N41937.
MANUFACTURER   Peninsula Blood Bank, Burlingame, California.
RECALLED BY    Manufacturer, by telephone May 26, 1992.  Firm-initiated recall  
               complete.
DISTRIBUTION   California.
QUANTITY       1 unit of each component.
REASON         Blood products which tested repeatedly reactive for the          
               antibody to the hepatitis C virus (anti-HCV) were distributed.

              
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-095/096-3.
CODE           Unit #M34154.
MANUFACTURER   Milton S. Hershey Medical Center, University Hospital Blood      
               Bank, Hershey, Pennsylvania.
RECALLED BY    Manufacturer, by letter October 2, 1992.  Firm-initiated recall  
               ongoing.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit of each component.
REASON         Blood products, which tested negative for the antibody to human  
               immunodeficiency virus type 1 (anti-HIV-1), but were collected   
               from a donor who previously tested repeatedly reactive for       
               anti-HIV-1, were distributed.

              
PRODUCT        Recovered Plasma.  Recall #B-097-3.
CODE           Unit #147619.
MANUFACTURER   York Hospital Blood Bank, York, Pennsylvania.
RECALLED BY    Manufacturer, by letter October 22, 1992.  Firm-initiated        
               recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit.

                                      -2-
REASON         Blood product, which tested negative for the antibody to human   
               immunodeficiency virus type 1 (anti-HIV-1), but was collected    
               from a donor who previously tested repeatedly reactive for       
               anti-HIV-1, was distributed.

              
PRODUCT        Recovered Plasma.  Recall #B-098-3.
CODE           Unit numbers:  9201530, 9201533, 9201535, 9201536, 9201537,      
               9201539, 9201543.
MANUFACTURER   Children's Hospital, Washington, D.C.
RECALLED BY    Manufacturer, by letters dated October 16, 1992, and November    
               30, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Maryland.
QUANTITY       7 units.
REASON         Blood products, incorrectly tested for the antibodies to human   
               immunodeficiency virus, types 1 and 2 (anti-HIV-1/2), were       
               distributed for further manufacture into in-vitro products.

              
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  Recall #B-099/100-3.
CODE           Unit numbers:  40GL46148, 40GL46150, 40GL46151, 40GL46152,       
               40GL46153, 40GL46154, 40GL46155, 40GL46157.
MANUFACTURER   American Red Cross Blood Services, Peoria, Illinois.
RECALLED BY    Manufacturer, by telephone October 23, 1992, followed by letter  
               November 4, 1992.  Firm-initiated recall complete.
DISTRIBUTION   (a) Illinois, Iowa, Washington, D.C., Missouri, Puerto Rico;
               (b) California.
QUANTITY       8 units of each component.
REASON         Blood products, incorrectly tested for hepatitis B surface       
               antigen (HBsAg), were distributed.

RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
              
PRODUCT        Recovered Plasma.  Recall #B-088-3.
CODE           Unit numbers:  C81448, C81466, C81473
               C81476, C81479, C81484, C81494, C81505, C81510,
               C81512, C81524, C81540, C81541, C81549, C81550,
               C81595, C81596, C81598, C81632, C81634, C81635,
               C81648, C81674, C81685, C81686, C81687, C81688,
               C81692, C81693, C81704, C81713, C81714, C81716,
               C81717, C81718, C81722, C81723, C81748, C81749,
               C81769, C81776, C81777, C81778, C81781, C81782,
               C81795, C81796, C81797, C81803, C81806, C81808,
               C81809, C81821, C81829, C81831, C81833, C81856,
               C81863, C81864, C81867, C81870, C81875, C81876
               C81879, C81880, C81888, C81890, C81898, C81900,
               C81912, C81916, C81922, C81930, C81931, C81934,
               C81935, C81938, C81940, C81966, C81969, C81979,
               C81989, C81990, C81991, C81993, C81996, C81997,
               C82002, C82005, C82007, C82009, C82013, C82014,

                                      -3-
               C82018, C82028, C82030, C82039, C82044, C82048,
               C82057, C82058, C82059, C82060, C82064, C82066,
               C82067, C82068, C82069, C82071, C82075, C82103,
               C82109, C82111, C82116, C82121, C82127, C82137,
               C82140, C82142, C82149, C82163, C82164, C82165,
               C82179, C82182, C82185, C82187, C82189, C82192,
               C82195, C82196, C82202, C82213, C82238, C82239,
               C82240, C82242, C82245, C82252, C82254, C82257,
               C82260, C82261, C82263, C82266, C82271, C82276,
               C82287, C82289, C82290, C82291, C82292, C82293,
               C82296, C82299, C82308, C82309, C82315, C82320,
               C82321, C82322, C82327, C82329, C82330, C82331,
               C82338, C82342, C82344, C82345, C82349, C82350,
               C82362, C82365, C82367, C82373, C82375, C82376,
               C82379, C82381, C82395, C82398, C82402, C82429,
               C82430, C82473, C82479, C82485, C82487, C82495
               C82496, C82516, C82520, C82534, C82539, C82549,
               C82577, C82579, C82586, C82603, C82604, C82613,
               C82615, C82627, C82628, C82629, C82672, C82673,
               C82708, C82717, C82753, C82754, C82781, C82784,
               C82786, C82788, C82799, C82803, C82804, C82805,
               C82809, C82812, C82818, C82829, C82831, C82834,
               C82842, C82872, C82889, C82891, C82892, C82899,
               C82904, C82905, C82908, C82911, C82924, C82929,
               C82931, C82933, C82936, C82942, C82951, C82955,
               C82956, C82957, C82958, C82965, C82970, C82971,
               C82972, C82977, C82992, C82994, C83009, C83013,
               C83014, C83023, C83028, C83032, C83034, C83035,
               C83036, C83042, C83052, C83078, C83096, C83099,
               C83100, C83108, C83112, C83114, C83120, C83130,
               C83145, C83177, C83184, C83186, C83187, C83188,
               C83189, C83195, C83199, C83200, C83202, C83203,
               C83204, C83208, C83210, C83218, C83224, C83226,
               C83231, C83233, C83249, C83257, C83259, C83263,
               C83267, C83268, C83271, C83275, C83276, C83277,
               C83280, C83282, C83284, C83285, C83286, C83290,
               C83293, C83295, C83298, C83301, C83307, C83310,
               C83324, C83335, C83337, C83339, C83341, C83343,
               C83344, C83345, C83346, C83351, C83352, C83358,
               C83362, C83364, C83366, C83368, C83369, C83374,
               C83375, C83380, C83385, C83389, C83399, C83400,
               C83404, C83405, C83406, C83411, C83412, C83413,
               C83421, C83422, C83426, C83427, C83439, C83442,
               C83455, C83456, C83463, C83476, C83480, C83482,
               C83484, C83485, C83494, C83496, C83497, C83499.
MANUFACTURER   Nueces County Medical Society Community Blood Bank, Corpus       
               Christi, Texas.
RECALLED BY    Manufacturer, by letter October 22, 1992.  Firm-initiated        
               recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       363 units.

                                      -4-
REASON         Blood products, which were unsuitable for use in the             
               manufacture of injectable products but were labeled as if        
               suitable for the manufacture of injectable products, were        
               distributed.

              
PRODUCT        (a) Cryoprecipitated AHF; (b) Recovered Plasma.
               Recall #B-089/090-3.
CODE           Unit #25GZ21838.
MANUFACTURER   American Red Cross Blood Services, Greater Upstate New York      
               Region, Syracuse, New York.
RECALLED BY    Manufacturer, by telephone September 3, 1992.  Firm-initiated    
               recall complete.
DISTRIBUTION   New York, Switzerland.
QUANTITY       1 unit of each component.
REASON         Blood products, which tested repeatedly reactive for hepatitis   
               B surface antigen (HBsAg), were distributed.

              
PRODUCT        Coombs Control Reagent Red Blood Cells.  Recall #B-091-3.
CODE           Lot #K129.
MANUFACTURER   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
RECALLED BY    Manufacturer, by letter July 13, 1992.  Firm-initiated recall    
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,795 packages were distributed.
REASON         Reagent Red Blood Cells which were found to be hemolyzed were    
               distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
              
PRODUCT        Pediatric Sample Cups, used to hold pediatric samples which are  
               used with the 550 Express or the Fast 4 Clinical Analyzers.
               Recall #Z-151-3.
CODE           Catalog #751063.  All cartons with manufacturing date codes of   
               9-10-91 through 10-21-91.
MANUFACTURER   Elyria Plastics, Elyria, Ohio.
RECALLED BY    Manufacturer, by telephone October 23, 1992, followed by         
               letter.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       158 cartons (1,000 cups per carton) were distributed.
REASON         The cups may leak.

              
PRODUCT        Lyphomed Empty Sterilized Vial, 10 ml and 30 ml.  Recall         
               #Z-154-3.
CODE           Product #SV10, lot #300939, and Product #SV30, lot #310984.
MANUFACTURER   Fujisawa USA, Inc., Grand Island, New York.
RECALLED BY    Fujisawa USA, Inc., Melrose Park, Illinois, by FAX October 23,   
               1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Italy.

                                      -5-
QUANTITY       183,700 vials were distributed; firm estimates 27,555 vials      
               remain on the market.
REASON         An improper crimp on the vial seal may result in a lack of       
               assurance of sterility of the product.

              
PRODUCT        Scanlan Taheri Assurance Valvulotome, used to excise valves in   
               veins to facilitate in-situ by-pass procedures.
               Recall #Z-159-3.
CODE           Catalog #1001-771.  All Lots.
MANUFACTURER   Scanlan International, St. Paul, Minnesota.
RECALLED BY    Manufacturer, by letter November 6, 1992.  Firm-initiated        
               recall ongoing.
DISTRIBUTION   Nationwide and Austria.
QUANTITY       199 units were distributed.
REASON         The device does not function as intended which could result in   
               injury to the vessel.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
              
PRODUCT        Sagittal Blade, Thick, No Offset, Part Number 2108-118, a        
               cutting accessory for orthopedic bone saw.  Recall #Z-130-3.
CODE           Lot #92080341.
MANUFACTURER   Stryker Corporation Instruments Division, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter September 21, 1992.  Firm-initiated      
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       120 boxes were distributed.
REASON         Mislabeled.  Part #2108-105, Sagittal Deep Cutting Blade may be  
               in the package instead of Part #2108-118.

              
PRODUCT        Sagittal Blade, Wide, Thick, Part #2108-110, a cutting           
               accessory for orthopedic bone saw.  Recall #Z-131-3.
CODE           Lot #92060231.
MANUFACTURER   Stryker Corporation Instrument Division, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter September 21, 1992.  Firm-initiated      
               recall ongoing.
DISTRIBUTION   Virginia, Washington state.
QUANTITY       6 boxes (3 blades per box) were distributed.
REASON         Mislabeled.  Part #277-88-10 may be in the package instead of    
               Part #2108-110.

              
PRODUCT        ALT (IFCC) Reagent in single vials, used on the Ciba Corning     
               400E Chemistry Analyzer.  Recall #Z-150-3.
CODE           Lot #G64MK.
MANUFACTURER   Ciba Corning Diagnostics Corporation, Oberlin, Ohio.
RECALLED BY    Manufacturer, by telephone October 9 and 11, 1991, and by        
               letter October 17, 1991.  Firm-initiated recall complete.
DISTRIBUTION   Texas, Illinois.
QUANTITY       130 vials were distributed.
REASON         Some of the vials were filled with an excess of ALT reagent.


                                      -6-
RECALLS AND FIELD CORRECTIONS:  VETERINARY -- CLASS II
              
PRODUCT        Neomycin products:
               (a) Bulk Neomycin Sulfite USP Powder; (b) Neomycin Unfinished;   
               (c) Neomycin Agricultural; (d) Neomycin Technical.
               Recall #V-053-2.
CODE           Lot numbers:  (a) 430HR, 073HT, 177FU, 178F7; (b) 889DP;
               (c) 501FX, 502FX, 503FX, 504FX, 505FX, 508FX, 509FX, 865HB,      
               866HB, 867HB, 887HB, 888HB, 038HP; (d) 358FU.
MANUFACTURER   The Upjohn Company, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter August 18, 1992.  Firm-initiated recall  
               ongoing.
DISTRIBUTION   New Jersey, Florida, Nebraska, Belgium, Germany, India, Mexico,  
               Argentina, Taiwan, Hong Kong, Canada.
QUANTITY       Undetermined amount.
REASON         Product contains elevated levels of sodium fluoride.


MEDICAL DEVICE SAFETY ALERTS:
              
PRODUCT        Hurst and Maloney Mercury-Filled Bougies, esophageal dilators.   
               Safety Alert #M-002-3.
CODE           All product marked "Don't Use After 4-30-95" which appears       
               directly on the product, not on the label.
MANUFACTURER   Pilling Company, Fort Washington, Pennsylvania.
ALERTED BY     Manufacturer, by letter August 22, 1992.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       Approximately 12,000 units were distributed.
REASON         A design change in which the tapered tip was made shorter,       
               which could possibly contribute to complaints of esophageal      
               perforation, puncturing or tearing.

              
PRODUCT        Renalin Cold Sterilant for dialysis use in 1 liter bottles and   
               2-1/2 gallon bottles.  Safety Alert #M-003-3.
CODE           All lots.
MANUFACTURER   Minntech Corporation, Minneapolis, Minnesota.
ALERTED BY     Manufacturer issued an Urgent:  Important Safety Alert dated     
               October 10, 1992.
DISTRIBUTION   Nationwide.
QUANTITY       37,000 cases of 2-1/2 gallon bottles (2 bottles per case) and    
               7,000 cases of 12 one-liter bottles per case were distributed.
REASON         Preliminary findings showed a statistical increase in the        
               relative risk of death for both manual and automated Renalin     
               reuse and for all glutaraldehyde reuse over the control (no      
               reuse); and showed a statistical increase in the risk of death   
               with manual Renalin reuse procedures compared to the control.

                                      -7-

END OF ENFORCEMENT REPORT FOR DECEMBER 23, 1992.  BLANK PAGES MAY