FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
05/14/1991      

FDA Enforcement for the week of 5/15/91



                            FDA ENFORCEMENT REPORT
                               FOR MAY 15, 1991 
                                                             May 15, 1991 

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

Product:          (a) Low Moisture Mozzarella Cheese; 
                  (b) Low-Moisture-Part Skim Mozzarella Cheese. 
                  Recall #F-373/374-1.
Code:             Cheese dated January 25, 1991 through February 12, 1991.
Manufacturer:     B.C. Acquisitions, Inc., doing business as Branch 
                  Cheese, Branch, Wisconsin.
Recalled by:      Manufacturer, by telephone February 14, 1991 and
                  March 6, 1991.  Firm-initiated recall ongoing.
Distribution:     Connecticut, Georgia, Illinois, Michigan, New York, 
                  Ohio, Pennsylvania, Texas, West Virginia, Wisconsin.
Quantity:         6,553 cases and 89/1,000 pound totes were distributed.
Reason:           Product is contaminated with Listeria monocytogenes.

                                     ---- 

                                      -1- 






                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 


Product:          Mini Jelly Cups, a gelatin like candy.  Recall #F-378-1.
Code:             None
Manufacturer:     Pokka Corporation (S) PTE. LTD, Jurong Town, Singapore. 
Recalled by:      Tan Duc Trading, Brooklyn, New York, by telephone January 
                  19, 1991.  Firm-initiated recall ongoing. 
Distribution:     New York, Massachusetts.
Quantity:         95 cartons were distributed; firm estimates 12 cartons
                  remain on the market. 
Reason:           Product contains undeclared FD&C Yellow No. 5 along with
                  non-permitted food color cochineal Red or Ponceau 4R. 

                                     ---- 

Product:          New York Bagel Boys brand Egg Bagels packed 12 bagels in
                  plastic bag.  Recall #F-377-1.
Code:             An orange bag closure is used.  Orange closure indicates
                  product was manufactured Monday, March 25, 1991.
Manufacturer:     General Bagel Corporation, Bothell, Washington. 
Recalled by:      Manufacturer, by telephone on March 27, 1991, followed
                  by visit on March 28, 1991.  Firm-initiated recall ongoing. 
Distribution:     Washington state. 
Quantity:         53 dozen were distributed.
Reason:           Product contains undeclared FD&C Yellow No. 5.

                                     ---- 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


Product:          (a) Silver Creek brand Canned Smoked Sturgeon, in 4 ounce 
                  cans; (b) Canned Smoked Salmon, in 4 ounce cans.
                  Recall #F-375/376-1.
Code:             (a) Not coded; (b) 26290104.
Manufacturer:     Truzzolino's Food Products, Butte, Montana. 
Recalled by:      Tunnel Creek, Inc., Twin Falls, Idaho, by telephone 
                  and visit October 26, 1990.  Firm-initiated recall ongoing. 
Distribution:     Washington state, Florida, Idaho. 
Quantity:         (a) 64 cans; (b) 64 cans were distributed.
Reason:           Product processed without having a field scheduled process. 

                                     ---- 

                                   COSMETICS

                                     NONE 

                                      -2- 





                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 
                  Class II -

Product:          Trilafon Tablets and Etrafon Tablets, immediate 
                  release coated tablets, packed in 100, and 500 tablet 
                  bottles.  The Etrafon tablets are also packed in 100
                  tablet/unit dose packs: 
                  (a-d) Trilafon Tablets (Perphenazine), USP - All strengths
                  2 mg, 4 mg, 8 mg and 16 mg, an Rx product used in the 
                  treatment of psychotic disorders and to control severe
                  nausea and vomiting in adults;
                  Etrafon is an Rx product prescribed for patients with 
                  severe anxiety and in a depressed mental state: 
                  (e) Etrafon Tablets, USP 2/25 - Each tablet contains
                  Perphenazine, USP - 2 mg; and Amitriptyline HCl, USP - 25 mg. 
                  (f) Etrafon 2-10 Tablets, USP 2/10 - Each tablet contains 
                  Perphenazine, USP - 2 mg; and Amitriptyline HCl, USP - 10 mg. 
                  (g) Etrafon Forte Tablets, USP 4/25 - Each tablet contains
                  Perphenazine, USP - 4 mg; and Amitriptyline HCl, USP - 25 mg. 
                  (h) Etrafon A Tablets, USP 4/10 - Each tablet contains
                  Perphenazine, USP - 4 mg; and Amitriptyline HCl, USP - 10 mg. 
                  Recall #D-402/409-1.
Code:             Batch Numbers:

(a) Trilafon (perphenazine) Tablets, 2 mg:
6-ADH-1, EXP  1/91     6-ADH-107, EXP  9/91     8-ADH-3, EXP  3/93
6-ADH-2, EXP  3/91     6-ADH-108, EXP  9/91     8-ADH-4, EXP  5/91
6-ADH-3, EXP  3/91     7-ADH-1, EXP  8/91       8-ADH-5, EXP  5/91
6-ADH-101, EXP  3/91   7-ADH-2, EXP  12/91      9-ADH-1, EXP  3/92
6-ADH-102, EXP  4/91   7-ADH-3, EXP  12/91      9-ADH-4, EXP  10/92 
6-ADH-103, EXP  5/91   7-ADH-4, EXP  3/92       0-ADH-1, EXP  3/93
6-ADH-104, EXP  5/91    7-ADH-101, EXP  3/92    0-ADH-2, EXP  4/93
6-ADH-105, EXP  7/91    8-ADH-1, EXP  1/93
6-ADH-106, EXP  9/91    8-ADH-2, EXP  1/93
(b) Trilafon (Perphenazine) Tablets, 4 mg:
6-ADK-1, EXP 1/91   7-ADK-1, EXP 10/92   8-ADK-2, EXP 3/93
6-ADK-2, EXP 3/91   7-ADK-2, EXP 3/92    8-ADK-3, EXP 3/93   9-ADK-1, EXP 3/92
6-ADK-3, EXP 3/91   7-ADK-101, EXP 9/92  8-ADK-4, EXP 5/91   9-ADK-2, EXP 3/92
6-ADK-4, EXP 4/91   7-ADK-102, EXP 3/92  8-ADK-5, EXP 5/91   9-ADK-3, EXP 9/92
6-ADK-5, EXP 7/91   7-ADK-103,EXP 11/92  8-ADK-6, EXP 7/91   0-ADK-1, EXP 3/93
6-ADK-6, EXP 9/91   7-ADK-104,EXP 12/92  8-ADK-7, EXP 9/91   0-ADK-3, EXP 5/92
6-ADK-101,EXP 3/91  8-ADK-1, EXP 1/93    8-ADK-8, EXP 9/91
                                         8-ADK-9,EXP 9/91 or 9/90 
                                         8-ADK-10,EXP 12/91 
(c) Trilafon (perphenazine) Tablets, 8 mg:
6-ADJ-1, EXP  1/91     7-ADJ-101, EXP  9/92    8-ADJ-6, EXP  7/91 
6-ADJ-2, EXP  6/91     7-ADJ-102, EXP  10/92   8-ADJ-7, EXP  9/91 
6-ADJ-3, EXP  7/91     7-ADJ-103, EXP  11/92   8-ADJ-8, EXP  9/91 
6-ADJ-101, EXP  7/91   7-ADJ-104, EXP  11/92   8-ADJ-9, EXP  9/91 or 9/90 

                                      -3- 





6-ADJ-102, EXP  9/91   8-ADJ-1, EXP  1/93      8-ADJ-10, EXP  8/91
6-ADJ-103, EXP 10/91   8-ADJ-2, EXP  3/93      8-ADJ-11, EXP  12/91 or 12/90
6-ADJ-104, EXP  1/91   8-ADJ-3, EXP  3/93    8-ADJ-12, EXP  11/91 
7-ADJ-1, EXP  12/91    8-ADJ-4, EXP  5/91      9-ADJ-10, EXP  9/92
7-ADJ-2, EXP  3/92     8-ADJ-5, EXP  5/91      9-ADJ-12, EXP  10/92 
                                               9-ADJ-1, EXP  3/92 
                                               9-ADJ-2, EXP  9/92 
                                               0-ADJ-1, EXP  4/93 
(d) Trilafon (perphenazine) Tablets, 16 mg: 
6-ADM-1, EXP  1/91      7-ADM-1, EXP  1/92        8-ADM-1, EXP  1/93
6-ADM-2, EXP  1/91      7-ADM-2, EXP  1/92        8-ADM-2, EXP  3/93
6-ADM-3, EXP  5/91      7-ADM-3, EXP  4/92        8-ADM-3, EXP  3/93
6-ADM-4, EXP  5/91      7-ADM-4, EXP  4/92        8-ADM-4, EXP  5/91
6-ADM-5, EXP  7/91      7-ADM-5, EXP  4/92        8-ADM-5, EXP  5/91
6-ADM-6, EXP  8/91      7-ADM-6, EXP  10/92       8-ADM-6, EXP  7/91
6-ADM-7, EXP  8/91      7-ADM-101, EXP  4/92      9-ADM-1, EXP  1/92
6-ADM-101, EXP  2/91                              9-ADM-2, EXP  4/92
6-ADM-102, EXP  2/91                              0-ADM-2, EXP  2/93
6-ADM-103, EXP  4/91
6-ADM-104, EXP  4/91
(e) ETRAFON 2-10, 2/10MG TABLETS: 
8-ANA-1, EXP  1/91
8-ANA-2, EXP  3/91
9-ANA-1, EXP  1/92
(f) ETRAFON, 2/25MG TABLETS:
8-ANC-1, EXP  2/91
8-ANC-2, EXP  6/91
9-ANC-1, EXP  9/92
(g) ETRAFON A, 4/10MG TABLETS:
9-ANB-1, EXP  1/92, 
9-ANB-2, EXP  6/92
9-ANB-3, EXP  9/92
0-ANB-1, EXP  2/92
(h) ETRAFON FORTE, 2/25MG TABLETS:
8-ANE-101, EXP  6/91
8-ANE-1, EXP  2/91
0-ANE-1, EXP  4/92
Manufacturer:     Schering Plough Corporation, Kenilworth, New Jersey.
Recalled by:      Schering Laboratories, Inc., Union, New Jersey, by
                  letter January 23, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide and international. 
Quantity:         Approximately (a) 28 million tablets; (b) 2 million tablets 
                  were distributed. 
Reason:           Some lots fail dissolution specifications, others lack
                  assurance of meeting specificatons through expiration date. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Platelets; (c) Cryoprecipitated
                  AHF; (d) Fresh Frozen Plasma; (e) Recovered Plasma. 
                  Recall #B-198/202-1.

                                      -4- 
Code:             Unit numbers: (a) 17F60075, 17G77735, 17H31920, 17H39657, 
                  17H44384, 17J41878, 17K72889, 17K79829, 17M76005, 17M84182, 
                  17P11630, 17Q25860, 17Q36163, 17R33969, 17R45737; 
                  (b) 17F60075, 17G77735, 17H31920, 17H39657, 17M76005, 
                  17M84182, 17P11630, 17Q25860, 17Q36163, 17R33969; 
                  (c) 17K72889; 
                  (d) 17K79829, 17M84182, 17Q36163, 17R45737; 
                  (e) 17F60075, 17G54452, 17G77735, 17H31920, 17H39657, 
                  17H44384, 17J33645, 17J41878, 17K72289, 17M76005, 
                  17P11630, 17P13748, 17Q25860, 17R33969. 
Manufacturer:     American Red Cross, St. Paul, Minnesota.
Recalled by:      Manufacturer, by letters of May 21, 25, 29 and 31, 1990,
                  June 1, 6 and 8, 1990, September 6 and 9, 1990; and 
                  November 9, 1990 and by telephone May 29, 1990, 
                  August 28, 1990 and September 6, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Minnesota, Massachusetts, Tennessee, California 
                  Wisconsin, Switzerland. 
Quantity:         (a) 15 units; (b) 10 units; (c) 1 unit; 
                  (d) 4 units; (e) 14 units.
Reason:           Blood products were collected and distributed from
                  donors who: 1) previously tested repeatably 
                  reactive for the antibody to the human immuno-
                  deficiency virus type-1 (anti-HIV-1); 2) had spouses
                  that participated in high risk behavior; 3) were IV 
                  drug users; or 4) had histories of hepatitis. 

                                     ---- 

Product:          Red Blood Cells.  Recall #B-224-1.
Code:             Unit #38T30383. 
Manufacturer:     American Red Cross Blood Services, Fort Wayne,
                  Indiana.
Recalled by:      Manufacturer, by telephone August 31, 1990.  Firm-
                  initiated recall complete.
Distribution:     Indiana.
Quantity:         1 unit. 
Reason:           Blood product collected from a donor who was
                  subsequently diagnosed as having a gum infection
                  at the time of donation was distributed.

                                     ---- 

Product:          (a) Red Blood; (b) Recovered Plasma.  Recall #B-234/235-1.
Code:             Unit #53R53398. 
Manufacturer:     American Red Cross, Baltimore Maryland. 
Recalled by:      Manufacturer, (a) by telephone August 28, 1990 followed 
                  by letter dated January 18, 1991; (b) by letter dated 
                  November 14, 1990. Firm-initiated recall ongoing. 
Distribution:     (a) Maryland; (b) Switzerland.
Quantity:         1 unit of each component. 

                                      -5- 
Reason:           Blood products which tested non-reactive for the antibody to
                  the human immunodeficiency virus type 1 (anti-HIV-1), but 
                  were collected from a donor who previously tested repeatably
                  reactive for anti-HIV-1, were distributed.

                                     ---- 

Product:          Recovered Plasma.  Recall # B-236-1.
Code:             Unit# 4742270.
Manufacturer:     Memorial Sloan Kettering Cancer Center, New York, New York. 
Recalled by:      Manufacturer, by telephone followed by letter October 2, 1990 
                  and by telephone and letter December 5, 1990.  Firm-
                  initiated recall complete.
Distribution:     New Jersey, Florida, Switzerland. 
Quantity:         1 unit. 
Reason:           Recovered Plasma, collected from a donor who previously 
                  tested repeatably reactive for Human Immunodeficiency Virus 
                  Type 1 (anti-HIV-1), was distributed. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Recovered Plasma.
                  Recall #B-237/238-1.
Code:             Unit numbers: 40FK21979, 40GR97973. 
Manufacturer:     American Red Cross, Peoria, Illinois. 
Recalled by:      Manufacturer, by letters of January 7 and 31, 1991. 
                  Firm-initiated recall ongoing.
Distribution:     (a) Illinois; (b) California. 
Quantity:         2 units of each component.
Reason:           Blood products collected from a donor diagnosed 
                  with cancer, were distributed.

                                     ---- 

CORRECTION:       Intal Nebulizer Solution, 2 ml ampules, Recall #D-401-1 
                  Which appeared in the May 1, 1991 Enforcement Report
                  should read:
                  Recalled by:  Fisons Corporation, Rochester, New York, by 
                  telephone March 15, 1991, FAX March 18, 1991 and by 
                  letter March 22, 1991.
                  Quantity:  27 boxes (120 ampules per box) were distributed. 

                                     ---- 
                  Class III - 

Product:          Fred Meyer Certified brand Children's Chewable Non- 
                  Aspirin Pain Relief, 80 mg, an OTC product. 
                  Recall #D-410-1.
Code:             Lot numbers H0080J EXP 8/92, J0135C EXP 9/92. 
Manufacturer:     Hall Laboratories, Inc., Portland, Oregon.
Recalled by:      Manufacturer, by telephone April 2 and 19, 1991.
                  Firm-initiated recall ongoing.

                                      -6- 





Distribution:     Oregon, Washington state, Alaska, Idaho.
Quantity:         13,963 boxes of 30 tablets each were distributed. 
Reason:           The unit carton incorrectly states dosage for 2-3 
                  year old as 3 tablets.  The correct dosage of 2 tablets 
                  appears on the label. 

                                     ---- 

Product:          Furosemide Tablets, an Rx diuretic and anti-hypertensive: 
                  (a) 20 mg Tablets, in bottles of 100 and 1,000; 
                  (b) 40 mg Tablets, in bottles of 100 and 500; 
                  (c) 80 mg Tablets, in bottles of 100.  Recall #D-411/413-1. 
Code:             Lot numbers: (a) 19579L EXP 9/91, 27899L EXP 9/91,
                  395D9L EXP 12/91 (100's); 13310L EXP 2/91, 16711L EXP 10/92 
                  19679L EXP 7/91, 21630L EXP 3/92, 27799L EXP 9/91,
                  397D9L EXP 1/92 (1000's); 
                  (b) 17169L  (100's); 13410L EXP 1/92, 17069L EXP 6/91 
                  17169L EXP 6/91, 18011L EXP 12/92, 20479L EXP 7/91, 
                  25540L EXP 6/92, 34650L EXP 5/92, 34750L EXP 5/92,
                  34950L EXP 6/92, 32309L EXP 10/91, 38060L EXP 6/92, 
                  57590L EXP 9/92;
                  (c) 13610L EXP 2/92, 16420L EXP 2/92, 43170L EXP 8/92,
                  27999L EXP 10/91. 
Manufacturer:     Warner Chilcott Labs, Lititz, Pennsylvania. 
Recalled by:      Warner Chilcott, Division of Warner Lambert, Morris 
                  Plains, New Jersey, by letter April 12, 1991.  Firm-
                  initiated recall ongoing. 
Distribution:     Nationwide. 
Quantity:         127,860 bottles were distributed. 
Reason:           Product may not meet dissolution specifications through 
                  expiration date.

                                     ---- 

Product:          Isovue 300 (Iopamidol Injection), in 75 ml clear
                  glass bottles, an Rx product for intravenous diagnostic 
                  imaging.  Recall #D-414-1.
Code:             Lot #1B48177 EXP 2/1/94.
Manufacturer:     Bristol-Myers Squibb Company, New Brunswick, New Jersey.
Recalled by:      Squibb Diagnostics, Princeton, New Jersey, by letter
                  March 26, 1991.  Firm-initiated recall ongoing. 
Distribution:     California, Georgia, Iowa, Illinois, Indiana, Massachusetts,
                  Maryland, Michigan, North Carolina, Nebraska, North Dakota, 
                  New Jersey, Oklahoma, Pennsylvania, Wisconsin.
Quantity:         810 bottles were distributed; firm estimates 320
                  bottles remain on the market. 
Reason:           The primary labeling panel correctly identifies the 
                  product as Isovue 300.  The secondary identification
                  which appears along the bottom of the label, incorrectly
                  identifies the product as Isovue-370. 

                                     ---- 

                                      -7- 





Product:          Abbott Pharmaceuticals brand Dicumarol Tablets, USP,
                  50 mg, an Rx oral coumarin anticoagulant for the
                  prophylaxis and treatment of blood clots, in bottles
                  of 100 and 1,000 tablets.  Recall #D-415-1. 
Code:             Lot numbers: 03-386-AY-40 EXP 5/1/92, 03-386-AY-21
                  EXP 5/1/92, 16-451-AY-21 EXP 6/1/93.
Manufacturer:     Abbott Laboratories, Barceloneta, Puerto Rico.
Recalled by:      Abbott Laboratories, Abbott Park, Illinois, by letter 
                  April 12, 1991.  Firm-initiated recall ongoing. 
Distribution:     Nationwide. 
Quantity:         4,240 bottles of 100 and 66 bottles of 1000 were
                  distributed; firm estimates very little of the
                  product remains on the market.
Reason:           Product may not meet dissolution specifications through 
                  expiration date.

                                     ---- 

Product:          Red Blood Cells.  Recall #B-183-1.
Code:             Unit #6016238.
Manufacturer:     LifeSource, Glenview, Illinois. 
Recalled by:      Manufacturer, telephone August 1, 1990.  Firm-initiated 
                  recall complete.
Distribution:     Illinois. 
Quantity:         1 unit. 
Reason:           Blood product labeled with an incorrect expiration date.

                                     ---- 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 
                                     NONE 

                  Class II -

Product:          Johnson & Johnson Reach Full Size Gentle Toothbrush.
                  Recall #Z-557-1.
Code:             Product #7118 - Lot numbers: DOP, EOP, GOP, HOP, JOP, KOP.
                  Corrugated shipper date codes first 3 digits 091 through
                  270.
                  Product #7117 - Lot numbers: 4DOP, 4EOP, 4GOP, 4HOP,
                  4JOP, 4KOP.  Shipper date codes first 3 digits 091
                  through 243.
Manufacturer:     Johnson & Johnson Dental Care Company, Las Piedras, 
                  Puerto Rico.
Recalled by:      Johnson & Johnson Consumer Products, Inc., Skillman,
                  New Jersey, by letters of February 15, 1991 and 
                  April 22, 1991.  Firm-initiated recall ongoing. 
Distribution:     Nationwide and Puerto Rico. 

                                      -8- 





Quantity:         1.7 million units (less 2,856  - code 7117) were
                  distributed; firm estimates 320,000 units remain
                  on the market.
Reason:           Broken handles. 

                                     ---- 

Product:          Intravascular (IV) Administration Sets, with piggyback
                  feature, used for Secondary Infusion of IV Fluids:
                  Recall Nos. Z-610/645-1 
                  1.  V1419-17 15 Drops/mL IV Set with 2 Injection Sites and 2
                  Check Valves; 
                  2.  V1443 15 Drops/mL ADDitIV Primary IV Set with 2 Injection 
                  Sites; 86" Length; w/V-Clamp; 
                  3.  V1445 Drops/mL ADDitIV Primary IV Set with 3 Injection
                  Sites; 96" Length; w/V-Roller Clamp;
                  4.  V1445-24 15 Drops/mL ADDitIV Primary IV Set with 4
                  Injection Sites;
                  5.  V1445-30 15 Drops/mL ADDitIV Primary IV Set with 3
                  Injection Sites; 118" Length; w/Roller Clamp; 
                  6.  V1446 15 Drops/mL ADDitIV Primary IV Set with 2 Injection 
                  Sites; 93" Length; w/V-Clamp; 
                  7.  V1448 60 Drops/mL ADDitIV Primary IV Set with 3 Injection 
                  Sites; 87" Length; w/V-Clamp; 
                  8.  V1449-01 60 Drops/mL ADDitIV Primary Mini-Drop Vented IV
                  Set wi th 3 Injection Sites; 89" Length; w/Roller Clamp;
                  9.  V1449-12 60 Drops/mL ADDitIV Primary Mini-Drop
                  vented IV Set with 2 Injection Sites; 
                  10.  V1451 60 Drops/mL ADDitIV Primary Mini-Drop IV Set 
                  with 3 Injection Sites; 89" Length; w/Roller Clamp; 
                  11.  V1453 15 Drops/mL ADDitIV Primary IV Set with 3
                  Injection Sites; 96" Length; w/V-Clamp; 
                  12.  V1464-15 15 Drops/mL ADDitIV Primary IV Set with 2 
                  Injection Sites; 88" Length; w/Roller Clamp;
                  13.  V1486 60 Drops/mL ADDitIV Primary IV Set with 3
                  Injection Sites; 87" Length; w/Roller Clamp;
                  14.  V1776 15 Drops/mL ADDitIV Primary IV Set with PMF Filter 
                  and 3 Injection Sites;
                  15.  V1776-40 15 Drops/mL ADDitIV Primary IV Set with 
                  pre-attached Secondary Set with PMF Filter and 2
                  Injection Sites;
                  16.  V1778 60 Drops/mL ADDitIV Primary Mini-Drop IV Set with
                  PMF Filter and 3 Injection Sites; 
                  17.  V1782 15 Drops/mL ADDitIV Primary IV Set with 3
                  Injection Sites and Low Pressure Trimed Filter (0.22) 
                  Pressure Rating: 20 psi; 87" Length; w/Roller Clamp and Side
                  Clamp;
                  18.  V1778 60 Drops/mL ADDitIV Primary Mini-Drop IV Set with
                  PMF Filter and 3 Injection Sites; 
                  19.  V1787 60 Drops/mL ADDitIV Primary IV Set with 3
                  Injection Sites and Low Pressure Trimed Filter (0.22) 
                  Pressure Rating: 20 psi; 87" Length; w/Roller Clamp & Side
                  Clamp;
                                      -9- 





                  20.  V1788 60 Drops/mL ADDitIV Primary Mini-Drop IV Set with
                  Low Pressure Trimed Filter (0.22) 3 Injection Sites 
                  Pressure Rating: 20 psi; 81" Length; w/V-Clamp; 
                  21.  V5405 Extension Set with 2 Injection Sites and Check 
                  Valve;
                  22.  V7210* 15 Drops/mL Intelligent Pump Primary Additive IV
                  Set with Check Valve, Nonvented; 130" Length; w/Roller
                  Clamp. (*May appear as Cat. Nos. C7210-83 or V7210-83 on
                  the label). 
                  23.  V7215 15 Drops/mL Intelligent Pump Measured Volume 
                  Primary Additive Set with Check Valve and Low Pressure
                  Trimed Filter (0.22), Nonvented; 129" Length; 
                  24.  V7220 15 Drops/mL Intelligent Pump Measured Volume 
                  Primary Additive Set with 150 mL Buret No Automatic 
                  Shutoff;
                  25.  V7225 15 Drops/mL Intelligent Pump Primary Additive IV 
                  Set with Check Valve, Nonvented;
                  26.  V7230 15 Drops/mL Intelligent Pump Primary Additive IV 
                  Set with Check Valve, Vented; 130" Length; w/Roller Clamp;
                  27.  V7230-14 15 Drops/mL Intelligent Pump Primary Additive 
                  IV Set with Check Valve, Vented; 130" Length; w/Roller
                  Clamp & Ball Valve; 
                  28.  V7235 15 Drops/mL Intelligent Pump Primary Additive IV 
                  Set with Check Valve and Low Pressure Trimed Filter 
                  (0.22), Vented; 129" Length; w/Roller Clamp;
                  29.  V7235-14 15 Drops/mL Intelligent Pump Primary Additive 
                  IV Set with Check Va1ve and Low Pressure Trimed Filter
                  (0.22), Vented; 129" Length; w/Roller Clamp & Ball Valve; 
                  30.  V8600 15 Drops/mL AccuPro Pump Additive IV Set with 2
                  Injection Sites;
                  31.  V8610 15 Drops/mL AccuPro Pump Additive IV Set with 3
                  Injection Sites;
                  32.  V8623 60 Drops/mL AccuPro Pump Mini-Drop Additive IV 
                  Set;
                  33.  V8631 15 Drops/mL AccuPro Pump Vented Additive IV Set
                  with 2 Injection Sites; 
                  34.  V8720 15 Drops/mL AccuPro Pump Additive IV Set with
                  Trimed Filter (0.22) and 3 Injection Sites Filter 
                  Pressure Rating: 50 psi;
                  35.  V8730 15 Drops/mL AccuPro Pump Primary Vented IV Set 
                  with Trimed Filter (0.22) and 3 Injection Sites Filter
                  Pressure Rating: 50 psi;
                  36.  V0218 20 Drops/mL Intellegent Pump Primary Additive IV 
                  Set w/Ball Drip Chamber, Nonvented. 
Code:             Lot numbers: 1.  F0C179 
                  2.  F9N133, F9P007, F0P241, F9P279, F9P005, F9S217, 
                  F9S218, F9S219, FOA104, F0A332, F0A340, FOB051, 
                  F0B091, F0B357, FOC086, FOC087, FOC121, F0C156, 
                  F0C157, F0D005, F0D033, F0D065, F0E111, F0E145, 
                  FOH038, F0H072, F0H102, F0J008, F0H050, F0H096, 
                  FOJ142, FOK004, F0K017, F0K066, F0K114, F0K161, 
                  FOL077, F0L177, F0L152, F0N050, F0N081, FON121, 
                  F0N157, F0P008, F0E079, F0H007, F9P321

                                     -10- 





                  3.  F9N134, F9P234, F9P294, F9S325, F0A105, F0A148, 
                  F0A182, F0B002, F0B093, F0C158, F0D006, F0D034, 
                  F0D066, F0E008, F0E047, F0E080, F0E112, F0E146, 
                  F0H008, F0H039, FOH073, F0H103, F0H292, FOH009, 
                  FOJ051, F0J097, F0K018, F0K067, F0K115, F0K162, 
                  F0L047, F0L118, F0L153, F0N051, F0N082, F0N122, 
                  F0N158, F0P009, F0L078
                  4.  F0L400
                  5.  F9S326, F0D021, F0S053, F0J133, F0K048, F0L104, 
                  F0N109
                  6.  F9N147, F9N148, F0S263, F0B053, F0D049, F0E066, 
                  F0J031, F0J161, F0L134, F0L397, F0N101, F0N144, 
                  F0P024, F0H055
                  7.  F9N112, F9P008, F0A106, F0B054, F0B132, F0B358, 
                  F0C013, F0C089, F0C123, F0C159, F0D067, F0E081, 
                  F0H009, F0H074, F0H010, F0H052, F0H098, F0J143, 
                  F0K019, F0K068, FOK116, F0K163, F0L079, F0L119, 
                  F0L154, F0N083, F0N123, F0N159, F0P042, F0E113, 
                  8.  F9P009, F9S009, F0A049, F0A107, F0A149, F0B003, 
                  FOB055, F0B094, FOB133, FOC014, FOC091, F0C124, 
                  F0C160, F0D035, F0D068, F0E048, F0E082, F0E114, 
                  FOE147, F0H075, F0H104, F0H304, F0J011, F0J053, 
                  F0J099, F0K100, F0K117, F0K164, F0L080, F0L120, 
                  F0L155, F0N052, F0N084, F0N124, F0N160
                  9.  F0H061, F0J168, F0A128
                  10.  F9P231, F9S010, F0B007, F0C015, F0C125, F0D069 
                  F0E009, F0E049, F0H076, F0H123, F0H293, F0H119, 
                  F0L099, F0N102, F0N179
                  11.  F9N136, F9P296, F9S265, F0A050, F0B056, F0C091,
                  F0C126, F0D007, F0D036, F0E010, F0E050, F0E115, 
                  F0H105, F0H305, F0H012, F0H054, F0H100, F0L121, 
                  F0L170, F0N069, F0N103, F0N145, F0N180, F0N025, F0E083
                  12.  F0J081 
                  13.  F9S285, F0E132, F0H134, F0K191, F0N110 
                  14.  F9N115, F9P238, F9S012, F9S225, F9S268, F0A064,
                  F0A110, F0A153, F0A186, F0B015, F0B113, F0E052, 
                  F0H057, F0H123, F0K039, F0K140, F0L063, F0L140, 
                  F0P026, F0J035
                  15.  F0J269, F0A233 
                  16.  F9P305, F0A205, F0K049, F0K145, F0P032 
                  17.  F9P229, F9P287, F9S013, F9S226, F0A053, F0A111,
                  F0A187, F0A333, F0A334, F0B027, F0B099, F0B382, 
                  F0C017, F0C129, F0D038, F0E053, F0H042, F0J014, 
                  F0J057, F0K022, F0K071, F0K120, F0L051, F0L123, 
                  F0N127, F0P012
                  18.  F0B061, F0C177, F0H036, F0K090, F0N104, F0E028 
                  19.  F0C037, F0H095, F0K050, F0K102, F0N147 
                  20.  F0A167, F0S019, F0E027, F0K051, F0L067 
                  21.  F9P189, F0A283, F0B219, F0C066, F0D191, F0E360,
                  F0H212, F0H223, F0K291, F0L376, F0N249
                  22.  F0E018, F03062, F0E092, F0E125, F0E158, F0H018,
                  F0H050, F0H086, FOH116, F0H316, F0H317, F0J397, 
                  F0J418, F0K030, F0K129, F0N135, F0N169, F0P021, 
                  FOD232, F0E420, F0H319, F0K007, F0K095, FON420
                                     -11- 





                  23.  F0D080, F0E019, F0E126, F0H051, F0H166, F0K008,
                  F0K096, F0K142, F0K189, F0N182
                  24.  F0E290, F0H215, F0H225, F0H312, F0K443, F0L362,
                  F0N237
                  25.  F0H029, F0K056, F0K150, F0N149 
                  26.  F0E029, F0E134, F0H059 
                  27.  F0K009, F0K057, F0K104, F0K151, F0N062, F0N094,
                  F0N136, F0N170
                  28.  F0E071, F0H030 
                  29.  F0K058, F0K105, F0K152 
                  30.  F9N145, F9P020, F9S027, F0A045, F0A120, F0A163,
                  F0B048, F0B088, F0B128, F0C028, F0C105, F0C141, 
                  F0C173, F0C013, F0C046, F0F081, F0E063, F0E093, 
                  F0E127, F0E159, F0H052, F0H087, F0H117, F0J025, 
                  F0H068, F0J113, F0J154, F0K031, F0K081, F0K130, 
                  F0K178, F0L059, F0L094, F0N063, F0N064, F0N095, 
                  F0N137, F0N171, F0L134, F0L166
                  31.  F9P021, F9P242, F9P243, F9P334, F9S029, F0A046 
                  F0A121, F0A164, F0A199, F0B024, F0B049, F0B089, 
                  F0B129, F0C106, F0C174, F0D082, F0E020, F0E064, 
                  F0E094, F0E160, F0H019, F0H088, F0H118, F0J026, 
                  F0H069, F0J114, F0J155, F0K032, F0K082, F0K131, 
                  F0K179, F0L095, F0L167, F0L384, F0N096, F0N138, 
                  F0N172, F0P022, F0P094
                  32.  F9P022, F0B090, F0C107, F0C175, F0D047, F0D014,
                  F0E095, F0H053, F0H302, F0J129, F0K190, F0L065, 
                  F0N185
                  33.  F9N146, F9P023, F9P047, F0A122, F0A200, F0B050,
                  F0C030, F0C108, F0C142, F0C048, F0E021, F0E096, 
                  F0E161, F0H054, F0H089, F0H119, F0J070, F0J115, 
                  F0H156, F0K033, F0K132, F0K180, F0L096, F0L135, 
                  F0L168, F0N097, F0N173, F0P095
                  34.  F0A201, F0B240, F0B241, F0B296, F0C061, F0C201,
                  F0C235, F0C279, F0C313, F0E345, F0K450, F0K451, 
                  F0L346, F0N326, F0N383
                  35.  F9P261, F0C070, F0E205, F0H315, F0N419 
                  36.  F0H323.
Manufacturer:     McGaw of Puerto Rico, Inc., Sabana Grande,
                  Puerto Rico.
Recalled by:      McGaw, Inc., Irvine, California, by letter February 13, 
                  1991.  Firm-initiated recall ongoing. 
Distribution:     Nationwide and Canada.
Quantity:         7,221,335 units were distributed. 
Reason:           Product uses a piggyback feature in which the rubber
                  check back valve may become tilted, allowing back flow
                  during a secondary (piggyback) infusion.

                                     ---- 

                                     -12- 





                  Class III - 

Product:          CBC-7 High Hematology Control in CBC-7 Dual Pack and
                  Tri-Level Pack of Whole Blood Hematology Controls for 
                  in-vitro diagnostic use.  Recall #Z-651-1.
Code:             Lot #R110H EXP 2/5/91.
Manufacturer:     R & D Systems, Inc., Minneapolis, Minnesota.
Recalled by:      Manufacturer, by letter October 19, 1990 and by 
                  FAX October 23, 1990.  Firm-initiated recall ongoing. 
Distribution:     Nationwide, France, Canada, Australia, New Zealand, 
                  Ireland.
Quantity:         2,713 vials of CBC-7 High Hematology Control in about 
                  600 packages of Dual Pack and Tri-Level controls
                  were distributed. 
Reason:           The product labeling for assay values and expected ranges 
                  gave an incorrect hemogologin concentration assay 
                  value/range.

                                     ---- 

                              VETERINARY PRODUCTS 

                                     NONE 

Seizure Actions Filed:

Product:          U.S.P Oxygen (90-575-171/173).
Charge:           Current good manufacturing practice deficiencies. 
Firm:             Price-Bass Company, Inc., Nashville, Tennessee. 
Filed:            February 1, 1991 - U.S. District Court for the Middle 
                  District of Tennessee; Civil #3-91-0086, FDC #66011/66013.