06/19/1990     

FDA Enforcement for the week of June 20, 1990 



                            FDA ENFORCEMENT REPORT
                               FOR JUNE 20, 1990

                                                          June 20, 1990 

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                  Class III - A situation in which the use of, or exposure to 
                  a violative product is not likely to cause adverse health 
                  consequences. 


                                     NONE 

                                      -1- 






                                   COSMETICS

                                     NONE 

                           HUMNA DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 
                  Class II -

Product:          Muromonab-CD3 Orthoclone OKT3 Sterile Solution, 
                  5 mg (1 mg/1ml), for IV use only.  Recall #B-149-0. 
Code:             Lot numbers: 19B007 EXP 11/89, 19D014 EXP 12/89,
                  68K111 EXP 7/89, 68L116 EXP 8/89, 68N129 EXP 9/89.
Manufacturer:     Ortho Pharmaceutical Corporation, Biotech Division, 
                  Raritan, New Jersey.
Recalled by:      Manufacturer, by telephone between August 7, 1989 and 
                  December 18, 1989.  Firm-initiated recall complete. 
Distribution:     Nationwide, Japan, Mexico, Switzerland. 
Quantity:         Firm estimates none remains on market.
Reason:           Product failed to maintain stability throughout the 
                  approved dating period. 

                                     ---- 

                  Class III - 
                                     NONE 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 

                                     NONE 

                  Class II -

Product:          Exactech Blood Glucose Testing System, consisting of
                  Blood Glucose Sensor and Blood Glucose test strips: 
                  (a) Catalog #97100, Contains Blood Glucose Sensor;
                  Disposable Lancets, Lancing Device, Travel Case,
                  User's Manual;
                  (b) Catalog #97134, Contains Blood Glucose Sensor,
                  Blood Glucose Test Strips, Disposable Lancets,
                  Lancing Device, Travel Case, User's Manual; 
                  (c) Catalog #97131, Exactech Hospital Pack -
                  Contains Blood Glucose Sensor, User's Manual. 
                  Recall #Z-493/495-0.
Code:             Lot numbers 1-198 and H-079.
Manufacturer:     Medisense, Inc., Cambridge, Massachusetts 
Recalled by:      Manufacturer, by letter on or about January 24, 1990
                  and June 6, 1990.  Firm-initiated field correction
                  ongoing.
Distribution:     Nationwide and international. 

                                      -2- 





Quantity:         41,536 systems were distributed.
Reason:           The user's manual did not explain that the "ERR" message
                  may indicate that the user's blood glucose level may have 
                  exceeded the electronic capability of the sensor. 

                                     ---- 

Product:          Hawkins Breast Localization Needles:
                  (a) Catalog No. 253125, Hawkins Breast Localization 
                  Needle, Version III, needle length 12.5 cm, sterile;
                  (b) Catalog No. 253075, Hawkins Breast Localization 
                  Needle, Version III, needle length 7.5 cm sterile.
                  Recall #Z-650/651-0.
Code:             Lot numbers: (a) 00521082; (b) 92280182.
Manufacturer:     National-Standard Medical Products Company, 
                  Gainesville, Florida. 
Recalled by:      Manufacturer, by telephone May 7, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     (a) Tennessee; (b) Arkansas.
Quantity:         20 units were distributed.
Reason:           Opened packages with defective pouch seals which may
                  compromise sterility. 

                                     ---- 

Product:          Fernoplex Ambulance Cot, Model 28, used for transporting
                  patients and has a load capacity of 4100 pounds.
                  Recall #Z-653-0.
Code:             Lot Numbers:        Serial Numbers
                   M472460            L53573, L53575, L53576, L53577, L53578, 
                                      L53580, L53581, L53582, L53583; 
                   M501730            L54371 through L54381;
                   M526340            L55740 through L55747;
                   M549030            L56956 through L56963;
                   M565260            L57440, L57441, L57443 through L57450;
                   M580340            L59076 through L59087.
Manufacturer:     Ferno-Washington, Inc., Wilmington, Ohio. 
Recalled by:      Manufacturer, by letter March 7, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         58 cots were distributed. 
Reason:           A design change was made, consisting of placing a rolling 
                  cam on the shaft which serves as the bearing for the
                  bracket.  However, the first 66 cots manufactured after the 
                  design modification were assembled without the rolling cam. 

                                     ---- 

Product:          Ohmeda check valve disc retainer: 
                  (a) Ohmeda GMS Absorber, Double Canister Model, 
                  Stock #309-0170-800;
                  (b) Ohmeda GMS Absorber, Single Canister Model, 
                  Stock #309-0171-800;

                                      -3- 





                  (c) Ohmeda GMS Absorber, Double Canister ISO Model, 
                  Stock #1400-8002-000; 
                  (d) Ohmeda Check Valve Disc Retainers, Stock #0219-1338-300.
                  Recall #Z-657/660-0.
Code:             All Ohmeda GMS Abosbers that have a ALPHA serial number of
                  "ACL------" (all serial numbers contain a total of nine 
                  digits).
Manufacturer:     Ohmeda, Madison, Wisconsin. 
Recalled by:      Manufacturer, by letter June 1, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         13,990 units were distributed.
Reason:           The inhalation and/or exhalation check valve retainers
                  may become dislodged on the Ohmeda GMS Absorbers allowing 
                  patient rebreathing to occur. 

                                     ---- 

Product:          Stryker Fold Down Siderails, catalog #399-31 and #940-32
                  which may have been used on the following Stryker beds and
                  stretchers:  Model 911 Transcare, Model 915 Transcare,
                  Model 916 Gurney, Model 917 Gurney; Model 920/921 Insta Care; 
                  Model 930/931/932/933 Consta Care; Model 935/936/937 Surgi
                  Care; Model 940 Surgibed; Model 942 LTD Series; Model 960 
                  Surgibed; Model 962 Surgibed; Model 974/976 VIP Table.
                  Recall #Z-661/662-0.
Code:             All fold-down siderails manufactured prior to 
                  January 1985. 
Manufacturer:     Stryker Corporation, Medical Division, Kalamazoo, Michigan. 
Recalled by:      Manufacturer, by maintenance alert May 1988.  Firm- 
                  initiated field correction complete.
Distribution:     Nationwide. 
Quantity:         Firm estimates as many as 2,500 units remain on market. 
Reason:           The siderails could collapse completely and cause injury
                  to patients or hospital staff.

                                     ---- 

Product:          Beacon Laboratories B-100 Argon Coagulator Disposable 
                  Footcontrol Pencil with 10 ft. Cord, a single-use, sterile, 
                  electrosurgical prescription device, used in the operating
                  room during patient surgery.  Recall #Z-666-0.
Code:             Catalog #4010, Stock #03002-001, Lot #935201. 
Manufacturer:     Beacon Laboratories, Inc., Broomfield, Colorado.
Recalled by:      Manufacturer, by letter February 12, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Michigan, New Jersey, Alabama, Nebraska, Colorado.
Quantity:         230 units were distributed; firm estimates 44 units 
                  remain on market. 

                                      -4- 





Reason:           Some of the package seals were found to be wrinkled 
                  which may compromise the sterility of the product.
                  In addition two units were noted to be missing electrical 
                  parts (banana plugs). 

                                     ---- 

Product:          Reusable Single Patient Self-Adhesive Electrodes with 
                  Snap Connectors for neuromuscular stimulation, an Rx
                  device designed to be connected to lead wires of a
                  battery powered transcutaneous electrical nerve stimulator, 
                  which disperses electrical charges from the instrument. 
                  distributed under the Versa-Stim and Eas-E-Stim labels. 
                  Recall #Z-669-0.
Code:             All lot numbers.
Manufacturer:     Conmed Corporation, Utica, New York.
Recalled by:      Manufacturer, by telephone May 16, 1990 followed by 
                  letter May 17, 1990.  Firm-initiated recall ongoing.
Distribution:     Illinois, Indiana, Kansas, Missouri, Florida, Texas,
                  Arizona, Oregon, California, Pennsylvania, New Hampshire, 
                  North Carolina, The Netherlands.
Quantity:         1,289 cases (4 electrodes per envelope, 10 envelopes
                  per box, 2 boxes per case) were distributed; firm 
                  estimates 50 percent remains on market. 
Reason:           Labeling did not warn users that the device is intended 
                  for use with battery powered T.E.N.S. devices only. 

                                     ---- 

CORRECTION:       Software Revision LVP36F-Intelligent (infusion) Pump
                  Model No. 521, Recall #Z-648-0 which appeared in the
                  June 13, 1990 Enforcement Report should read: 

                  Quantity:  2,027 pumps with software version LVP36F 
                             were distributed.

                                     ---- 
                  Class III - 

Product:          Marquest Medical Products Arterial Blood Gas Kit, used
                  for pH and blood gas analysis, for in-vitro diagnostic
                  use (for arterial puncture).  Recall #Z-652-0.
Code:             Lot 31554 EXP 1/92. 
Manufacturer:     Marquest Medical Products, Inc., Englewood, Colorado. 
Recalled by:      Manufacturer, by telephone April 17, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Ohio, Colorado, California, Virginia. 
Quantity:         100 cases (10 per case) were distributed; firm estimates
                  20 cases remain on market.
Reason:           The tip and plunger rod of the plastic barrel syringe 
                  may separate. 
                                     ---- 

                                      -5- 





Product:          Centocor CA 15-3 RIA Radioimmunoassay, an in-vitro
                  diagnostic kit used for the measurement of cancer antigen 
                  15-3 in serum plasma.  Recall #Z-654-0. 
Code:             8K0609 JAN 1, 90; 8K0629 JAN 15, 90; 8K0659 FEB 6, 90;
                  8K0699 FEB 13, 90; 8K0020 MAR 10, 90; 8K0010 MAR 13, 90;
                  8K0040 MAR 21, 90; 8K0070 MAR 23, 90. 
Manufacturer:     Centocor, Inc., Malvern, Pennsylvania.
Recalled by:      Manufacturer, by letter dated May 25, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, Canada, Japan, England, The Netherlands.
Quantity:         1,635 kits were distributed.
Reason:           Product standards are out of specifications.  Degradation 
                  resulted in patient serum levels to be biased by
                  approximately 11 percent to 14 percent higher than the
                  actual value. 

                                     ---- 

Product:          High Flow 50 Oxygen System, Product #406333, a portable 
                  liquid oxygen device for providing a controlled rate of 
                  oxygen flow to a patient.  Recall #Z-656-0. 
Code:             All serial numbers. 
Manufacturer:     Cryogenic Associates, Medical Division of Minnesota,
                  Valley Engineering, Inc., New Prague, Minnesota.
Recalled by:      Manufacturer, by telephone May 7, 1990 followed by
                  letter May 21, 1990.  Firm-initiated recall ongoing.
Distribution:     Colorado, Georgia, Maryland, Minnesota, Ohio, 
                  Pennsylvania, Canada, France. 
Quantity:         101 units were distributed. 
Reason:           The oxygen flow rate may be as much as 20 percent less
                  than the flow rate setting during the first half hour of use. 

                                     ---- 

Product:          Vacutainer brand Tube containing 0.1 molar buffered sodium
                  citrate, a blood collection tube.  Recall #Z-667-0. 
Code:             Lot numbers: 9L019, Reorder #6418 and 9L020 Reorder #6419 
                  EXP 6/91. 
Manufacturer:     Becton Dickinson, Broken Bow, Nebraska. 
Recalled by:      Becton Dickinson Vacutainer Systems, East Rutherford, 
                  New Jersey, by telephone between March 7 and 9, 1990
                  followed by letter letters mailed between March 8 and 12, 
                  1990.  Firm-initiated recall ongoing. 
Distribution:     Nationwide, Canada, England.
Quantity:         302,000 tubes of lot 9L019 and 361,000 tubes of lot 9L020.
                  were distributed; firm estimates 3,808 tubes of lot 9L019 
                  and 12,765 tubes of lot 9L020 remain on market. 
Reason:           Some of the tubes are contaminated with Tripotassium EDTA 
                  (K3EDTA), which is another anticoagulant instead of the 
                  intended anticoagulant, sodium citrate. 

                                     ---- 

                                      -6- 





                              VETERINARY PRODUCTS 

                                      NONE