FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
06/24/1992

RECALLS AND FIELD CORRECTIONS: June 24, 1992

                                    FOODS -- CLASS II
                                                                  92-26

PRODUCT        Style Shampoo with Pro-Vitamin Complex, Body Building Formula    
               for Fine Hair, in 15 ounce bottles.  Recall #F-338-2.
CODE           B18.
MANUFACTURER   DowBrands L.P., Fridley, Minnesota.
RECALLED BY    Manufacturer, by telephone May 16, 1992, followed by letter May  
               18, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       3,455 cases (12 bottles per case) were distributed.
REASON         Product is contaminated with Pseudomonas aeruginosa.

              
PRODUCT        Mission Valley brand Sweet Pickle Relish, in 1 gallon jars.      
               Recall #F-339-2.
CODE           Jar code 10 1491, 11 1391.
MANUFACTURER   Valley's Pride Food Products, Inc., Stockton, California.
RECALLED BY    Manufacturer, by visit beginning December 17, 1991.              
               Firm-initiated recall ongoing.
DISTRIBUTION   California.
QUANTITY       666 cartons were distributed; FDA estimates none remains on the  
               market.
REASON         Product is contaminated with wood slivers.

                                      -1-
                                       

              
PRODUCT        Hotel Del Coronado Hair Conditioner, in 1.3 fluid ounce plastic  
               bottles.  Recall #F-341-2.
CODE           C1 stamped on shipping cases.
MANUFACTURER   Midik Packaging Corporation, Elkhart, Indiana.
RECALLED BY    Manufacturer, by telephone March 2, 1992 and by FAX April 26,    
               1992.  Firm-initiated recall ongoing.
DISTRIBUTION   California.
QUANTITY       8,476 bottles were distributed; firm estimates none remains on   
               the market.
REASON         Product is contaminated with Pseudomonas aeruginosa.

              
PRODUCT        Cheese Bread Sticks, in 5 ounce cellophane trays and wrapped in  
               plastic, under the following labels:  Orowheat, Gai's, Ralphs,   
               The Bakery (Lucky's), Raley, Safeway, and Rainbo, Baker's        
               Choice, Old San Francisco.  Recall #F-342-2.
CODE           11611CP, 11621CP through 29811CP, 29821CP (April 26 through      
               October 25, 1991).
MANUFACTURER   California Pretzel Company, Visalia, California.
RECALLED BY    Manufacturer, by letter November 8, 1991.  Firm-initiated        
               recall ongoing.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       10,500 to 12,500 cases.
REASON         Product contains undeclared FD&C Yellow No. 5.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
              
PRODUCT        Thrifty Maid Mandarin Oranges, in 11 ounce cans.
               Recall #F-340-2.
CODE           All product which bears a top line code with "P" in it.
MANUFACTURER   The Monterey Canning Company, San Mateo, California (importer).
RECALLED BY    Winn Dixie Stores, Inc., Jacksonville, Florida, by letter        
               February 26, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Florida, Alabama, South Carolina, North Carolina, Georgia,       
               Kentucky, Louisiana, Texas.
QUANTITY       Undetermined.  However, 4,974 cans were retrieved as of 5/5/92.
REASON         Product contained in swollen cans.


RECALLS AND FIELD CORRECTIONS: DRUGS -- CLASS II
              
PRODUCT        Adenocard (adenosine) 3 mg/ml, a Rx sterile cardiac solution,    
               in 2 ml single dose vials, 10 vials per shelf carton.
               Recall #D-398-2.
CODE           Lot #311319.
MANUFACTURER   Lyphomed, Division of Fujisawa USA, Inc., Grand Island, New      
               York. 

                                      -2-

RECALLED BY    Fujisawa Pharmaceutical Company, Division of Fujisawa USA,       
               Inc., Deerfield, Illinois, by letter June 9, 1992.               
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       73,010 vials were distributed; firm estimates 11 percent         
               remains on the market.
REASON         The vials have a low fill volume.

               
UPDATE         Various injectable drug products, manufactured by Royce          
               Laboratories, Recall #D-365/371-2, which appeared in the June    
               10, 1992 Enforcement Report has been extended as follows:        
               Heparin Sodium Injection, 5,000 USP units has been extended to   
               include lot #320179 EXP 2/94 and Sodium Chloride Injection,      
               USP, 0.9% has been extended to include lot #320140 EXP 1/94.


RECALLS AND FIELD CORRECTIONS: DRUGS -- CLASS III

              
PRODUCT        Darvon brand Propoxyphene Hydrochloride, USP, packaged in        
               blister packages with 10 capsules to a strip and 10 strips per   
               carton (100 capsules per carton).  Recall #D-396-2.
CODE           Lot number 5CZ50M blister packs in cartons.  Other similar lot   
               numbers in bottles are not under recall.
MANUFACTURER   Eli Lilly Industries, Inc., Carolina, Puerto Rico.
RECALLED BY    Eli Lilly and Company, Indianapolis, Indiana, by letter May 8,   
               1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,039 packages were distributed
REASON         The cartons are correctly labeled.  The blister strips are       
               incorrectly labeled as Propoxyphene Napsylate.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
              
PRODUCT        Red Blood Cells.  Recall #B-319-2.
CODE           Unit #50M01376.
MANUFACTURER   American Red Cross Blood Services, Toledo, Ohio.
RECALLED BY    Manufacturer, by telephone August 20, 1991 followed by letter    
               August 23, 1991.  Firm-initiated recall complete.
DISTRIBUTION   Ohio.
QUANTITY       1 unit.
REASON         A unit of autologous Red Blood Cells, which tested repeatedly    
               reactive for the antibody to hepatitis C virus encoded antigen   
               (anti-HCV), was distributed without the required biohazard       
               labeling.


RECALLS AND FIELD CORRECTION:  BIOLOGICS -- CLASS III
              
PRODUCT        Red Blood Cells.  Recall #B-320-2.
CODE           Unit numbers:  NA00220, NA00242, NA00259, NA00282.

                                      -3-

MANUFACTURER   Northampton-Accomack Memorial Hospital, Nassawadox, Virginia.
RECALLED BY    Manufacturer, by telephone September 11, 1991.  Firm-initiated   
               recall complete.
DISTRIBUTION   Virginia.
QUANTITY       4 units.
REASON         Red Blood Cells labeled with the incorrect expiration date were  
               distributed.

RECALLS AND FIELD CORRECTION:  DEVICES -- CLASS I

              
PRODUCT        Sodasorb PrePak Carbon Dioxide Absorption Units, in a one size   
               cartridge form.  Recall #Z-711-2.
CODE           Batch numbers: AA11-B720-13 through AA11-B720-25 (includes 5     
               batches) AA12-B720-3 through AA12-B720-31 (includes 13 batches)  
               AB01-B720-2 through AB01-B720-31 (includes 24 batches)           
               AB02-B720-3 through AB02-B720-27A (includes 18 batches)          
               AB03-B720-3A (includes 1 batch).
MANUFACTURER   W.R. Grace & Company, Atlanta, Georgia.
RECALLED BY    W.R. Grace & Company, Lexington, Massachusetts, by letter May    
               7, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       Approximately 184,000 units were distributed.
REASON         There may be occlusions in the top and bottom holes of the       
               PrePak panels, which can cause unexpected high gas flow          
               resistance during inhalation anesthesia.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
              
PRODUCT        Infant Resuscitator Kits:  (a) with two oxygen lines;
               (b) with one oxygen line.  Recall #Z-773/774-2.
CODE           Product #14543318, lot numbers:  201157, 112206, 201327, 202423.
MANUFACTURER   Gibeck-Dryden Corporation, Indianapolis, Indiana.
RECALLED BY    Manufacturer, by telephone and by letter March 27, 1992.         
               Firm-initiated recall ongoing.
DISTRIBUTION   Utah, Idaho.
QUANTITY       2,460 kits were distributed;  firm estimated that 60 percent of  
               the product remained on market at time of recall on 3/27/92.
REASON         Some kits were assembled with the wrong elbow component.

              
PRODUCT        Bipolar Atrial J Permanent Pacer Leads:
               (a) Model ZY 52 JBV; (b) Model ZY 52 PJBV;
               (c) Intermedics 406-04.  Recall #Z-775/777-2.
CODE           All active fixation bipolar atrial J leads shipped prior to      
               December 11, 1991.
MANUFACTURER   Oscor Medical Corporation, Palm Harbor, Florida.
RECALLED BY    Manufacturer, by telephone December 11, 1991.  Firm-initiated    
               recall complete.

                                      -4-

DISTRIBUTION   Nationwide, Canada, Australia.
QUANTITY       602 leads were distributed.  Firm estimates none remains on the  
               market.
REASON         Pacer leads may short out.

              
PRODUCT        Soft Contact Lenses, Daily Wear (DW) and/or Extended Wear (EW):
               (a) Hydrocurve II - DW & EW;
               (b) Hydrocurve II - DW (Toric);
               (c) Hydrocurve II - EW (Toric);
               (d) Softmate B - DW;
               (e) Softmate I -DW & EW;
               (f) Softmate II - DW & EW;
               (g) Signature - DW (Toric);
               (h) CSI - DW;
               (i) CSI-T -EW;
               (j) CTL Lite - DW & EW;
               (k) CTL Aquamarine - DW & EW.  Recall #Z-778/788-2.
CODE           Lot numbers:  (a) BBLHB, BBNMZ, YBIUF; (b) BAWKT, BAWRU, BAXCX,  
               BBNMC, BBNOA; (c) YAXHU, YAXHW, YAXXE, YAYBW, YBHWD; (d)
BBDTX;  
               (e) BBEWP; (f) YBDCP, YBFRH, YBMAQ, YBFXF; (g) BBIDW, BBIQL,     
               BBOIP; (h) CBCYD, CBDDY, CBDFD; (i) CBCRE; (j) PAIMW; (k)        
               PAKIQ, PANEL.
MANUFACTURER   Sola/Barnes-Hind, Sunnyvale, California.
RECALLED BY    Sola/Barnes-Hind, San Diego, California, by letter July 17,      
               1991.  Firm-initiated recall complete.
DISTRIBUTION   California, Florida, North Dakota, New York, Virginia, Georgia,  
               Hawaii, Indiana, Louisiana, New Mexico, New York, Connecticut,   
               Maine, Rhode Island, South Carolina, Washington state, North     
               Carolina, Mississippi, Maryland, Texas.
QUANTITY       (a) 1,020; (b) 961; (c) 4,563; (d) 939; (e) 405; (f) 843; (g)    
               747; (h) 669; (i) 177; (j) 6 units; (k) 56 units were            
               distributed.
REASON         Lenses are labeled with incorrect parameters for base curve      
               and/or diameter measurements.

              
PRODUCT        Concord/Portex 23/28 Microcatheter Systems, for continuous       
               spinal anesthesia:
               (a) Catalog #4985-23/28; (b) Catalog #4987-23/28;
               (c) Catalog #4989-23/28; (d) Catalog #A1011-22.
               Recall #Z-789/792-2.
CODE           All lots.
MANUFACTURER   Concord/Portex, Keene, New Hampshire.
RECALLED BY    Manufacturer, by letter May 13, 1992.  Firm-initiated recall     
               ongoing.
DISTRIBUTION   Nationwide, Canada, England.
QUANTITY       Undetermined.
REASON         Adverse performance of the microcatheters.

                                      -5-

               
PRODUCT        Nortech Microelectrode Electrohydraulic Lithotripter Probe -     
               1.9 FR, Tip - 1.2 FR, Shaft - 120 cm, a sterile Rx device for    
               use with Northgate Technologies SD-1 and SD-100, ACMI EL-115     
               and AEH-2 EHL generators for the breakup of kidney stones.       
               Recall #Z-793-2.
CODE           Catalog #9-190-12, lot #985.
MANUFACTURER   Northgate Technologies, Inc., Arlington Heights, Illinois.
RECALLED BY    Manufacturer, by telephone, followed by letter May 7, 1992.      
               Firm-initiated recall complete.
DISTRIBUTION   California, New York, Colorado.
QUANTITY       39 boxes were distributed; firm estimates none remains on the    
               market.
REASON         Improper seals on the Tyvek pouch containing the device          
               compromises the sterility of the device.

              
PRODUCT        Laser Diode Module, Model LDM-135, used generally as a light     
               source in engineering and experimental applications.
               Recall #Z-794-2.
CODE           Serial numbers:  None.
MANUFACTURER   Imatronic Limited, Berkshire, United Kingdom (Imatronic, Inc.,   
               Batavia, Illinois).
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action plan    
               June 2, 1992.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       213 units were distributed.
REASON         Noncompliance with the performance standards for laser products  
               in that it failed to incorporate an emission indicator and beam  
               attenuator.

              
PRODUCT        Mobile X-Ray Systems used in Fluoroscopy and Radiography:
               (a) Model SXT-650A; (b) SXT-900A.  Recall #Z-795/796-2.
CODE           All serial numbers.
MANUFACTURER   Toshiba America Medical Systems, Inc., Tustin, California.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action plan    
               May 19, 1992.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       19 units.
REASON         Noncompliance with performance standards for x-ray systems in    
               that they failed to meet all the requirements for source-skin    
               distance (SSD) and fluoroscopic timer, as well as certification  
               and identification requirements.

              
PRODUCT        Intertech Anesthesia Breathing Circuit for pediatric use.        
               Recall #Z-797-2.
CODE           Catalog #225-3720-800, lot #01929.
MANUFACTURER   Intertech Resources, Inc., Fort Myers, Florida.
RECALLED BY    Manufacturer, by letter May 4, 1991.  Firm-initiated recall      
               ongoing.
DISTRIBUTION   Nationwide.

                                      -6-

QUANTITY       180 cases (20 units/case) were distributed.
REASON         The tubing is subject to splitting which can result in leakage.

              
PRODUCT        IV Administration Sets:  (a) Model S75211; (b) Model S75310.
               Recall #Z-798/799-2.
CODE           Lot numbers:  (a) 112445 and 201421; (b) 111540.
MANUFACTURER   (a) IVAC Corporation, Creedmoor, North Carolina;
               (b) Ensambles Medicos S.A. de CV, Tijuana, B.C. Mexico. 
RECALLED BY    IVAC Corporation, San Diego, California, by letter February 17,  
               1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       33,081 sets were distributed.
REASON         The IV sets were distributed with the wrong sensor support       
               flanges, which may result in miscount of IV solution drops.

              
PRODUCT        Ampro Sterile Latex Surgeons Gloves, 50 outerwrapped pairs per   
               box.  Recall #Z-800-2.
CODE           Lot #323B12, size 6-1/2.
MANUFACTURER   American Pro Latex, Monmouth, Illinois.
RECALLED BY    Manufacturer, by telephone March 30, 1992, followed by letter    
               March 31, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Oklahoma, Illinois, Puerto Rico.
QUANTITY       2 cases (4 boxes per case) were distributed; firm estimates 1    
               case remains on the market.
REASON         The devices failed the FDA leak test.

              
PRODUCT        Contact lenses under the following names:
               (a) NewVues; (b) Focus; (c) Cibasoft Softcolors and Visitint;
               (d) Cibathin.  Recall #Z-801/804-2.
CODE           Lot numbers:  (a) 114569, 120066; (b) 115278, 115281, 112524;
               (c) 116558, 115210; (d) 115210.
MANUFACTURER   Ciba Vision Corporation, Atlanta, Georgia.
RECALLED BY    Manufacturer, by letter June 5, 1991.  Firm-initiated recall     
               ongoing.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       28,013 lenses were distributed; firm estimates 27,000 lenses     
               remain on the market.
REASON         NewVues are mislabeled as having a base curve of 8.4 when the    
               actual curve was 8.8, Focus lenses indicate the wrong power on   
               the carton, and some broken glass vial fragments were found in   
               vials of Cibasoft Softcolors, Visitint and Cibathins.

              
PRODUCT        Contact Lens Cleaners, in 1 fluid ounce white plastic bottles    
               with tamper resistant seal:  (a) Fred Meyer Certified            
               Thimerosal Free Daily Cleaner; (b) Brite-Lite Thimerosal Free    
               Daily; (c) Good Neighbor Pharmacy Daily Cleaner Thimerosal       
               Free; (d) Fays Quality Daily Cleaner Easy Rinsing Formula 

                                      -7-

               Thimerosal Free; (e) Smith's Food and Drug Centers Thimerosal    
               Free Daily Cleaner.  Recall #Z-807/811-2.
CODE           Lot number 1021 EXP 4/94.
MANUFACTURER   Sherman Laboratories, Inc., Abita Springs, Louisiana.
RECALLED BY    KC Pharmaceuticals, Pomona, California, by telephone and by      
               letter March 5, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Colorado, Michigan, New York, Oregon, Utah, Virginia.
QUANTITY       15,022 bottles were distributed; firm estimates that little      
               product remains on the market.
REASON         The methods used in, and the facilities and controls used for    
               their manufacture, packing and storage were not in conformity    
               with the good manufacturing practice regulations for medical     
               devices.

              
UPDATE         American Sterilizer 308ORL Surgical Table, Z-742-2 which         
               appeared as an UPDATE in in the June 17, 1992 Enforcement        
               Report should read firm-initiated field correction ongoing.


RECALLS AND FIELD CORRECTIONS:  VETERINARY -- CLASS II
              
PRODUCT        Filaribits Plus (Diethylcarbamazine Citrate/Oxibendazole)        
               Chewable Tablets, 180 mg/136 mg, packaged in 100 and 200 tablet  
               bottles, a Rx product for prevention of heartworm and hookworm   
               and for removal and control of whipworm and mature and immature  
               stages of intestinal ascarids.  Recall #V-038-2.
CODE           Lot numbers:  39957020 FEB 1992 (100-tablet bottles),
               40738020 FEB (200-tablet bottles).
MANUFACTURER   Smithkline Beecham Animal Health, Lincoln, Nebraska.
RECALLED BY    Smithkline Beecham Animal Health, West Chester, Pennsylvania,    
               by telephone April 24, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Colorado, Missouri, Texas, Florida, Minnesota, Georgia,          
               California, Wisconsin, Texas, Arkansas.
QUANTITY       347 100-tablet bottles and 12 200-tablet bottles were            
               distributed.
REASON         Product is contaminated with salmonella.

              
PRODUCT        Betavet Soluspan (Betamethasone Acetate and Betamethasone        
               Sodium Phosphate) packed in 5 ml glass vials, recommended for    
               the treatment of various inflammatory conditions involving the   
               joints.  Recall #V-044-2.
CODE           Lot numbers:  1-AMQ-1 EXP JAN. 93, 1-AMQ-2 EXP JUL. 93,
               1-AMQ-3 EXP OCT. 93.
MANUFACTURER   Schering Plough Products, Inc., Manati, Puerto Rico.
RECALLED BY    Manufacturer, by telephone April 24, 1992.  Firm-initiated       
               recall ongoing.
DISTRIBUTION   Nebraska, New Jersey, Georgia, California.
QUANTITY       89,350 vials were distributed; firm estimates 4,000 vials        
               remain on the market.
REASON         Product found to contain crystal formations.

                                      -8-

              
PRODUCT        Booster Plus NT Medicated (medicated feed) containing            
               oxytetracycline and neomycin, in 50 pound bags.
               Recall #V-045-2.
CODE           3/17/92.
MANUFACTURER   Farmers Cooperative Company, Creston, Iowa.
RECALLED BY    Manufacturer, by flyer mailed or Faxed March 27, 1992.           
               Firm-initiated recall complete.
DISTRIBUTION   Iowa.
QUANTITY       129 bags were distributed; firm estimates none remains on the    
               market.
REASON         Incorrect formulation - Neo-Terramycin 50/50 was used together   
               with Neo-Terramycin 100/50 making the product superpotent.

                                      -9-

END OF ENFORCEMENT REPORT FOR JUNE 24, 1992.  BLANK PAGES MAY FOLLOW.

                                     ###