FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
08/21/1990   

FDA Enforcement for the week of August 22, 1990.



                            FDA ENFORCEMENT REPORT

                                                           August 22, 1990
Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                                      -1- 






                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


Product:          Various Mexican food products:
                  (a) Fresh Bean Dip, 10 ounces;
                  (b) Beef Burrito Mix, 8 ounces; 
                  (c) Mexican Gravy, 8 ounces and in one gallon containers. 
                  Recall #F-596/598-0.
Code:             All lots. 
Manufacturer:     Mexican Village Wholesale Foods, Inc., Grand Forks, 
                  North Dakota. 
Recalled by:      Manufacturer.  Firm's products in stores were labeled 
                  with the supplemental "Keep Refrigerated" label stickers
                  starting on March 30, 1990.  Firm-initiated field correction
                  complete. 
Distribution:     North Dakota, Minnesota.
Quantity:         (a) 180 retail containers; (b) 80 containers; (c) 120 
                  containers were distributed.
Reason:           Retail containers require refrigeration but are not labeled 
                  "Keep Refrigerated."

                                     ---- 

                           HUMAN DRUGS AND BIOLOGICS

                                     NONE 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 
                                     NONE 

                  Class II -

Product:          Model LS 1001 LASTAC Cardiovascular Laser System, used with 
                  2CM830 LASTAC Control Monitors for cardiac monitoring.
                  Recall #Z-745-0.
Code:             Serial numbers: JM000148, JM000135, JM000137. 
Manufacturer.     GV Medical, Inc., Minneapolis, Minnesota. 
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan July 24, 1990.  Firm-initiated field
                  correction ongoing. 
Distribution:     Georgia, Louisiana, Alabama.
Quantity:         3 units were distributed. 
Reason:           Noncompliance with performance standard for laser 
                  products in that maintenance procedures at the optical
                  coupler could cause access to Class IV laser radiation and
                  no interlocks were provided to preclude such access.

                                     ---- 

                                      -2- 





Product:          Laserflo Blood Perfusion Monitors used for continuous 
                  monitoring of local microvascular blood flow: 
                  (a) Laserflo Blood Perfusion Monitor, Model BPM403; 
                  (b) Laserflo Blood Perfusion Monitor, Model BPM403A.
                  Recall #Z-785/786-0.
Code:             Serial numbers: (a) 1101 through 1175;
                  (b) 2000 through 2267.
Manufacturer:     TSI, Inc./Vasamedics, Inc., St. Paul, Minnesota.
Recalled by:      Manufacturer, by letter January 19, 1990.  Firm-initiated 
                  field correction complete.
Distribution:     Nationwide and international. 
Quantity:         (a) 75 units; (b) 268 units were distributed. 
Reason:           A grounding fault may occur in the chassis of the monitor 
                  which may result in continuous or intermittent errors in
                  blood flow measurements.

                                     ---- 

Product:          Providone Iodine Cleansing Gel, used to cleanse continuous
                  ambulatory peritoneal dialysis (CAPD) exchange sets.
                  Recall #Z-778-0.
Code:             Lot #K2-10 EXP 5/10/90. 
Manufacturer:     Huntington Laboratories, Inc., Huntington, Indiana. 
Recalled by:      Kendall McGaw Laboratories, Inc., Irvine, California, by
                  letter February 8, 1990.  Firm-initiated recall complete. 
Distribution:     Nationwide, Canada. 
Quantity:         4,095 cases (40 tubes/case) were distributed; firm
                  estimates none remains on market. 
Reason:           Product contains 0.77% available iodine instead of
                  1.00% as stated on the label. 

                                     ---- 

Product:          Schneider Open-Ended Guidewires for negotiating the 
                  vascular system and allowing fluid delivery to a
                  predetermined site.  Recall #Z-779-0. 
Code:             Lot numbers: 14958, 15034, 15595, 17026, 17027 of 
                  Model 0ESM35145IT7, lot numbers: 14959 and 15035 of 
                  Model 0ESM38145IT7. 
Manufacturer:     Schneider (USA), Inc., formerly Medrad Inc., Pittsburgh,
                  Pennsylvania. 
Recalled by:      Schneider (USA), Inc., Minneapolis, Minnesota, by 
                  letter July 13, 1990.  Firm-initiated recall ongoing. 
Distribution:     Nationwide, Canada, Switzerland.
Quantity:         1,040 units were distributed. 
Reason:           The product's instructions for use state that "during 
                  injection do not exceed the maximum recommended pressure
                  of 500 PSI."  They should read not to exceed 200 PSI. 

                                     ---- 

                                      -3- 





Product:          MR MAX Systems with head holders: 
                  (a) G.E. MR MAX  with Head Holder, Model P9300HE; 
                  (b) G.E. MR MAX Head Holder with Operator's Console,
                  Model 46-282269G1;
                  (c) G.E. MR Max Head Holder with Operator's Console 
                  Model P3901AE.  Recall #Z-780/782-0.
Code:             Serial numbers: (a) 24762YM7 through 24820YM3;
                  (b) 28273YM1, 28285YM5, 28286YM3, 15860YM0 through
                  15868YM3; 
                  (c) 24261YM0, 24278YM4, 24288YM3 through 24303YM0.
Manufacturer:     Yokogawa Medical Systems, Hino-Shi, Tokyo, Japan. 
Recalled by:      General Electric Company, Medical Systems, Waukesha,
                  Wisconsin, by Field Modification Instruction 61054, 
                  April 1990.  Firm-initiated field correction ongoing. 
Distribution:     Nationwide and international. 
Quantity:         89 units were distributed.
Reason:           The microphone is covered with a thin film of rubber
                  which may deteriorate.  A radio frequency burn may
                  occur if the metal housing of the microphone is touched 
                  during the scan.

                                     ---- 

Product:          Fiber-Optic Blood Perfusion Probes, used in conjunction 
                  with a blood perfusion monitor (BPM).  The BPM is used for
                  continuous monitoring of local microvascular blood flow:
                  (a) Model PR-434 Implantable Prism Probe; 
                  (b) Model PD-434 Implantable Disk Probe;
                  (c) Model P-430 Right Angle Probe;
                  (d) Model P-431 Hard Tip Pencil Probe;
                  (e) Model P-433 Needle Probe; 
                  (f) Model P-435 Soft-Tip Pencil Probe;
                  (g) Model PR-436 Endoscopy Probe Side-View; 
                  (h) Model PE-436 Endoscopy Probe End-View.
                  Recall #Z-787/794-0.
Code:             Products manufactured between June 1, 1985 and
                  December 31, 1989 and have sequence numbers between 
                  1,000 and 2,244.
Manufacturer:     TSI, Inc./Vasamedics, Inc., St. Paul, Minnesota.
Recalled by:      Manufacturer, by letters of March 15, 1990, April 16, 1990
                  and May 21, 1990.  Firm-initiated field correction ongoing. 
Distribution:     Nationwide and international. 
Quantity:         903 probes were distributed.
Reason:           The ambient light which is not absorbed by the optical
                  filter may be transmitted through the fiber-optic cable 
                  and into the optical detector.  This provides erroneous 
                  information which could lead to inappropriate diagnosis or
                  therapeutic intervention. 

                                     ---- 

                                      -4- 





CORRECTION:       Recall #Z-761/763-0, Bard Woven Phillips Follower Urethral
                  Catheters which appeared in the August 8, 1990 Enforcement
                  Report is a Class II recall.  The listing for product (c) 
                  should read:

                  (c) Urethral Catheter Gift Set, No. 021519, 19 Fr.

                                     ---- 
                  Class III - 

                                     NONE 

                              VETERINARY PRODUCTS 

                  Class I - 
                                     NONE 

                  Class II -

Product:          Oxytetracycline Hydrochloride (HCl) Injection, in 500 
                  ml vials, a broad spectrum antibiotic for use in beef 
                  cattle, beef calves, non-lactating dairy cattle and dairy 
                  calves for the treatment of a variety of diseases caused
                  by pathogens sensitive to Oxytetracycline HCl, under the
                  Fermenta, IBA and Tech America labels.  Recall #V-053-0.
Code:             Lot #9251B EXP 9/92, 9101A EXP 4/92.
Manufacturer:     Fermenta Animal Health Company, Elwood, Kansas. 
Recalled by:      Manufacturer, by letter June 19, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         12,666 vials were distributed.
Reason:           The lots under recall were not manufactured in accordance 
                  with current good manufacturing practice regulations. 

                                     ---- 

Product:          Sodium Iodide 20% Sterile Injectable Solution, in 250 ml
                  vials, for use as an aid in the treatment of Actinomycosis
                  (lumpy jaw), and Actinobacillosis (wooden tongue, necrotic
                  stomatitis) in cattle, an Rx drug under the Lextron and Tech
                  America labels.  Recall #V-054-0. 
Code:             Lot #9136A EXP 5/92.
Manufacturer:     Fermenta Animal Health Company, Elwood, Kansas. 
Recalled by:      Manufacturer, by letter June 19, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         10,764 vials were distributed.
Reason:           The lot under recall was not manufactured in accordance 
                  with current good manufacturing practice regulations. 

                                     ---- 


                                      -5- 





Product:          50% Dextrose Sterile Injectable Solution, in 500 ml 
                  vials, for use as an aid in the treatment of uncomplicated
                  primary ketosis in cattle, under the Durvet, Lextron, and 
                  Tech America labels.  Recall #V-055-0.
Code:             Lot #9128B EXP 5/92.
Manufacturer:     Fermenta Animal Health Company, Elwood, Kansas. 
Recalled by:      Manufacturer, by letter June 19, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         12,787 vials were distributed.
Reason:           The lot under recall was not manufactured in accordance 
                  with current good manufacturing practice regulations. 

                                     ---- 

Product:          Tylosin Injection, sterile, in 200 mg/ml vials, an
                  antibiotic for use in beef cattle, calves, nonlactating 
                  dairy cattle, and swine, for the treatment of a variety 
                  of diseases, under the Agrilabs, Coop, Butler, Fermenta,
                  RXV and Tech America labels.  Recall #V-056-0.
Code:             Lot #9137A EXP 5/91.
Manufacturer:     Fermenta Animal Health Company, Elwood, Kansas. 
Recalled by:      Manufacturer, by letter June 19, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         7,199 vials were distributed. 
Reason:           The lot under recall was not manufactured in accordance 
                  with current good manufacturing practice regulations. 

                                     ---- 

Product:          Vitamin A-D Injectable Sterile Solution, in 100 ml and
                  250 ml vials for use as a supplemental source of vitamins 
                  A and D3 in cattle, 100 ml, under the Agrilabs, Butler, 
                  Durvet, and Tech America labels, 250 ml under the 
                  Durvet, Lextron, RXV, and Tech America labels.
                  Recall #V-057-0.
Code:             Lot #9135A EXP 5/92.
Manufacturer:     Fermenta Animal Health Company, Elwood, Kansas. 
Recalled by:      Manufacturer, by letter June 19, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         9,935/100 ml vials and 12,619/250 ml vials were 
                  distributed.
Reason:           The lot under recall was not manufactured in accordance 
                  with current good manufacturing practice regulations. 

                                     ---- 

                  Class III - 
                                     NONE