FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
01/08/1992

Recalls and Field Corrections: January 8, 1992

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or
                  death.

                                     NONE

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the
                  probability of serious adverse health consequences is
                  remote.

                                     NONE

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse
                  health consequences.

                                     NONE

                                      -1-
                                       



                                   COSMETICS

                                     NONE

                           HUMAN DRUGS AND BIOLOGICS

                  Class I -

                                     NONE
                  Class II - 

Product:          DDAVP Injection (Desmopressin Acetate), 4 mgc/ml, in 1 ml
                  ampule, ten  1 ml ampules per unit carton, shipper
                  carton contains 50 unit cartons, an Rx antidiuretic
                  hormone.  Recall #D-140-2.
Code:             Lot #QM9598 EXP 1/93.
Manufacturer:     Ferring Pharmaceuticals, Malmo, Sweden.
Recalled by:      Rhone Poulenc Rorer, Fort Washington, Pennsylvania, by
                  telephone December 6, 1991, followed by letter December
                  10, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide.
Quantity:         Approximately 1,949 unit cartons were distributed.
Reason:           Front panel of carton states "Preservative Free."  Rear
                  panel declares chlorobutanol as preservative which the
                  product does contain.

                                     ----

Product:          Whole Blood.  Recall #B-074-2.
Code:             Unit numbers:  4256873, 4256885.
Manufacturer:     Blood Bank Naval Hospital Long Beach, Department of
                  the Navy, Long Beach, California.
Recalled by:      Manufacturer, by telephone December 24, 1990.  Firm-
                  initiated recall complete.
Distribution:     New Jersey.
Quantity:         2 units.
Reason:           Blood components, which tested repeatably reactive for
                  the antibody to the human immunodeficiency virus, type 1
                  (anti-HIV-1), were distributed.

                                     ----

Product:          (a) Red Blood Cells; (b) Platelets;
                  (c) Cryoprecipitated AHF; (d) Fresh Frozen Plasma;
                  (e) Recovered Plasma.  Recall #B-075/079-2.
Code:             Unit numbers:  (a) 31C01142, 31C02790, 31C04046, 31C14148,
                  31C16315, 31F23510, 31G06328, 31G68272, 31G71807, 31G91100,
                  31H05395, 31S07348, 31S09633, 31S11959, 31S14017, 31S15953,
                  31T53794; (b) 31C16315;
                  (c) 31C02790, 31C14148, 31S09633, 31S15953;
                  (d) 31G91100, 31H05395;
                  (e) 31C01142, 31C02790, 31C04046, 31C14148, 31C16315,
                  31S14017, 31S15953, 31T53794.

                                      -2-





Manufacturer:     American Red Cross Blood Services, Buffalo, New York.
Recalled by:      Manufacturer, by letters in June and July 1991.
                  Firm-initiated recall ongoing.
Distribution:     New York, California, Georgia, Illinois, Pennsylvania,
                  North Carolina, Switzerland.
Quantity:         (a) 17 units; (b) 1 unit; (c) 4 units; (d) 2 units;
                  (e) 8 units.
Reason:           Blood products, collected from donors who had histories
                  of self-injected drug use, unspecified hepatitis, or
                  jaundice, were distributed.

                                     ----

Product:          Platelets.  Recall #B-081-2.
Code:             Unit numbers:  19GG26079, 19GG26082, 19GG26084, 19KK77530,
                  19KK77531, 19KK77535, 19KK77537, 19KK77538, 19KK77540,
                  19KK77542, 19KK77547, 19Z43458, 19Z43460, 19Z43464,
                  19Z43465, 19Z43467, 19Z43468, 19Z43470, 19Z43472, 19Z43473,
                  19Z43478, 19Z43480, 19Z43482, 19Z43483, 19Z43485, 19Z43496.
Manufacturer:     American Red Cross Blood Services, Paducah, Kentucky.
Recalled by:      Manufacturer, by telephone April 9, 1991.  Firm-initiated
                  recall complete.
Distribution:     Kentucky.
Quantity:         26 units.
Reason:           Blood products labeled with incorrect expiration dates
                  were distributed.

                                     ----

                  Class III -

Product:          Tylenol, OTC acetaminophen, 80 mg, chewable tablets for
                  children, one chewable tablet per unit dose packette with
                  250 unit dose packettes per folding carton (intermediate
                  carton), 4 folding cartons per shipping carton.
                  Recall #D-139-2.
Code:             Lot #HDM573 EXP 5/94.
Manufacturer:     McNeil Consumer Products, Fort Washington, Pennsylvania.
Recalled by:      Manufacturer, by letter December 13, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide.
Quantity:         2,384 folding cartons (596 cases) were distributed.
Reason:           Some units of 250 unit dose tablets (80 mg) were
                  packaged in cartons labeled Tylenol Tablets (325 mg).

                                     ----

Product:          Phenergan Suppositories, Rx, 12.5 mg, packaged 12 units in
                  a unit carton with 12 unit cartons making up a shelf pack
                  (intermediate packaging) and 12 shelf packs comprising a
                  shipper carton.  Recall #D-141-2.
Code:             Lot #1910171.
Manufacturer:     Wyeth-Ayerst Laboratories, Frazier, Pennsylvania.

                                      -3-





Recalled by:      Wyeth-Ayerst Laboratories, St. Davids, Pennsylvania, by
                  telephone followed by visit.  Firm-initiated field
                  correction complete.
Distribution:     Nationwide.
Quantity:         47,856 unit cartons (3,988 shelf packs) were distributed.
Reason:           Some correctly labeled unit packs of 12.5 mg strength were
                  packed in shelf packs labeled 25 mg.

                                     ----

Product:          Whole Blood.  Recall #B-080-2.
Code:             Unit #27FC07342.
Manufacturer:     American Red Cross Blood Services, Johnstown, Pennsylvania.
Recalled by:      Manufacturer, by telephone February 1, 1991.  Firm-
                  initiated recall complete.
Distribution:     Pennsylvania.
Quantity:         1 unit.
Reason:           Blood product, labeled with an incorrect expiration
                  date and the incorrect anticoagulant, was distributed.

                                     ----

EXTENSION         Dyazide Capsules (Triamterene 50 mg, Hydrochlorothiazide
NOTICE:           25 mg), Recall #D-035-2, manufactured by Smithkline
                  Beecham Pharmaceuticals Co., Cidra, Puerto Rico,
                  which appeared in the November 13, 1991, Enforcement Report
                  has been extended to include additional lots.  These lots
                  can be identified in that the last 4 characters of the lot
                  numbers being recalled are as follows:  6E90, 7E90, 8E90.

                                     ----

                       DEVICES AND RADIOLOGICAL PRODUCTS

                  Class I -

Product:          Reflex DDD, Model 8222, Dual Chamber, Multiprogrammable
                  Pulse Generator with Telemetry.  Recall #Z-101-2.
Code:             All serial numbers.
Manufacturer:     Telectronics Pacing Systems, Inc., Englewood, Colorado.
Recalled by:      Manufacturer, by memorandum to sales representatives
                  October 18, 1991 regarding unimplanted units and by
                  letter November 11 and 26 to physicians and hospital
                  administrators.  Firm-initiated recall ongoing.
Distribution:     Nationwide and international.
Quantity:         2,364 units were distributed.
Reason:           Cracking or fracture of the ceramic terminal assembly
                  post.  The cracking or fracturing causes a loss of
                  hermeticity and subsequent loss of output.

                                     ----

                                      -4-





                  Class II -

Product:          A.T.H.S. Latex Disposable Examination Gloves, 100 per
                  box, 2,000 per case.  Recall #Z-102-2.
Code:             None.
Manufacturer:     Phoenix Medical Service, Inc., St. Paul, Minnesota.
Recalled by:      Manufacturer, by letter October 28, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Minnesota.
Quantity:         459 cases were distributed.
Reason:           Product failed FDA leak test.

                                     ----

Product:          Del Med brand Blood Component Recipient Set, packaged 1
                  unit per box, 48 boxes per case, used for intravenous
                  administration of blood or blood components;
                  (b) Blood Set with Large Filter, packaged 1 unit per box,
                  48 boxes per case, used for intravenous administration of
                  blood or blood components;
                  (c) Non-Vented IV Administration Set, packaged 1 unit
                  per box, 48 boxes per case, used for intravenous
                  administration of IV fluids;
                  (d) Plasma Transfer Set with Two Spikes, packaged 2 units
                  per box, 48 boxes per case, used for the transfer of
                  plasma from the blood bag to transfer packs;
                  (e) Plasma Transfer Set with Needle Adapter and Spike
                  packaged 2 units per box, 48 boxes per case, used to
                  transfer plasma from blood bag to plasma collection
                  container with rubber stopper.  Recall #Z-105/109-2.
Code:             Product Nos.      Lot Nos.
                  (a) 03-110-00     8K-002-04
                  (b) 03-118-00     8K-007-04
                  (c) 03-131-00     8J-016-04
                  (d) 03-220-00     8J-017-04
                  (e) 03-220-30     8J-018-04.
Manufacturer:     DelMed, Inc., Ogden, Utah (sterilizer, shipper, release
                  tester); DelMed, Inc., SA, Soyapango, San Salvador,
                  El Salvador, California (clamp, unit).
Recalled by:      CharterMed, Inc., Lakewood, New Jersey, by letter mailed
                  out on September 30, 1991.  Firm-initiated recall ongoing.
Distribution:     (a) California, Kansas, New Jersey, Ohio, Maryland,
                  New York, Georgia; (b) Florida; (c) Massachusetts;
                  (d) Washington state, Maryland, Ohio, Tennessee, Florida,
                  New Jersey, Mississippi, California, New York;
                  (e) Florida, Pennsylvania, Maryland, New York, California,
                  Texas.
Quantity:         (a) 7,440 units; (b) 240 units; (c) 2,400 units;
                  (d) 6,816 units; (e) 3,360 units were distributed.

                                      -5-





Reason:           Products contain defective roller clamps which may break
                  during use or not completely prevent flow through the
                  tubing.  This may result in interrupted or unrestricted flow
                  of solution through the tubing.

                                     ----

Product:          Tape Electrodes, with Medtronic and OEM labeling:
                  (a) Part #1650 - Holter/Stress Transparent Tape Electrode;
                  (b) Part #1620 - Huggable (R) Tape Electrode;
                  (c) Part #1675 - High Performance ECG Monitoring Electrode
                  Recall #Z-116/118-2.
Code:             Medtronic Product Nos.        Lot Nos.
                  1650-005 N                    30 MAY 92 D, 30 MAY 92 E
                                                07 JUN 92 B, 28 JUN 92 C
                                                28 JUN 92 E, 04 JUL 92 A
                  1620-001 N                    07 JUN 92 D, 07 JUN 92 E
                                                04 JUL 92 C, 04 JUL 92 D
                  1620-003 N                    30 MAY 92 E, 07 JUN 92 A
                  1675-030                      14 JUN 92 D
                  5201                          28 JUN 92
                  5202                          07 JUN 92
                  OEM Label Product No.         Lot Nos.
                  18366                         14 JUN 92 B, 14 JUN 92 C
                  220041G2                      21 JUN 92 C
                  24938-401                     07 JUN 92 B
                  392015-001                    21 JUN 92 B, 21 JUN 92 C
                  H997                          21 JUN 92 D, 21 JUN 92 E
                                                28 JUN 92 E.
Manufacturer:     Medtronic Andover Medical, Haverhill, Massachusetts.
Recalled by:      Manufacturer, by letter April 4, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide and international.
Quantity:         793,053 units were distributed.
Reason:           Non-uniform silverplating on stud of ECG may cause poor
                  trace quality.

                                     ----

Product:          Stryker Power Cord, Part Number 277-702-19, a component
                  distributed with the following endoscopy equipment:
                  System SE3, arthroscopy system consoles, part numbers
                  266-702, 266-792; cameras 777, 590, and 578; light source
                  220-130; and insufflator 620-20.  Recall #Z-134-2.
Code:             No specific codes are identified.  All of the above
                  products which were shipped between March 15 and May 3,
                  1991, may contain a cord which is suspected as being
                  improperly wired.
Manufacturer:     Stryker Corporation, Endoscopy Division, San Jose,
                  California.
Recalled by:      Stryker Corporation, Kalamazoo, Michigan, by letter
                  May 20, 1991.  Firm-initiated recall complete.

                                      -6-





Distribution:     Nationwide, Canada, Hong Kong, Argentina, Brazil, Japan,
                  Venezuela.
Quantity:         Power cords in 531 finished product units remained in
                  commerce at time of recall.
Reason:           The clear-colored (female) connector end of the cord
                  assembly has live (brown) and neutral (blue) wires
                  reversed.
                                     ----

Product:          GE Computers for computerized tomographic (CT) systems:
                  (a) Models 46-301394G4 & G14 Computers for CT
                  Independent Consoles (CT-IC);
                  (b) Models 46-301394G5 & G15 Computers of CT Operator
                  Consoles (CT-OC);
                  (c) Model 46-301394G31 Computers for (CT-OC);
                  (d) Model 46-301394G33 Computers for (CT-IC).
                  Recall #Z-150/153-2.
Code:             All serial numbers of the CT system computers made prior
                  to October 14, 1991.
Manufacturer:     GE Medical Systems, Waukesha, Wisconsin.
Recalled by:      Manufacturer, by letter October 25, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide and international.
Quantity:         413 units were distributed.
Reason:           The imaging, patient, and other hospital information for
                  one patient exam may be mismatched with the images from
                  different CT exams.

                                     ----

Product:          Baxter Temperature Probes, Catalog numbers MR560, MF561,
                  MR565; and Accessory Packs with Catalog #MR746.
                  Recall #Z-201-2.
Code:             All lot numbers from March 1989 to October 1991.
Manufacturer:     Fisher and Paykel Electronics, Ltd., Auckland, New Zealand.
Recalled by:      Baxter Healthcare Corporation, Pharmaseal Division,
                  Valencia, California.  Firm-initiated recall ongoing.
Distribution:     Nationwide
Quantity:         756 probes and 4 accessory packs which contain probes
                  were distributed.
Reason:           The temperature probe may short circuit, which may result
                  in incorrect airway temperature display on humidifier heater
                  base.

                                     ----

Product:          Companion 2801 Portable Volume Ventilator, used to
                  provide support for ventilator dependent patients in both
                  the hospital and home care settings.  Recall #Z-216-2.
Code:             All serial numbers.
Manufacturer:     Puritan-Bennett, Portable Ventilator Division, Boulder,
                  Colorado.
                                      -7-





Recalled by:      Manufacturer, by letter December 16, 1991.  Firm-initiated
                  field correction ongoing.
Distribution:     Nationwide, Canada, England, Puerto Rico, Argentina, Italy.
Quantity:         1,200 units were distributed.
Reason:           Various failure modes, which appear to be associated
                  with breakage of the roll pin which holds the crank arm to
                  the motor output shaft, can result in the device failing
                  to function as intended.

                                     ----

Product:          Healthy Kleaner, The Good for Your Skin Kleaner, in 2, 4,
                  and 8 ounce sizes, labeled uses are for removal of
                  permanent ink, adhesive, heavy oil, tar, glue, caulking
                  and other oily grime from skin and many other surfaces.
                  Recall #Z-220-2.
Code:             None.
Manufacturer:     The GreenSpan, Inc., Boulder, Colorado.
Recalled by:      Manufacturer, by telephone in mid-December 1991, to be
                  followed by letter.  Firm-initiated recall ongoing.
Distribution:     Nationwide.
Quantity:         29 cases of 8 ounce bottles and 7 cases of 4 ounce
                  bottles were distributed.
Reason:           The device makes medical claims in that it can be used
                  to remove bandage adhesive, and is distributed without
                  an approved 510(k).

                                     ----

Product:          SynchroMed Infusion Pump:
                  (a) Model 8611H; (b) Model 8615; (c) Model 8610H/8502;
                  (d) Model 8611H/8502, designed to contain and to
                  administer parenteral drugs to a specific site.
                  Recall #Z-221/224-2.
Code:             Serial range from DAA000004R to DAA00084OR, SJ1203199R
                  to SJ120444OR and VZ1201342R to VZ1201486R.  Not all
                  serial numbers in these ranges are the numbers of
                  recalled devices.
Manufacturer:     Medtronic Neurological, Minneapolis, Minnesota.
Recalled by:      Medtronic, Inc., Minneapolis, Minnesota, by letter
                  dated October 24, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide, Argentina, Australia, Canada, Hong Kong, Japan,
                  Mexico, The Netherlands.
Quantity:         1,181 pumps were distributed.
Reason:           The devices may exhibit flow rates in excess of accuracy
                  claims of plus or minus 15 percent due to pump tubing
                  deviating from required inner diameter size.

                                     ----

                                      -8-





Product:          Stryker Compact Endoscopy Organizer, Part #240-75,
                  a cart with 4 shelves which holds a complete endoscopy
                  system including a camera, VCR and monitor.
                  Recall #Z-227-2.
Code:             All product ever distributed.
Manufacturer:     Practical Design, Safety Harbor, Florida.
Recalled by:      Stryker Corporation, Kalamazoo, Michigan, by letter
                  July 25, 1991.  Firm-initiated recall complete.
Distribution:     Nationwide and Canada.
Quantity:         301 carts were distributed.
Reason:           The caster (wheels) on the Endoscopy Organizer have the
                  propensity to fall off (over time) at the rivet flange joint.
                  This joint, responsible for holding the wheel housing to
                  the frame of the compact Endoscopy Organizer, is being
                  worn by the sharp edge of the column bolt head.

                                     ----

Product:          Omni-Fit Series II Cup Insert, a modular acetabular cup
                  component of total hip replacement system, packaged in
                  cartons of one, a Rx sterile device.  Recall #Z-228-2.
Code:             Catalog #2041-2850, case code CADW (43 units);
                  Catalog #2041-2862, case code CADY (18 units).
Manufacturer:     Disanto Machine & Tool Company, Stamford, Connecticut.
Recalled by:      Osteonics, Corporation, Allendale, New Jersey, by
                  telephone September 25, 1991, followed by letter
                  October 11, 1991.  Firm-initiated recall ongoing.
Distribution:     Arkansas, Florida, Massachusetts, Michigan, Minnesota,
                  North Carolina, Nebraska, Texas, Virginia, Wisconsin.
Quantity:         61 units were distributed.
Reason:           There is a manufacturing error in the location of the
                  positioning groove on the metal shell face.

                                     ----

                  Class III -

Product:          IVION Burette IV Sets.  Recall #Z-133-2.
Code:             Catalog No.      Lot No.
                  IV3A03           F7018A, F7019B, F7032, F7053A, F7054,
                                   F7076, F7095, F7106, F7107, F7128, F7129,
                                   F7131, F7161, F7162A
                  IV3A03#603       F7094, F7123
                  IV3A05           F6971A, F6997, F6998, F7051B, F7057,
                                   F7058, F7092, F7093, F7119, F7120, F7158,
                                   F7159, F7165, F7166, F7233A
                  IV3A06           F6973, F6974, F6993, F6994A, F7006,
                                   F7007, F7039A, F7060A, F7061, F7078,
                                   F7086, F7097, F7098, F7103, F7104, F7110,
                                   F7111, F7125, F7134, F7195
                  IV3A08           F7023A, F7085, F7146

                                      -9-





                  IV3A09#601       F6982, F6983, F7064A, F7065, F7147,
                                   F7148, F7180, F7181
                  IV3A11#273       F7114, F7143
                  IV3A12           F7024A, F7025, F7074, F7171
                  IV3A13           F7003A, F7068, F7157
                  IV3D05           F7170
                  IV3D06           F7005, F7069, F7070A
                  IV3D07           F7113, F7155
                  IV3D08           F7028A, F7091, F7156
                  IV3D08#618       F7082
                  IV3D10           F7000, F7009, F7016, F7029, F7030, F7080,
                                   F7141
                  IV3D11#611       F6981, F7001A, F7034A, F7067A, F7116,
                                   F7142.
Manufacturer:     Ivion Corporation, Broomfield, Colorado.
Recalled by:      Manufacturer, by telephone November 27, 1991, followed
                  by letter November 30, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide, Canada.
Quantity:         Firm estimates 3,000 cases (20 sets per case) remain
                  on the market.
Reason:           The device may leak between the burette cylinder and the
                  bottom cap juncture.

                                     ----

Product:          COBE Waste Handling Option (WHO) Accessory.
                  Recall #Z-217-2.
Code:             Catalog numbers: 018737-000 and 333250-000, all serial
                  numbers.
Manufacturer:     CGH Medical, Inc., Lakewood, Colorado 
Recalled by:      Manufacturer, by letter November 1, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide, Australia, Belgium, Canada.
Quantity:         1,051 units were distributed.
Reason:           A higher than anticipated failure rate of two components
                  result in leaking dialysate around the "boot" and valve
                  of the assembly.

                                     ----

Product:          Baxter Fenwal Blood Component Infusion Set, a Rx sterile
                  13" infusion set with a female luer used for the
                  syringe push of small volume blood components.
                  Recall #Z-218-2.
Code:             Product #4C2223, lot numbers:  U139337, U140863, U141663,
                  U143347, U145301, U145292.
Manufacturer:     Baxter Healthcare Corporation of Puerto Rico,
                  Aibonito, Puerto Rico.
Recalled by:      Baxter Healthcare Corporation, Deerfield, Illinois, by
                  letter October 29, 1991.  Firm-initiated field correction
                  ongoing.
Distribution:     Nationwide, Canada, Colombia, England, Belgium, Spain,
                  The Netherlands, Germany, Mexico.

                                     -10-





Quantity:         67,488 sets were distributed; firm estimates 30 percent
                  or the product remains on the market.
Reason:           The labeling states that the infusion set contains a 80
                  micron filter when it actually contains a 170-260 micron
                  standard blood filter.

                                     ----
Medical Device Safety Alerts:

Product:          In-vitro diagnostic products.  Items (a-f) under the
                  IMMCO label, items (g-i) under Behring Diagnostics label,
                  and item (j) under Whittaker Bioproducts label:
                  (a) Antinuclear Antibody (ANA) Test (HEp-2 cells);
                  (b) Antinuclear Antibody (ANA) Test (Mouse Liver Sections);
                  (c) Anti-Native DNA Antibody Test (Crithidia luciliae);
                  (d) Autoantibody Test System (Mouse Kidney/Stomach
                  Sections);
                  (e) Anti-Skin Antibody Test (Monkey/Guinea Pig
                  Esophagus Sections);
                  (f) Anti-Skin Antibody Test (Monkey Esophagus Sections);
                  (g) AFT Systems I Kit (small);
                  (h) AFT System I Kit (large);
                  (i) AFT System I Reagent Package for Metpath;
                  (j) Anti-human IgG conjugate (fluorescein labeled) 2 liter
                  bulk plasma packs.  Safety Alert #M-009/018-2.
Code:             All lot and catalog numbers.
Manufacturer:     IMMCO Diagnostics, Inc., Buffalo, New York.
Alerted by:       Manufacturer, by letter October 25, 1991.
Distribution:     Nationwide, Germany, Ireland, Austria, Switzerland,
                  Canada, Hong Kong.
Quantity:         79 in-vitro diagnostic kits bearing the IMMCO label;
                  923 AFT System I Kits (small);
                  3,150 AFT System I Kits (large);
                  28 AFT System I Reagent Packages for Metpath;
                  16.2616 liters Anti-human IgG conjugate (fluorescein
                  labeled) in 2 liter bulk plasma packs were distributed;
                  firm estimates 75 percent of the product remains on
                  the market.
Reason:           Product contain fluorescein labeled anti-human IgG
                  conjugate which becomes unstable when stored in direct
                  light for prolonged periods of time.  This may result in
                  autoantibody test results with a reduced staining
                  intensity, reduced titers, or false negative reactions.

                                     ----

Product:          Thora-Seal III (TS-III) underwater chest drainage systems:
                  (a) Thora-Seal III Underwater Chest Drainage System,
                  Product #8884-713308;
                  (b) Thora-Seal III Replacement Collection Bottle,
                  Product #8884-713900;
                  (c) Thora-Seal III Autotransfusion Chest Drainage System,
                  Product #8884-713192;

                                     -11-





                  (d) Thora-Seal III Autotransfusion Chest Drainage Unit,
                  Product #8884-713176.  Safety Alert #M-022/025-2.
Code:             All lots.
Manufacturer:     Sherwood Medical, St. Louis, Missouri.
Alerted by:       Manufacturer, by letter November 13, 1991.
Distribution:     Nationwide, Canada, Argentina, Australia, Hong Kong,
                  UK, France, Germany, Italy, Hollans.
Quantity:         178,000 units were distributed.
Reason:           Both the collection bottle and the suction control bottle
                  in the devices may develop cracks in the area beneath the
                  manifold which may not be visible unless the bottles are
                  removed from the manifold, and which could result in an
                  air leak.

                                     -12-

                                     ###