09/10/1991

Enforcement Report for the Week of September 11, 1991 


                                                September 11, 1991

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

Product:          Chinese Fortune Cookies in cardboard boxes of 50 cookies, 
                  net wt. 1 lb. 8 oz.  Recall #F-522-1. 
Code:             Not coded.  All products manufactured before 8/8/91 are 
                  involved. 
Manufacturer:     Super Foods, Inc., Washington, D.C. (most responsible 
                  firm).
Recalled by:      Manufacturer by telephone August 9, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Virginia, Maryland, Washington, D.C.
Quantity:         200 cases of cookies were manufactured weekly.
Reason:           Product contains undeclared FD&C Yellow No. 5 and No. 6.

                                     ---- 

                                      -1- 





Product:          Holly Tropical Pineapple Soft Drink in 20 oz. glass bottles.
                  Recall #F-523-1.
Code:             All lots. 
Manufacturer:     Holly Beverage Company, Youngstown, Ohio. 
Recalled by:      Manufacturer by letter July 1, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     New York, Pennsylvania, Ohio. 
Quantity:         Firm estimates 50 cases remain on market. 
Reason:           Product contains undeclared FD&C Yellow No. 5.

                                     ---- 

Product:          Contadina Seasoned Bread Crumbs packaged in a 6-3/4 inch
                  tall, 3 inch diameter cardboard container.  Recall #F-524.
Code:             Lot number LD013. 
Manufacturer:     Quality Bakery, Detroit, Michigan.
Recalled by:      Carnation Company, Glendale, California by letter August
                  30, 1990.  Firm-initiated recall complete.
Distribution:     Arizona, California, Nevada, New Mexico.
Quantity:         827 cases were distributed.  Firm estimates none remains
                  on market.
Reason:           Product contains glass fragments. 

                                     ---- 

Product:          Orange flavored frozen dessert on a stick.  Recall #F-525-1.
Code:             Not coded.
Manufacturer:     Jocel Manufacturing Corporation, Cayey, Puerto Rico.
Recalled by:      Manufacturer by telephone August 19, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Puerto Rico.
Quantity:         84 cartons (24 cans per case) were distributed; firm
                  estimates none remains on the market. 
Reason:           Product contains undeclared FD&C Yellow No. 5.


                                     ---- 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


                                     NONE 

                                   COSMETICS

                                     NONE 

                                      -2- 





                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 
                  Class II -

Product:          Methyltestosterone 10 mg in 100 tablet bottles, an oral Rx
                  drug used for replacement therapy in males with 
                  testosterone, under the following labels: Richlyn 
                  Laboratories, Inc.; Schein Pharmaceutical, Inc.; Goldline 
                  Laboratories; Rugby Laboratories, Inc.; Major Pharmaceutical
                  Corporation; United Research Laboratories, Inc. 
                  Recall #D-627-1.
Code:             Lot number 44116 EXP 4/94.
Manufacturer:     Richlyn Laboratories, Inc., Philadelphia, Pennsylvania. 
Recalled by:      Manufacturer by letter August 23, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         11,321 bottles were distributed.
Reason:           A bottle of Methyltestosterone Tablets contained a foreign
                  tablet identified as a Richlyn meclizine 25 mg. tablet. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen Plasma.
                  Recall #B-370/372-1.
Code:             Unit numbers: (a) 8807877; (b) 8808095; (c) 8808095.
Manufacturer:     St. Luke's Hospital, Denver, Colorado.
Recalled by:      Manufacturer by telephone July 3, 1991 and letter 
                  July 9, 1991.  Firm-initiated recall complete.
Distribution:     Colorado. 
Quantity:         Units were distributed and transfused.
Reason:           Blood products, which were initially reactive for the 
                  hepatitis B surface antigen (HBsAg) but were improperly 
                  interpreted as nonreactive, were distributed. 

                                     ---- 

Product:          STS-M Red Cell Diluent, Enzyme, in 250 ml vials.
                  Recall #B-373-1.
Code:             Lot number EM108-1. 
Manufacturer:     Gamma Biologicals, Inc., Houston, Texas.
Recalled by:      Manufacturer, by telephone October 23, 1990 and 
                  November 5, 1990, followed by letter.  Firm-initiated 
                  recall ongoing. 
Distribution:     Oklahoma, Florida, Iowa, Idaho, Nebraska, Kentucky, 
                  South Carolina. 
Quantity:         492 vials were distributed. 

                                      -3- 





Reason:           Product fails to maintain reactivity throughout the 
                  approved dating period. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Recovered Plasma.
                  Recall #B-374/375-1.
Code:             Unit #21G06575. 
Manufacturer:     American Red Cross Pacific Northwest Region, Portland,
                  Oregon. 
Recalled by:      Manufacturer by telephone April 2 and 11, 1991. 
                  Firm-initiated recall complete. 
Distribution:     Oregon, Kentucky. 
Quantity:         1 unit. 
Reason:           Blood products, incorrectly tested for the antibody to the
                  human immunodeficiency virus type 1 (anti-HIV-1), were
                  distributed.

                                     ---- 

Product:          Abbott Laboratories Tray Cover Seal, Comm. #93-2311,
                  used to test blood samples for hepatitis.  Recall #B-378-1. 
Code:             Lot numbers: 47095P100, 47-044P200, 49-068P100, 50-066P100
                  50-215P200, 51-421P100, 52-242P100, 53-285P100. 
Manufacturer:     Green Bay Packing Inc., Green Bay, Wisconsin. 
Recalled by:      Abbott Laboratories Diagnostic Division, Abbott Park, 
                  Illinois by telephone May 28 and 29, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         1,236,000 seals were distributed; firm estimates none 
                  remains on the market.
Reason:           The use of the Abbott standard tray cover seals caused
                  false positive test results when the AUSAB EIA test was 
                  performed using the Commander Dynamic Incubator (CDI) in
                  conjunction with the Commander Parallel Processing Center 
                  (CPPC). 

                                     ---- 

Product:          Recovered Plasma.  Recall #B-379-1. 
Code:             Unit #9431. 
Manufacturer:     Sewickley Valley Hospital Blood Bank, Sewickley,
                  Pennsylvania. 
Recalled by:      Manufacturer by telephone June 25, 1991.  Firm-initiated
                  recall complete.
Distribution:     Pennsylvania. 
Quantity:         1 unit. 
Reason:           Blood product, which tested initially reactive for the
                  hepatitis B surface antigen (HBsAg) and not retested in 
                  duplicate, was distributed. 

                                     ---- 

                                      -4- 





Product:          Red Blood Cells.  Recall #B-380-1.
Code:             Unit numbers: T63518, T69635, T69707. 
Manufacturer:     Tulane University Hospital (Administrators of the Tulane) 
                  Educational Fund, New Orleans, Louisiana. 
Recalled by:      Manufacturer by telephone on or about July 30, 1991.
                  Firm-initiated recall ongoing.
Distribution:     Louisiana.
Quantity:         3 units.
Reason:           Blood product, which tested initially reactive for the
                  hepatitis B surface antigen (HBsAg), was distributed. 

                                     ---- 

Product:          (a) Platelets; (b) Platelets, Pheresis. Recall #B-383/384-1.
Code:             Unit numbers: (a) 484460, 604830, 604831, 604832, 604835, 
                  604843, 611472, 611489, 611492, 611497; (b) 606768. 
Manufacturer:     Belle Bonfils Memorial Blood Center, Denver, Colorado.
Recalled by:      Manufacturer, by telephone May 17, 1991.  Firm-initiated
                  recall complete.
Distribution:     Colorado. 
Quantity:         11 units. 
Reason:           Blood products labeled with incorrect expiration dates
                  were distributed. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Recovered Plasma.
                  Recall #B-385/386-1.
Code:             Unit numbers E94093, K07786.
Manufacturer:     Rhode Island Blood Center, Providence, Rhode Island.
Recalled by:      Manufacturer, by telephone, followed by letter May 10, 1991.
                  Firm-initiated recall complete. 
Distribution:     Rhode Island, New York, North Carolina. 
Quantity:         4 units were distributed. 
Reason:           Blood products which tested non-reactive for the
                  antibody to the human immunodeficiency virus type 1 
                  (anti-HIV-1), but collected from a donor who previously 
                  tested repeatably reactive for anti-HIV-1, were distributed.

                                     ---- 

                  Class III - 

Product:          Liquigel-AMS MLX Type Aluminum Hydroxide/Magnesium
                  Hydroxide Sorbitol-Fluid Gel for manufacturing use only 
                  filled in Net wt. 450 lbs. drums.  Recall #D-625-1. 
Code:             Lot number 1043406. 
Manufacturer:     Reheis, Inc., Berkeley Heights, New Jersey. 
Recalled by:      Manufacturer by letter February 18, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Pennsylvania. 
Quantity:         14 drums were distributed; firm estimates none
                  remains on market.

                                      -5- 





Reason:           Fourteen drums of Liquigel AMS MLX Type were mislabeled as
                  AMS MYL Type. 

                                     ---- 

Product:          Ventolin (Albuterol Sulfate, USP) Solution for inhalation 
                  0.5% w/v, in 20 ml bottles with calibrated stopper, a 
                  multi-use vial containing solution for use with nebulizer.
                  Recall #D-626-1.
Code:             All lots. 
Manufacturer:     Glaxo Holdings, Ltd., Ware, England.
Recalled by:      Glaxo Inc., North Carolina by letter August 22, 1991. 
                  Firm-initiated recall ongoing.
Distribution:     Nationwide. 
Quantity:         2,907,687 bottles were distributed. 
Reason:           Product does not meet pH specifications.

                                     ---- 

Product:          Red Blood Cells.  Recall #B-381-1.
Code:             Unit numbers: 15M02022, 15M02023, 15M02024. 
Manufacturer:     American Red Cross Montana Region, Great Falls, Montana.
Recalled by:      Manufacturer by telephone March 27, 1991 with follow-up 
                  letter April 13, 1991.  Firm-initiated recall complete. 
Distribution:     Montana.
Quantity:         3 units.
Reason:           Blood product labeled with incorrect expiration dates 
                  was distributed.

                                     ---- 

Product:          Platelets.  Recall #B-382-1.
Code:             Unit number 38F89238. 
Manufacturer:     American Red Cross Blood Services, Fort Wayne, Indiana. 
Recalled by:      Manufactuer by telephone March 18, 1991 followed by letter
                  March 25, 1991.  Firm-initiated recall complete.
Distribution:     Indiana.
Quantity:         1 unit. 
Reason:           Blood products, untested for the antibody for the human 
                  immunodeficiency virus type 1 (anti-HIV-1), was distributed.

                                     ---- 

Product:          Red Blood Cells.  Recall #B-387-1.
Code:             Unit number: 25GP01055. 
Manufacturer:     American Red Cross Greater Upstate New York Region, Syracuse
                  New York. 
Recalled by:      Consignee notified recalling firm and returned unit 
                  August 7, 1991.  Firm-initiated recall complete.
Distribution:     New York. 

                                      -6- 





Quantity:         1 unit. 
Reason:           Blood product containing anti-E antibodies, but labeled as
                  negative for unexpected antibodies, was distributed.

                                     ---- 

                       DEVICES AND RADIOLOGICAL PRODUCTS

                  Class I - 
                                     NONE 
                  Class II -

Product:          ECG Casette Recorder, is a portable recorder with two or
                  three channels to provide a continuous diagnostic quality 
                  ECG recording of a patient over a 24 hour time span.
                  Recall #Z-959/961-1.
Code:             Model 1100-6 serial numbers 3454 through 3499 (Quiktrak); 
                  Model 1100-7 serial numbers 1101 through 1103, 1109, and 1120 
                  (Pacetrak); Model 1100-8 serial numbers 1001 through 1071 
                  (Tritrak).
Manufacturer:     Zymed Puerto Rico, Inc., Ponce, Puerto Rico.
Recalled by:      Zymed, Inc., Camarillo, California, by telephone May 17,
                  1990.  Firm-initiated recall complete.
Distribution:     Nationwide. 
Quantity:         123 units were distributed. 
Reason:           Inadequate drive belt tension promotes clutch failure and 
                  eventual slipping of belt from either motor pully or capstan
                  pully.

                                     ---- 

Product:          320 Micron Fiber Assemblies (Part #N 8075-26-0310) used 
                  with Candela MDL-1 and MDL2000 Lasertripter System. 
                  Recall #Z-965-1.
Code:             Lot number SUR-858-S043-26, SUR-858-S043-24.
Manufacturer:     Candela Laser Corporation, Wayland, Massachusetts.
Recalled by:      Manufacturer by letter April 4, 199l.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         50 systems were distributed.
Reason:           The outer polyamide coating of the fibers may come off when 
                  exposed to Cidex or alcohol.

                                     ---- 

Product:          Flow Sensor Cassette with 60" Infusion Set, Catalog No. 
                  810152, Model No. 2004; Infusion Set Cassette/with 60"
                  Infusion Set-Catalog No. 8810295, Model No. 2100; 
                  Programmer-Catalog No. 8810276, Model No. 2000; Non-
                  Universal (PCA) Programmer, Catalog No. 8810206, Model No.
                  2500; Primer Catalog No. 8810168, Model No. 102 
                  Recall #Z-969/973-1.

                                      -7- 





Code:             All serial numbers. 
Manufacturer:     Strato Medical Corporation, Beverly, Massachusetts. 
Recalled by:      Manufacturer by letter July 23, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         33,388 cassettes were distributed.
Reason:           The product may not accurately deliver drug therapy, due
                  to leakage and under infusion.

                                     ---- 

Product:          Lifeport Vascular Access System, Catalog Number LPS-1045. 
                  Recall #Z-974-1.
Code:             All lots. 
Manufacturer:     Strato Medical Corporation, Beverly, Massachusetts. 
Recalled by:      Manufacturer by letter July 17, 1991.  Firm-initiated 
                  recall ongoing. 
Distiribution:    Nationwide and international. 
Quantity:         73 units were distributed.
Reason:           The device was marketed and distributed to unapproved 
                  clinical investigators. 

                                     ---- 

Product:          Collared PERFECTA Stems, (a) Perfecta Stem, Standard, 
                  Collared, Plasma Sprayed, size 12, 30mm Neck Catalog
                  No. 3708-1200; (b) Perfecta Stem, Standard, Collared, 
                  Plasma Sprayed, Size 13.5, 30mm Neck, Catalog No. 
                  3708-1350; (c) Perfecta Stem, Reduced, Collared, Plasma 
                  Sprayed, size 15.0, 35mm Neck, Catalog No. 3716-1500. 
                  Recall #Z-975/977-1.
Code:             Lot numbers: (a) 001501001, (b) 001502001, (c) 001147001. 
Manufacturer:     Orthomet, Inc., Minneapolis, Minnesota. 
Recalled by:      Manufacturer by telephone July 22,23, 31, 1991.  Firm-
                  initiated recall ongoing. 
Distribution:     Michigan, Florida.
Quantity:         6 units were distributed.  Two units of each lot number.
Reason:           The labeling on the outer and interior packaging and
                  the etch on the device does not properly identify the 
                  actual device packaged. 

                                     ---- 

Product:          Inspiron brand Universal Manifold and Nebulizer with
                  39" Aerosol Tube and 2 Supply Lines, an Rx device 
                  containing a mouthpiece 6" length, 22 mm I.D. Flextube
                  and Drive Line Adapters for use with the Bird I.P.P.B.
                  Respirator, product #002208-A; product #002218-A. 
                  Recall #Z-978/979-1.
Code:             Lot numbers: (a) C92071, C92072, C92143, C92160, C92242,
                  C92243, D92212, D92213, D92269, E92078; B92172, B92173, 
                  B92174, B92201, B92254, B92255, B92256, C92068, C92096, 
                  C92098, C92129, C92161, C92162, C92186, C92271, D90059
                  D90060, D90082, D92083, D92084, D92134, E90041. 

                                      -8- 





Manufacturer:     Specialty Packaging Products, Inc., El Paso, Texas. 
Recalled by:      Intertech Resources, Inc., Bannockburn, Illinois by letter
                  July 5, 1989.  Firm initiated recall complete.
Distribution:     Nationwide. 
Quantity:         122,934 units were distributed. 
Reason:           The exhalation valve on the devices manufactured by Specialty 
                  Packaging Products, Inc., in use with certain respirators 
                  may not fully close.

                                     ---- 

Product:          Abbott TestPack Chlamydia, an in-vitro diagnostic enzyme
                  immunoassay for the rapid detection of Chlamydia
                  trachomatis from endocervical swab specimens. 
                  Recall #Z-981-1.
Code:             All lots. 
Manufacturer:     Abbott Laboratories, Inc., Abbott Park, Illinois. 
Recalled by:      Manufacturer by mailgram August 5, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, international.
Quantity:         19,637 kits were distributed. 
Reason:           The product performance may fall below the package insert 
                  sensitivity claims. 

                                     ---- 

Product:          BISNARE PolypectomySnare with Handle, used to remove
                  colon polyps.  Recall #Z-983/984-1. 
Code:             Model numbers 5125, 5128. 
Manufacturer:     Everest Medical, Minneapolis, Minnesota.
Recalled by:      Manufacturer by letter August 6, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, Japan, France, The Netherlands. 
Quantity:         1,904 units were distributed. 
Reason:           The product has the potential for electrical shorting during
                  use.

                                     ---- 

Product:          Beckman Synchron Control Multi-Level Assayed Kit, Part
                  657365.  Recall #Z-982-1. 
Code:             Lot number M91158.
Manufacturer:     Beckman Instruments, Inc., Brea, California.
Recalled by:      Manufacturer by letter November 12, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, Australia, Austria, Taiwan, The Netherlands,
                  Mexico, Hong Kong, Sinapore.
Quantity:         2,850 units were distributed; firm estimates none remains 
                  on the market.

                                      -9- 





Reason:           AST and total Bilirubin analytes in this lot of product 
                  are recovering below the published insert range on the
                  Level 3 assayed control in the kit on some SYNCHRON AS
                  CX4 and CX5 Analyzers.

                                     ---- 

Product:          Misti Gold Inflatable Mammary Prosthesis Kit. 
                  Recall #Z-985-1.
Code:             All lots. 
Manufacturer:     Bioplasty, Inc., St. Paul, Minnesota. 
Recalled by:      Manufacturer by letter August 6, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         4,668 (domestic) and 1,898 (foreign) units were distributed.
Reason:           The product does not have an appropriate marketing
                  authorization and because the product labeling contains 
                  unsubstantiated claims which make it false and misleading.

                                     ---- 

Prdoct:           Universal Adapters, used on ventilator patient circuits,
                  under brand names Neotech Brand, Catalog No N1001 and 
                  N1002 individually and bulk packaged; Acura Brand 
                  Catalog Nos. A-101 and A-102, individually and bulk 
                  packaged.  Recall #Z-986/989-1. 
Code:             Lot numbers beginning with 89 and 90. 
Manufacturer:     Wright Manufacturing, Burbank, California.
Recalled by:      Neotech Products, Inc., Chatsworth, California by letter
                  dated March 12, 1991.  Firm-initiated recall ongoing. 
Distribution:     Nationwide. 
Quantity:         368,750 units were distributed; FDA estimates little
                  product remains on the market.
Reason:           Due to defective mold, the adapter may be partially or
                  fully blocked and may impede the flow of respiratory gases
                  in the air circuit. 

                                     ---- 

Product:          Silicone Gel Filled Mammary Prostheses (a) Gel-Filled 
                  Mammary Prostheses, Catalog Nos. 350-7100 through 
                  30-7600/one style, round sizes (volume): 100 cc through 
                  600cc; Single Lumen Filled with Gel;
                  (b) Combination Gel-Saline/Style 6000 Mammary Prostheses, 
                  Catalog Nos. 360-6175M through 360-6550M/Style: Oval; Sizes 
                  (total volume): 175cc through 550cc; Double Lumen, One Gel
                  Filled, the other for Saline Fill;
                  (c) Combination Gel-Saline/Style 7000 Mammary Prostheses, 
                  Catalog No. 360-7150M through 360-7525M/Style: Round; Sizes 
                  (total volume): 150cc through 525cc; Double Lumen, One Gel
                  Filled the other for Saline Fill, used for implantation 
                  to augment or replace diseased breast tissue. 
                  Recall #Z-990/992-1.
                                     -10- 





Code:             All lots. 
Manufacturer:     Mentor Corporation, Goleta, California. 
Recalled by:      Manufacturer by letter July 11, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         9,255 units were distributed; firm estimates less than
                  1 percent of product remains on the market. 
Reason:           The silicone gel-filled breast protheses were on the market 
                  after July 9, 1991 without a premarket approval application 
                  (PMA), as required by the final rule of April 10, 1991, 
                  pursuant to Section 5.15(b) of the FD&C Act.

                                     ---- 

Product:          Silicone Gel-Gilled Mammary Prostheses of Various Styles: 
                  1. Style 42: Round (single lumen), Mammary Prostheses,
                  120cc to 560cc envelope size, Catalog nos. 25-42121 
                  through 25-42561; 
                  2. Style 100 Round, Intrashiel (low bleed), Biocell 
                  (textured, Posterior leaf valve, Gel-Filled Mammary 
                  Prostheses, 200cc to 360cc sizes, Catalog nos.
                  27-100200 through 27-100360;
                  3. Style 80, Round, Gel-Filled Mammary Prostheses, 80cc 
                  to 560cc sizes, Catalog nos. 25-80080 through 25-80561; 
                  4. Style 81, Teardrop, Gel-Filled Mammary Prostheses, 
                  105cc to 375cc sizes, Catalog nos. 25-81105 through 
                  25-81376; 
                  5. Style 82, Oval, Low Profile, Gel-Filled Mammary
                  Prostheses, 130cc to 400cc sizes, Catalog nos.25-82130
                  through 25-82401; 
                  6. Style 85, Submuscular, High Profile, Gel-Filled
                  Mammary Prostheses, 120cc to 800cc sizes, Catalog 
                  nos. 25-85120 through 25-85801; 
                  7. Style 56, Round, Double Lumen, Gel-Saline Filled,
                  High Profile Mammary Prostheses, 200/220cc to 700/800cc 
                  sizes, Catalog nos. 25-56201 through 25-56701.; 
                  8. Style 57, Oval, Double Lumen, Gel-Saline Filled, High
                  Profile, Mammary Prostheses, 210/230cc to 600/650cc 
                  sizes, Catalog nos. 25-57211 through 25-57601;
                  9. Style 73: Round, Intrashiel (low bleed), Double Lumen
                  RTV Gel-Saline Filled, Mammary Prostheses, 160cc to 500cc 
                  Sizes, Catalog nos. 25-73160 through 25-73501;
                  10. Style 76: Round, Double Lumen, Gel-Saline Filled
                  Mammary Prostheses, 100/120cc to 500/560cc Sizes, Catalog 
                  nos.25-76100 through 25-76501;
                  11. Style 77: Teardrop, Double Lumen, Gel-Saline Filled 
                  Mammary Prostheses, 105/125cc to 345/375cc Sizes, Catalog 
                  nos. 25-77105 through 25-77346; 
                  12. Style 78: Oval, Low Profile, Double Lumen, Gel-Saline 
                  Filled Mammary Prostheses, 130/150cc to 350/400cc Sizes,
                  Catalog nos 25-78130 through 25-78351, Recall #Z-993/1004-1.

                                     -11- 





Code:             All lots. 
Manufacturer:     McGhan Medical Corporation, Santa Barbara, California.
Recalled by:      Manufacturer by letter July 10, 11 and 12, 1991 and by
                  visit on or about August 5, 1991.  Firm-initiated recall
                  ongoing.
Distribution:     Nationwide and international. 
Quantity:         133,388 units were distributed.  Firm estimates 6,822 
                  remain on the market. 
Reason:           The silicone gel-filled breast protheses were on the
                  market after July 9, 1991 without a premarket approval
                  application (PMA), as required by the final rule of 
                  April 10, 1991, pursuant to Section 5.15 (b) of the 
                  FD&C Act. 

                                     ---- 

                  Class III-

Product:          Sagittal Surgical Saw Blade, Angled 5.5 x 18.0 mm.
                  Recall #Z-980-1 
Code:             Lot number "NOV 90".
Manufacturer:     Hall Surgical Division of Zimmer Inc., Carpinteria, 
                  California. 
Recalled by:      Manufacturer by letter February 18 and March 19, 1991.
                  Firm-initiated recall ongoing.
Distribution:     Nationwide, Canada, Belgium.
Quantity:         41 blades were distributed; firm estimates 15 remain
                  on the market.
Reason:           The package labels the product as 14 x 15.5 mm and the
                  actual dimensions are 5.5 x 18.0 mm.
                                                       ---- 

                              VETERINARY PRODUCTS 

                  Class I - 
                                     NONE 
                  Class II -

Product:          Sulfa Urea Solution in 1 gallon containers, topical liquid, 
                  an OTC for treating deep and superficial wounds.
                  Recall #V-078-1.
Code:             900148, 900086. 
Manufacturer:     Chesterfield Laboratories, Inc., Riverside, Missouri. 
Recalled by:      Manufacturer by letter March 28, and April 15, 1991.
                  Firm initiated recall ongoing.
Distribution:     Ohio, Arizona, Georgia, California, Alabama, Pennsylvania,
                  Texas.
Quantity:         204 1-gallon containers were distributed. 
Reason:           Subpotency. 
                                     ---- 

                                     -12- 

                                     ####