FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
09/16/1992

RECALLS AND FIELD CORRECTIONS: September 16, 1992

                                FOODS -- CLASS I
                                                            92-38 
PRODUCT        Pewter Baby Cups 2-1/2" high with hand, boy & girl cups.         
               Recall #F-522/523-2.
CODE           Not coded.
MANUFACTURER   Shirley Pewter Company, Williamsburg, Virginia.
RECALLED BY    Manufacturer, by Fax June 15, 1992.  Firm-initiated recall       
               ongoing.
DISTRIBUTION   Virginia and South Carolina, retail stores, tourist sales and    
               catalog sales may be nationwide.
QUANTITY       Approximately 650 cups.
REASON         Products contain leachable lead.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS I
             
PRODUCT        Clozaril (Clozapine) 100 mg SandoPak (unit dose pack), 100       
               tablets, a Rx anti-psychotic drug.  Recall #D-497-2.
CODE           All lots on the market as of 7/31/92.
MANUFACTURER   Sandoz Pharmaceuticals Corporation, East Hanover, New Jersey.
RECALLED BY    Manufacturer, by letters dated July 24 and 31, 1992.             
               Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Firm estimates 1,000 SandoPaks remain on the market.
REASON         Some blister cavities of the unit dose strips contained two      
               tablets instead of one.

                                       

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
              
PRODUCT        Various Rx and OTC drug products requiring refrigerated storage  
               conditions:
               (a) Repository Corticotropin Injection, USP, ACTH, in 5 ml       
               vials: 
               (i) 40 units/ml, Major item #179606;
               (ii) 80 units/ml, Major item #184143;
               (b) Major Naphazole A Ophthalmic Solution, Naphazoline
               Hydrochloride 0.025% and Pheniramine Maleate 0.3%
               Ophthalmic Solution, USP (Sterile), in 15 ml dropper container,  
               Major item #135418;
               (c) Major Sulfacetamide Sodium Ophthalmic Solution, USP, 10%
               (Sterile), in 15 ml bottle, Major item #129155;
               (d) Acetaminophen Suppositories:
               (i) Mapap Pediatric Rectal Suppositories, Acetaminophen 120 mg,  
               12 suppositories per box, Major item #127449, M-11 on box;
               (ii) G & W Acephen Acetaminophen Rectal Suppositories,           
               Acetaminophen 325 mg, 12 suppositories per box, Major item       
               #184200;
               (iii) Mapap Rectal Suppositories, Acetaminophen 650 mg, 12       
               suppositories per box, Major item #127464, M-11 on box;
               (e) G & W Truphylline Aminophylline Suppositories, 500 mg, 10    
               suppositories per box, Major item #184150;
               (f) Major Phenameth Promethazine HCl Suppositories, 50 mg, 12    
               and 25 suppositories per box, Major item #153866;
               (g) Major Erythromycin Ethylsuccinate Oral Suspension, in 1      
               pint bottles:
               (i) 200 mg/5 ml: Major item #140293;
               (ii) 400 mg/5 ml: Major item #140301;
               (h) Major Erythromycin Estolate Oral Suspension, USP, in 1 pint  
               bottles, store in refrigerator to preserve taste:
               (i) 125 mg/5 ml: Major item #140319;
               (ii) 250 mg/5 ml: Major item #140335.  Recall #D-482/494-2.
CODE           Ellis Division Ormond Beach, FL, Woburn, MA and Winston Salem,   
               NC locations:  All products, all lots within expiration date;
               Murray Division, Murray, KY and Houston, TX locations:  items    
               a, b, c, g, h; all lots within expiration date;
               Crown Division Chicago, IL and Omaha, NE locations:  items b,    
               e, f (12's):  All lots;
               item c, lot #0693 and 0398 only;
               item g - ii, lot #0020 only;
               item a - i, all lots, Omaha customers only;
               Michigan Division, Auburn Hills, MI location:
               item b, lot 0874 only;
               item d - ii, all lots;
               item h - ii, lot #13097 only;
               Ultra Division, Seattle, WA and San Diego, CA locations:
               items b, f:  all lots;
               item c, all lots for Seattle and lot 9764 only for San Diego.

                                      -2-
MANUFACTURER   (a) Organics/LaGrange, Chicago, Illinois;
               (b, c, g) Pharmafair, Inc., Tampa, Florida;
               (b, c) Bausch & Lomb Pharmaceuticals, Clearwater, Florida;
               (d, e, f) G & W Laboratories, Inc., South Plainfield, New        
               Jersey; 
               (g, h) Barre-National, Inc., Baltimore, Maryland.
RECALLED BY    Major Pharmaceuticals, Chicago, Illinois (responsible firm), by  
               letter August 25, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       189 vials of 40 u/ml and 417 vials of 80 u/ml repository         
               corticotropin; 10,540 bottles of Naphazole A; 48,011 bottles of  
               Sulfacetamide; 1,108 125-mg, 420 325-mg, & 1,119 650-mg boxes    
               of acetaminophen suppositories; 214 boxes of aminophylline
               suppositories; 801 x 12 and 121 x 25 boxes of promethazine
               hydrochloride suppositories; 713 200-mg and 562 400-mg
               bottles of erythromycin ethylsuccinate; and 168 125-mg
               and 495 250-mg bottles of erythromycin estolate were
               distributed which had not been refrigerated.  The firm
               estimates that 10% of the product remains on the market.
REASON         The products were not held at the labeled storage temperatures,  
               which could affect the products' potency and/or physical         
               properties.

              
PRODUCT        Rx small volume parenterals in prefilled syringes:
               (a) Heparin Lock Flush Solution, USP, 100 units/ml, in 3 ml      
               syringes, catalog #10773 and 5 ml syringes, catalog #10775,      
               with the 3 ml syringe also packaged in the Lok-Pak-N Heparin     
               Lock Flush Procedure Pack, catalog #11873;
               (b) Sodium Chloride Injection, USP, 0.9%, Preservative Free, in  
               2 ml syringe, catalog #11272, packaged in the Lok-Pak-N Heparin  
               Lock Flush Procedure pack, catalog #11771.  Recall #D-495/496-2.
CODE           Lot numbers:  (a) 91M003B (5 ml), 91M004D (3 ml), 91M004C (Kit   
               #11873); (b) 91M002 (2 ml), 91M001B (kit #11771.
MANUFACTURER   Smith & Nephew SoloPak, Franklin Park, Illinois.
RECALLED BY    Smith & Nephew SoloPak, Elk Grove Village, Illinois, by letters  
               dated August 18, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 55,560 3-ml syringes, 115,200 3-ml syringes in kits, and     
               92,280 5-ml syringes; (b) 163,320 2-ml syringes in kits were     
               distributed; firm estimates that 10 percent of the product       
               remains on the market.
REASON         Lack of assurance of sterility.

              
PRODUCT        Ritalin (Methylphenidate Hydrochloride), 20 mg, 100 tablets, an  
               Rx product for oral administration as a central nervous system   
               stimulant.  Recall #D-498-2.
CODE           Lot numbers:  1B005856 EXP 11/96, 2B005856 EXP 11/96, 1B006379   
               EXP 12/96.
MANUFACTURER   Ciba-Geigy Corporation, Pharmaceutical Division, Summit, New     
               Jersey.

                                      -3-
RECALLED BY    Manufacturer, by letter July 14, 1992, followed by telephone.    
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       28,476 units of lot 1B005856, 6,814 units of lot 2B005856, and   
               33,354 units of lot 1B006379 were distributed; firm estimates    
               20,500 units remain on the market.
REASON         An incorrect National Stock Number (NSN) intended for            
               Ritalin-SR 20 mg was an error printed on the Ritalin 20 mg       
               label. 

              
PRODUCT        Various Lyphomed Rx injectable drug products:
               (a) Aminophylline Injection, USP, 25 mg/ml, in 20 ml vials;
               (b) Atropine Sulfate Injection, USP, 0.5 mg/ml, in 1 ml vials;
               (c) Cyanocobalamin Injection, USP, 1,000 mcg/ml, in 1 ml vials; 
               (d) Diazepam Injection, USP, 5 mg/ml, in 2 ml vials;
               (e) Heparin Lock Flush Solution, 10 USP units/ml, in 1 ml vials;
               (f) Heparin Sodium Injection, USP, 10,000 USP units/ml, in 1 ml  
               vials;
               (g) Lidocaine Hydrochloride Injection, USP, 2%, 20 mg/ml, in 2   
               ml vials;
               (h) Mannitol Injection, USP, 25%, 250 mg/ml, in 50 ml vials;
               (i) Multilyte-20 Multi-Electrolyte Concentrate, in 25 ml vials;
               (j) Phenytoin Sodium Injection, USP, 50 mg/ml, in 2 ml and 5 ml  
               vials;
               (k) Potassium Chloride for Injection Concentrate, USP,           
               Preservative Free, 2 mEq/ml, in 10 ml vials, and in 20 ml vials;
               (l) Potassium Chloride for Injection Concentrate, USP, 2         
               mEq/ml, preserved, in 20 ml vials;
               (m) Pyridoxine Hydrochloride Injection, USP, 100 mg/ml, in 1 ml  
               vials;
               (n) Sodium Chloride Injection, USP, 23.4%, in 30 ml vials;
               (o) Sterile Water for Injection, USP, in 5 ml vials, 50 ml       
               vials, and 100 ml vials;
               (p) Thiamine Hydrochloride Injection, USP, 100 mg/ml, in 2 ml    
               vials.  Recall #D-503/518-2.
CODE           Lot numbers:  (a) 320114, 320290, 320291; (b) 311244;
               (c) 311250, 320371; (d) 320324; (e) 311199, 311210, 311307,      
               320024, 320120, 320186, 320158, 320168, 320208, 320214, 320354,  
               (f) 311213, 311299, 311302, 320105, 320111, 320195, 320451;
               (g) 311238; (h) 310131, 310133, 310132, 310968, 320012, 320011;  
               (i) 320344; (j) 311288, 320044, 320129; (k) 311277, 320078,      
               311248, 320072, 320312; (l) 311340; (m) 311245, 320155; (n)      
               320403, 320185; (o) 311366, 320172, 311258, 320322;
               (p) 320159, 320331.
MANUFACTURER   Lyphomed, Division of Fujisawa USA, Inc., Grand Island, New      
               York. 
RECALLED BY    Lyphomed, Division of Fujisawa USA, Inc., Melrose Park,          
               Illinois, by letter August 28, 1992.  Firm-initiated recall      
               ongoing.
DISTRIBUTION   Nationwide, The Bahamas, Nicaragua, Hong Kong, Peru, Malta.

                                      -4-
QUANTITY       Vials distributed         Estimated to remain on market
               (a) 87,100                 43,450
               (b) 56,250                 11,250
               (c) 233,150                47,295
               (d) 88,790                 48,660
               (e) 1,927,625              269,868
               (f) 1,448,150              260,667
               (g) 42,350                 4,930
               (h) 124,775                36,491
               (i) 8,750                  4,275
               (j) 108,182                59,540
               (k) 292,800                38,521
               (l) 53,150                 6,798
               (m) 77,325                 31,199
               (n) 86,475                 24,213
               (o) 116,325                29,994
               (p) 119,975                48,306
REASON         The vials have a low fill volume.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
              
PRODUCT        Ben-Aqua Gel 5% (Benzoyl Peroxide 5%), in 4 ounce tubes.         
               Recall #D-499-2.
CODE           Lot #H282 EXP 12/93.
MANUFACTURER   Syosset Laboratories Company, Inc., Syosset, New York.
RECALLED BY    Manufacturer, by letter August 6, 1992, followed by telephone    
               August 7, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   California, New Mexico, Minnesota, New York, New Jersey.
QUANTITY       918 units were distributed.
REASON         Some tubes have leakage at the crimp.

               
PRODUCT        Triamcinolone Acetonide Cream, 0.1%, 80 gram size.
               Recall #D-500-2.
CODE           Lot #H208 EXP 10/93.
MANUFACTURER   Syosset Laboratories Company, Inc., Syosset, New York.
RECALLED BY    Manufacturer, by letter August 13, 1992.  Firm-initiated recall  
               ongoing.
DISTRIBUTION   Florida, Illinois.
QUANTITY       1,208 tubes were distributed.
REASON         Some tubes have leakage at the crimp.

               
PRODUCT        (a) Thorazine (Chlorpromazine Hydrochloride) Concentrate 30      
               mg/ml, in 4 fluid ounce bottles; (b) Thorazine Syrup, 10 mg/5    
               ml in 4 fluid ounce bottles.  Recall #D-501/502-2.
CODE           Lot numbers:  (a) X1-1T47 EXP 6/30/95;
               (b) X3-OT72 EXP 11/30/93, X4-OT72 EXP 11/30/93, X5-OT72 EXP      
               11/30/92, X6-1T72 EXP 9/30/94, X7-1T72 EXP 9/30/94, X8-1T72 EXP  
               9/30/94, 2-2T72 EXP 2/28/95, 3-2T72 EXP 3/31/95.
MANUFACTURER   Smithkline Beecham Pharmaceuticals Company, Cidra, Puerto Rico.
RECALLED BY    Manufacturer, by letter mailed August 11, 1992.  Firm-initiated  
               recall ongoing.
                                      -5-
DISTRIBUTION   Nationwide.
QUANTITY       (a) 4,120 bottles were distributed; firm estimates none remains  
               on the market; (b) 35,905 bottles were distributed; firm         
               estimates 4,000 bottles remain on the market.
REASON         (a) Product does not meet pH specifications; (b) Potency not     
               assured if continuously stored at upper temperature limit.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
              
PRODUCT        Patient Circuit Body used with Neonatal High Frequency           
               Oscillatory Ventilators:
               (a) Model 3100, part #766898; (b) Model 3100A, part #766895.
               Recall #Z-1254/1255-2.
CODE           Lot numbers:  (a) 56920091; (b) 55728191.
MANUFACTURER   Surgical Technologies, Inc., Brea, California.
RECALLED BY    Sensormedics Corporation, Yorba Linda, California, by            
               interoffice memorandum dated January 10, 1992.  Firm-initiated   
               recall complete.
DISTRIBUTION   Colorado, Ohio, Michigan, California, South Carolina,            
               Pennsylvania, Maryland, Massachusetts, New York, Florida, Utah,  
               Washington state, Texas, Georgia, Tennessee, Iowa, Virginia,     
               Idaho.
QUANTITY       Approximately (a) 80 2-unit boxes; (b) 60 4-unit boxes were      
               distributed.
REASON         The patient circuit may separate at either end of the rigid      
               inspired tube or at the patient "wye", due to insufficient glue  
               and ultra-violet curing.

              
PRODUCT        Above Knee Prosthetic Wood Block, "TWB-4", used as a component   
               of an artificial leg.  Recall #Z-1259-2.
CODE           Not coded.
MANUFACTURER   Teh Lin, Taiwan.
RECALLED BY    Daw Industries, Inc., San Diego, California, by telephone        
               December 31, 1991, and by letter January 23, 1992.               
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       166 units were distributed.
REASON         The TWB-4 component was showing movement which could lead to     
               failure of the finished device.

              
PRODUCT        Barrier Ultra Protection and Extra Protection Surgical Gowns 
               1.  Reorder Nos. 0579 and 7-0579;
               2.  Reorder Nos. 20-9579 and 27-9579;
               3.  Reorder Nos. 0575 and 7-0575;
               4.  Reorder Nos. 0576 and 7-0576;
               5.  Reorder Nos. 0580 and 7-0580;
               6.  Reorder Nos. 20-9675 and 27-9675;
               7.  Reorder Nos. 20-9676 and 27-9676;
               8.  Reorder Nos. 20-9680 and 27-9680;

                                      -6-
               9.  Reorder Nos. 0230 and 7-0230;
               10. Reorder Nos. 0244 and 7-0244;
               11. Reorder Nos. 0263 and 7-0263;
               12. Reorder Nos. 0284 and 7-0284;
               13. Reorder Nos. 1041 and 7-1041;
               14. Reorder Nos. 1062 and 7-1062;
               15. Reorder Nos. 1156 and 7-1156;
               16. Reorder Nos. 1223 and 7-1223;
               17. Reorder Nos. 1237 and 7-1237;
               18. Reorder Nos. 1238 and 7-1238;
               19. Reorder Nos. 1239 and 7-1239;
               20. Reorder Nos. 1263 and 7-1263;
               21. Reorder Nos. 1267 and 7-1267;
               22. Reorder Nos. 1269 and 7-1269;
               23. Reorder Nos. 1273 and 7-1273;
               24. Reorder Nos. 1291 and 7-1291;
               25. Reorder No. 1051.  Recall #Z-1263/1287-2.
CODE           Lot Numbers with first four digits of 1811 through 3651 and      
               0012 through 1182;
MANUFACTURER   Johnson & Johnson Medical, Inc., El Paso, Texas.
RECALLED BY    Johnson & Johnson Medical, Inc., Arlington, Texas, by letter     
               July 20, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       218,168 cases (2,957,531 gowns) were distributed; firm           
               estimates 10,100 cases remain on the market.
REASON         The gown liner can tear during donning and use, compromising     
               its integrity and function as an impervious barrier, resulting   
               in strike through.

              
PRODUCT        Hulka Uterine Mobilizer, used for mobilizing the uterus.         
               Recall #Z-1288-2.
CODE           Product #433-245, lot #31-609.
MANUFACTURER   G.F. Instrumente, Heidelberg, Germany.
RECALLED BY    Zinnanti Surgical Instruments, Chatsworth, California, by        
               letter April 9, 1990.  Firm-initiated recall complete.
DISTRIBUTION   Texas, Arizona.
QUANTITY       7 units were distributed; firm estimates none remains on the     
               market.
REASON         The uterine sound part at the tip of the tenaculem may break     
               off and be retained inside the patient.

               
PRODUCT        Inferno Moist Heat Therapy Unit, eight hot pack capacity, Model  
               917, a water bath used to heat up and maintain the hot packs at  
               a desired temperature.  Recall #Z-1289-2.
CODE           Serial numbers range from 183133 to 183137.
MANUFACTURER   Polar Ware Company, Shoboygen, Wisconsin.
RECALLED BY    Ferno Washington, Wilmington, Ohio, by telephone January 29,     
               1992.  Firm-initiated recall complete.
DISTRIBUTION   Oregon, Florida, Texas, Tennessee, Canada.
QUANTITY       5 units were distributed; firm estimates none remains on the     
               market.
                                      -7-
REASON         The capillary tube of the primary thermostat can come into       
               contact with a terminal on the safety thermostat, causing a      
               short circuit.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
              
PRODUCT        Calibration Gas 2 Cylinder used to calibrate a cutaneous gas     
               system (CGS) sensor.  Recall #Z-1256-2.
CODE           Part #550902, lot #WK12L79.
MANUFACTURER   Puritan Bennett, City of Industry, California.
RECALLED BY    SensorMedics Corporation, Yorba Linda, California, by letter     
               January 16, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania, California, Maryland, Ohio, South Carolina,        
               Hawaii, Utah, Georgia, North Carolina, Colorado, Tennessee,      
               Kansas, Washington state, Florida.
QUANTITY       130 units were distributed.
REASON         The calibration gas may have been blended incorrectly.  The      
               incorrect mixture will cause a "Cal Gases May be Reversed"       
               message to be displayed on the station when trying to calibrate  
               a Cutaneous Gas System (CGS) Sensor.

              
PRODUCT        Tandem-R Pap Kit, an immunoradiometric assay indicated for the   
               quantitative measurement of prostatic acid phosphatase (PAP) in  
               serum.  Recall #Z-1260-2.
CODE           Catalog #3073, Kit lot numbers:  290789 EXP 6/22/92, 290936 EXP  
               7/13/92, 290998 EXP 7/28/92, 291009 EXP 7/28/92, 291110 EXP      
               8/11/92.  Zero diluent component lot #260333 EXP 3/20/93.
MANUFACTURER   Hybritech, Inc., San Diego, California.
RECALLED BY    Manufacturer, by letters of  July 29, 1992 and August 5, 1992.   
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,139 kits were distributed.
REASON         The firm neglected to put enough preservative in the calibrator  
               which has led to fungus growth in the vials.

              
PRODUCT        Hemaflex Absorbable Collagen Hemostat rectangular textured pad,  
               2.5 cm x 5 cm (small), indicated for use in surgical procedures  
               (other than neurological, urological and ophthalmological        
               surgery) as an adjunct to hemostasis when control of bleeding    
               by ligature or other conventional methods is ineffective or      
               impractical.  Recall #Z-1261-2.
CODE           Catalog #2201-10, lot #19111C EXP 12/94.
MANUFACTURER   Bioplex Medical B.V., a subsidiary of Bioplex Corporation,       
               Vaals, Holland.
RECALLED BY    Bioplex Corporation, a subsidiary of Datascope Corporation,      
               Montvale, New Jersey, by letter March 31, 1992.  Firm-initiated  
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       693 pieces were distributed.

                                      -8-
REASON         The outer pouch may not contain the lot numbers or expiration    
               date.


MEDICAL DEVICE SAFETY ALERTS
              
PRODUCT        Aluminum Pump used on an Acidified Bicarbonate Dialysis          
               System.  Safety Alert #M-092-2.
CODE           All codes.
MANUFACTURER   Community Dialysis Centers, Laguna Hills, California.
ALERTED BY     Manufacturer, by letter July 31, 1992.
DISTRIBUTION   Community Dialysis Centers.
QUANTITY       Unknown.
REASON         Patients may have been exposed to bicarbonate dialysate which    
               came into contact with a pump containing aluminum components.    
               This may result in elevated aluminum in patients' blood.

               
PRODUCT        Medex C-Fusor Pressure Infusors, inflatable polyurethane cuffs   
               which are placed around solution bags to produce counter         
               pressure on arterial monitoring or central pressure lines.  The  
               device may also be used to rapidly infuse blood products or IV   
               solutions:
               (a) Catalog #MX4805; (b) Catalog #MX4806; (c) Catalog #MX4810;
               (d) Catalog #MX4811; (e) Catalog #5104366; (f) Catalog           
               #5000359.  Safety Alert #M-093/098-2.
CODE           All lots.
MANUFACTURER   Medex, Inc., Hilliard, Ohio.
ALERTED BY     Manufacturer, by letter February 10, 1992.
DISTRIBUTION   Nationwide and international.
QUANTITY       19,300 units were distributed.
REASON         It came to the firm's attention that some customers used the     
               devices by removing the inflation bulb which is sold with the    
               device and by connecting the pressure infusor to a compressed    
               air line at the facility.  The firm advised users they should    
               discontinue use of this method because the infusor may balloon   
               to the point of failure resulting in complete rupture of the     
               infusor's bladder.

                                      -9-


END OF ENFORCEMENT REPORT FOR SEPTEMBER 16, 1992.  BLANK PAGES MAY
FOLLOW.

                                     ###