FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
04/22/1992

Recalls and Field Corrections: April 22, 1992

                                      FOODS

                Class I - A situation in which there is a reasonable
                probability that the use of, or exposure to, a violative
                product will cause serious adverse health consequences or
                death.

                                      NONE

                Class II - A situation in which the use of, or exposure to
                a violative product may cause temporary or medically
                reversible adverse health consequences or where the
                probability of serious adverse health consequences is
                remote.


Product:        Hungry Jack Buttermilk Complete Pancake & Waffle Mix, in
                32 ounce boxes.  Recall #F-225-2.
Code:           C2M13 and C2M16.
Manufacturer:   The Pillsbury Company, Murfreesboro, Tennessee.
Recalled by:    The Pillsbury Company, Minneapolis, Minnesota, by facsimile
                March 24, 1992.  Firm-initiated recall ongoing.
Distribution:   Indiana, Kentucky, Michigan, New York, Ohio.

                                       -1-
                                        



Quantity:       896 cases (12 boxes per case) were distributed.
Reason:         Product contains undeclared corn flour.

                                     ----

Product:        Fruit syrups, in 750 ml and 187 ml glass bottles:
                (a) Lime Flavored Fruit Syrup; (b) Lemon Flavored Fruit Syrup;
                (c) Kiwi Flavored Fruit Syrup; (d) Mandarino Flavored Fruit
                Syrup; (e) Mango Flavored Fruit Syrup; (f) Orange Flavored
                Fruit Syrup; (g) Peach Flavored Fruit Syrup; (h) Apricot
                Flavored Fruit Syrup; (i) Pineapple Flavored Fruit Syrup;
                (j) Passion Fruit Flavored Fruit Syrup.  Recall #F-247/256-2.
Code:           None.
Manufacturer:   R. Torre & Company, Inc., San Francisco, California.
Recalled by:    Manufacturer, by letter January 21, 1992.  Firm-initiated
                recall ongoing.
Distribution:   Nationwide.
Quantity:       (a) 1,720; (b) 770; (c) 825; (d) 580; (e) 275; (f) 1499;
                (g) 1,485; (h) 430; (i) 385; (j) 1,040 cases were distributed.
                FDA estimates none remains on the market.
Reason:         Product contains undeclared FD&C Yellow No. 5.

                                     ----

Product:        Vacuum packed Smoked Whitefish, gutted, unskinned fish
                individually packed in a vacuum type wrap.  Recall #F-257-2.
Code:           None.
Manufacturer:   Goode Foods, Otto, New York.
Recalled by:    Southtowns Seafoods, Inc., Orchard Park, New York.
                Firm-initiated recall complete.
Distribution:   New York, Pennsylvania.
Quantity:       Firm estimates none remains on the market.
Reason:         Product was in swollen packages and contained sodium
                nitrites.
                                     ----

                Class III - A situation in which the use of, or exposure
                to a violative product is not likely to cause adverse
                health consequences.

                                     NONE

                                   COSMETICS

                                     NONE

                                  HUMAN DRUGS

                Class I -
                                     NONE

                                      -2-



                Class II -

Product:        Prescribing Information booklet accompanying the
                Copper T Model TCU 380U Intrauterine Copper Contraceptive.
                Recall #D-295-2.
Code:           Lot numbers:  89569 through 91630.  Only portion shipped or
                potential of being shipped to Arabic speaking countries.
Manufacturer:   Finishing Enterprises, Inc. (FEI), North Tonawanda,
                New York.
Recalled by:    Manufacturer, by letter March 30-31, 1992.  Firm-initiated
                field correction ongoing.
Distribution:   Egypt, Morocco, Jordan, Israel.
Quantity:       Approximately 3,473,100 units were distributed.
Reason:         There is an inaccurate translation in the Arabic language
                portion of the prescribing information booklet.

                                     ----

Product:        Artificial Tears with Phenylepherene Ophthalmic Solution,
                in 5 ml opaque white plastic squeezable bottles.
                Recall #D-298-2.
Code:           Lot #18660.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Abita Springs, Louisiana, by
                letter December 31, 1991.  Firm-initiated recall ongoing.
Distribution:   Colorado.
Quantity:       5,392 bottles were distributed.
Reason:         Use of unapproved grade of polyvinyl alcohol.

                                     ----

Product:        Artificial Tears Sterile Ophthalmic Solution, in 15 ml
                bottles, a clear solution in white opaque plastic vials with
                dropper fitting, under Phoenix Pharmaceutical, Safety Supply
                America, Continental Imports, and Maurry Biological labels.
                Recall #D-299-2.
Code:           Lot #18823 EXP 11/92.
Manufacturer:   Walnut Pharmaceuticals, Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter January 10, 1992.  Firm-initiated recall ongoing.
Distribution:   Missouri, California, Arizona.
Quantity:       18,390 bottles were distributed.
Reason:         Current good manufacturing practice deficiencies.

                                     ----

Product:        Rose Bengal Ophthalmic Solution 1%, in 10 ml opaque white
                squeezable plastic bottles with a 5 ml fill, indicated for use
                as a diagnostic reagent under Akorn, Eagle Vision, and Wilson
                Ophthalmic labels.  Recall #D-300-2.
Code:           Lot numbers:  18845 EXP 11/91, 18978 EXP 6/92.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.

                                      -3-

Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 13, 1991.  Firm-initiated recall ongoing.
Distribution:   Illinois, Tennessee, Louisiana.
Quantity:       15,736 bottles were distributed.
Reason:         Product is hypertonic.

                                     ----

Product:        Sulfa-Prednisolone Acetate Ophthalmic Suspension (predamide),
                a white aqueous suspension in an opaque white plastic bottles,
                under Akorn, Occusoft, Medical Ophthalmic and Wilson
                Ophthalmics labels.  Recall #D-301-2.
Code:           Lot numbers: 18948 EXP 5/92, 18859 EXP not provided,
                19019 EXP 6/92, 19024 EXP 6/92, 19035 EXP 7/92,
                19142 EXP 11/92, 19150 EXP 11/92, 19297 EXP 4/93.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 13, 1991.  Firm-initiated recall ongoing.
Distribution:   Louisiana, Oklahoma, Florida, Texas.
Quantity:       96,408 bottles were distributed.
Reason:         Use of unapproved manufacturing process.

                                     ----

Product:        Sodium Sulfacetamide Ophthalmic Solution 10%, 2 ml, a
                clear colorless solution in a opaque white plastic bottle,
                under Akorn label.  Recall #D-302-2.
Code:           Lot numbers:  18980 EXP 5/93, 19011 EXP not provided.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana,
                by letter January 10, 1992.  Firm-initiated recall ongoing.
Distribution:   Louisiana.
Quantity:       24,486 bottles were distributed.
Reason:         Lack of assurance of sterility.

                                     ----

Product:        Fluorescein/Proparacaine Ophthalmic Solution, a clear
                orange to pale yellow solution in a 5 ml amber glass
                screw cap bottle, under the Akorn label.  Recall #D-303-2.
Code:           lot numbers:  18986 EXP 5/92, 19010 EXP 5/92, 19014 EXP 6/92,
                19105 EXP 9/92, 19009 EXP 5/92, 19013 EXP 9/92, 19104 EXP 9/92.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 31, 1991.  Firm-initiated recall ongoing.
Distribution:   California, Louisiana.
Quantity:       65,697 bottles were distributed.
Reason:         Product does not meet pH specifications.

                                     ----

Product:        AK-Nefrin with Phenylephrine HCl Sterile Ophthalmic
                Solution, in 15 ml white squeezable plastic bottles, under
                Akorn, Ocusa Eye Drops, De La Cruz PDTS; Ojo De Aguila,
                Grandall Distribution labels.  Recall #D-304-2.

                                      -4-



Code:           Lot #18908 EXP 2/93.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 13, 1991.  Firm-initiated recall ongoing.
Distribution:   Louisiana, California.
Quantity:       15,224 bottles were distributed.
Reason:         Lack of assurance of sterility.

                                     ----

Product:        Sodium Chloride Ophthalmic Solution 5%, a clear colorless
                solution in 15 ml opaque white plastic bottles, under Akorn,
                Medical Ophthalmics labels.  Recall #D-305-2.
Code:           Lot #19124 EXP 10/92.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter January 10, 1992.  Firm-initiated recall ongoing.
Distribution:   Louisiana, Florida.
Quantity:       19,743 bottles were distributed.
Reason:         Lack of assurance of sterility.

                                     ----

Product:        Methylprednisolone Acetate Suspension 40 mg/ml and 80 mg/ml
                injection in a clear glass 5 ml vial, under Americal
                Pharmaceuticals, URL, In Source, WE Hauck, Stat
                Pharmaceuticals, Rugby, Maurry Biological, and Merit
                Pharmaceuticals labels.  Recall #D-306/307-2.
Code:           Lot numbers:  19191 EXP 06/92, 19316 EXP 12/92,
                19354 EXP 11/92, 19385 EXP 12/92, 19388 EXP 01/93,
                19436 EXP 01/93.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 13, 1991.  Firm-initiated recall ongoing.
Distribution:   Virginia, Georgia, Pennsylvania, California.
Quantity:       45,190 bottles were distributed.
Reason:         Product does not meet pH specifications.

                                     ----

Product:        Magnesium Sulfate Injection, 10% ampules, a clear
                colorless solution in 20 ml clear glass ampules.
                Recall #D-308-2.
Code:           Lot #19129 EXP 10/92.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 31, 1991.  Firm-initiated recall ongoing.
Distribution:   California.
Quantity:       6,100 ampules were distributed.
Reason:         Lack of assurance of sterility.

                                     ----

                                      -5-



Product:        Fluorescein Injection USP 25%, a deep reddish brown
                solution in a 2 ml clear glass ampule.  Recall #D-309-2.
Code:           Lot numbers:  19146 EXP 10/92, 19156 EXP 10/92.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana,
                by letter December 30, 1991.  Firm-initiated recall
                ongoing.
Distribution:   Louisiana, Canada.
Quantity:       17,724 ampules were distributed.
Reason:         Lack of assurance of sterility.

                                     ----

Product:        Sodium Heparin Injection, USP, 40,000 units per ml, a
                clear colorless solution, in 5 ml clear glass vials.
                Recall #D-310-2.
Code:           Lot #19110 EXP 8/93.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 13, 1991.  Firm-initiated recall ongoing.
Distribution:   California.
Quantity:       1,888 vials were distributed.
Reason:         Product packaged in unapproved vials.

                                     ----

Product:        Heparin Sodium Injection 1000 USP units/ml, a clear
                colorless solution in clear 30 ml glass vials.
                Recall #D-311-2.
Code:           Lot numbers:  19092 EXP 9/93, 19128 EXP 10/93.
Manufacturer:   Walnut Pharmaceuticals, Inc., Los Angeles, California.
Recalled by:    Walnut Pharmaceuticals, Inc., Abita Springs, Louisiana, by
                letter December 31, 1991.  Firm-initiated recall ongoing.
Distribution:   Louisiana.
Quantity:       14,214 vials of lot 19091 and 13,738 vials of lot 19128.
Reason:         Use of unapproved manufacturing process.

                                     ----

Product:        Effervescent Potassium Tablets, 25 mEq, lime flavored,
                individually wrapped in foil packets, in boxes of 30,
                100 and 250, an Rx preparation of an oral solution of
                potassium for prevention of potassium depletion.
                Recall #D-316-2.
Code:           Lot LPS10710-1.
Manufacturer:   Tower Laboratories Ltd., Centerbrook, Connecticut.
Recalled by:    Bajamar Chemical Company, Inc., St. Louis, Missouri
                (responsible firm), by telephone April 9, 1992, followed by
                letter.  Firm-initiated recall ongoing.
Distribution:   California, New Jersey, Florida, Rhode Island, Michigan,
                Texas, Virginia, South Dakota.
Quantity:       456 30-tablet boxes, 12 100-tablet boxes, and 1 250-tablet
                box was distributed.

                                      -6-



Reason:         Labeling does not comply with respect to declaration of
                Yellow No. 5.

                                     ----
                Class III -

Product:        E-Z-Paque, oral barium sulfate for suspension, in 176 g
                cups, used as a testing indicator of gastro-intestinal
                problems.  Recall #D-296-2.
Code:           Lot #TPO-006.
Manufacturer:   E-Z-EM, Inc., Westbury, New York.
Recalled by:    Manufacturer, by telephone March 23-24, 1992, followed by
                letter March 25, 1992.  Firm-initiated recall ongoing.
Distribution:   Great Britain.
Quantity:       234 cases (36 cups per case) were distributed.
Reason:         Some units of the E-Z-Paque were mislabeled as E-Z-HD
                (another E-Z-EM barium product).

                                     ----

Product:        Dexamethasone Tablets, 0.75 mg, in 100 and 1000 tablet
                bottles, under the following labels:  Richlyn, Parmed
                Pharmaceutical, IDE-Interstate, Genetco, URL.
                Recall #D-312-2.
Code:           Lot numbers:  41920 EXP 7/92, 42812 EXP 10/93.
Manufacturer:   Richlyn Laboratories, Philadelphia, Pennsylvania.
Recalled by:    Manufacturer, by letter March 25, 1992.  Firm-initiated
                recall ongoing.
Distribution:   Nationwide.
Quantity:       Firm estimates 93,400 tablets remain on market.
Reason:         Use of unapproved supplier of active ingredient.

                                     ----

Product:        (a) Methyltestosterone Tablets, 10 mg, oral;
                (b) Methyltestosterone Tablets, 10 mg, sublingual;
                (c) Methyltestosterone Tablets, 25 mg, oral, under the
                following labels:  Rugby - 10 mg - oral and sublingual
                (100's only); 25 mg - oral
                Goldline - 10 mg - oral and sublingual; 25 mg - oral
                (all 100's only)
                Schein Pharmaceutical- 10 mg - oral and sublingual; 25 mg -
                oral (all 100's only)
                United Research Laboratories - 10 mg - oral and sublingual;
                25 mg - oral (all 100's only)
                IDE-Interstate, Inc. - 10 mg - oral and sublingual; 25 mg -
                oral (all 100's only)
                Major Pharmaceutical - 10 mg - oral and sublingual; 25 mg -
                oral (all 100's only)
                C.O.Truxton, Inc. - 10 mg - sublingual (1000's only)
                Richlyn - 10 mg - oral and sublingual; 25 mg - oral
                (all in 100's & 1000's).  Recall #D-313/315-2.

                                      -7-

Code:           (a) 43084 43005 42865 42643 42546 42307, 42205 42084 
                41906, 41793, 41770
                (b) 43085, 43003, 42551, 42463, 42228, 42147, 41907, 41792
                (c) 44084, 44048, 43083, 43004, 42956, 42864, 42716
                42642, 42587, 42545, 42498, 42308, 42269, 42204, 42148, 42082
                42033, 41939, 41909, 41794, 41690.
Manufacturer:   Richlyn Laboratories, Philadelphia, Pennsylvania.
Recalled by:    Manufacturer, by letter March 25, 1992.  Firm-initiated
                recall ongoing.
Distribution:   Nationwide.
Quantity:       Undetermined.
Reason:         Use of unapproved supplier of active ingredient.

                                     ----

Product:        Various Rx drugs.  Recall #D-317-2.
Code:           All lots.
Manufacturer:   Various.
Recalled by:    Morris and Dickson Company, Ltd., Shreveport, Louisiana
                (distributor), by letter March 5, 1992.  Firm-initiated recall
                ongoing.
Distribution:   Louisiana, Texas, Arkansas, Oklahoma, Mississippi, Missouri,
                Kansas.
Quantity:       Unknown.
Reason:         Repackaging of products without accompanying prescribing
                information inserts.

                                     ----

                                   BIOLOGICS

                Class I -
                                     NONE
                Class II -

Product:        Elisa Computer Software, Version 4.6C.  Recall #B-250-2.
Code:           Product #937988.
Manufacturer:   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
Recalled by:    Manufacturer, by letter February 13, 1991.  Firm-initiated
                recall complete.
Distribution:   Nationwide.
Quantity:       73 pieces were distributed.
Reason:         Design defect in computer software that could cause the
                inadvertent distribution of untested blood and blood
                products.
                                     ----

Product:        Elisa Computer Software, Versions 4.6C, 4.6D, 5.0, and
                5.1.  Recall #B-251-2.
Code:           937988.
Manufacturer:   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
Recalled by:    Manufacturer, by telephone February 28, 1992, followed
                by letters of March 3, 1991 and by telefax March 9, 1992.
                Firm-initiated recall ongoing.

                                      -8-



Distribution:   Nationwide and international.
Quantity:       379.
Reason:         Design defect in computer software that could cause the
                inadvertent distribution of blood and blood products
                inadequately tested for the antibody to human immunodeficiency
                virus type 1 (anti-HIV-1).

                                     ----

Product:        (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen
                Plasma; (d) Recovered Plasma.  Recall #B-255/258-2.
Code:           Unit numbers:  (a) 6200452, 6253797, 6312769, 6359583,
                6374953, 6393719, 6414431;
                (b) 6200452, 6253797, 6312769, 6359583, 6374953, 6393719;
                (c) 6374953, 6393719;
                (d) 6200452, 6253797, 6312769, 6359583.
Manufacturer:   Broward Community Blood Center, Inc., Lauderhill, Florida.
Recalled by:    Manufacturer, by letter October 5, 1990.  Firm-initiated
                recall ongoing.
Distribution:   Florida, New York.
Quantity:       (a) 7 units; (b) 6 units; (c) 2 units; (d) 4 units.
Reason:         Blood products, which tested nonreactive for the
                antibody to human immunodeficiency virus type 1
                (anti-HIV-1), but were collected from a donor who
                previously tested repeatably reactive for anti-HIV-1,
                were distributed.

                                     ----
                Class III -

Product:        Red Blood Cells.  Recall #B-209-2.
Code:           Unit numbers 4412564 and 4412721.
Manufacturer:   Florida Georgia Blood Alliance, Jacksonville, Florida.
Recalled by:    Consignees notified firm by telephone June 13, 1991 and
                July 3, 1991.  Recall complete.
Distribution:   Florida.
Quantity:       2 units.
Reason:         Blood products, labeled with incorrect expiration dates,
                were distributed.

                                     ----

Product:        (a) Red Blood Cells; (b) Red Blood Cells, Leukocytes
                Removed.  Recall #B-211/212-2.
Code:           Unit numbers:  (a) 1611479, 1620674; (b) 167959, 1676660.
Manufacturer:   Palm Beach Blood Bank, West Palm Beach, Florida.
Recalled by:    Manufacturer, by telephone March 27, 1991, April 16, 1991,
                and by telephone and follow-up letter on December 27, 1991.
                Firm-initiated recall ongoing.
Distribution:   California, Washington, D.C.
Quantity:       2 units of each component.

                                      -9-


Reason:         Blood products, labeled with incorrect expiration dates,
                were distributed.

                                     ----

Product:        Red Blood Cells.  Recall #B-231-2.
Code:           Unit #68479.
Manufacturer:   Metropolitan Washington Blood Banks, Inc., Silver Spring,
                Maryland.
Recalled by:    Manufacturer, by retrieving the recalled product from the
                consignee on August 15, 1991.  Recall complete.
Distribution:   Washington, D.C.
Quantity:       1 unit.
Reason:         Blood product which was labeled with an incorrect expiration
                date was distributed.

                                     ----

Product:        Red Blood Cells.  Recall #B-252-2.
Code:           Unit #45G77095.
Manufacturer:   American Red Cross Blood Services, Waterloo, Iowa.
Recalled by:    Manufacturer, by telephone January 16, 1992.  Firm-initiated
                recall complete.
Distribution:   Iowa.
Quantity:       1 unit.
Reason:         A unit of Red Blood Cells, labeled with the incorrect
                expiration date, was distributed.

                                     ----

Product:        Anti Human Globulin (anti-IgG), an in-vitro diagnostic
                used in the Coombs test.  Recall #B-253-2.
Code:           Lot #MSGG42-1B.
Manufacturer:   Baxter Diagnostics, Inc., Miami, Florida.
Recalled by:    Manufacturer, by letter February 6, 1992.  Firm-initiated
                recall complete.
Distribution:   Nationwide, Canada, Australia, Brazil, Spain.
Quantity:       196 packages were distributed.
Reason:         Anti-Human Globulin labeled with the incorrect expiration
                date was distributed.

                                     ----

                       DEVICES AND RADIOLOGICAL PRODUCTS

                Class I -
                                     NONE

                Class II -

Product:        Cordis Shuttle Catheters, sterile single lumen catheters
                designed to reach cardiovascular system arterial stenoses.
                Recall #Z-455-2.

                                     -10-


Code:           Catalog numbers:  525-402 and 525-402S - All lots.
Manufacturer:   Cordis Corporation, Miami Lakes, Florida.
Recalled by:    Manufacturer, by letter June 7, 1991.  Firm-initiated
                recall ongoing.
Distribution:   Nationwide, Canada.
Quantity:       Firm estimates none remains on the market.
Reason:         The tips may separate from the catheters.

                                     ----

Product:        Palmaz-Schatz Balloon-Expandable STENT, for the biliary
                system, 20 cm length, 2.5 mm diameter, used to maintain patency
                of a major bile duct which is obstructed by scar tissue or
                impinging tumor.  Recall #Z-516-2.
Code:           Product #PS20M, Lot #CG068 EXP Use before July '93.
Manufacturer:   Johnson & Johnson Interventional Systems Company, Warren,
                New Jersey.
Recalled by:    Manufacturer, by letter January 23, 1992.  Firm-initiated
                recall ongoing.
Distribution:   New Jersey, Pennsylvania, Maryland, Ohio, Michigan, Illinois,
                Wisconsin, Florida, Oklahoma, Minnesota, Kansas, Nebraska,
                Nevada, Texas, Colorado, Arizona, California, Washington state,
                Oregon, Ireland, Chile.
Quantity:       80 boxes (4 units per box) were distributed.
Reason:         Product was distributed without premarket approval authority
                for use in the renal arteries.

                                     ----

Product:        Bateman UPF II Outer Head BiPolar Shell, a component in a
                modular hip replacement system, and is produced in a
                number of sizes.  Recall #Z-554-2.
Code:           Catalog #10644, lot #004084.
Manufacturer:   Kirschner Medical Corporation, Timonium, Maryland.
Recalled by:    Manufacturer, by letter February 13-14, 1992.  Firm-initiated
                recall ongoing.
Distribution:   Illinois, Massachusetts, Tennessee, Florida.
Quantity:       12 units were distributed.
Reason:         The labeling on the outside box is incorrect.  The label on
                the box stated that the product was 46 mm, but the device
                was actually 43 mm.

                                     ----

Product:        Performance Knee Tibial Component Modular Insert.
                Recall #Z-555-2.
Code:           Catalog #7404-08-072, lot #006475.
Manufacturer:   Kirschner Medical Corporation, Timonium, Maryland.
Recalled by:    Manufacturer, by letter February 13-14, 1992.  Firm-initiated
                recall ongoing.
Distribution:   California, Pennsylvania, Oregon, Texas, Alabama, France,
                Belgium, Holland.
Quantity:       47 units were distributed.

                                     -11-



Reason:         The color coding labeling on the outside box is incorrect.
                The label on the box stated that the product was 46 mm,
                but the device was actually 43 mm.

                                     ----

Product:        Cordis Hakim Hydrocephalus Valve; Pediatric Shunt System,
                an implantable device which serves as a parallel flow pathway
                to divert cerebrospinal fluid (CSF) from the cerebral
                ventricles to an appropriate drainage site in patients with
                hydrocephalus or other conditions in which CSF flow and/or
                absorption is impaired.  Recall #Z-558-2.
Code:           Catalog #908-162, serial numbers:  52894, 52889, 52888.
Manufacturer:   Cordis-Biot, Valbonne, France.
Recalled by:    Cordis Corporation, Miami Lakes, Florida, by telephone
                January 25, 1991.  Firm-initiated recall complete.
Distribution:   Texas, Florida.
Quantity:       3 units were distributed; firm estimates none remains
                on the market.
Reason:         Certain body parts are incorrectly fabricated from 316
                stainless steel (instead of 304 stainless steel).

                                     ----

Product:        Karlin Crank Frame Retractor Blades, components of the spinal 
                retractor set, designed for posterior lumbar and thoracic
                surgery.  Recall #Z-559-2.
Code:           Lot numbers:  60L, 049L, 060L, 90L, 30L, 129L, 109L, 020L,
                089L, 69L, and all product that does not bear a lot number.
Manufacturer:   T. Koros Surgical Instruments, Moorpark, California.
Recalled by:    Karlin Technology, Inc., Torrance, California, by
                letter August 23, 1991.  Firm-initiated recall ongoing.
Distribution:   Nationwide, Canada.
Quantity:       6,619 sets were distributed.
Reason:         Some of the welds which hold the blades together may not be
                adequate, with the possibility for failure and patient
                injury.
                                     ----

Product:        PCA Two Piece Cup - Low Profile Acetabular Shell, used
                for total hip replacement orthopedic implant surgery.
                Recall #Z-560-2.
Code:           Catalog #6289-5-064, Lot #ACCUAA 9129017.
Manufacturer:   Howmedica Division Pfizer Hospital Products Group, Inc.,
                Rutherford, New Jersey.
Recalled by:    Manufacturer, by telephone January 10, 1992.  Firm-initiated
                recall complete.
Distribution:   New York, Michigan, Nebraska, New Mexico, Georgia, Texas,
                Ohio, Canada.
Quantity:       9 units were distributed.  Firm estimates none remains
                on the market.

                                     -12-


Reason:         The package was incorrectly labeled Catalog #6289-5-046,
                46 mm size shell, instead of Catalog #6289-5-064, 64 mm size
                shell.
                                     ----

Product:        XT Fixed Height Single Monitor Support, Model 46-240838G1,
                used with x-ray systems as a monitoring tool.  Recall #Z-563-2.
Code:           Supports used on B2056FF (oyster gray) and on B2057FF
                (mist gray) systems were affected.
Manufacturer:   General Electric Company, Inc., Waukesha, Wisconsin.
Recalled by:    Manufacturer, by letter March 16, 1992.  Firm-initiated
                field correction ongoing.
Distribution:   Nationwide.
Quantity:       135 units were distributed.
Reason:         The "pivot stud" of the suspension system could shear off
                allowing the system to fall onto a patient or operator.

                                     ----

Product:        Percutaneous Cardiovascular Catheter, Pigtail Curve,
                80 cm, 4F, disposable, used for injecting a dye to
                diagnose blockage.  Recall #Z-565-2.
Code:           Catalog #0001-10-1545, lot #134117.
Manufacturer:   Universal Medical Instruments Corporation, Ballston Spa,
                New York.
Recalled by:    Manufacturer, by telephone March 3, 1992, followed by
                letter March 4, 1992.  Firm-initiated recall ongoing.
Distribution:   Michigan, New York, California, Massachusetts.
Quantity:       161 units were distributed; firm estimates 103 units
                remain on the market.
Reason:         Product may leak at the proximal end of the assembly under
                the shrink tube when fluids are injected through the
                catheter.

                                     ----

Product:        Fixed Wing Huber Infusion Sets, a non-coring needle
                administration set with a "Y" injection site, under
                Medical Specialties and Churchill labels:
                (a) Product Nos. MS-425-192, MSNS-425-192 and CHNS-425-192,
                19GA x 3/4" long needle;
                (b) Product Nos. MS-425-193, MSNS-425-193 and CHNS-425-193,
                19GA x 1" long needle;
                (c) Product Nos. MS-425-202, MSNS-425-202 and CHNS-425-202,
                20GA x 3/4" long needle;
                (d) Product Nos. MS-425-203, MSNS-425-203 and CHNS-425-203,
                20GA x 1" long needle;
                (e) Product Nos MS-425-206, MSNS-425-206 and CHNS-425-206,
                20GA x 1 1/4" long needle.  Recall #Z-566/570-2.
Code:           All lots.
Manufacturer:   Nipro Medical Industries, Ltd., Tatebayashi, Gumma, Japan.
Recalled by:    Medical Specialties Co., Inc., South Easton, Massachusetts,
                by letter March 16, 1992.  Firm-initiated recall ongoing.

                                     -13-


Distribution:   Nationwide.
Quantity:       21,387 pieces were distributed.
Reason:         Cannula separation can occur.

                                     ----

Product:        Hewlett-Packard Sonos 100, 500 and 1000 Ultrasound Imaging
                Systems, used with software revisions B.0 through B.0.2 and
                M.0 through M.3.1.  Recall #Z-571/573-2.
Code:           Model numbers:  HP series (a) 77000A; (b) 77025A; (c) 77030A.
Manufacturer:   Hewlett-Packard Company, Andover, Massachusetts.
Recalled by:    Manufacturer, by letter September 6, 1991.  Firm-initiated
                field correction ongoing.
Distribution:   Nationwide, Canada, Europe, Asia, Latin America.
Quantity:       (a) 219 units; (b) 1,721 units; (c) 1,720 units were
                distributed.  As of 3/5/92 the firm had corrected 1,850
                units.
Reason:         Software design error may lead to misdiagnosis
                resulting in unnecessary invasive procedures when
                analyzing mitral valve flow rates.

                                     ----

Product:        LaserCat L2503 CNC Laser Cutting Center, used for cutting
                sheet metal.  Recall #Z-575-2.
Code:           Serial numbers:  Not applicable.
Manufacturer:   Trumpf, Inc., Farmington, Connecticut.
Recalled by:    Manufacturer.  FDA approved the firm's corrective action plan
                April 3, 1992.  Firm-initiated field correction ongoing.
Distribution:   Nationwide.
Quantity:       19 units.
Reason:         Noncompliance with the performance standard for laser
                products in that the operational and adjustment controls
                were not adequately located to prevent the operator from
                being exposed to laser beams.

                                     ----

Product:        MISTI Mammary Prostheses:
                (a) MISTI (silicone) Gel-Filled Textured Mammary Prostheses,
                all sizes (80-980 cc);
                (b) MISTI (silicone) Double-Lumen Textured Mammary Prostheses,
                all sizes;
                (c) MISTI GOLD Bio-Oncotic Gel Single-Lumen Mammary Prostheses,
                all sizes (80-620 cc).  Recall #Z-610-612-2.
Code:           All lots.
Manufacturer:   Biomanufacturing, Inc., Minneapolis, Minnesota.
Recalled by:    Bioplasty, Inc., St. Paul, Minnesota, by letter dated
                March 10, 1992.  Firm-initiated recall ongoing.
Distribution:   Nationwide and international.
Quantity:       (a & b) Approximately 50,000 units were distributed;
                (c) Unknown.

                                     -14-



Reason:         The devices are not approved for marketing.

                                     ----

                Class III -
                                     NONE

                              VETERINARY PRODUCTS

                Class I -
                                     NONE
                Class II -
                                     NONE
                Class III -

Product:        Banminth 48 Type A Medicated Article, in 50 pound bags,
                an anthelmintic for use in swine, under the Pfizer label.
                Recall #V-030-2.
Code:           Lot numbers:  K10735539 and K10735540.
Manufacturer:   Pfizer, Inc., Lee's Summit, Missouri.
Recalled by:    Manufacturer, by letter February 14, 1992.  Firm-
                initiated recall complete.
Distribution:   North Carolina, Iowa, South Dakota, Nebraska, Tennessee,
                Minnesota, Utah.
Quantity:       236 bags were distributed; firm estimates none
                remain on the market.
Reason:         Product is subpotent.

                                     ----

Seizure Actions Filed:

Product:        Herbal Shampoo and Herbal Cosmetic Scalp Cleanser
                (92-649-597).
Charge:         Products are unapproved new drugs.
Firm:           Catherine Michael's Hair Salon.
Filed:          March 20, 1992 - U.S. District Court for the District of
                Massachusetts; Civil #92-10684MA, FDC #66370.
Seized:         April 10, 1992 - products and pamphlets valued at
                approximately $1,600.

                                     ----

                                     -15-

                                     ###