FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
04/10/1990    

FDA Enforcement for the week of April 11, 1990



                            FDA ENFORCEMENT REPORT
                              FOR APRIL 11, 1990
                                                           April 11, 1990 
Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.


Product:          All L-tryptophan products not covered by 21 CFR 172.320 
                  or by a current IND, from all levels of distribution. 
                  Recall #F-456-0.  An extension of Recall #F-140-0 which 
                  appeared in the January 17, 1990 Enforcement Report.
Code:             All lots. 
Manufacturer:     All manufacturers, distributors and repackers.
Recalled by:      On March 22, 1990, the Food and Drug Administration issued
                  a press release (P90-21) and a video release which announced
                  that the Agency is requesting manufacturers, repackers, or
                  own-label distributors to recall all manufactured 
                  L-tryptophan containing products in tablet, capsule,
                  caplet, powdered or liquid form, along with multi-
                  ingredient, non-protein supplements that also contain 
                  L-tryptophan. 
                                      -1- 





                  distribution.  FDA requested recall ongoing.
Distribution:     Nationwide. 
Quantity:         Unknown large quantity. 
Reason:           Eosinophilia-myalgia syndrome attributed to L-tryptophan
                  containing products.

                                     ---- 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 


Product:          Noah's Cheezettes packed in 1 ounce and 7 ounce poly bags.
                  Recall #F-457-0.
Code:             The firm uses a 21 day freshness expiration date.  Any
                  package that does not declare the presence of FD&C Yellow 
                  #5 and #6 is under recall.
Manufacturer:     Noah's Potato Chip Company, Inc., Alexandria, Louisiana.
Recalled by:      Manufacturer, by letter November 30, 1989.  Firm-initiated
                  field correction complete.
Distribution:     Louisiana.
Quantity:         Firm estimates none remains on market.
Reason:           Product lacks a color declaration and contains undeclared 
                  Yellow No. 5. 
                                     ---- 

                  Class III - A situation in which the use of, or exposure to 
                  a violative product is not likely to cause adverse health 
                  consequences. 


Product:          Nature's Sunshine Products brand Tofu Moo Carob, a
                  non-dairy dried tofu beverage mix, in 27.6 ounce cans.
                  Recall #F-458-0.
Code:             Lot numbers 9100E, 9100F. 
Manufacturer:     Great American Foods, Inc., Salt Lake City, Utah. 
Recalled by:      Nature's Sunshine Products, Inc., Spanish Fork, Utah, 
                  by telephone February 14, 1990 followed by letter February
                  16, 1990.  Firm-initiated recall ongoing. 
Distribution:     Nationwide. 
Quantity:         420 cans were distributed.
Reason:           Product is contaminated with small round metal particles. 

                                     ---- 

                                   COSMETICS

                                     NONE 

                                      -2- 





                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 

                  Class II -

Product:          Source Plasma.  Recall #B-122-0.
Code:             Unit numbers: 9810, 9914, 10072, 10218, 10467, 10597, 
                  10827, 10945, 11152, 11301, 11534, 11728, 11898, 12151, 
                  12427, 12557. 
Manufacturer:     Yale Blood Plasma, Inc., Albuquerque, New Mexico. 
Recalled by:      Manufacturer, by letter March 29, 1989.  Firm-initiated 
                  recall ongoing. 
Distribution:     New York, Japan.
Quantity:         16 units were distributed.
Reason:           Units of Source Plasma drawn from a donor who previously
                  tested repeatably reactive for the hepatitis B surface
                  antigen (HBsAg) were distributed. 

                                     ---- 

                  Class III - 

Product:          Recovered Plasma.  Recall #B-117-0. 
Code:             Unit #06GF51181.
Manufacturer:     American Red Cross Blood Services, Los Angeles, California. 
Recalled by:      Manufacturer, by telephone June 26, 1989.  Firm-initiated 
                  recall complete.
Distribution:     California. 
Quantity:         1 unit was distributed and destroyed. 
Reason:           Recovered Plasma, which tested nonreactive for the
                  Hepatitis B Surface Antigen (HBsAg) but was collected from
                  a donor who previously tested reactive for HBsAg, was 
                  distributed.
                                     ---- 

                   MEDICAL DEVICES AND RADIOLOGICAL PRODUCTS

                  Class I - 
                                     NONE 

                  Class II -

Product:          Datascope Intra-Aortic Balloon Pump (IABP)- System 90,
                  the pump, a critical device is coupled with an intra-aortic 
                  balloon catheter which is inserted into the patient's aorta.
                  Recall #Z-138-0.
Code:             Product #0998-00-0010-XX, serial numbers range from 1001
                  through 1614, inclusive; Product #0998-00-0058-XX, serial 
                  numbers range from 2000 through 3181, inclusive.

                                      -3- 





Manufacturer:     Datascope Corporation, Paramus, New Jersey. 
Recalled by:      Manufacturer, by letter February 7, 1990.  Firm-initiated 
                  field correction ongoing. 
Distribution:     Nationwide and international. 
Quantity:         Approximately 2,200 units were distributed. 
Reason:           The System 90 (IABP) power supply board may fail. 

                                     ---- 

Product:          Biosearch Pedi Tube Pediatric Nasal Gastric Feeding Tube. 
                  Recall #Z-458-0.
Code:             Product #14-7318, batch lot #8821, G-80.
Manufacturer:     Biosearch Medical Products, Inc., Somerville, New Jersey. 
Recalled by:      Manufacturer, by telephone between August 25, 1989 and
                  August 29, 1989.  Firm-initiated recall ongoing.
Distribution:     California, Maine, Wisconsin, New Jersey, Florida,
                  Pennsylvania, Michigan, Massachusetts, South Carolina,
                  Canada. 
Quantity:         700 units consisting of 10 tubes each were distributed. 
Reason:           There is a potential for a leak to develop at the bond
                  between the luer connector and the feeding tube which 
                  could weaken the connection and allow the luer to detach. 

                                     ---- 

Product:          Hygroscopic Cervical Dilator - Dilapan (tm) 
                  4 mm x 55 mm, a sterile Class III medical device
                  indicated for dilation of the cervix prior to termination 
                  of pregnancy up to 16 weeks gestation.  Recall #Z-460-0.
Code:             Lot numbers:  D070389, D092989. 
Manufacturer:     Gynotech, Inc., Middlesex, New Jersey.
Recalled by:      Manufacturer, by telex December 8, 1989 and by telephone
                  December 18, 1989.  Firm-initiated recall ongoing.
Distribution:     Minnesota, Texas, California, Ohio, Japan.
Quantity:         1,725 dilators were distributed.
Reason:           The device is labeled as sterile, however, firm failed
                  to satisfactorily document that lots D070389 and D092989
                  had in fact been sterilized.

                                     ---- 

Product:          Instascreen CHAP Kit, an in-vitro diagnostic test system
                  designed to detect the presence of cocaine, heroin/morphine,
                  amphetamines, PCP and their metabolites in urine. 
                  Recall #Z-461-0.
Code:             Lot numbers: 61-04049, 65-04279, 65-06129, 65-05159,
                  65-08189, 65-09089, 61-10070. 
Manufacturer:     Drug Screening Systems, Inc., Blackwood, New Jersey.
Recalled by:      Manufacturer, by letter on or about October 26, 1989. 
                  Firm-initiated recall ongoing.
Distribution:     Nationwide, Canada, The Phillipines.
Quantity:         5,402 kits were distributed.

                                      -4- 





Reason:           The kits exhibit variability in product performance over
                  time.  As a consequence the use of this product may give a
                  false negative. 

                                      ----

Product:          Quad Ferm +, for in-vitro diagnostic use, for the rapid 
                  determination carbohydrate utilization, DNase activity
                  and Penicillinase activity by organisms belonging to the
                  genera Neisseria and Branhamella.  Recall #Z-463-0. 
Code:             Product #8886-301020, Lot #1807F9.
Manufacturer:     API Analytab Products, Plainview, New York. 
Recalled by:      Manufacturer, by letter August 30, 1989.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, France, Italy.
Quantity:         456 kits were distributed.
Reason:           The substrates DNase and Beta-Lactamase were filled in
                  reverse order such that DNase was filled in the Beta- 
                  Lactamase well and vice-versa.

                                      ----

                  Class III - 

Product:          Amerlite TBG (Thyroid Binding Globulin) EIA Kit, an 
                  in-vitro diagnostic for the quantitative measurement of 
                  thyroxine-binding globulin (TBG) in human serum or
                  plasma.  Recall #Z-457-0. 
Code:             Product codes: LAN.0005 and LAN.2005, lot numbers 30 and 32 
                  EXP 7/23/90.
Manufacturer:     Amersham International plc, Cardiff, Wales, UK. 
Recalled by:      Amersham Corporation, Arlington Heights, Illinois, by 
                  telephone March 5, 1990 followed by 
                  visit.  Firm-initiated recall ongoing.
Distribution:     Massachusetts, Connecticut, Maryland, Florida, Tennessee, 
                  California. 
Quantity:         26 kits were distributed; firm estimates 14 kits remain 
                  on market.
Reason:           Some of the "D" Standard 20 (mg/l) vials were mislabeled
                  with "F" Standard (80 mg/l) labels. 

                                      ----

Product:          Instascreen Cannabinoid Kit, an in-vitro diagnostic test
                  system designed to detect the presence of marijuana,
                  hashish, THC and their derivatives in urine.
                  Recall #Z-459-0.
Code:             Lot numbers: 51-05159, 55-05049, 55-05029, 55-05039,
                  51-01279, 51-01069, 55-02009, 51-01199, 51-01269, 
                  51-02169, 51-03019, 51-04209, 51-03309, 51-04269. 
Manufacturer:     Drug Screening Systems, Inc., Blackwood, New Jersey.
Recalled by:      Manufacturer, by telephone July 1989 followed by letter.
                  Firm-initiated recall ongoing.

                                      -5- 





Distribution:     Nationwide, Canada, Phillipines.
Quantity:         Approximately 5,416 kits were distributed.
Reason:           The device may exhibit variability in product performance 
                  over time.  As a consequence the product may give a false 
                  positive test result. 
                                      ----

                              VETERINARY PRODUCTS 

                  Class I - 
                                      NONE

                  Class II -

Product:          Chloramphenicol Capsules, 250 mg. each, packaged 8
                  capsules in foam cut-out insert inside a clear plastic box, 
                  an OTC drug for aquarium use only, labeled to control and 
                  cure bacterial infections in marine fish.  Recall #V-037-0. 
Code:             Lot numbers 673 through 681 (all lots). 
Manufacturer:     Barr Laboratories, Northvale, New Jersey. 
Recalled by:      Silco Pet Products, Inc., Alexandria, Virginia, by letters
                  the week of March 13, 1990.  Firm-initiated recall ongoing. 
Distribution:     Washington, D.C., Maryland, Virginia, New York, Florida.
Quantity:         Firm estimates less than 1,000/8 capsule packs remain on
                  market. 
Reason:           Chloramphenicol for OTC aquarium use is an unapproved 
                  new animal drug.

                                      ----

                  Class III - 
                                      NONE

Seizure Actions Filed:

Product:          Frozen Shrimp (90-510-972). 
Charge:           Misbranded - Product is short weight. 
Firm:             Macau Frozen Food Company, Ltd., Pereira Marquez, Macau.
Filed:            February 7, 1990 - U.S. District Court for the District 
                  of Puerto Rico; Civil #90-1186(JAF), 65806. 

                                      ----

Product:          Ventilation Tubes for Otology (89-457-548). 
Charge:           Adulterated - Products were prepared, packed and held 
                  under conditions not in conformity with good manufacturing
                  practice regulations; Misbranded - Labeling fails to bear 
                  adequate directions for use.
Firm:             Anthony Products, Inc., Indianapolis, Indiana.
Filed:            February 15, 1990 - U.S. District Court for the Southern
                  District of Indiana; Civil #1P90149C, FDC #65781. 

                                      ----

                                      -6- 





Product:          Neomycin Sulfate Sterile Powder (90-601-545). 
Charge:           Misbranded antibiotic drug -- No certificate or 
                  release has been issued.
Firm:             Pharma-Tek Inc., Northport, New York. 
Filed:            February 26, 1990 - U.S. District Court for the Eastern 
                  District of New York; Civil #CV-90-0655, FDC #65783.

                                      ----

Product:          Spices (90-600-019).
Charge:           Products were held under insanitary conditions
                  whereby they may have become contaminated with filth. 
Firm:             Schiff Food Products Company, Inc., Brooklyn, New York. 
Filed:            March 2, 1990 - U.S. District Court for the Eastern 
                  District of New York; Civil #CV 90-0735, FDC #65834.

                                      ----

Product:          All foods in rodent and insect susceptible containers.
                  (90-577-515). 
Charge:           Adulterated - Products were held under insanitary conditions. 
Firm:             Economy, Inc., Chicago, Illinois. 
Filed:            March 9, 1990 - U.S. District Court for the Northern
                  District of Illinois; Civil #90C-1375; FDC #65840.

                                      ----

Product:          Personal Size Oxygen and Personal Oxygen Systems
                  (89-545-562). 
Charge:           Products are unapproved new drugs; fails to bear prescription 
                  legend; labeling fails to bear adequate directions for use; 
                  manufactured at non-registered/non-listed establishment.
Firm:             Pro-Line International, Inc., Rochester, New York.
Filed:            March 14, 1990 - U.S. District Court for the Western
                  District of New York; Civil #90-274-T, FDC #65684.