FDA ISSUES A NATIONWIDE WARNING AGAINST THE USE OF "ABTOX" PLAZLYTE STERILIZATION SYSTEM

FDA is warning hospitals and physicians against the use of the Abtox Plazlyte Sterilization System, because FDA is aware of at least 10 injuries associated with ophthalmic surgical instruments which had been sterilized with this system. These reports involved serious injuries to the cornea with damage to vision, and at least two cases have required corneal transplantation.

The product's manufacturer, Abtox, Inc., of Mundelein, Ill. has initiated a worldwide recall for this model of device. This device has never been cleared by FDA.

The problem appears to be formation of toxic salt from the chemical sterilants when the device is used on surgical tools made of copper, brass or zinc.

The firm has notified each of the 155 institutions, numerous foreign accounts and individuals who received these unapproved sterilization systems about the recall.

Users of these systems should be aware of the following:

FDA will actively monitor the progress of this recall and will take whatever appropriate action is needed to assure its successful completion.

####