FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
10/23/1990  

FDA Enforcement for the week of October 24, 1990



                            FDA ENFORCEMENT REPORT
                             FOR OCTOBER 24, 1990 


                                                        October 24, 1990

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.


Product:          Del Monte brand French Style Sliced Green Beans, in 
                  16 ounce cans.  Recall #F-681-0.
Code:             0201J6F82 as one-line ink jet code on can bottom. 
Manufacturer:     Del Monte Corporation, Markesan, Wisconsin. 
Recalled by:      Del Monte Corporation, Walnut Creek, California, by 
                  telephone August 29, 1990.  Firm-initiated recall complete. 
Distribution:     Connecticut, Delaware, New Jersey, New York, Pennsylvania.
Quantity:         843 cases (24 cans/case) were distributed.
Reason:           Potential Clostridium Botulinum toxin hazard. 

                                     ---- 

NOTE:             Mt. Olive brand Fresh Kosher Dill Pickles, in 1 
                  gallon glass jars, Recall #F-609-0 which appeared in
                  the October 10, 1990 Enforcement Report is a completed
                  recall. 
                                     ---- 
                                      -1- 






                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


Product:          (a) Sugar-Free Almond Butter Crisp Candy, packaged in 
                  4 pound boxes; (b) Sugar-Free Almond Bark Candy, packaged 
                  in 5 pound boxes.  Recall #F-686/687-0. 
Code:             (a) Product #451700, lot #OC01 and #2B06; 
                  (b) Product #452400, lot #2D95. 
Manufacturer:     American Confectionery Company, Chino, California.
Recalled by:      The Sweetery, Minneapolis, Minnesota, by letter dated 
                  August 10, 1990 or by telephone during week of
                  August 27, 1990.  Firm-initiated recall complete. 
Distribution:     Iowa, Minnesota, North Dakota, South Dakota.
Quantity:         Approximately (a) 25 boxes; (b) 22 boxes were distributed.
Reason:           Product is insect infested. 

                                     ---- 

Product:          Kiko Maple Syrup, packaged in 24 fl ounce polyethylene
                  containers.  Recall #F-699-0. 
Code:             Not coded.
Manufacturer:     Kiko, Inc., New Orleans, Louisiana. 
Recalled by:      Manufacturer, by telephone on or about August 21, 1990. 
                  Firm-initiated recall ongoing.
Distribution:     Louisiana.
Quantity:         120 cases were distributed. 
Reason:           Imitation-flavored maple syrup not labeled as such. 

                                     ---- 

Product:          Tetley Iced Tea Sweetened and with Natural Lemon Flavor,
                  packed in 12 ounce aluminum cans.  Recall #F-706-0. 
Code:             K022N.
Manufacturer:     Royal Crown Bottling of Virginia, Norfolk, Virginia.
Recalled by:      Manufacturer, by contacting sales personnel August 29 
                  and 30, 1990.  Firm-initiated recall ongoing. 
Distribution:     Virginia, North Carolina. 
Quantity:         72,360 cans were distributed. 
Reason:           Product contains undeclared added saccharin.

                                     ---- 

                                      -2- 





Product:          Nature's Formula Aloe Vera Drinks:
                  (a) Nature's Formula Spirits Natural Juice with Aloe
                  Herbal & Beta carotene, in 10-2/3 fluid ounce plastic 
                  bottles, packaged 24 bottles per case and in eight 4-packs
                  per case (32 bottles);
                  (b) Nature's Formula Aloe Herbal Liquid Formula, in 
                  32 fluid ounce plastic bottles packed 12 per case and 
                  in 64 fluid ounce bottles packed 8 bottles per case.
                  Recall #F-715/716-0.
Code:             Lot numbers: (a) 9187, 9193, 9197, 9211, 9212, 9216,
                  9227, 9234, 9236, 9254; 
                  (b) 9182 (32 oz.),        9206 (32 & 64 oz.)
                      9183 (64 oz.),        9218 (32 & 64 oz.)
                      9188 (32 & 64 oz)     9250 (32 & 64 oz) 
                      9198 (32 & 64 oz)     9256 (32 & 64 oz).
Manufacturer:     Aloe Vera Trading Company, Dallas, Texas. 
Recalled by:      Manufacturer, by memorandum June 18, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Arizona, California, Colorado, Florida, Indiana,
                  Maryland, Mississippi, New York, Ohio, Texas, Washington
                  state, Wisconsin. 
Quantity:         34,776 bottles (10-2/3 oz), 2,880 bottles (32 oz),
                  and 1,880 (64 oz) bottles were distributed. 
Reason:           Mislabeled - High fructose corn syrup substituted for honey.

                                     ---- 

                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 

                                     NONE 
                  Class II -


Product:          Chewable Aspirin for Children, Orange Flavored Tablets, 
                  an OTC product in bottles of 1,000.  Recall #D-080-1. 
Code:             Lot numbers: 90E008 EXP 5/92, 90F007 EXP 6/92, 90G002 
                  EXP 9/92. 
Manufacturer:     Advance Pharmaceutical, Inc., Ronkonkoma, New York. 
Recalled by:      Manufacturer, by letter August 28, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     New York, Florida, New Jersey, California, Illinois.
Quantity:         416 bottles were distributed. 
Reason:           Product packaged in containers greater than 36 tablets
                  each. 
                                     ---- 
                                      -3- 





Product:          Platelets.  Recall #B-001-1.
Code:             Unit #04K02810. 
Manufacturer:     American Red Cross-NE Region, Dedham, Massachusetts.
Recalled by:      Manufacturer, by telephone February 12 and 16, 1990.
                  Firm-initiated recall complete. 
Distribution:     Maine, Massachusetts. 
Quantity:         1 unit was distributed. 
Reason:           Platelets from donor who developed signs and symptoms 
                  of viral infection, within 24 hours of donation, were 
                  distributed.
                                     ---- 

Product:          (a) Red Blood Cells; (b) Cryoprecipitated AHF;
                  (c) Recovered Plasma.  Recall #B-008/010-1. 
Code:             Unit #13C54414. 
Manufacturer:     American Red Cross Blood Services, Detroit, Michigan. 
Recalled by:      Manufacturer, by telephone on or about June 22, 1988. 
                  Firm-initiated recall complete. 
Distribution:     Michigan. 
Quantity:         1 unit of each component was distributed. 
Reason:           Blood components, from a donor who previously tested
                  repeatably reactive for the human immunodeficiency
                  virus, type 1 (anti-HIV-1), were distributed. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Platelets; (c) Recovered 
                  Plasma.  Recall #B-014/016-1. 
Code:             Unit numbers: (a) 38F92740, 38F95049, 38H07576, 38M12547; 
                  (b) 38F92740, 38F95049, 38J11428; 
                  (c) 38F92740, 38F95049, 38H07576, 38M12547. 
Manufacturer:     American Red Cross, Fort Wayne, Indiana.
Recalled by:      Manufacturer, by letters of June 22, 1990, August 14 and
                  22, 1990.  Firm-initiated recall ongoing. 
Distribution:     Indiana, California, Illinois, Ohio.
Quantity:         (a) 4 units; (b) 3 units; (c) 4 units were distributed. 
Reason:           Blood products which tested non-reactive for the
                  antibody to the human immunodeficiency virus type 1 
                  (anti-HIV-1) but were collected from donors who 
                  previously tested repeatably reactive for anti-HIV-1, 
                  were distributed. 
                                     ---- 

                  Class III - 

Product:          Oxygen USP, Medicinal Gas, packed in green steel C, D and 
                  E size cylinders, an Rx product.  Recall #D-082-1.
Code:             Lot #121-9003 and Serial numbers: 702, 736, 795, 811, 
                  946, 10105, 10111, 10132. 
Manufacturer:     Tri-County Oxygen and Medical Supplies, Inc., Brooksville,
                  Florida.
Recalled by:      Manufacturer, by visit on or about September 2, 1990. 
                  Firm-initiated recall complete. 

                                      -4- 





Distribution:     Florida.
Quantity:         13 units were distributed; firm estimates none remains
                  on the market.
Reason:           Failure to test purity. 

                                     ---- 

Product:          Baxter Fenwal Plasmacell-C Disposable Plasma Collection 
                  Bag with In-line Anticoagulant drip chamber.
                  Recall #B-002-1.
Code:             Lot numbers: 9001 through 9365, 0001 through 0151.
Manufacturer:     Baxter Healthcare Corporation, Fenwal Division, 
                  Deerfield, Illinois.
Recalled by:      Manufacturer, by letter June 25, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         Unknown.  All products manufactured between January 1,
                  1988 and June 25, 1990. 
Reason:           Possibly adulterated with plastic shavings (particulate 
                  matter).

                                     ---- 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 
                                     NONE 

Product:          Enema Administration Kits with Latex Retention Cuffs: 
                  (a) Enema Tips with Inflatable Cuffs sold for 
                  attachment to Enema Kits, such as the Miller Air
                  Retention Tip;
                  (b) Enema Tips with Inflatable Cuffs attached to Empty
                  Enema Bags, such as the Super XL Enema Kit; 
                  (c) Enema Tips with Inflatable Cuffs attached to
                  prefilled Barium Enema Bags.  Recall #Z-932/934-0.
Code:             All lots of the following E-Z-EM products:
                  Filled/unfilled Enema Kits-Catalog Numbers: 
                  110/     416/     1316/     P650PP
                  112/     420/     2212      P650PPK 
                  114/     510/     2216/ 
                  116/     512/     910/
                  310/     514/     920/
                  312/     516/     930/
                  314/     520/     935/
                  316/     524/     925S
                  318/     812/     925K
                  320/     816/     925/
                           1312/    M25 
                                    M25K
                                    M20 

                                      -5- 





                  Retention Cuff Enema Tip-Catalog Numbers: 
                  913    915    931    932
                  933    934    938    939
                  940    955    966    9308 
                  9317   9309   9325   9301 
                  9310   9328   9302   9311 
                  9335   9305   9314   9336 
                  9306   9315   9338   9307 
                  9316   9345 
                  All lots of the following Lafayette Model numbers 
                  (Filled/unfilled enema bags and unit dose kits):
                  1112    6616    1260    4404    1837
                  1116    6642    1802            1840
                  1117    1815    1805            1841
                  1120    1197    1806    1835    1842
                  1125    1242    1811    1836    1844. 
                  All lots of Enterocylsis Catheters-Catalog Numbers
                  ET412, 5510, 5563.
Manufacturer:     E-Z-EM Inc., Westbury, New York.
Recalled by:      E-Z-EM Inc., Westbury, New York or Lafayette Division 
                  of E-Z-EM, Inc., Lafayette, Indiana, by letter October 8, 
                  1990.  Firm-initiated recall ongoing.  See also FDA Talk
                  Paper T90-49, October 15, 1990. 
Distribution:     Nationwide and international. 
Quantity:         Firm estimates between 110 and 150 thousand units remain
                  on the market.
Reason:           In some cases the devices may trigger an allergic reaction
                  causing injury or possible death if not treated in time.

                                     ---- 

                  Class II -

Product:          BRT 2001 and BRT 2010 Laser Bar Code Scanners, used 
                  for inventorying warehouse stock.  Recall #Z-888-0. 
Code:             Serial numbers:  Model 2001 - 11028701; Model 2010 - 07158701 
                  07158702, 02258901, 03258701, 03038701, 03038702, 
                  01298701, 01298702, 01168601, 01168602, 09308701, 
                  09308702, 09308703, 09308704, 10018701, 10018702, 
                  11248601, 11248602, 08028601, 08028602, 04218602, 
                  08158601, 02199003, 02108601, 06048701, 08248701. 
Manufacturer:     BRT Corporation, Doylestown, Pennsylvania.
Recalled by:      Manufacturer, by letter October 19, 1990.  Firm-initiated 
                  field correction ongoing. 
Distribution:     Connecticut, Delaware, Florida, New Jersey, Pennsylvania, 
                  Utah, Wisconsin.
Quantity:         1 unit of model 2001 and 23 units of model 2010 were
                  distributed.
Reason:           Noncompliance with performance standards for laser
                  products in that labeling and user manuals were inadequate. 

                                     ---- 

                                      -6- 





Product:          Orient brand latex gloves:
                  (a) "Sterile" Surgical Gloves; (b) "Non-sterlie"
                  Examination Gloves, all sizes, 12 pairs per unit, 
                  pre-powdered, use once.  Recall #Z-007/008-1. 
Code:             No codes listed on inner carton. Individual packages: 
                  Size 6, batch (lot) No 880071, EXP 8/91; Size 6 1/2,
                  batch (lot) No. 880073 EXP 8/91.
Manufacturer:     Orient Medical Supply Company, Ltd., Seoul, Korea.
Recalled by:      Surgix, Inc., Ronkonkoma, New York, by letter August
                  28, 1990.  Firm-initiated recall ongoing. 
Distribution:     New York. 
Quantity:         205 master cartons with 42 inner cartons of 12 pairs each.
Reason:           (a) "Sterile" glove packets had partially incomplete pack 
                  seals which would compromise the sterility of the product;
                  (b) The gloves failed FDA leak test.

                                     ---- 

Product:          Tourniquet Cuff Pressure Monitors, used to restrict 
                  blood flow in patient's arm or leg during surgery:
                  (a) Model TCPM; (b) Model Medi-Quet 2740. 
                  Recall #Z-010/011-1.
Code:             All lots and serial numbers:
Manufacturer:     Electromedics, Inc., Englewood, Colorado. 
Recalled by:      Manufacturer, by letter September 6, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, Turkey, Italy.
Quantity:         (a) 188; (b) 630 units were distributed.
Reason:           A limited number of units may have been manufactured with 
                  an incorrect microprocessor timing chip that will cause 
                  the device to show incorrect time by allowing the unit
                  to show 99 minutes on the visual display before changing
                  to 1:00 hour. 

                                     ---- 

Product:          Single Patient 480 Dialysis Delivery Systems, except
                  High Flow and TUEV Models:
                  (a) Standard 480 (non UF Control): Acetate, Bicarbonate,
                  and Variable Sodium;
                  (b) Ultrafiltration Control 480s: Acetate, Bicarbonate, and 
                  Variable Sodium;
                  (c) 480 Dialysate control Spare Modules: Non UF Control,
                  UF Control.  Recall #Z-012/014-1. 
Code:             Lot numbers:
                  (a) Acetate      Bicarbonate       Variable Sodium
                  N480A-0A0        N480B-0A0         N480V-1A0
                  N480A-1AR        N480B-1A0         N480V-1AR
                  N480A-1A0        N480B-1AR         N480V-1B0
                  N480A-1BR        N480B-1B0         N480V-1BR
                  N480A-1B0        N480B-1BR         N480V-1C0
                  N480A-1CR        N480B-1C0         N480V-1CR

                                      -7- 





                  N480A-1C0        N480B-1CR         N480V-1D0
                  N480A-1DO        N480B-1D0         N480V-2A0
                  N480A-2A0        N480B-1DR         N480V-2AR
                  N480A-3A0        N480B-2A0         N480V-3A0
                  N480A-5A0        N480B-2AR         N480V-5A0
                                   N480B-3A0
                                   N480B-5A0
                  (b) Acetate      Bicarbonate       Variable Sodium
                  N485A-1A0        N485B-0A0         N485V-0A0
                  N485A-1AR        N485B-1A0         N485V-1A0
                  N485A-1B0        N485B-1AR         N485V-1AR
                  N485A-1BR        N485B-1B0         N485V-1B0
                  N485A-1C0        N485B-1BR         N485V-1BR
                  N485A-1CR        N485B-1CO         N485V-1C0
                  N485A-2A0        N485B-1CR         N485V-1CR
                  N485A-3A0        N485B-1D0         N485V-1D0
                  N485A-5A0        N485B-2A0         N485V-2A0
                                   N485B-2AR         N485V-2AR
                                   N485B-3A0         N485V-3A0
                                   N485B-5A0         N485V-5A0

                  (c) For lot numbers beginning with M1... through M9 
                  and MA... through MB13560.

                  Non UF Control      UF Control
                  N-1004-0E-00        N1104-4FS 
                  N-1004-0F-00        N-1004-0E-UF
                  N-1004-0F-02        N-1004-0F-UF
                  N112A-4E0           N113A-4E0 
                  N-1104-RE-00        N-1104-RE-UF. 
Manufacturer:     CD Medical, Inc., Subsidiary of the Dow Chemical Company, 
                  Portland, Oregon. 
Recalled by:      Manufacturer, by letter August 20, 1990.  Firm-initiated
                  field correction ongoing. 
Distribution:     Nationwide and international. 
Quantity:         6,233 machines/modules were distributed.
Reason:           A short allowed by an alarm lamp design may disable 
                  conductivity or temperature monitoring, alarm, and/or 
                  bypass circuit. 
                                     ---- 

Product:          Candela LaserTripter System (a) Model MDL-1;
                  (b) Model MDL-2000, used for urinary/biliary laser
                  lithotripsy.  Recall #Z-015/016-1.
Code:             MDL-1 (modified) serial numbers 0001 through 0151,
                  non-inclusive; MDL-1 (modified) serial numbers 1152 
                  through 1179, inclusive; MDL-2000 serial numbers 1180 
                  through 1238, inclusive.
Manufacturer:     Candela Laser Corporation, Wayland, Massachusetts.
Recalled by:      Manufacturer, by letter May 14, 1990.   Firm-initiated
                  recall ongoing. 

                                      -8- 





Distribution:     Nationwide and international. 
Quantity:         l85 finished devices are subject to recall. 
Reason:           The 320 micron fiber may break, causing fragments 
                  to remain in the body.

                                     ---- 

Product:          Cobe Spectra Apheresis System Centrifuge Collar Holder, 
                  a component of the Spectra Apheresis Machine designed to
                  hold blood tubing sets in a configuration during
                  centrifugation.  Recall #Z-017-1. 
Code:             Catalog Nos. 950000-000 through 950000-003, and 
                  950000-005.  Serial numbers 1S00028 through 1S00192.
Manufacturer:     Cobe Laboratories, Inc., Lakewood, Colorado.
Recalled by:      Manufacturer, by letter on or about September 24, 1990. 
                  Firm-initiated recall ongoing.
Distribution:     Nationwide and international. 
Quantity:         415 units remain on the market. 
Reason:           The centrifuge collar holder may not stay closed
                  during a procedure, due to wear of components.  The 
                  centrifuge collar can come out of the holder, cause the 
                  tubing to break, and create a saline or blood leak during 
                  a procedure.
                                     ---- 

Product:          CeraMed brand Osteograf/P Periodontal Hydroxylapatite,
                  0.5g Prefilled Syringes, a sterile high purity, high
                  density, nonresorbable, radiopaque, permanent particulate 
                  implant (bone grafting) material used for repair of 
                  periodontal lesions, packaged 8 syringes per carton.
                  Recall #Z-018-1.
Code:             Lot #9219262. 
Manufacturer:     CeraMed Corporation, Lakewood, Colorado.
Recalled by:      Manufacturer, by letter January 31, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, Egypt, Canada.
Quantity:         144 syringes were distributed; firm estimates none
                  remains on market.
Reason:           The sterility of the product may be compromised 
                  because of faulty pouch seals.

                                     ---- 

Product:          Cordis II Super Torque Angled Pigtail Ventricular 
                  Catheter.  Recall #Z-021-1. 
Code:             Catalog #532-533A, lot #R-689501. 
Manufacturer:     Cordis Europa N.V., The Netherlands.
Recalled by:      Cordis Corporation, Miami Lakes, Florida, by telephone
                  May 3, 1990.  Firm-initiated recall complete. 
Distribution:     Nationwide, Hong Kong.

                                      -9- 





Quantity:         775 units were distributed. 
Reason:           The sterility of the product may be compromised 
                  due to open package end seals.

                                     ---- 
                  Class III - 

Product:          Companion 2801 Portable Volume Ventilator a device
                  which provides support for ventilator-dependent patients in 
                  both the hospital and home care settings.  Recall #Z-741-0. 
Code:             Serial numbers 000040PI through 000057PI, 
                  000101DE through 000297DE.
Manufacturer:     Puritan-Bennett, Portable Ventilator Division,
                  Boulder, Colorado.
Recalled by:      Manufacturer, by letter August 2, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, Canada, Italy, England. 
Quantity:         210 units were distributed. 
Reason:           The transient electrical fluctuation of short duration
                  may cause the fuse in the Companion 2801 Portable Volume
                  Ventilator to blow unnecessarily.  Should this occur, the 
                  ventilator will switch to battery operation with the
                  normal alarm to signal the fact.

                                     ---- 

Product:          Lifescan Hospital One Touch Test Strips for Blood 
                  Glucose Meter, an in-vitro diagnostic device. 
                  Recall #Z-760-0.
Code:             Lot #910464A EXP 6/91.
Manufacturer:     Lifescan, Inc., Mountain View, California.
Recalled by:      Lifescan, Inc., Milpitas, California, by telephone
                  April 3-4, 1990.  Firm-initiated recall ongoing.
Distribution:     Nationwide, Canada. 
Quantity:         176 boxes were distributed. 
Reason:           The High Control Range for product testing was printed
                  as 272-280 mg/dL instead of the correct range 272-380 
                  mg/dL.

                                      --- 

                              VETERINARY PRODUCTS 

                                     NONE 

                                     -10- 





Seizure Actions Filed:

Product:          GSL Cellular Products (90-581-561). 
Charge:           Unapproved new drugs and the articles' labeling fails to
                  bear adequate directions for use. 
Firm:             TransAmerica Mailings, Inc., Berea, Ohio. 
Filed:            September 28, 1990 - U.S. District Court for the
                  Northern district of Ohio, Eastern Division;
                  Civil #1:90CV1737, FDC #65889.